Three lead detection on ADS1298ECGFE-PDK is not working

i wanted to use the analog front end kit to record my ecg signal. I am using a three lead detection system. I tested the connections and they are fine but my signal continues to be noisy. Has anyone used it for real time ecg signal detection?

  • Sneha,


    I am moving your post to the correct forum.

  • Hey Sneha,

    The EVM was built for performance evaluation and was not intended to be used in end-equipment. As a result, it is not optimized for real-time detection or measurement. The "out of the box" version of the EVM will not be able to collect and stream data in real time. I have heard of people changing the design in some way in order to fit their needs, but of course TI will not be able to provide EVM support if you make a change that is not specified in the user's guide. For this device, the firmware and software source code are included in the installation package on the product page.

    Regards,
    Brian Pisani
  • Hey Brian,

    Thanks for the reply! I manged to figure it out using the filters and hitting a good posture for ECG detection