I think this should be in the Medical forum, but for some reason I can only select the High Speed forum.
I have a couple of questions regarding the offset spec of the ADS5263. In the Electrical Characteristics table, it lists the Offset as +-10 mV typ, +-30mV max. Assuming fullscale is 4V diff, and 16 bit resolution, 30 mV offset results in code error of +/- 491 codes (164 codes for 10 mV offset). Is this correct?
Figure 34 shows the output code histogram for the input shorted. The main bin of the histogram is 202 LSBs less than midscale (equivalent to about 12 mV). Can this be directly related to offset? I'm not sure if this is really equivalent to the offset spec since this is measured at midscale which might include gain errors on top of offset error.
Also, is Figure 34 measured from one device on one input? or the average of multiple inputs or devices?
Some data was collected on each input of 4 devices with no input signal in the ADC inputs (not shorted). Each ADC input is AC coupled with VCM connection through 50 ohms on each input. Collecting multiple output measurements on each ADC output at 100 MSPS across temp and averaging, the offset from midscale varied from about -140 LSBs to -460 LSBs. Most of the variation was from ADC to ADC not so much from temp. Does this much variation make sense? Some ADCs are much higher than what Figure 34 implies. Maybe the measurement technique of leaving the inputs open is adding errors?