Hello experts,
I'm using an ADS124S08 for NTC measurements, and while going thru noise calculations, I ended up with some rather bad results. I hope you can help me understand what's really going on and how to improve the system resolution.
My design currently implements a Two-Wire Low-Side topology, and my main concern is related to the precision required by the reference resistor, and noise impact due to resistance tolerance and temperature drift.
I'm considering using a 0.05% tolerance resitor and 10 ppm/ºC coefficient. This presents a good cost/benefit ratio. It seems higher precision resistor are much more expensive.
As for the noise calculations, I'm using the equations from reference slau520a - 3-Wire RTD Measurement System Reference Design.
For the calculations below, I have the following parameters:
- Vin(max) = 0,735V (result from Rref, NTC and IDAC current calculation)
- Gain = 1
The resistance tolerance error I get, using equations 38-40, is:
- Gain error = 0.04998%
- Gain error noise = 367,45 uV
The temperature drift error I get, using equations 56-60, is:
- Temperature Delta = 65 - 15 = 50ºC
- Resistance drift = 0.05%
- Gain Error due to Resistance drift = 367,45 uV
Total Noise from those two sources alone equals 525,05 uV (square-root sum), and using equation for effective resolution from the component datasheet (eq 1), yields an ENOB of 11,479.
In addition, this ENOB give me a resistance resolution of 5,25 ohms, when my system needs 3,08 ohms.
The questions are:
- Are those calculations correct? The final ENOB and resolution numbers seems correct?
- How can I improve the situation, without recurring to a very expensive reference resistor?
- Is it possible to calibrate during manufacturing process to remove the tolerance error? Is yes, how can that be done? (what kind of setup is needed, what tools are needed, how long it could take, etc..)
Best regards.