Dear TI Experts:
We have configured a 64 channel system with 8 TI chips in a referential montage. We are using 64 P inputs, single reference electrode using SRB1 connected to all chips, bias electrode with PD_BIAS enabled only on one chip and BIAS_SENSP/BIAS_SENSN enabled on all channels.
We have configured DC lead off detection at 24 nA, set the comparator thresholds to 70%/30%, and have enabled LOFF_SENSP for all channels. We're logging the status bits from each channel for lead off status.
1) Even though we're seeing the expected behavior when an electrode is removed, the comparators do not consistently indicate lead off events (see figures below). Is this expected?
2) Since the mux disconnects the SRB1 from the lead-off current source, can we physically connect the SRB1 to one of the N inputs to access the current sink from that channel and use the comparators on that channel for lead-off status of the reference electrode simultaneously with all other electrodes?
3) Shouldn't the bias be trying to correct and DC offset introduced by the lead off current sources? Would there be any difference if we disable BIAS_SENSP/BIAS_SENSN?