Hello,
I am playing with EVM ADS1298RECGFE-PDK with channels assignment as follows: LA = IN2P; RA = IN2N, IN3N; LL = IN3P; C1 = IN8P; N = ECG_RL. I use unipolar 2.5V supply voltage. Which bits from RLD_SENSP and RLD_SENSN registers should I use as "RLD derivation"? Is there any general rule which inputs should be taken for "RLD derivation"? I was not able to find such guidelines, therefore I have set "RLD derivation" from all inputs I use, but I am not sure if this is correct (see question below regarding lead-off).
Another question (somehow related to RLD) is about lead-off detection. I have configured lead-off detection to DC current mode 12nA and selected LOFFP and LOFF bits for all inputs I use:
I noticed that in such configuration I am not able to detect when RA electrode (IN2N, IN3N) is connected. It is permanently detected as "LEAD OFF" (even if shorted directly with RLD). Lead-off detection for LA , LL and C1 works correctly. Lead-off detection for RA works correctly if I set all bits from RLD_SENSP and RLD_SENSN to 0 (no RLD derivation). What am I doing wrong? What should I do to have RLD and lead-off detection working correctly at the same time? Should I make any changes in LOFF_FLIP register?
Thanks & regards, Maciej