• Join
  • Sign In with my.TI Login
Texas Instruments
  • Products
  • Applications
  • Tools & Software
  • Support & Community
  • Sample & Buy
  • About TI
Sample & Purchase Cart Sample & Purchase Cart
  • Search
  • Advanced
TI E2E™ Community
  • Support Forums
  • Blogs
  • Groups
  • Videos
  • 简体中文
  • More ...
TI Home » TI E2E Community » Support Forums » Data Converters » Precision Data Converters » Precision Data Converters Forum » ADS1298 using single-ended inputs
Share
Precision Data Converters
  • Forum
  • Files
  • E2E Wiki
Options
  • Subscribe via RSS
Check out
The Signal blog
  • $core_v2_blog.Current.Name

    Handy Gadgets and Resistor Divider Calculations

    Posted 4 days ago
    by Bruce Trump
    Handy gadgets make our engineering life easier—the little...
  • $core_v2_blog.Current.Name

    Chopper Op Amps—are they really noisy?

    Posted 12 days ago
    by Bruce Trump
    Chopper op amps offer very low offset voltage and dramatically...
  • $core_v2_blog.Current.Name

    Bypass Capacitors… yes, but why?

    Posted 24 days ago
    by Bruce Trump
    Everyone knows that op amps should have power supply bypass capacitors...

ADS1298 using single-ended inputs

ADS1298 using single-ended inputs

This question is answered
Naveen
Posted by Naveen
on Aug 08 2012 04:54 AM
Prodigy10 points

Hi,

This is our first project in medical instrumentation. We are designing a educational medical measurement unit using ADS1298 device. We are using single ADS1298 device to capture 1 channel ECG (PGA inputs 1,2,3,4) and 4 channel EMG (PGA inputs 5,6,7,8). In our first version we saw both sections working together, however, the gain (set to 12) was inadequate for EMG and signal levels were very low and we were observing noise in the captured EMG signal. We wanted to include a pre-amp for EMG before feeding the signal to ADS1298 device. In this regard I had few questions (I am looking for answers for specific questions which I could not find in the E2E community postings):

The whole system is battery operated and works with single supply of +3.3V.

1. For EMG, each electrode (passive) will have 2 leads which we connect to ADS1298 PGA inputs, along with that there will be shield which is required for reference. Can we use the same RLD reference generated in the device for ECG for  the EMG shield or it must be seperate shield signal which needs to be generated for each electrode pair?

2. We are planning to use a instrumentation amplifier such as INA128 or INA333 in pre-amp stage. The output of INA will be single ended. Since ADS1298 input can take differential or single ended, we wanted to use PGA inputs 5,6,7,8 in single ended mode. I will be connecting the output of INA128 to IN5P of ADS1298. What to do with IN5N? As per datasheet, "When the input is single-ended, the INN input is held at the common-mode voltage,preferably at mid-supply". Is this mid-supply referring to VDD/2 generated through resistor divider?

3. In single-ended mode, what will be the effect on max gain setting of the PGAs?

4. Since the device is non-isolated, where do we need to insert the isolation? Is this isolation required for battery operated devices?

Please clarify our doubts.

Regards, Naveen

ads1298 single-ended input
Report Abuse
  • Reply
You have posted to a forum that requires a moderator to approve posts before they are publicly available.
All Replies
  • Greg Hupp
    Posted by Greg Hupp
    on Aug 09 2012 17:57 PM
    Genius9310 points

    Naveen -

    We are working on a reply and should be getting back with you shortly

    Greg Hupp

    Hardware and Software Platform Development

    Precision Data Converters

    Report Abuse
    • Reply
    You have posted to a forum that requires a moderator to approve posts before they are publicly available.
  • Tony Calabria
    Posted by Tony Calabria
    on Aug 10 2012 10:02 AM
    Verified Answer
    Verified by Tom Hendrick
    Mastermind42750 points

    Hi Naveen,

    I will do my best to try and answer your questions below -

    1. The shield drive interest and questions have had more to do with ECG and the RLD. We are not intimately familiar with a lot of EMG applications where the requirements for the shield drive may be different. ECG shield drive for the RLD, we invert the RLD signal externally and route it as the shield for the RLD signals.

    2. Depending on what you are measuring, you could either use the WCT or RLD signal set to mid supply. These two signals are configurable using internal registers which can be read about in the data sheet. Setting the negative supply to mid rail is ideal because it lets you maximize the code range you are using.

    3. Depends on the gain stage from the INA and what you have the INN pin set to. You cannot exceed the full scale differential input voltage of +/-Vref / GAIN OF ADC. The analog input to the ADS1298 including common mode restrictions is explained in detail on pages 24 and 25.

    4. You will need patient protection on the front end to limit current that can shock the body. Where and how you implement any patient protection will depend on the classification of the equipment and the safety standards your end equipment needs to pass.

    Regards,

    Tony Calabria

    Report Abuse
    • Reply
    You have posted to a forum that requires a moderator to approve posts before they are publicly available.
TI E2E™ Community
  • Support Forums
  • Blogs
  • Videos
  • Groups
  • Site Support & Feedback
  • Settings
TI E2E™ Community Groups
  • TI University Program
  • Make the Switch
  • Microcontroller Projects
  • Motor Drive & Control
Other Communities
  • Deyisupport
  • Designsomething.org
  • beagleboard.org
  • TI on Element 14
  • TI on TechXchangeSM
Other Technical & Support Resources
  • WEBENCH® Design Center
  • Product Information Centers
  • Technical Documents
  • TI Design Network
  • TI Technical Articles
  • TI Training

All content and materials on this site are provided "as is". TI and its respective suppliers and providers of content make no representations about the suitability of these materials for any purpose and disclaim all warranties and conditions with regard to these materials, including but not limited to all implied warranties and conditions of merchantability, fitness for a particular purpose, title and non-infringement of any third party intellectual property right. TI and its respective suppliers and providers of content make no representations about the suitability of these materials for any purpose and disclaim all warranties and conditions with respect to these materials. No license, either express or implied, by estoppel or otherwise, is granted by TI. Use of the information on this site may require a license from a third party, or a license from TI.

Content on this site may contain or be subject to specific guidelines or limitations on use. All postings and use of the content on this site are subject to the Terms of Use of the site; third parties using this content agree to abide by any limitations or guidelines and to comply with the Terms of Use of this site. TI, its suppliers and providers of content reserve the right to make corrections, deletions, modifications, enhancements, improvements and other changes to the content and materials, its products, programs and services at any time or to move or discontinue any content, products, programs, or services without notice.

Follow Us Texas Instruments on Facebook Texas Instruments on Twitter Texas Instruments on LinkedIn Texas Instruments on Google+
TI Worldwide | Contact Us | my.TI Login | Site Map | Corporate Citizenship | mobile m.ti.com (Mobile Version)

TI is a global semiconductor design and manufacturing company. Innovate with 100,000+ analog ICs and
embedded processors, along with software, tools and the industry’s largest sales/support staff.

© Copyright 1995-2013 Texas Instruments Incorporated. All rights reserved.
Trademarks | Privacy Policy | Terms of Use