Would enabling LeadOff Detection affect the Respiration Impedance signal on Channel1?
Do I need to disable Lead Off Detection in order to get a clear Respiration Signal?
Thanks,
This thread has been locked.
If you have a related question, please click the "Ask a related question" button in the top right corner. The newly created question will be automatically linked to this question.
Would enabling LeadOff Detection affect the Respiration Impedance signal on Channel1?
Do I need to disable Lead Off Detection in order to get a clear Respiration Signal?
Thanks,
Hi Eric,
Thanks for your question.
When Channel 1 is configured for a respiration impedance measurement, the inputs are AC coupled - therefore, DC lead-off detection would not work on this channel.
AC lead-off detection might still work, but I'm not sure what effect it might have on the respiration signal. I would imagine that you could do it because these signals are in different frequency ranges: RESP is between 0.1Hz - 0.5Hz; AC lead-off = fDR/4; and RESPMOD is either 32kHz or 64kHz.
That said, typically the same electrodes connected to the respiration path on Channel 1 are also connected to Channel 2 for ECG measurements. It would probably make more sense to disable lead-off on Channel 1 and monitor the lead-off status on Channel 2 instead.
Please let me know if you still have any questions.