Hello,
I am working with the ADS1248 to take measurement. I developped the driver and it's ok.
Now I have to make comparison between ADS1248 measures and a DAQ (DT9826) to prove that ADS1248 measures are good. I made two tests and the results was the following:
- Test 1:
I connected a calibrator to both ADS1248 (ch0,ch1) and DT9826 inputs (same calibrator output).
The calibrator outputs a voltage equal to 500mV.
The result is that I get exactly 500mV on the DT9826, but I get 497mV on the ADS1248.
The strange part is that when I only connect the ADS1248 to the calibrator I get 500mV like I am supposed to receive.
When I put back the DT9826 in parallel the ADS measures 498mV again.
-Test 2:
I connected a PT100 sensor in 4 wires configuration. On the ADS1248 the measurement channels are ch0 and ch1, and the excitation channel is ch4.
On the DT9826 I connect ch0 input at the same points ch0 and ch1.
on the DT9826 I get around 110 mV (110ohm*1000µv excitation) wich is good, on the ADS1248 I get full scale measurement.
When I disconnect the DT9826, the ADS1248 measures around 110 mV.
Could this problem be related to the input impedance of one of the DAQs? Otherwise why am I getting full scale when I connect the DT9826?