I have to measure resistance with step change of 0.010ohm & max upto 2 ohm. Attached is the circuit which I made.
ADS1248 analog & digital voltage is 3.3V. Will be suing internal reference of 2.048V
1. Max current source of ADS1248 will be 1.5mA so I am uisng this value. A 1000ohm(1%) resistor is used to level shift the voltage in CMIR of ADS1248.
2. For step change of 0.010ohm & with 1.5mA current source, step voltage change will be 15uV (10mohm * 1.5mA).
IF I use PGA=128, AVDD=3.3V, AVSS=0V, Internal Vref=2.048V & data rate =5sps, I think I can measure the value accurately.
CMIR range with ganin=128 & max vin=2ohm*1.5mA = 0.003V.
0.292V < Vin < 3.008V.
Full scale range = +-2.048/128 = 16mV.
Also page13 of datasheet says peak noise with AVDD=5V & PGA=128, data=5SPS is 0.29uV.
Will ADS1248 will work as predicted in above calculation?