Other Parts Discussed in Thread: ADS1298,
I am currently developing medical holter with ADS1292IRSMT front end. I have problems with complying with "frequency response" test according to IEC 60601-2-47. The response to a 3mV 100ms rectangular pulse shall not show a baseline amplitude displacement more than 0,1mV and the slope sholud be less than 0,3mV/s. With fs=125Hz we achieved 0,8mV displacement and with fs=500Hz 2,1mV. From my calculations it looks like there is a high passs filter with 3dB frequency of 0.5Hz. In order to comply with IEC 60601-2-47 we should have a 3dB frequency at 0.05Hz not 0.5Hz. I did not have problems with our other holters based on ADS1298.
For me it looks like ADS1292 could comply requirements of 60601-2-27 (monitoring ECG) where they require 0,67Hz-40Hz badndwidth. For ambulatory recorders with diagnostics I cannot achieve small amplitude displacement after pulse.
What could cause a problem?
Thank you in advance for your help.