DLPNIRSCANEVM: Hi, I currently have DLP NIRSCAN EVM and I think I've got an faulty item !! I don't see any scanned data whatsoever !please help

Part Number: DLPNIRSCANEVM

Hi I have recently bought a DLP NIR scan EVM. I am trying to test the device and have some problem measuring NIR using liquid samples. I first did the referencing by inserting nothing in the holder (not even cuvette) and reference measured. After that, I inserted plastic cuvette nothing contained and obtained the plot below

After that, I filled the cuvette with water and obtained the plot below, which has zero absorbance as shown below. Why It shows no data? Is this device's problem or am I missing something? I also tried referencing with empty cuvette and scanned with water-filled cuvette and it showed no data, which is same as the previous run. Can anyone please help me with this issue? Thank you.

Saved data.

Method: Line Scan
Host Date-Time: 22/11/2017 @ 21:42:32
Sample Name: default_scan_1
Spectral Range Start (nm): 1350
Spectral Range End (nm): 2450
Number of Wavelength Points: 200
Digital Resolution (nm): 5.5
Number of Scans to Average: 1
Total Measurement Time (ms): 300
PGA Gain: 32
Wavelength (nm) Absorbance (AU) Reference Signal (unitless) Sample Signal (unitless)
1353.75 -3.3062 1886390 -4.8E+08
1359.75 -3.3522 1696788 -4.8E+08
1365.76 -3.3011 1908539 -4.8E+08
1371.76 -3.3441 1728795 -4.8E+08
1377.75 -3.2943 1938802 -4.8E+08
1383.74 -3.3433 1731926 -4.8E+08
  • Hi Young Ho Shin,

    From the data you've attached, it looks like you were able to get a good reference sample taken, which is great. Something that would help us understand what is going on would be to see sample scans (with excel data) of the following:

    * No sample (open beam)
    * Empty Cuvette
    * Cuvette with Water

    One thing to keep in mind is that water has a very strong absorption curve past about 1400nm, so not much light will hit the sensor. It looks like your cuvette also has a strong absorption at higher wavelengths as well.

    Best,
    Paul