LDC1000 inductance measuremet hysteresis

Hello all;

I spend some time with LDC1000EVK and I make some measurement with various coils.

My requirements:

measurement distance: 6cm

resolution and repeatability: better than 0,5mm

size of coil and steel plate in not critical, and both may be quite big.

Tested coils:

air coil - diameter ~40mm, ~100 turns/~850uH, Ct=1nF, f=180kHz, Cf=220pF, rod located parallel to steel plate (best results in distance range)

ferrite 3F3 coil1 - ~200 turns on ferrite rod, 5cm long, Ct=1nF, f=95kHz, Cf=470pF, rod located parallel to steel plate (best results in distance range)

ferrite 3F3 coil2- ~80 turns on ferrite rod, 5cm long, Ct=1nF, f=198kHz, Cf=470pF

ferrite coil3 - WE coil for QI, but this have very short measurement distance

photo shows my test bench

 

my test bench allows to set distance with 0,05mm or better repeatability (micrometer screw), this micrometer test bench my company buy at farnell. all others components was designed specially for test LDC1000EVK

 

PROBLEM DESCRIPTION

Problem is with coils with ferrites.

Example, ferrite coil1:

1) set distance to 5,00cm - L=470,000uH

2) set distance to 0,30cm - L= I'm not remember now, i think this is not important now

3) set distance back to 5,00cm - L=469,970uH

4) if I repeat 5 times steps 1)...3) then every final result in step 3 was smaller. Of course error is big only at first step, all next step has smaller errors, ~0,003uH

It looks like some hysteresis.

In my implementation this problem is critical, it makes measurement error quite big - at about 0,5mm or worst.

I have not tested yet this problem with air coils.

My observation: current which flows via coil looks like half-sine wave. this makes some DC offset/component which may magnetize ferrite. Similar problems are with flyback converter, but in SMPS has little importance.

Does have someone similar experience?

Is it possible to avoid this problem? I think that problem is hard to solve, because avg current depends on distance.

Regards

Wojciech Błędziński

  • Hello Wojciech,

    Thank you very much for the description. I would need more information to assist. What is the composition of the target you are using? Are you using L measurement only? Yes, half sine wave is what you usually see on the LC tank if you measure INA or INB to GND. The signal is differential between INA and INB.

    Best Regards,
    Natallia Holubeva
  • Hello Natalia,

    thank you very much for your support.

    Same problem (i call it "hysteresis" but i'm not sure if this is a good term) I observe for L and R.

    But still my question is actual - why I see hysteresis in measurements.

    Hysteresis is big with measurement of smallest distances - <10mm.

    With larger distances I do not notice hysteresis.

    About signals INA and INB - it is only my suggestion, that if there is some DC offset, then this phenomenon may exist. But I'm not sure for seriously.

    Please help or confirmation of the existence of such a phenomenon.

    Regards,

    Wojciech Błędziński

  • Hello Wojciech,

    Could you please provide a scope shot of what you call hysteresis? Your help is greatly appreciated.

    Best Regards,
    Natallia Holubeva