This thread has been locked.

If you have a related question, please click the "Ask a related question" button in the top right corner. The newly created question will be automatically linked to this question.

Calibration, Signal Strength

Part Number: AFE4400

while performing calibration of our custom afe4400 board using Fluke  "ProSim SPOT Light SpO2 Pulse Oximeter Analyzer".

The afe4400 register configuration is same as of afe4400spo2_evm. Using this same configuration(Led1,2 = 4mA), the signal strength shown on  ProSim is around 30%.

So to improve signal strength(60-70%) we increased the LED1,2 current from 4mA to 10mA . Due to this, RED and IR adc data is saturated.

How to prevent adc saturation while maintaining good signal strength?

What should be the required signal strength ?