Hello, I was wondering if you can shed some insight into two issues we have been observing with our OpAmp circuit topology. We have a Transimpedance amplifier circuit design (schematic drawn below), that essentially converts a picoamp current from ionized electrons from a VOC gas, into a voltage that we read with an ADC. The gas is ionized via a UV lamp and the ionized electrons induce a current when they travel through the electric field. One plate of the electrode is at -200V and the other plate is connected to the inverting input of the OpAmp.
1. The first issue we have been seeing is that the output signal of the OpAamp, when responding to gas, has some oscillation behavior. Initially almost all the sensors' output signals when responding to gas seem ideal (picture attached below of ideal response), but then after some prolonged time powered on, most sensors begin to show this oscillation effect. See photo below of oscillation behavior. I am wondering if this could be related to some noise issues on our GUARD net that could be causing an elevated output and nonideal signal response. Technically our Baseline or output of the OpAmp is supposed to be close to ~100mV but when we ignite the lamp, the output Baseline significantly increase to somewhere between 200mV and 350mV.
2. The second issue we have observed is that sometimes some sensors will show an "inverse" response to gas. So when the output of the OpAmp is supposed to swing high, the output will instead drop, but once it reaches a random value above the baseline ref 100mV, we see the signal swing back high. Please see attached photo below of this behavior, PINK signal is showing this problem.
I am wondering if perhaps both of these problems we are observing are related? Possibly noise issues. I was wondering if you had any recommendations as to what to look out for, or any methods you'd suggest for testing and verifying some of these behaviors and how to improve our design. I have written some notes about the circuit and can provide more information if you have any questions.
Notes:
a. This oscillation behavior tends to happen to sensors that have been powered on over time and dont really show this behavior initially. Also, when the sensor was typically reaching say 1.5V when responding to a certain gas concentration, the oscillation behavior will cause the output signal to NOT be able to reach that same level of output voltage and will be lower.
b. BaseLine is supposed to be at 100mV, but output is always around ~300mV when lamp is ignited. Tends to be roughly closer to the ~100mV when lamp is NOT ignited.
c. Have observed when C15 is broken, missing, or not soldered properly the output signal suffers and doesnt reach the proper voltage levels its supposed to. This is one of the reasons why i believe it could be related to noise issues. Do I need better filtering?
Schematic: link and photo attached
IDEAL response
Inverted pink response signal
Oscillation behavior