ADS1298: ADS1298, the 27 bytes in the first frame are all 0 after the first falling edge of DRDY

Part Number: ADS1298

Hi,

    My device uses ADS1298, and everything runs OK. The data collect is OK. I have used it for several years.

    Recently, I find that when the ADS1298 is started by writing the ADS_START (0x08) command, the first data read after start is "0" for all of the 8 channels.

    After that, everything runs OK forever.

    That only happens at the first frame I got after writing the ADS_START (0x08) command.

And the status word in the first frame after falling edge of DRDY pin is not "1100", what I get is "0000".

In fact, the 27 bytes in the first frame are all 0 after the first falling edge of DRDY.

Is that common for ADS1298?

    Thank you.

  • Hi Feng,

    The status word in the ADS129x data readback will always start with 0b1100, as shown in Figure 61 in the datasheet. Does this happen if you issue a STOP command (0x0A) following by the START command? It may be helpful if you can capture the SPI waveform with a logic analyzer or oscilloscope.

    Thanks

    -TC