ADS1147: ADS1147IPWR

Part Number: ADS1147

Dear Sir/Madam,

In our application we are using ADS1147 for measuring temperature from Pt100 sensor in a 4-wire measurement system. Our micro controller code is as per the pseudo-code example mentioned in page 65 of ADS1147 datasheet.

We have written the following values into configuration registers.

REGISTER

VALUE

 

MUX0

0x0A

 

MUX1

0x20

 

SYSO

0x3F

 

IDAC0

0x06

IDAC Excitation Current Magnitude set to 1mA

IDAC1

0x0F

 

 

As mentioned in the above table we have configured the excitation current source as 1mA. Also by using RREG command the read back data from IDAC0 register is 0x96. This confirms our setting of 1mA. Our 4-wire RTD measurement is based on Four-Wire RTD Measurement, Low-Side reference mentioned in page 23 of application note sbaa275

Our RREF is 1.65k Ω and so the expected voltage across RREF must be 1.65V( 1.65k* 1 mA).

However, we are getting 2.46 V. So this indicates that the current is 1.5 mA( 2.46/1.65k) but not 1mA.

Why ?.  Can you help us what is wrong ?

  • Hi Prakash,

    Are you just measuring the voltage across RREF with a DMM, or are you back calculating this from the measurement data?

    Can you try taking some measurements to see if the output date is correct? You can just provide the hex code values you receive from the ADC, there is no need to convert them to a voltage.

    Also, can you send us a schematic to review?

    -Bryan