Dear Sir,
I have doubt regards DDC112U can support to measure -100nA to 100nA with 1pA precision. This doubt i asked in Texas customer support and the case number is CS2211553.
Reply from Customer Support: The precision is too small for this IC. The device cannot support a -100nA reading with 1pA per LSB. DDC112U can read 24.4pA per LSB for -100nA measurement.
As per my understanding of DDC112U, precision is 0.383pA/LSB.
Explanation:
ADC Resolution: 20 bit -> 2^20 = 10,48,576 ADC Count.
Range 4: 200pC
Tint : 500us
Negative FSR is -0.4% = -0.8pC
Total Measuring Range = 200.8pC, In terms of Measuring Total current is 200.8pC/500us = 4,01,600pA or 401.600nA.
Precision pC/LSB = 200.8pC/1048576 = 0.0001915pC
Precision pA/LSB = 4,01,600pA/1048576 = 0.38299pA.
To measure Bipolar input, simply current source circuit added in input side. This recommended by Texas support, link Creating_a_Bipolar_Input_Range_for_the_DDC112.
In this Bipolar configuration, we are simply shifting the measuring scale.
Without Bipolar Config circuit:
-Full Scale(pC) is -0.8
+Full Scale(pC) is 200
In terms of current -1600pA to 400000pA or -1.6nA to 400nA
With Bipolar Config circuit:
-Full Scale(pC) is -103.2
+Full Scale(pC) is 97.6
In terms of current -206400pA to 195200pA or -206.400nA to 195.200nA
So, it can be able to measure -100nA to 100nA.
Please kindly check my above comments and refer the attached schematic and doc.
Range:200pC
Range:50pC
Schematic:
And eagerly waiting for your reply.
please check and reply, is this design possible to measure -100nA to 100nA with 1pA precision.
One more question, could you please suggest ACF2101 or DDC112U which one is best suitable for my application.
Thanks & Regards,
Gowtham R,
Panacea Medical Technologies