测试发现通过DI05=1的条件下去PD,会在当前帧的第16个SCLK下降沿PD,而不会在按照SPEC所说的下一帧的第16个SCLK PD,下面示波器第一张是15个SCLK下降沿没有PD成功,而第二张16个SCLK下降沿便PD成功了
This thread has been locked.
If you have a related question, please click the "Ask a related question" button in the top right corner. The newly created question will be automatically linked to this question.
Hi - welcome to our e2e forum. Please post your query in English.
The test found that PD under the condition of DI05=1 would fall along PD at the 16th SCLK in the current frame, but not at the 16th SCLK PD in the next frame according to SPEC. Below, the first oscilloscope picture showed no PD success at the falling edge of 15 SCLK, while the second one achieved PD success at the falling edge of 16 SCLK
Hi,
What in the two pictures you provided shows that in one case the device did not power down while in the other it did? What is the signal on the yellow trace (CH1)?
OK - so what are you using as a definitive case for the device powered down? I don't see that in the screen shots you provided.
So in the first screen shot, yellow as /CS is not valid low for the first 3-4 SCLK cycles. Can you fix that?
From the last two screen shots, note how /CS (blue trace this time) is going high slightly ahead of the last SCLK, while in the second picture, there is maybe 200nS delay before /CS goes high. Can you keep the /CS time constant observing at least tSU1 and tD3 timings (see page 17 of the datasheet)?
OK - I believe I understand now what your issue is. The text in section 8.3.2 is not correct when it says "in the next data frame". Tables 1, 2 and 5 are correct. We'll get that fixed in the next datasheet revision.