Part Number: ADS131M08
Hello TI Technical Support,
We are experiencing a reproducible synchronization issue with two ADS131M08 devices configured per datasheet section 9.1.5 "Multiple Device Configuration."
## PROBLEM SUMMARY
**Hardware Setup:**
- Two ADS131M08 ADCs sharing CLKIN (7.81 MHz), SYNC/RESET pin tied together
- OSR = 4096, Sample rate = 954 Hz
- Separate CS pins, shared SPI bus
**Issue:**
After exactly 60 seconds from power-on/reset, ADC2 DRDY signal becomes random while ADC1 remains stable. This happens once (not periodically) and is 100% reproducible.
**Timeline:**
- 0-60s: Both ADCs synchronized, stable operation
- At 60s: ADC2 DRDY becomes random, so measurements oscillate
- Observation: Issue does NOT occur again after 60s if left running
## VERIFICATION
When we manually pulse SYNC/RESET at 61 seconds (after the desync occurs), the system becomes stable for 30+ minutes. This confirms that proper post-configuration synchronization resolves the issue.
## QUESTIONS
1. Is post-configuration SYNC pulse required for multiple device setup? Datasheet section 9.1.5 is unclear on timing.
2. Is there any particular timing needed apart from Datasheet requires: tREGACQ = 5µs minimum , in init sequence cause this exact 60-second desynchronization?
3. Is there a known internal timeout/counter in ADS131M08 that manifests at 60 seconds when devices are improperly synchronized?
4. What is the recommended initialization sequence for multiple ADS131M08 devices?
## ATTACHED DOCUMENTATION
- Oscilloscope captures of DRDY signals
- When both SYNC ADS1 and ADS2 chip.
when ADS2 is desync inbetween.