Hello all,
i want to design a simple ECG device. Is it nessecary to protect the patient with a extern protection circuit
or does the ADS1298 has an internal protection??
Kind regards
Sebastian
This thread has been locked.
If you have a related question, please click the "Ask a related question" button in the top right corner. The newly created question will be automatically linked to this question.
Hello all,
i want to design a simple ECG device. Is it nessecary to protect the patient with a extern protection circuit
or does the ADS1298 has an internal protection??
Kind regards
Sebastian
Hi Sebastian,
The ADS1298 does not offer any internal 'patient protection' circuitry. You would have to develop your own protection scheme based on the end product classification of the product/design you wish to build. Sports/fitness versus clinical/medical application - each are governed by different standards, so we can't provide you with a definitive answer here.
Hi Tom,
thank you for your answer. That is all i want to know so it is a definitive answer. =)
Regards,
Sebastian