TLA2021: 0 - 3V signal input - HW and SW configuration
Part Number: TLA2021
Other Parts Discussed in Thread: TLA2022, TLA2024, ADS1115
Hello E2E,
I would like to consult how to configure HW and SW on TLA2021 for 0-3V input signal.
for example, minimum bias voltage, setting, so on.
Jason
  • Jason,
    For a measurement of 0V to 3V, you would need to use the PGA, and therefore use the TLA2022. The TLA2021 only can be set to a full scale range of ±2.048V. For something larger in the TLA2022, you could use the ±4.096V range. Note that the ADC is fully differential and that it always measures AINP-AINN, so that 0V to 3V is represented as codes from 0000h to 05DCh so you have less than 11 bits of resolution. You can get the conversion information in the Data Format section in 8.5.3 of the datasheet.
    The schematic that you might use looks something like this:
    There an input RC filter, and bypass capacitance for the supplies. There are also pull-ups for the SCL and SDA lines for the I2C communication.
    The TLA2022 uses a 16-bit configuration register. I would generally use the single-shot conversion mode. Here's a table of the register values for setting a data rate of 3300 SPS and an FSR of ±4.096V.
    Bit Field Value Description
    15 OS 1 Start a single conversion
    14:12 MUX[2:0] 000 Default input
    11:9 PGA[2:0] 001 FSR = ±4.096 V
    8 MODE 1 Single-shot conversion mode or power-down state (default)
    7:5 DR[2:0] 111 DR = 3300 SPS
    4:0 Reserved 00011 Always write 03h
    Additionally, here is some basic pseudocode to show how you might repeatedly read from the device.
    Configure microcontroller for I2C communication, I2C address=1001000 (48h)
    Loop
    Send 90h 01h 43h E3h //
    // Start write to address 48h, write bit 0 (90h)
    // Configuration register 01h
    // Set 43E3h, AIN0-AIN1, FSR=±6.144V, Single-shot conversion, DR=3300SPS, stop
    Wait 353us // Wait for data period, +10% for internal oscillator variation, +20us
    Send 90h 00h 91h xxh xxh // Read back ADC conversion data
    // Conversion register 00h, stop
    // Start read from address 48h, read bit 1 (91h)
    // Read back 2 bytes, stop
    That should get you started. I took some of this from a cookbook circuit that you can find here:
    http://www.ti.com/lit/an/sbaa353/sbaa353.pdf
    While the cookbook circuit is for the TLA2024, I've changed some of it to reflect the use of the TLA2022.
    Joseph Wu
  • Joseph,
    Thanks for your reply.
    Customer finally chooses TLA2024 to detect 4 single-end 0-3V input.
    Please help to review the schematic below.
    Jason
  • Jason,
    From your last post, their application is similar to the following cookbook (except they are using the TLA2024 instead of the ADS1115):
    www.ti.com/.../sbaa408.pdf
    The ADS1115 has a few extra features that aren't in the TLA2024, but they aren't used in the cookbook. Again, they can read through it and see if helps.
    The schematic looks fine. With ADDR connected to GND, the address would be 1001 000. One thing that I wanted to know is what data rate they intend on using. Here the RC input filtering is set to 3.4kHz. Generally, I would set a bandwidth that is about 10x to 20x of the data rate. If they intend on using the 250SPS data rate this is fine. If they intend on using the 3300SPS data rate, I would reduce the input filtering capacitance to 47nF.
    Joseph Wu