Hi,
We are using ADS1298 for EEG monitoring. We are biasing the body using RLD electrode during EEG monitoring.
To measure the impedance of each channel we use the ac current source and sink of the channels. For example, if we want to measure channel 1's impedance, we turn on the current source on its INP as well as the current sink on its INN and measure the voltage across the inputs. Dividing the voltage over the amplitude of the injected current gives us the impedance, which is fairly accurate.
- My question is can we turn off the RLD while measuring the impedance of the input channels?
- Turning off the RLD during impedance measurement, the body should still be biased to mid-supply through the current source and sink. As they have high input impedance (500M-ohm), the input common mode voltage should still be close to mid-supply even when connecting a big load to channel's input, right?
- What is your suggestion on this? I'm asking because in my test, the measured impedances seems to be more accurate when RLD is OFF. Thanks!
Regards,
Milad