I could not find a spec for the # of kV supported by the ADS1248's ESD protection on the analog inputs and GPIOs.
Can someone help me with these numbers?
Thanks.
This thread has been locked.
If you have a related question, please click the "Ask a related question" button in the top right corner. The newly created question will be automatically linked to this question.
Steve,
2kV HBM and 500V CDM.
Best regards,
Bob B
Joseph
Thanks again for your continuing support. If I do add ESD diodes, I have seen two possible configurations: 1) the diodes are arranged back to back and placed across the thermocouple inputs or 2) one diode goes to ground and the other goes to AVDD. In case 1, technically one diode is forward biased, but the thermocouple voltage maximum is only 20mV so not likely to be much leakage. Which arrangement would you recommend? Keep in mind that our thermocouples are wired single ended but we have a +/- 2.5V supply for ADS1248.
I would like to use AVDD as the reference voltage using the LM4132A-2.5, so connecting REFP0 to AVDD and REFN0 to ground. The -2.5V will come from an small inverter such as MAX1720. Any concern with tieing REFP0 to AVDD? I don't think the internal reference is quite good enough to give sub 1C accuracy.
Regards,
Tom
Joseph,
I did want to use the +2.5V as both the external reference and as the positive supply. I can choose a much better reference such as LT6654 or REF5020. The ADS1248 uses only 225mA for analog and REF is 30nA so even 5 of them only need about 1mA. Of course, I can use a separate +2.5V supply and 2.048V reference if you advise that. If I use a separate external reference can the reference connect COM to REFN1 and OUT to REFP1? The eval board seems to have the reference COM to AVSS. Also, I will have 5 ADS1248, so how much capacitance should I put at each one. The eval board shows 10uF, but hard to believe I need that much capacitance at every REFP1 input and every AVDD and DVDD.
I am going to put the ESD diodes between AVDD and GND, but take a look at Fig 44 on the last page of LTC2484 datasheet and you will see both diodes connected to GND. I suppose this is okay because the thermocouple voltage is so low, the forward biased diode does not get turned on.
Thanks again,
Tom
Joseph
I will follow your advice and use an external reference but still provide capacitance at REFOUT just in case I need it. Sorry if I was not clear, but I intend to have a external reference which is 2.048 above GND so would be connecting REFN1 to GND and REFP1 to the external reference output. I assume this is okay, rather than having the external reference have its negative lead to AVSS. As for the capacitance, I was asking how much capacitance I need at AVDD and at REFP1 for supply bypassing. The eval board has 10uF, which seemed like quite a lot, particularly for the reference input. Also, I will have 5 ADS1248 close together.
Again,
Thanks,
Tom
Joseph
Here is a portion of the schematic.
There are 32 thermocouples which are wired single ended. They are wired with 1k + 10uF filter and ESD diodes. The ADS1248 gets +5V from ADuM5401 isoPower with 10uF output capacitor. The AVDD, AVSS and REF supplies are shown below. There are a total of 5 ADS1248 which all share the same analog supply and REF.
My question is do I have the right values of capacitance? Anything else I need?
Thanks,
Tom
Joseph
You have a good point about the ripple. I chose the TPS60403 because it runs at 250kHz and figured the ripple would be beyond the sampling frequency. I think I might add a 10 ohm in series with each ADS1248. The voltage drop is tiny and combined with the 1uf should provide a good deal of filtering of the 250kHz ripple. I agree that the 10uF on the thermocouple inputs is much larger than is needed. My customer used it in a previous design so I did not want to remove it.
Thanks again for all your help.
Tom