hi, i am designing a ads1192 based ecg sensor. can anyone help me to verify this schematic is correct?
This thread has been locked.
If you have a related question, please click the "Ask a related question" button in the top right corner. The newly created question will be automatically linked to this question.
Hi John,
I would stick to 1% resistors and 5% or 10% capacitors in the signal path. This is not a requirement that is specific to the device, just a general guideline. Also, signal chain capacitors should have an NP0/C0G temperature coefficient. The more mismatched your signal path components are, the worse your CMR will be. On the other hand, power supply decoupling capacitors can use 10% or 20%.