Hi,
I am an engineer designing medical hardware.
I have a question about the ADC pin of your product "TM4C1294KCPDT".
We have conducted actual measurements of the circuit under the following conditions, but the results were not as expected.
・Input voltage +3.0V
・Rs = 1.8kΩ
▸Results considered to be expected
・Input impedance RADC: 2.5kΩ, so the input power supply 3.0V causes a voltage drop and the input voltage is about 1.9V.
▸Measured results
・No voltage drop occurred and 3.0V was output.
We would like to know the cause of the above result.