Dears,
Whether the termination configuration parameters, in addition to voltage and current, are related to time, SOC,RM, etc.
Thanks.
This thread has been locked.
If you have a related question, please click the "Ask a related question" button in the top right corner. The newly created question will be automatically linked to this question.
Dears,
Whether the termination configuration parameters, in addition to voltage and current, are related to time, SOC,RM, etc.
Thanks.
Hello Batt,
My customer appears when the RSOC is 10% in the actual test process, but the voltage and current have met the condition of terminating charging(Charge Term Voltage and Charge Term Taper Current), then RSOC jumps to 100% directly.
The customer's cell is lithium iron phosphate, and the full voltage setting is low.
My question is, in addition to Charge Term Voltage and Charge Term Taper Current, what conditions do RSOC,RM and FCC need to meet to terminate charging?
Thank you for your support. I'm looking forward to your reply.
Have they characterized their LFP cell? They need to get the appropriate CEDV parameters first. Otherwise a jump is expected. If you want it to terminate at current taper please set FC set % = -1% (minus 1%).