Hi
My customer discharge the battery voltage to 13.0V (4S1P), set the AFE Shutdown Voltage = 4000, then send Manufacturer Access command Shutdown (0x0010) in the unseal mode, the battery still cannot enter Shutdown mode.
Then discharge the battery voltage lower than 7.5V, the battery can automatically enter Shutdown mode.
Operation Cfg C 0007
[Power(Power)]
Flash Update OK Voltage = 7500
Shutdown Voltage = 11400
Shutdown Time = 10
Cell Shutdown Voltage = 2850
Cell Shutdown Time = 10
AFE Shutdown Voltage = 4000
Sleep Current = 10
Bus Low Time = 5
Cal Inhibit Temp Low = 5.0
Cal Inhibit Temp High = 45.0
Sleep Voltage Time = 5
Sleep Current Time = 20
Wake Current Reg = 03
Why the battery can not in shutdown mode?
Thanks