Other Parts Discussed in Thread: AFE4404,
Hello
We are using the AFE4403 in our device and during EMC Radiated Immunity testing we are seeing low-frequency signals being picked up.
The signals are not at the carrier or modulation frequency, they are at a much lower frequency, around 2-3Hz.
The Carrier is in the range of 2.3GHz with a modulation of 80% 1kHz.
When the interference is enabled, a near perfect 2Hz sign wave is seen on the PPG signal.
On closer inspection of the data sheet, buried on page 30 is the following paragraph:
TI highly recommends that a single clock source be used to generate the clock required by the AFE as well as the clock needed by the microcontroller (MCU).
If an independent clock source is used by the MCU, then any energy coupling into the AFE supply or ground or input pins can cause aliased spurious tones close to the heart rate being measured.
This is not the default set up of the AFW4403EVM, which is surprising as it is "highly recommended", we modified our circuit to drive the clock from the MSP430 and sure enough, the noise frequency appears to have been reduced.
However, we are still picking up level shifts in the signal when the noise is enabled and disabled. We are also still picking up burst modulated noise as well.
Is anyone from TI able to explain the noise coupling effect and why modifying the circuit todrive both Micro and AFE chip from the same clock source has this effect.
We would like to understand the effect, so we can attempt to further mitigate the noise pick-up on the signals.
The explanation is the datasheet, doesn't make sense as the AFE handles the A/D conversion and only SPI interface goes between the Micro and the AFE.
If anyone can explain why it makes a difference we would be very grateful.
Thanks.