Using TDC1000-7200EVM to test the concentration, the temperature and concentration did not change. Why did the TOF value change from 125348 to 125051 after 3 hours?
This thread has been locked.
If you have a related question, please click the "Ask a related question" button in the top right corner. The newly created question will be automatically linked to this question.
Using TDC1000-7200EVM to test the concentration, the temperature and concentration did not change. Why did the TOF value change from 125348 to 125051 after 3 hours?
Hello Hsiao,
Can you please provide some more information about the configuration? Are you using 1Mhz transducer with water as the medium? Do you have a log of the data? I would like to see if this was a sudden shift or slowly changing over time. Last, can you provide your GUI configuration?
| TIME | TOF | Temperature | |
| 1 | 09:58 | 125348 | 19.99 |
| 2 | 10:15 | 125337 | 20.14 |
| 3 | 10:58 | 125319 | 20.24 |
| 4 | 11:30 | 125278 | 20.45 |
| 5 | 12:08 | 125194 | 21.09 |
| 6 | 13:33 | 125015 | 21.14 |
| 7 | 14:14 | 124947 | 21.15 |
| 8 | |||
| transducer | 1MHZ | 15mm*1mm | |
| water |
The distance from Transducer to reflector is 90mm
The medium is water
Transducer 1MHZ diameter15mm*thickness1mmm
Another test record TOF will also decrease.
| TIME | TOF | Temperature | |
| 1 | 13:40 | 124834 | 20.87 |
| 2 | 14:31 | 124823 | 21.9 |
| 3 | 15:16 | 124802 | 21.1 |
| 4 | 15:44 | 124790 | 21.07 |
| 5 | 16:08 | 124782 | 20.94 |
| 6 | 17:02 | 124755 | 20.88 |
| 7 | |||
| 8 | |||
| transducer | 1MHZ | 15mm*1mm | |
| water |
Hsiao,
Thanks for the information. This ToF decrease seems to be related to the temperature. In the first data set, the temperature increase by 1.15C. As the temperature increases, the speed of sound in water will increase resulting in a shorter time of flight. The link below shows a chart of speed of sound vs. temperature. You could consider using the temperature of the medium to adjust your ToF calculation.
https://www.engineeringtoolbox.com/sound-speed-water-d_598.html