Tool/software: WEBENCH® Design Tools
Hello,
I have been done some experiments for testing on a pair of Audiowell senor the ultrasonic sensor using MSP430FR6047 board, to check the reliability of these sensors for non-invasive flow metering application. While observing the result it came to my notice that the DTOF shown by MSP430 GUI is off by a constant factor of 3.35 approx. when compared with actual DTOF.
Velocity (m/sec) | Theoretical DTOF (psec) | MSP GUI DTOF (psec) | Off by factor |
0.6287 | 24168.57 | 6986.5 | 3.459324411 |
0.569 | 21867.73 | 6418.58 | 3.406942034 |
0.536 | 20601.33 | 5818.8 | 3.540477418 |
0.442 | 16972.12 | 5009 | 3.388325015 |
0.3794 | 14583.3 | 4361.64 | 3.343535918 |
0.334 | 12840.53 | 3918.2 | 3.277150222 |
0.29832 | 11466.59 | 3375.93 | 3.396572204 |
0.2241 | 8611.91 | 2647.7 | 3.25260037 |
For calculating Actual DTOF = 2 *L* V / C^2 is used
where, L - acoustic length traveled by the wave i.e. 0.103 m as per my setup.
V - Flow Velocity
C - Velocity of the sound wave in water i.e. 1450 m/sec.
Is there some inbuilt constant setting in GUI of MSP430FR6047 which is effecting the DTOF readings? What the way to change the constant?