I'm also using the ADS2198FE board in this manner (bypassing the J1 connector and sending signals directly to JP30-37 for EMG). I have been successful in generating square waves using the internal signals following the flow diagram on page 70 of the ADS1298 user document, but when I apply an external square wave from a signal generator, the output signal is very noisy. It also seems to have an offset for some reason, as it does not pass through zero (it is either a very large degree above zero or a large degree below zero, both for the positive and negative portions of the wave depending on the signal strength I input). How should I be interpreting the values of the output and is using the RLDOUT for the ground of my signal generator applicable here? Thanks, Adam

  • **TI EDIT February 20th 2019; FOR GENERAL ADS129x ASSISTANCE, PLEASE SEE NEW BIOPOTENTIAL FAQ LOCATED HERE**

    Hello Adam, Can you capture the data using the "Save" tab? Also, try applying your signal directly to pins 2/4 of the input jumpers JP-30/37, like the one below:

  • Hi Michael,

    I am not using the software associated with the ADS1298FE, but I have included some data files for you to inspect.  The channel 1 output data (wvch1.txt) from the internally generated signals as described on p.70 of the manual and its associated plot is shown.  The data has not been converted from 2's complement.  I have also included a set of data from the external signal generator (sqwv.txt) and its associated plot. 

    The leads for the signal generator are soldered to the underside of the board at pins 2 and 4 of JP30 as you described.  My oscilloscope shows the signals to be stable and without noise at entry point on the board, although they are distorted when I probe the jumper pins on the upper side of the board.  I currently have the jumpers in place on the 3-4 pins of JP30-JP37, as well as on JP20, however, removing them does not seem to change the results.  I am also setting all registers to default values, except for CONFIG3, where I am setting it to use the internal reference.

    111111111000000110100100
    111111111000000110101011
    111111111000000101111001
    111111111000000110000010
    111111111000000110010111
    111111111000000110011110
    111111111000000101111011
    111111111000000110000100
    111111111000000101101110
    000000000010010000011011
    000000000010010011101100
    000000000010010011100110
    000000000010010011000110
    000000000010010000010110
    000000000010010011000001
    000000000010010011000011
    000000000010010010111011
    000000000010010011000101
    000000000010010000011100
    000000000010010010110100
    000000000010010010110100
    000000000010010010011011
    000000000010010000010110
    000000000010010011000011
    111111111000000100010010
    111111111000000101110100
    111111111000000101101110
    111111111000000100010000
    111111111000000100001101
    111111111000000101110110
    111111111000000101110110
    111111111000000101001011
    111111111000000101110001
    111111111000000101100010
    111111111000000100010000
    111111111000010011011110
    000000000010010010110111
    000000000010010011000010
    000000000010010011010101
    000000000010010000010100
    000000000010010000011010
    000000000010010011000111
    000000000010010000011101
    000000000010010010111110
    000000000010010011000010
    000000000010010010111010
    000000000010010011000011
    000000000010010010101111
    000000000010010010111001
    000000000010010010110110
    000000000010010010111011
    111111111000000100001101
    111111111000000101010010
    111111111000000100001111
    111111111000000101101110
    111111111000000101100011
    111111111000000101011000
    111111111000000101011111
    111111111000000100001111
    111111111000000100001111
    111111111000000101010100
    111111111000000101001100
    000000000010010000010101
    000000000010010011000001
    000000000010010010101011
    000000000010010010110100
    000000000010010010111101
    000000000010010011000010
    000000000010010000010101
    000000000010010010110100
    000000000010010010110010
    000000000010010010110101
    000000000010010011000100
    000000000010010010110101
    000000000010010011010010
    000000000010010011000101
    000000000010010010101011
    000000000010010010111100
    111111111000000101110011
    111111111000000100010010
    111111111000000101011101
    111111111000000101100000
    111111111000000100001111
    111111111000000101110011
    111111111000000100001111
    111111111000000101011111
    111111111000000101101111
    111111111000000101100000
    111111111000000101101110
    000000000010000101010100
    000000000010010000010111
    000000000010010011000011
    000000000010010010111101
    000000000010010011000000
    000000000010010010101010
    000000000010010011000110
    000000000010010011000001
    000000000010010010111110
    000000000010010010100000
    000000000010010010110010
    000000000010010011000101
    000000000010010010110110
    000000000010010010101011
    000000000010010010111011
    000000000010010010110010
    111111111000000101100010
    111111111000000101011110
    111111111000000100001100
    111111111000000100001101
    111111111000000101111000
    111111111000000100001111
    111111111000000101110100
    111111111000000101110011
    111111111000000101011011
    111111111000000101010101
    111111111000000101100010
    000000000010000101011101
    000000000010010011010010
    000000000010010011000100
    000000000010010000011010
    000000000010010011010010
    000000000010010011010010
    000000000010010010110110
    000000000010010000010110
    000000000010010000010110
    000000000010010010110100
    000000000010010000011001
    000000000010010000011011
    000000000010010000010100
    000000000010010011100001
    000000000010010011000110
    000000000010010010111111
    111111111000000101011010
    111111111000000101000100
    111111111000000101101110
    111111111000000101011001
    111111111000000101101110
    111111111000000100010000
    111111111000000101001101
    111111111000000101111000
    111111111000000101100001
    111111111000000101101110
    111111111000000101111110
    000000000010010011000011
    000000000010010011101000
    000000000010010011011011
    000000000010010010110100
    000000000010010011000001
    000000000010010011010101
    000000000010010010110000
    000000000010010000010110
    000000000010010011000110
    000000000010010011000101
    000000000010010010110011
    000000000010010010111111
    000000000010010011000101
    000000000010010000011101
    000000000010010011000010
    000000000010010000011001
    111111111000000100010010
    111111111000000101001110
    111111111000000100001110
    111111111000000101100000
    111111111000000101100010
    111111111000000101011011
    111111111000000101010110
    111111111000000101011000
    111111111000000101101110
    111111111000000100001010
    111111111000000100001010
    111111111000000101011011
    000000000010010011000110
    000000000010010010111100
    000000000010010011010011
    000000000010010010111100
    000000000010010010101001
    000000000010010010111110
    000000000010010011000110
    000000000010010010111111
    000000000010010010110001
    000000000010010010111001
    000000000010010011010110
    000000000010010000010100
    000000000010010011000111
    000000000010010010111100
    000000000010010011000000
    111111111000000110000100
    111111111000000101011000
    111111111000000110000011
    111111111000000101110110
    111111111000000100010010
    111111111000000101010111
    111111111000000101011000
    111111111000000100001010
    111111111000000101110010
    111111111000000101110001
    111111111000000101111001
    111111111101001011111100
    000000000010010011000010
    000000000010010011010010
    000000000010010000011011
    000000000010010011000101
    000000000010010011000001
    000000000010010000010100
    000000000010010010110110
    000000000010010010110010
    000000000010010010101010
    000000000010010000010100
    000000000010010010101111
    000000000010010010111101
    000000000010010010100110
    000000000010010010110100
    000000000010010010111100
    111111111000000100001111
    111111111000000101010110
    111111111000000101001011
    111111111000000101110010
    111111111000000100001011
    111111111000000100010010
    111111111000000101010111
    111111111000000101010011
    111111111000000101110010
    111111111000000101011000
    111111111000000101011000
    000000000010010011010100
    000000000010010010100101
    000000000010010010110001
    000000000010010011000101
    000000000010010010101000
    000000000010010011000101
    000000000010010000010110
    000000000010010010110011
    000000000010010010111001
    000000000010010011010101
    000000000010010011010101
    000000000010010010111001
    000000000010010011000001
    000000000010010010110001
    000000000010010011011000
    111111111101001100110011
    111111111000000100001010
    111111111000000101011000
    111111111000000101110000
    111111111000000101011100
    111111111000000101010101
    111111111000000101100011
    111111111000000101110110
    111111111000000100001010
    111111111000000101010001
    111111111000000101100000
    111111111000000101010000
    000000000010010010110110
    000000000010010011000101
    000000000010010010111110
    000000000010010011000010
    000000000010010010101000
    000000000010010010101111
    000000000010010010101011
    000000000010010010100000
    000000000010010010100000
    000000000010010010111000
    000000000010010010110010
    000000000010010010111100
    000000000010010011000000
    000000000010010010100011
    000000000010010010110001
    111111111000000101111101
    111111111000000101001000
    111111111000000101001011
    111111111000000101100001
    111111111000000101010000
    111111111000000101000101
    111111111000000100001011
    111111111000000101011110
    111111111000000101011000
    111111111000000101010001
    111111111000000101011010
    111111111000000101001000
    000000000010010010100000
    000000000010010010110011
    000000000010010010111111
    000000000010010010110110
    000000000010010010100111
    000000000010010010110001
    000000000010010010110010
    000000000010010010100110
    000000000010010010011101
    000000000010010010100111
    000000000010010010101010
    000000000010010010110011
    000000000010010010101101
    000000000010010010100010
    000000000010010010011100
    111111111000000101011000
    

    000101000010000000011111
    000101000001100000100101
    000101000010000100110011
    000101000000110111011000
    000101000001101011101111
    000101000000111010111010
    000101000001100001111100
    000101000001001110011010
    000101000010000100001011
    000101000001001101001010
    000101000010101101010011
    000101000010000110000000
    000101000010111000001100
    000101000001111100101101
    000101000011000101101110
    000101000001011010100000
    000101000010000011000111
    000101000001001110011110
    000101000001111101011101
    000101000001110001001011
    000101000010010101111010
    000101000010001010010010
    000101000010100001001101
    000101000010111111101010
    000101000010000000110000
    000101000010001010001100
    000101000001010011100100
    000101000010001010000100
    000101000001100100111000
    000101000010011101011111
    000101000001100010101111
    000101000010011101011001
    000101000010010110011100
    000101000011010011000100
    000101000010100010000101
    000101000011110010100001
    000101000010111000100101
    000101000011011100011011
    000101000010101011111110
    000101000010000100101101
    000101000010000001001111
    000101000001011111110100
    000101000010011100011001
    000101000010011011010111
    000101000011001100111100
    000101100010100000010011
    000101100101101110101010
    000101100011110111011111
    000101100101100000111011
    000101100011110000100111
    000101100101001011000010
    000101100011101110001011
    000101100100001100111110
    000101100011011100011011
    000101100100110101110000
    000101100100000100001010
    000101100101010111100000
    000101100100111001110001
    000101100101001011010110
    000101100101001010010000
    000101100101001110111001
    000101100100110101010000
    000101100100111000101001
    000101100101000100001010
    000101100011111100001110
    000101100100101100010000
    000101100011101000001001
    000101100100111001111011
    000101100100010000011101
    000101100101001110000101
    000101100100111110101111
    000101100100111111111111
    000101100101010100000011
    000101100100100010110110
    000101100101001010011010
    000101100100000100111010
    000101100101011100011011
    000101100100000100001110
    000101100100110100000100
    000101100011111100010000
    000101100101000010010100
    000101100100000101100011
    000101100101011011011101
    000101100100101101011010
    000101100110000111100101
    000101100100101000001010
    000101100100110111111101
    000101100101100011101111
    000101100011110110011011
    000101100101010101000111
    000101100100000000101011
    000101100101001000010110
    000101100011101100100100
    000101100101000001011000
    000101100011101110001100
    000101100101010010101101
    000101100101000101011010
    000101100101100110101011
    000101100101011010100000
    000101100101011011010010
    000101100101010111111001
    000101100100110111110111
    000101100101000100001101
    000101100100101000001101
    000101100101000000001110
    000101100100100001100011
    000101000101010111100100
    000101000011000101011001
    000101000011110001001011
    000101000010101010011010
    000101000100000010000111
    000101000010100100001101
    000101000100000111100000
    000101000010110011100011
    000101000011100100101110
    000101000010100101001101
    000101000010110110101010
    000101000010001101001001
    000101000011010010000001
    000101000010101010101111
    000101000010110110110010
    000101000011011010110111
    000101000011010010000111
    000101000011011000001101
    000101000011101000111010
    000101000011111111110011
    000101000011010101010011
    000101000011100000010111
    000101000010100100000110
    000101000011001001001001
    000101000011000111110000
    000101000011001010000011
    000101000011001011111011
    000101000010101000011101
    000101000011100100001000
    000101000011001100110101
    000101000100010000001101
    000101000011001101110101
    000101000100010100001100
    000101000011011111110011
    000101000100000001010101
    000101000010101100011111
    000101000100000110010000
    000101000010011101011100
    000101000011001101001011
    000101000010110110010111
    000101000011010000111000
    000101000010100101111100
    000101000011101100010010
    000101000010111101101111
    000101000011011110010000
    000101000011111001010000
    000101000011000000100001
    000101000011110011011111
    000101000010001001011100
    000101000011001011100110
    000101000010010010010111
    000101000011000000001010
    000101000010100000110001
    000101000011001010101110
    000101000010100000110011
    000101000010111111111110
    000101000010111111110100
    000101000011010100001101
    000101000100010000001000
    000101100011111011111010
    000101100101100000011100
    000101100101000110001011
    000101100100101100001100
    000101100100001101001010
    000101100101000100001011
    000101100100000010110010
    000101100100110001110001
    000101100100100100010000
    000101100101001110010000
    000101100100010101000011
    000101100101011000010001
    000101100100101100110110
    000101100101011100011001
    000101100100101100111101
    000101100100100001000000
    000101100100100000010010
    000101100100100010000010
    000101100100011010100100
    000101100100001011111111
    000101100100000010101000
    000101100100011001011001
    000101100100101011000110
    000101100100010101110100
    000101100101101000010110
    000101100100010011101100
    000101100101000101001100
    000101100100101000000111
    000101100100010011011110
    000101100100001100110100
    000101100011011110101101
    000101100100001100000000
    000101100011100101010100
    000101100100010100110101
    000101100011111100101001
    000101100100110110010010
    000101100100000000011011
    000101100101100000010010
    000101100100011110001000
    000101100101100010100001
    000101100100000110100010
    000101100101010110111011
    000101100100000110001011
    000101100100100001001000
    000101100100001111100000
    000101100100110010001100
    000101100011111000011001
    000101100100010011010011
    000101100100100000110001
    000101100100000100011000
    000101100100111110011111
    000101100100001100000010
    000101100101100111101001
    000101100100100000110010
    000101100100110100101001
    000101100100110111101011
    000101100100101000110111
    000101100100100000010101
    000101100010101001111111
    000101000010101010000100
    000101000010010100011101
    000101000010101110100001
    000101000010010111100011
    000101000011001010001011
    000101000010001001000101
    000101000011011010111101
    000101000010011100110111
    000101000011001011100100
    000101000010110011100101
    000101000100000011111001
    000101000010000010011010
    000101000011010000101010
    000101000010101111011001
    000101000011001010011111
    000101000010001000011101
    000101000011100110000010
    000101000011010101100010
    000101000011011100100001
    000101000100001110100001
    000101000011001100001111
    000101000010110000101100
    000101000011001110111000
    000101000011101100000001
    000101000011000101110010
    000101000011011100101000
    000101000010011100001111
    000101000011000100010101
    000101000010010001100011
    000101000011000010011100
    000101000010010100000110
    000101000011110110011110
    000101000010010000101111
    000101000011010000001111
    000101000100100010001110
    000101000011010000100111
    000101000011101010110100
    000101000010010011100001
    000101000011101011000011
    000101000010100100010100
    000101000011011000111101
    000101000001111100010000
    000101000010110001010111
    000101000010100000010101
    000101000011011010100111
    000101000010111100101001
    000101000100010111010111
    000101000011011111110101
    000101000011110000100010
    000101000011010110001000
    000101000011100111100001
    000101000011010100001010
    000101000011010100101101
    000101000011001100111110
    000101000010100001010000
    000101000011100001010001
    000101000010110110001010
    000101000011011101010011
    000101000100101001011100
    000101100101100110100011
    000101100101010011000101
    000101100101101000111011
    000101100110001011011111
    000101100101110000011101
    000101100101100011111101
    000101100101000010001010
    000101100101011001010101
    000101100100111111011100
    000101100101011101000000
    000101100100100110000000
    000101100101111111111011
    000101100100101100001111
    000101100101111000011011
    000101100100111000010011
    000101100101110101001110
    000101100101001000011000
    000101100101010100011011
    000101100100011111011100
    000101100100111110001110
    000101100100110100001010
    000101100100110011111101
    000101100100011000010110
    000101100100100001111010
    000101100100110001110110
    000101100100011010101010
    000101100101000101110011
    000101100100101000110000
    000101100110000000000100
    000101100100111011101010
    000101100101110011000110
    000101100100110100001010
    000101100100111011100100
    000101100100010100100001
    000101100101100100111101
    000101100100110110100010
    000101100101000011000100
    000101100100100011000111
    000101100110000000010111
    000101100100111101111010
    000101100110001010011111
    000101100101101000000010
    000101100101101000101100
    000101100101101100101101
    000101100101110100111010
    000101100101010111000010
    000101100101001011011110
    000101100101001101110100
    000101100100001011111111
    000101100101101010101100
    000101100100101000010001
    000101100101011011111011
    000101100101000101101110
    000101100101101010011010
    000101100100111110001001
    000101100110000100110111
    000101100101100111101101
    000101100110111010100100
    000101100101101011111011
    000101000010101100111100
    000101000011100010100111
    000101000010110011010010
    000101000011100100101010
    000101000010100011011011
    000101000100001111111001
    000101000011001011011111
    000101000100000010001111
    000101000011100010011010
    000101000100110111111111
    000101000011101100101101
    000101000100010001110011
    000101000011001100001110
    000101000011001100101010
    000101000010111000000010
    000101000011011010000110
    000101000011000001100000
    000101000010110110011000
    000101000011110000110011
    000101000011100000001100
    000101000011011011000010
    000101000011100111011011
    000101000011111100001111
    000101000011001000110100
    000101000100010110101100
    000101000010011001111110
    000101000011110100000101
    000101000010111100011101
    000101000011011011000010
    000101000010011011011100
    000101000011011010110100
    000101000010101000000110
    000101000011010111110000
    000101000011001100000010
    000101000011111110011101
    000101000011010111011010
    000101000011110000101110
    000101000011110001100001
    000101000011001100011011
    000101000011011111010110
    000101000010001010101000
    000101000011011110101001
    000101000001110001110000
    000101000011011010101001
    000101000010011100000000
    000101000011111000111011
    000101000010100111100010
    000101000100000111101111
    000101000010111110011111
    000101000011100000111110
    000101000011010000010000
    000101000011010111011100
    000101000011010011010010
    000101000011001011100101
    000101000010111100010100
    000101000010101100010001
    000101000010100001010011
    000101000010010100010011
    000101000010111010110010
    000101010011101010111101
    000101100101011000100001
    000101100100110100001011
    000101100101010011010101
    000101100100110000010100
    000101100101000111110000
    000101100100101000011010
    000101100100010100001000
    000101100100010100010011
    000101100100000100000000
    000101100100100010101110
    000101100100001101010110
    000101100100110101010011
    000101100100000110111000
    000101100100111001110101
    000101100100011101100011
    000101100101101011101101
    000101100101000000000001
    000101100101101011111100
    000101100100100000100000
    000101100101001111010100
    000101100100010111100100
    000101100101000000011010
    000101100011100000011100
    000101100100101100110010
    000101100100100001000011
    000101100100100111110100
    000101100100110100011100
    000101100100111100100001
    000101100101011110011111
    000101100100110011010101
    000101100101100000110110
    000101100100100000111110
    000101100100111010011010
    000101100100000010010010
    000101100100110110000000
    000101100100001000001111
    000101100101011001011101
    000101100100001000000101
    000101100101011111110000
    000101100100101010000101
    000101100001110000000001
    000100101010010001000111
    000100101000101111110010
    000100101000100011101111
    000100101000100110101011
    000100101000110100011000
    000100101000100100101001
    000100101001010100101111
    000100101000000110010011
    000100101001000010011010
    000100100111110011011101
    000100101001000111110000
    000100100111011000010111
    000100101000100101000100
    000100101000000100011101
    000100101001001001100010
    000100101000001001000001
    000100101000110011110101
    000100001011100000110101
    000100001011011110100100
    000100001010011100001011
    000100001011001011110100
    000100001010100101110111
    000100001011111010111001
    000100001011000110011101
    000100001011100101111000
    000100001011001100100010
    000100001010101110111010
    000100001011111000100110
    000100001011001100011011
    000100001011101011011011
    000100001010010110100111
    000100001011100000000100
    000100001010001111111111
    000100001011110110001110
    000100001001110100001001
    000100001011001100011010
    000100001001011111100100
    000100001010011110111001
    000100001001110011111101
    000100001010111000000110
    000100001001111001001101
    000100001010011100001111
    000100001001100111010111
    000100001010000100011111
    000100001001100111000011
    000100001010001100011011
    000100001010111000100010
    000100001001111101011010
    000100001010111000010110
    000100001001001101001011
    000100001010101110010000
    000100001001010000000111
    000100001010111100011011
    000100001001001100001000
    000100001010010110000111
    000100001001001010011101
    000100001010100110001100
    000100001001100001111111
    000100001010111011011101
    000100001001011110100001
    000100001010100010111011
    000100001001010101000011
    000100001001111011100110
    000100001010010010111010
    000100001010001001010011
    000100001010010100011101
    000100001001111111101110
    000100001001111011100101
    000100001001110111011000
    000100001010110011100100
    000100001001111110100100
    000100001011100100001100
    000100001001110110000010
    000100001011010010011001
    000100001011000000011100
    000100101001011010101000
    000100101000010010011010
    000100101001100011111110
    000100101000001010010110
    000100101001000100100111
    000100101000000100001010
    000100101001010111111000
    000100101000011111110010
    000100101000111000011000
    000100101000110011100111
    000100101000010110110010
    000100101000001011100011
    000100101000001110111000
    000100101000100101010101
    000100101000011011100011
    000100101001000101001001
    000100100111111100101100
    000100101000000111000100
    000100100111110100010000
    000100101000010110001010
    000100101000001011101000
    000100101000000010010001
    000100100111100000011001
    000100100111010010010000
    000100101000110101001101
    000100100111100011111010
    000100101000110010101000
    000100100111101100011001
    000100101000100100001110
    000100100111001010000110
    000100101000000011010100
    000100100111011101110110
    000100101000101010111100
    000100100111110011100000
    000100101000100011000111
    000100100111011010000000
    000100101000000001001001
    000100101000010100011010
    000100101000010010000000
    000100100111111101010100
    000100100111101010000101
    000100100111111101111001
    000100100111011010111001
    000100101001000001110100
    000100101000010011110000
    000100101000100101001100
    000100100111111111011011
    000100101000101010010101
    000100100111110010011100
    000100101001001111010010
    000100100111111100001001
    000100101001101101110011
    000100101000011011100010
    000100101001100011000101
    000100100111111101001001
    000100101000110000101111
    000100101000000110111011
    000100101000011101000010
    000100101000011010110110
    000100001010011000101100
    000100001001101100010111
    000100001001111001111000
    000100001010001010111101
    000100001001110100011101
    000100001001111110001011
    000100001001001010000110
    000100001001100001010011
    000100001000111000010000
    000100001010011101011111
    000100001001010101000101
    000100001010001110111110
    000100001000111011111000
    000100001010001000110010
    000100001000111110001000
    000100001010011011010010
    000100001001010110101101
    000100001010100110000111
    000100001001111000001110
    000100001001110001000011
    000100001001000111100100
    000100001001111101100010
    000100001001100100100010
    000100001001110000001110
    000100001001110100111101
    000100001001000100001111
    000100001010000111100011
    000100001001001011010111
    000100001010001000100011
    000100001001000001001010
    000100001001110011101000
    000100001000111000001101
    000100001001110000100010
    000100001000110111111110
    000100001010000000011000
    000100001001010101001101
    000100001010011001010000
    000100001000110110100001
    000100001011000001110000
    000100001001111100010110
    000100001010011111010010
    000100001010100100111100
    000100001010011101010111
    000100001010111001010001
    000100001010011100011101
    000100001011100100010010
    000100001010101001110001
    000100001011000100001111
    000100001010001100010110
    000100001011011001000110
    000100001010010000010010
    000100001011100011110110
    000100001010000111011011
    000100001100001010100100
    000100001010011001111111
    000100001011100100011110
    000100001001101111100000
    000100001010111000001010
    000100001001101111111100
    000100010100010111101111
    000100100111111111110000
    000100101000101000100111
    000100100111101000011111
    000100101000100111110001
    000100101000110000110101
    000100101000011100001111
    000100101000100001111100
    000100100111101101001010
    000100101000011100010011
    000100100111011110000100
    000100101000001010010110
    000100100111100100111100
    000100101000011010111000
    000100100110111101000011
    000100101000001100001100
    000100100111001011110110
    000100101000010001111001
    000100100111111101010010
    000100101000010100110101
    000100100111110110010100
    000100100111101001000001
    000100100111110011101110
    000100100111011010001100
    000100100111110100011101
    000100101000000100101010
    000100100111110100010111
    000100100111011110000100
    000100101000001111111011
    000100100111110100001010
    000100101000100010011000
    000100101000000000001000
    000100101000100110100001
    000100100111100010111100
    000100101000100010011101
    000100100111101111110100
    000100101001101001001111
    000100101000011111000001
    000100101001010100001110
    000100100111111100101000
    000100101000110001111001
    000100101000010100101010
    000100101001000101110001
    000100101001000001110101
    000100101001001110101110
    000100101001000001001111
    000100101000001001011101
    000100101000111101001001
    000100101001000011111000
    000100101000101001011000
    000100101000111010110100
    000100100111110110101100
    000100101001000101110100
    000100100111011100110100
    000100101000110010101010
    000100100111110011101100
    000100101000110011110000
    000100100111010000001111
    000100101000110000000010
    000100010110111100101111
    000100001001111100100100
    000100001000111101011011
    000100001010100001010010
    000100001001001010011010
    000100001001110011100100
    000100001001100011010011
    000100001001100100000110
    000100001001100100101000
    000100001001101011010100
    000100001001111010110110
    000100001000110000000011
    000100001010001011101101
    000100001000101000101000
    000100001010001100011101
    000100001001001010100011
    000100001001111100111000
    000100001000011001010110
    000100001010000110010110
    000100001000111110101111
    000100001010011100011001
    000100001001100001101111
    000100001010011111010101
    000100001001010010011011
    000100001010010011010110
    000100001000111111000000
    000100001010000000101111
    000100001010000000101011
    000100001010010111110111
    000100001001100000010101
    000100001001111000010001
    000100001010100110010100
    000100001001011001001110
    000100001010101000100010
    000100001010010010011010
    000100001011000011011000
    000100001001111101001101
    000100001010010111000001
    000100001010101110011010
    000100001010111110000101
    000100001011110011101100
    000100001011001000111011
    000100001011110110101110
    000100001010000101000011
    000100001011000000001110
    000100001010010001000110
    000100001100001110100000
    000100001010100101011011
    000100001011100111111000
    000100001001100001100010
    000100001010111011101001
    000100001001110000000111
    000100001011100101111011
    000100001001001010110100
    000100001001110101100011
    000100001000111100100010
    000100001010011110010001
    000100001001111001111010
    000100001010010101110101
    000100100111010111101110
    000100100111101100011000
    000100101000011001001101
    000100101000000001000000
    000100100111111111101111
    000100100111100101000110
    000100101000100110110011
    000100100111100111111001
    000100101000001010110111
    000100100111010010000110
    000100100111111101100000
    000100100111001000100110
    000100101000100100001110
    000100101000001101100011
    000100101000011000111000
    000100100111010100010101
    000100101000101010011011
    000100101000000000101110
    000100101000011111011100
    000100101000010000100111
    000100101000010011111101
    000100101000100100101001
    000100101000000001001001
    000100101001001111100010
    000100101000011010101001
    000100101001001000010011
    000100101000001100001010
    000100101001001101010100
    000100101000000111000010
    000100101001100110101100
    000100101000110101001101
    000100101010010110001101
    000100101000011000011000
    000100101001010110110001
    000100101000000100001100
    000100101001100000101011
    000100101010001010010111
    000100101001001000100001
    000100101001000011111110
    000100100111011110101011
    000100101001001111101010
    000100101000001110010011
    000100101001111101011000
    000100101000100100100100
    000100101010000110100111
    000100101000101101110100
    000100101001010100110001
    000100100111111111101010
    000100101000111110111011
    000100101000011101011010
    000100101001011010111100
    000100101000000001111101
    000100101000001010011100
    000100101000000000100110
    000100100111010100101000
    000100101000100001011011
    000100101000010001110010
    000100101000101001010000
    000100010100001000000000
    000100001010001101111101
    000100001001000100010100
    000100001010100111000010
    000100001000110000101011
    000100001010000110001111
    000100001000111101010010
    000100001010001101011101
    000100001001101011011000
    000100001010001011010111
    000100001001111001001010
    000100001010010000100110
    000100001010001011010010
    000100001010000110111100
    000100001010000011100101
    000100001010001010001101
    000100001010101110000011
    000100001010110000101000
    000100001011001100110001
    000100001001110111000100
    000100001011001111101111
    000100001010001011011101
    000100001011111011110101
    000100001010001001011111
    000100001011100101010110
    000100001001111011100011
    000100001011101000010011
    000100001010101001110101
    000100001100001101000100
    000100001010110000011001
    000100001011011110010111
    000100001001111010111101
    000100001010110001000111
    000100001010100001010100
    000100001011010100001011
    000100001011000001010100
    000100001010100101110011
    000100001010110010100011
    000100001001100000010111
    000100001001100000011100
    000100001000100011101001
    000100001010101000001000
    000100001001110100100011
    000100001001010100001011
    000100001001111100011010
    000100001000110111110101
    000100001010100010000101
    000100001001101000011101
    000100001010000001110111
    000100001000110100111101
    000100001001101010011000
    000100001000011100001010
    000100001010100000010011
    000100001001000100011011
    000100001010011101011111
    000100001000011100010100
    000100001001110100001110
    000100001001000000011010
    000100001001110110110111
    000100001001010000010110
    000100001010100111100111
    000100100001001110011110
    000100100111110011011000
    000100101000010110011110
    000100101000100000111001
    000100101000110100010111
    000100101000000000111001
    000100101000011011000101
    000100101000010100101000
    000100101000110000010101
    000100101000001001001101
    000100101001010000010110
    000100101000010010010000
    000100101001100100000011
    000100101000001000000100
    000100101001100111101101
    000100101000011011111100
    000100101001111101000011
    000100101001011000111111
    000100101001111011111100
    000100101000101010111110
    000100101001100010000010
    000100101000100111101100
    000100101000111101010100
    000100101001000010101111
    000100101000111001011110
    000100101000011100010011
    000100101000100100001101
    000100101000110010110011
    000100101000001110000100
    000100101001010100011110
    000100100111111001110111
    000100101000011001001001
    000100100111011101110010
    000100101001000110001111
    000100100111101100010010
    000100101001000000111001
    000100100111100101100001
    000100101000011000001101
    000100100111001001110111
    000100101000011001010100
    000100101000001000101011
    

    Please let me know if you require any more information.

    Regards,

    Adam

  • Hi Adam,

    By page 70 of the manual, are you referring to the data sheet for the ADS1298 revision 'F'?  The data from your 3113_wvch1 does seem to have a little more offset than what I am used to seeing but the magnitude seems reasonable.  What hardware are you using to capture this data?  How do you have the registers of the ADS1298 configured?

    Your input square wave has some strange DC levels/offsets.  When the entire txt file is plotted out, I see something like this:

     

    Can you possibly grab a ground referenced oscilliscope shot of the signal you are trying to apply to the ADS1298 inputs?

  • Hi,

    Sorry for taking so long to reply, but I didn't get any email notifications that you had made this last post.  Anyhow, I have taken a step back and acquired a new ADS1298FE setup. I removed the jumpers on JP30, then applied my signal (a square wave, input.bmp).  I have attached the save files from the ADS1298FE software. At this stage, I would like to see the proper output with the ADS1298FE and MMB0 for both my signal generator and then actual EMG before I go back to my original system design.

    To answer your questions, yes, I was referring to ADS1298 revision 'F'.  In my main system design, I am using a PIC24F mcu on am Explorer 16 board, saving the data to a SD memory card.  The registers were all set to default values, except I had internal reference powered on (config3).

    Not sure why you have half the data dropped down on your plot, I don't remember seeing that.

    Thanks,

    Adam

    6201.Device_0_Analysis.xls

    5488.Device_0_Codes.xls

    4152.Device_0_Volts.xls

  • FYI, the signal generator does not generate differential signals.  I have a signal and a ground from a coax output.

  • Hi Adam,

    I've been going over you recent and previous posts again and I really don't see where the problem might be.  Your scope shot (input.bmp) implies that you've got a +/-7V square wave at 500Hz applied to the input.  I'm going to assume there's something off about the 500.0us/ being the time scale and that perhaps you have the probe attenuation incorrectly set.  I've set up my board(s) the same way you have, here is my register setting tab:

     

    After plotting your Device_0_Volts.xls file, I figured you were inputting something on the order of a 1Hz square wave.  I see 2 cycles at 500SPS over 1000 samples, so here is your data plotted:

    I also put in a 1Hz square and sine wave, directly to pins 2 and 4 of JP37 (with no other jumpers loaded) from my signal generator (a Stanford Research DS345) and got the following results:

    So - where do we go from here to get to the bottom of this issue?  I think the easiest thing to do is to try and arrange a quick call to discuss exactly how you are applying your input signal as well as how you have the voltage rails configured on the ADS1298ECGFE-PDK board.  I'll have a quick chat with the administrators of this forum and get in touch with you directly.

     

    adam_Volts.xls
  • Hey Adam,

    I was hoping you could help me out with the step by step process you took when using the ADS1298ECGFE-PDK for EMG use. I know it has been a while, but let me know if you can help me out.

    Thanks,