In addition to the thread: https://e2e.ti.com/support/data-converters-group/data-converters/f/data-converters-forum/691122/ads1299-electrode-impedance-measurement-with-ads1299
I want to ask a few questions:
1. As stated by Ryan Andrews in the above thread that ADS1299 definitely won't give accurate impedance. How much error percentage does it give generally?
I analysed impedance by attaching a 100kOhm resistance on the 8th channel. It gave me an impedance of 192k ohm. For 1k ohm, it gave an impedance of 5kOhm. I have setup the circuit and calculated the impedance as stated by an openBCI blog: http://eeghacker.blogspot.com/2014/04/openbci-measuring-electrode-impedance.html
2. During AC impedance setting, at any frequency (31.2Hz or 7.2Hz) as per datasheet there should be a square wave of 6nA Vpp (if 6nA is set in LOFF register). But on the digital multimeter and Tektronix oscilloscope, it showed only 5.65nA Vpp. Is that fine? Is that what is causing the errors in the impedance results I am getting.
3. In the impedance testing, is the Lead-off comparison performed by the same PGA which is used for EEG or those comparators are separate? I couldn't see anything about this in the datasheet. The only thing visible in the block diagram is PGA.