AFE4960: Single channel; 2 electrode configuration help?

Part Number: AFE4960

Seeking configuration help; desired configuration is 2 electrode with single channel ECG and respiration.  Seems like it should work just fine, but configuring the options is challenging.

Lots of settings and very confusing to configure.  Example: should this configuration use RLD to one of the electrodes?

Thanks!

Bill

  • Hello Bill, 

    We highly recommend using RLD in order to properly bias the body. Without biasing the body, CMRR will suffer resulting in reduced ECG signal fidelity. 

    For a recommended configuration, I'll be emailing you offline as this technically a restricted device. 

    Kind regards,
    Nick