We are designing an ECG monitoring device where the requirements a.o. are:
- all limb leads (I, II, III, aVR, aVL, aVF) are derived in analog domain
- respiration measured by impedance pneumography is provided
For this reason we are evaluating ADS1298R on ADS1298RECGFE-PDK board.
It is connected (jumpers rearranged and additional connections added) in the following way:
- IN1N and IN1P - on-board respiration circuitry
- IN2N - R (RA)
IN2P - L (LA) - IN3N - R (RA)
IN3P - F (LL) - IN4N - L (LA)
IN4P - F (LL) - IN5N - N/C
IN5P - L (LA) - IN6N - N/C
IN6P - F (LL) - IN7N - N/C
IN7P - R (RA)
Augmented leads (channels 5, 6, and 7) are configured according to section 9.3.1.7.3.1 Augmented Leads of the ADS129x data sheet.
WCT amplifiers are set as follows:
In general the circuitry provides satisfactory results.
However, when the patient (or the patient simulator) is not connected, channel 4 (F-L measurement, thus lead III) measures enormous noise signal while other leads (CH2, CH3, CH5, CH6, and CH7) are flat lines - what can be the result of such inconsistent behaviour?
In such configuration (lead I - CH2, lead II - CH3, lead III - CH4) how should I configure RLD registers for the best results, as RLDN2, RLDP2, and RLDP3 only or rather RLDN2, RLDP2, RLDN3, RLDP3, RLDN4, and RLDP4? From what I checked it does not have any influence on the lead III behaviour.
Thank you.