ADS131M08: ADS131M08 for ECG application

Part Number: ADS131M08
Other Parts Discussed in Thread: ADS1296, ADS1294, ADS1292, ADS1293, ADS1291, ADS1278, ADS1298, , ADS1198, ADS1258, TLV4333, ADS131E08,

I an designing a low cost 12 lead ECG machine . For that option from TI is ADS 1298 , ADS 1198 ,ADS 1258 , ADS1278 (8 DIFF CHANEL) , ADS1293( 3 nos) and ADS1292,ADS1291, ADS1294 , ADS1296 ( 1,2,4,6) Channel.

ADS1298 , ADS1258, ADS 1278 are  expensive and ADS1198 is 16 bits resolution & other multiple combination of 1,2,3 , 4 channel A/D alo proves costly. One option is to us low cost ADS131M08 for ecg purpose. Same question you have answered as follow 3 years back

quote


That being said, my concern with choosing the ADS131M08 is the low input impedance. ECG applications are usually challenged by large skin-to-electrode contact impedance. You want the input impedance of the ADC to be very high in order to preserve the signal amplitude and to minimize leakage currents.

They would need an external buffer with high input impedance in front of the ADS131M08. For ECG applications, I don't think this is the right approach as it may lead to CMRR degradation.

Has the customer considered some of our dedicated bio potential front-end ADCs like the ADS1298?

unquote

But what if I  choose a op-amp buffer like TLV4333IDR for buffering input (unit gain buffer) ecg electrodes before ADS131M08.

TLV4333 - input impedance 10 G ohms

                    CMRR = 110-125  d

                      input offset < 30 uv

                     Gain−Bandwidth   350 kHz (

            Low Offset Drift: 0.07 V/°C

Please advice

  • Hi ,

    You can use an amplifier to achieve a high input impedance, or you can directly use ADS131E08 which has a higher input impedance. Some of customers used ADS131E08 and ADS129x in their design.

    BR,

    Dale 

  • I mean can I use ADS131M08  /  ADS131E08  for ECG purpose? Because previously on this forum it is advised  not to use  ADS131E08 even with buffer. See below from reply from TI for same question

    "They would need an external buffer with high input impedance in front of the ADS131M08. For ECG applications, I don't think this is the right approach as it may lead to CMRR degradation.

    Has the customer considered some of our dedicated bio potential front-end ADCs like the ADS1298?"

    Please confirm once again I can use ADS131M08  /  ADS131E08 for EGC (medical) purpose ?

  • Hi ,

    As I said, some of customers used ADS131E08 and ADS129x in their design including ECG application. I did not see ADS131M08 was used for this application, but maybe I was not aware of it. You could check with ADS131M08EVM + Op amp EVM.

    - Dale