Dear all,
I am designing an ECG device which is using metal electrodes (the resistance is much higher than gel electrodes).
Also, the circuit embeds a defibrillation protection circuit between the electrode and the anti-aliasing filters (see. circuit attached).
The used values are those from the evaluation board for anti-aliasing filters (two RC filters in a row :10kohm+47pF and 22.1kohm+47pF) and RLD (1Mohm // 1.5nF).
When the contact is getting slightly poorer between the electrode and the patient, the baseline is moving extremely quickly and the ECG becomes uninterpretable.
Please note that the slow baseline drift will be corrected in software and is NOT an issue at all.
I suspect the RLD : do you also share this feeling ? If so, would you decrease the R and/or C value to accelerate the response ? How much ?
Thank you,
Paul-Alexandre,