ADS114S08EVM: Data collection not done after data analysis

Part Number: ADS114S08EVM

Tool/software:

When I enter the following data in the register map and I do the data analysis I don't gett any result and the programme stops responding. I am working with a Windows 11. Does anyone know what may be happening?

  • Hi Sara Arcelus,

    Welcome to the E2E forum! The most common issue we have seen is not waiting long enough for the total number of conversions to complete.  Looking at the configuration settings you are using the default data rate which is 20sps.  If you are also using the default settings in Data Analysis (2048 samples), you need to patiently wait until all conversions complete (102 seconds).  You should see one of the LEDs flashing while the conversions are taking place (about once a second).  If you try and send commands, or press buttons during the conversion time you will lock up the device as the firmware/software does not have the ability to store commands in a FIFO.  Also, the data will not display in the Data Analysis window until all conversion data are collected.

    Best regards,

    Bob B