We are currently using the TX7332 as part of our prototype validation for a portable ultrasound device, and we have encountered an issue while configuring the channel delay.
Our setup involves the following basic settings:
- On-chip beamforming mode
- BF_CK: 100 MHz
- GBL_DELAY: 2 µs (TX_START_DEL: 13)
- TROFF_DELAY: 0.2 µs (TR_SW/OFF_DEL_G1/2/3/4: 2)
- TRON_DELAY: 1.1 µs (TR_SW_DEL_G1/2/3/4: 26)
- TR switch turn-on delay:TR switch turn on delay is counted from last pulse transmission (TR_SW_DEL_MODE: 1)
In the attached echo diagrams of the 32 channels, I set the delay for channel 11 to 30,000 ns, while the delays for the other 31 channels are set to 0 ns. It appears that channels 17 to 32 are receiving two echo signals as expected, but channels 1 to 16 all seem to exhibit a delay of 30,000 ns, even though only channel 11 was configured with this delay. In addition, when the delay of 32 channels is set to be the same, the ultrasound echo signals are normal
We would greatly appreciate your assistance in troubleshooting this issue.