Tool/software:
I am putting ADS8694 in one of my design.
The input is bipolar from -2V to 2V. and i am using the internal reference 4.096V with input range setting 0.625 x 4.096= +/-2.54V.
out of 18 bit data output 16 bit are changes as per input changes. but the 2 bits remains unchanged as input varies.
I am not getting why these two bits remains unchanged.
Please suggest the solution.
As per the last discussion,
I am attached here, the schematic and the logic analyzer plot for the 2.255mV and 1500mV input for your reference.
In the logic analyzer data channel 0 is SDO, Channel 1 is SCLK, Channel 3 is SDI.
Please let me know which 18 bits should I consider as data.
Hi Tom,
See the attached logic analyzer plots with CS line at channel 4, SDO at channel 0, SCK at channel 1 and SDI at channel 3 for the voltage -0.4960 V and 0.5024 V.
Please let me know which 18 bits should I consider as data output.
So...
What mode are you trying to use with the ADS8689? You have 3 16-bit transfers on MOSI with data on the 3rd 16-bit transfer that I don't understand. On MISO, it looks like the ADS8689 is responding to what it thinks is a NOP command with SDO starting on the first clock of the second 16-bit transfer. There appears to be a glitch on MISO (@19.5uS) as well.