Tool/software:
Hello,
How could I extend capture time to Save for instance 5 minutes long ECG signal for ADS1298ECGFE-PDK? Is it possible to modify related application software?
regards
Elzbieta
Hello Elzbieta,
Thank you for your post.
Unfortunately, any modifications to the EVM software cannot be supported by our team. Having said that, 5 minutes should be possible with the existing software. For example, at 1kSPS, 5 minutes equates to 300,000 samples, which is well within the memory limits of the MMB0 motherboard.
Regards,
Ryan
Hello Ryan,
How to obtain the mentioned time of capturing ECG signal with existing software?
Regards
Elzbieta
Set the number of samples to collect to 300,000 and channel data rate for 1kSPS. Then the EVM should run for 5 minutes once you click Capture. More time may be needed for the data to transfer to the host via USB.
Regards,
Ryan
Thank you a lot Ryan. Where could I find the option defining the number of samples?
Regards
Elzbieta