I am using ADS1292R as 2 lead ecg(LA & RA) and respiration sensor. How do I bypass the third lead (RL) and make necessary changes in the register values. I want to extract both ECG and Respiration
Tool/software:
I am using ADS1292R as 2 lead ecg(LA & RA) and respiration sensor. How do I bypass the third lead (RL) and make necessary changes in the register values. I want to extract both ECG and Respiration
Hello Pallavi,
Thank you for your post!
In short, using only two electrode leads without the RLD means there is no established DC bias voltage on the body. You will then need to AC-couple the ECG inputs (CH2) as well as the respiration receive channel (CH1).
For example register configurations, please refer to the ADS1292R data sheet and related EVM User Guide.