Other Parts Discussed in Thread: AFE4960P
Tool/software:
development environment: 1.Windows OS11 PC
2.AFE4960P EVM
3.Bio-Sensing
4.ECG simulator
May I ask
This thread has been locked.
If you have a related question, please click the "Ask a related question" button in the top right corner. The newly created question will be automatically linked to this question.
Tool/software:
development environment: 1.Windows OS11 PC
2.AFE4960P EVM
3.Bio-Sensing
4.ECG simulator
May I ask
Hi, Apologies for the delay. Our main technical support is out of office currently. I will try to get back to you asap, but we will answer your query next week at the latest.
Hello Deng,
Thank you for your patience. Do you still need help with this post?
Regards,
Ryan
Hi Deng,
The raw data is available in the GUI by default. To plot this data in raw ADC codes, choose the "ECG_CHx_RAW" data from the drop-down menu on the left side. Once the raw data is plotted, you can export to a CSV file by right-clicking on the plot window:


There is no such "1-mV" calibration through the EVM GUI. Your patient simulator should be calibrated to output the expected amplitude. There may be some signal loss between your simulator output and the input pins of the AFE. Can you replace the simulator with a known function generator and give a 100-mVp sine wave input? How much deviation are you seeing?
Regards,
Ryan
Hi Ryan