I have a problem to get an ECG from ADS1198
The processor works great in terms of programming and reading the data, but the gain is one which does not work for me.
The maximum gain for the ECG signal is 12 and a ECG skin electrodes voltage is 0.5 - 4 mV which means with a gain a 12 the maximum ECG voltage is 48 mV.
The ADC voltage reference is 4000mV and with 16 bit resolution I get almost 0.
It is no active signal getting into the ADC.
This is a real problem; how can I get a better signal resolution with such small gain?
Is this component really used in medical devices?
If I use a high ECG signal up to 300mV than is fine, but the skin ECG isn’t that low.
What's the work around?
It looks that a middle gain stage is missing from the device.
Can you advice me, please?
I have read on another post: http://e2e.ti.com/support/data_converters/precision_data_converters/f/73/t/161968.aspx
About using a pre-amp stage in front of the ADS1198. Is the pre-amp the best way to address the issue?
Are there any white papers from TI?
Thanks
Jamie