We want to use B-type thermocouple to mesure temperature with ADS1248. How to estimate accuracy/resolution(using ADS1248 internal ref and temperature is 1800c)?
This thread has been locked.
If you have a related question, please click the "Ask a related question" button in the top right corner. The newly created question will be automatically linked to this question.
We want to use B-type thermocouple to mesure temperature with ADS1248. How to estimate accuracy/resolution(using ADS1248 internal ref and temperature is 1800c)?
Hello Frank,
The accuracy that you may achieve is dependent on mainly three factors: Accuracy of the Cold Junction compensation Measurement, accuracy of the Type B thermocouple, and accuracy of the Analog to Digital converter performing the TC measurement. There are many factors in your system that will affect each of the 3 items above; the following is an estimate/approximation based on some assumptions about the system and typical specifications of the ADS1248 device.
A quick estimate/example:
1) Accuracy of the Cold Junction Compensation. Let's say that the thermocouple is connected through a thermocouple connector/isothermal block and the cold junction compensation temperature measurement is performed using class A PT-100 RTD. Class A RTDs can achieve accuracies of +/-0.15C.
2) Accuracy of the B-Type Thermocouple. Lets assume that the Class B thermocouple guarantees +/-0.5C accuracy. (This info provided by the thermocouple manufacturer).
3) Accuracy of the A to D system. The DC accuracy will be primarily a function of the Offset error, Gain error and INL error, and accuracy of the voltage reference.
For example:
A Type B thermocouple @ 700C produces a voltage of 2.431mV at the inputs of the ADS1248. The user sets up the ADS1248 with the extenal Reference REF5020=2.048V and with a PGA=128V/V. We can do an estimate of errors at a few points in temperature. The seebeck coefficient (uV/C) of the thermocouple changes as function of temperature:
Errors due to the ADC (based on t typical specs):
offset error = ~2uV (typ)
Gain error = 0.005% => 2.431mV*0.005%= ~0.1215uV
INL error = 6 ppm => 6 ppm * VREF/PGA = 0.096uV
Reference REF5020 Initial Accuracy Error = 0.05% => 2.431mV*0.05%=~1.215uV
Measurement Error of ADC = square root( INL Error2+Gain Error2+offset Error2 + Reference Error2)
Measurement Error ADC = 2.345uV at 700C
We know that the Seebeck coefficient of a B-type thermocouple at 700C is approximately 6.8uV/C; using this coefficient we can do a conversion to temperature:
Measurement Error ADC in Celsius = 2.3415uV/6.815uV = 0.344 C
Finally the total error in the system:->
Total Error @ 700C = square root( ADC Errror2 + CJC Error2 + TC Error2)
Total Error @ 700C = ~+/-0.6C
We can repeat the same estimate calculations using a spreadsheet for different temperatures to obtain an extimate, please see below. At 1200C, the calculated error in this example is approximately +/-0.8C.
Thank you and Best regards,