Hi
I would like to buy a ECG cable fit in the DB9 connector of ADS1293 evm.
I've looked similar evms like 1298 and 1292. The cables of corresponding evms are specified in the user's guides. But not in 1293.
Is there any source to get it?
This thread has been locked.
If you have a related question, please click the "Ask a related question" button in the top right corner. The newly created question will be automatically linked to this question.
Hi
I would like to buy a ECG cable fit in the DB9 connector of ADS1293 evm.
I've looked similar evms like 1298 and 1292. The cables of corresponding evms are specified in the user's guides. But not in 1293.
Is there any source to get it?
Hi,
The ECG cable shown in the 1292 would also fit on the 1293 evm; the only drawback is that the cable does not have the chest electrode (V1). Here is the pinout of the DB9 connector on the EVM; you may find a different cable that matches these connections:
PIN1: RA
PIN2: LA
PIN3: LL
PIN4: RL
PIN5: V1
PIN6-9: NC
Regards,
Sergio
Hi Marjan,
You can use shielded wires. The front-end of the ADS1293 consists of an Instrumentation Amplifier with a high input impedance.
The difference in impedance between cables should be kept minimal, otherwise, common-mode signals (to some extent rejected by the CMRR of the ADS1293) would be converted to into differential signals and go down the signal path.
Regards,
Sergio
Hi Sergio. We've implemented cc2541 with ads1293 reference design and we used not shielded wires for connecting body electrodes to a PCB, approximately 1m. We used standard ECG conductive jel fitted pads. The problem we are facing is that we are getting somewhat noisy input on channel 3 and channel 2 and 1 only work if pads are hard pressed by hand. I was wondering what are specific requirements for the actual pads and the wires? Should they be shorter? Should they be really ground shielded? What kind of pads were used with 1293 for testing? Are there really specific requirements for them? What to do to minimise noise and boost channel 3 output? Thanks!
Hi Oleg,
I don't have any specific requirements for the cables/pads. In general, shielded cables are better for interference. The ADS1293EVM is not intended to be connected to patients, so the testing was done with standard bench equipment and ecg simulators.
To minimize noise, you could adjust the settings of the digital filters to reduce the bandwidth and make sure the right-leg drive feedback circuit is used correctly, having a dedicated electrode for the right-leg as in the example applications. You could also configure the INA and SDM for high-resolution and high-clock in register 0x13.
Hope this helps.
Regards,
Sergio
Hi Alexandros,
It looks like pins 10-11 are the two mounting pins found on the sides of the DB9 connector and they are connected to the chassis of the DB9 connector.
Regards,
Sergio
Hi Sergio, Hi Alexandros,
i am having the same problem with the ADS1293 EVM.
I am using Pads ECG and cant measure any ECG signal.
The only thing i see is a 50Hz Noise.
How did you get it working?
Thank you,
Michael
Hello Michael,
The EVM is not intended to be connected to patients because it does not have patient protection. For legal reasons, I cannot comment or give advise on this topic, but the community may be able to help.
Regards,
Sergio