Hello,
we have a device which is using the ADS1299 and a ADS1292 and it is already working, we have checked it works and we can dectect biosignals but we are seeing a weird signal of about 20hz in our samples and we are wondering if the bias is working as it should.
We tested a prototype before making our custom hardware using a ADS1299 Eval KIT but we didn't pay too much attention to the quality of the signal since we just wanted too see if it worked and that was it.
Now in our custom hardware we are using 8 single ended inputs which have the negative input of the amplifiers connected to a reference electrode through SRB1.
What we have tried to test the BIAS is connecting the BIAS, REF and two inputs electrode to the patient and then we tried disconnecting BIAS and REF electrodes to see what happened with the signal.
If we don't disconnect anything, we get the EEG signal with the 20hz noise in the expected channels and we can see when the patient blinks. The signal in the channels with no electrodes are basically a constant line.
If we disconnect the REF electrode we just get noise in all channels which is logical.
If we disconnect just the BIAS electrode, the signal in the channels with electrodes connected to the patient doesn't seem to have any change but the channels with no electrodes connected we get noise.
Is that the expected behavior? Is there any way to make sure the BIAS is working ok? We have tried reading with one of the channels the BIAS signal configuring its input and we get something like a noise that one we disconnect the BIAS cable it dissapears and if we connect it again we don't get that "noise" back.
We are filtering the signals with software filter to get rid of the DC because there is a lot of offset. How much offset should we be typically expecting?
The inputs in the PCB are basically the same as in the Eval KIT with the same filter in the input. The BIAS drive loop has a 1Meg in parallel with 1.5nF.
Also the DC lead-off feature isn't working at all when using the same configuration we had with the eval KIT which was working perfectly. We only get zeros in the lead-off status as all the leads where always connected. We've tried changing the lead-off configuration with no luck but we are still investigating.
Please, any help will be greatly appreciated.