Real time monitoring with ADS1293 wireless heart monitor

HI TI team

I am in need of some assistance regarding real time monitoring  with ads1293.i have connected the electrode with the body. The ecg signal were converted as ADC in ads1293and transferred to android via ble.

The signal received is full of noise and didn't receive any pqrst signal. where as when connected to the simulator i have get smooth signal PQRST wave.

Could  you guide me the parameters to check  for real time monitoring.

Thanks!!!

  • Hi Mari,

    The simulator provides you an ideal ECG source and most likely an ideal connection. You may have to check for proper contact of the electrodes and keep the cables close together to minimize interference pickup.

    Regards,
    Sergio