First I would like to thank you for the great Application Report SBAA211. I am designing a very similar system and the document has been really helpful. I am using ADS1248 for reading RTD, thermocouples, current source (0-20mA) and voltage sources (0-10V) from the same connector.
I'm having some trouble understanding the Analog Input Connection (0-10V) on Application Report SBAA211 - Item 3.6, Figure 9, page 9.
To respect the common-mode input range, 0.1mA from AN1+IDACn will result on 200mV bias on pin 3 of the connector (200mV drop on R3) and 0.1mA from AN2+IDACn will result on 200mV bias on pin 2 of the connector (200mV drop on R2).
With the voltage source connected to pin2 (Vsource+) and pin 3 (Vsource-), if Vsource- is 0V and the ground is not isolated/floating (Vsource- = AVSS), it would connect the 200mV bias with the 0V.
What would happen then? In my understanding the IDACn connected to AN1 will keep the 0.1mA sourcing but the voltage would drop to 0V, not respecting the common-mode input range of AVSS + 0.1V + (Vin)(Gain)/2 (ADS1248 datasheet Page 3).
On pin2 there is a 18k resistor that separate Vsource+ from the 200mV bias, so here I think the voltage would add to the 200mV bias, and for a 10V input I would see a 1.2V after R1 (there is a voltage divisor (2k/(18k+2k))=0.1). Is that assumption right?
I would also like to point some differences in the schematic figures and the switch position table. In Figure 10, SW3 and SW2 are closed, but Table 1 says that they should be open. If SW3 is closed, R3 will be in parallel with R18 with Req=95Ohm, and a 1mA current from AN1 would make 95mV, lower than 100mV common-mode input range. With SW2 closed, depending on the 4-20mA source voltage, there will be some current leakage trough R2.
Would you please clarify that?
And what about circuit protection, would you please recommend a solution? I am only placing zener diodes (SMBJ10CA and SMBJ5.0A).