Hi,

What can be the max SCLK speed for ADS1292?

After I initialize the registers at 1MHz and I increase the SCLK to 12 MHz I see a sine wave kind of noise on the input data with a frequency of around 25 Hz and a Pk-Pk voltage of 0.5 mV.

The data rate is 125 and the clk of the ADS is 512 KHz, The controller is MSP430F5529 LP.

Thanks & Regards

Viney Chaddha

  • Hey Viney,

    The maximum SCLK speed varies based on your DVDD voltage and can be found on page 10 of the datasheet. A frequency of 12 MHz should be ok. Is the signal you mentioned what you are getting as output from the converter? If so, why do you think there is a problem communicating? What input are you forcing which you expect to see at the output?

    Regards,
    Brian Pisani
  • Thanks Brian,
    As expected that you would the first person to respond - it just proved

    The signal that i am talking if is a output of the ADS after filtering through the IIR to remove (same as that is used in eval board. subsequently this is filtered by 161 FIR filter. the filtered data shows the problem. In fact i see it like a sine wave getting imposed on the signal. The frequenct is approx 25 Hz so that can not be the naiin freqency which is 50 Hz

    If i get the ADS to run at 2 MHz would that help.
    Could you send me the regiter setting to run that at 2KHz

    Their slopes is creating a problem in further processing _ Could you suggest a way ot. I also though 12 K is ok but the output that i see is not promissong (As I would expect after filter)
    I would like to run it ar 125 SPS - ADS
    24 MHz - MSP430
    thanks and regards
    Viney Chaddha
  • Hey Viney,

    What is leading you to believe that the SCLK frequency is responsible for the 25 Hz interference? Is that signal present in the data before filtering?

    When you say you want to get the ADS to run at 2 MHz, are you talking about the modulator clock on the ADS1292? That cannot be changed unless you provide your own external clock.

    Regards,
    Brian
  • Brian,
    I feel that the "interference" that i am referring to is probably the input referred noise as that increases with the increasing the sampling rate.

    Besides this I would like to have your suggestion on the following:

    What would be a better device if I would like to capture Lead 2 and V1 and/or V6 ADS1294 or ADS1293? As I understand that ADS1293 already has the Wilson terminal output while it would need to be created for ADS1294, which in your opinion be a better choice in terms of accuracy and noise.

    I am tending towards ADS1294 because if I were to create the wilson terminal then I could possibly use the channels of 1294 for Lead 1, Lead 2, V1 and V6.

    Can I use pin 31 and 32 of ADS1292 (Channel 3, as mentioned on the datasheet) for this - is that possible?

    Regards
    Viney Chaddha
  • Hey Viney,

    The ADS1294 can produce a WCT output! Check out page 36 on the ADS1294's datasheet (www.ti.com/.../sbas459k.pdf). This is an integrated amplifier which you can flexibly take the average of any 3 inputs to create a WCT output.

    Using the "3rd channel" on the ADS1292 is a little tougher than it sounds. I wouldn't even really call it a channel as much as I would call it another set of inputs which can be multiplexed to one of the existing channels. This creates some serious timing challenges especially since you will have to wait 3 sample periods to data to be settled after muxing in a different channel. I would personally recommend you switch to the ADS1294 for 4 simultaneous sampling channels as well as a whole bunch of other integrated circuitry, including the WCT amp.

    Regards,
    Brian Pisani
  • Thanks Brian for your help.
  • Brian,

    Strangly I am seeing a lot of noise on the output. attached is the screenshot. The out put shown is cascade filtered:

    IIR - FIR (0.67 - 40 Hz) - FIR (0-20 Hz) and despite that I see this noise. Could you let me know what could be the possible reason for this. I have changed the caps on the pin 1-pin 2 and pin 7- pin 8 of the ads1292.

    Regards

    Viney Chaddha

  • Hey Viney,

    It seems like there is really just one signal which is disrupting the ECG waveform. Just by examining it, can you tell at which frequency it is occuring? I'm not asking for a Fourier analysis; I just want you to estimate the time between wave troughs and divide 1 by that time.

    Regards,
    Brian Pisani
  • Brian,
    I did that and it around 26 Hz. the time is approx 38 mS.
    Regards
    Viney Chaddha
  • Brian,

    It is this 25 Hz that I was talking of earlier as well. That time I reduced the sampling rate from 500 to 125 and i got a clean signal and when I reverted back i got that back and hence I concluded it is the noise arising due to the sampling rate being high and i resorted to remove it by software filters.

    This time the filter didn't seem to remove it much though it reduced the pk-pk values. 

    Another thing that surprises me is that when I took the reading without the 4.7 nF cap on the two PGAs (pin 1,2 and pin 7,8) i got quite a clean signal, the graph is attached. I am repeating it now with the caps back on them and post the results.

    Does this give you any clue on its source?

    Regards

    Viney Chaddha

  • Brian,

    I repeated it with the caps on the pins. The 25 Hz signal seems to be absent but the cleanliness of the signal seems to have deteriorated a bit. Attached is the graph. The magnitude is small as can be seen in the other graph with the brown line which is the graph after IIR and FIR (0.67-40 Hz) and the Blue one is further filtered using FIR (0-20 Hz). I will do the FFT and send you the results of all the 3 data sets.

    Regards

    Viney Chaddha

  • Brian,

    Screen shots of FFT Histograms.

    1. The FFT of the data with the caps

    2. The FFT of the data without caps

    3. The FFT of the data with the noise (the first graph that I posted)

    the noise range is 24 to 26 Hz

    The last data is of the FFT of the noisy signal for your review.

    Regards

    Viney Chaddha

    Sampling Rate 500
    Signal Excel Output of FFT Magnitude
    0 6 -10203 10203
    0.12207 6 -2246.27529912118+3267.41649625545i 3965.067878
    0.244141 3 178.953043337821+2894.50766358813i 2900.034277
    0.366211 -1 -2457.67010662417+1391.95625672811i 2824.479523
    0.488281 -8 -2739.20630882843-287.384821136032i 2754.240592
    0.610352 -17 -2549.77566472955-966.938092757297i 2726.962636
    0.732422 -25 -2469.81317934233+111.318963463701i 2472.32058
    0.854492 -32 -1360.98066104253-657.726804810316i 1511.579607
    0.976563 -38 -1543.15370509023-3486.28676207188i 3812.54754
    1.098633 -42 -5758.18008507712-2811.82757929813i 6408.042777
    1.220703 -45 -10047.1710081391-2524.46338116374i 10359.4672
    1.342773 -47 -16527.5343712662-18191.7596560366i 24578.43591
    1.464844 -49 16846.9391958412+20761.3310767955i 26736.7206
    1.586914 -51 7814.38459559439+3590.42201281259i 8599.752138
    1.708984 -52 3967.16560899754+3860.2336577853i 5535.323555
    1.831055 -51 3922.18070219328+4340.38879612067i 5849.99798
    1.953125 -46 4288.86391032229+2507.46784512022i 4968.07293
    2.075195 -37 1700.99907961502+1131.1245362741i 2042.753188
    2.197266 -12 1115.46493306524+1989.35033724425i 2280.740402
    2.319336 0 2617.25894952417+2303.57354231556i 3486.616623
    2.441406 6 5978.18350947642-921.520699309519i 6048.791489
    2.563477 7 4725.04595437393+2789.19559896992i 5486.863527
    2.685547 1 -3284.44791441725+9640.36509351122i 10184.50967
    2.807617 -9 -35090.2204640618-19688.1251885894i 40236.12613
    2.929688 -22 -2977.59589826086+12594.0586097631i 12941.26692
    3.051758 -36 307.312501380713+563.608538324266i 641.946694
    3.173828 -48 2687.86855910977-3105.3395851936i 4107.039241
    3.295898 -57 -1719.12002644843+730.127242575392i 1867.741806
    3.417969 -62 3226.85618412974+3103.18087343577i 4476.866356
    3.540039 -64 3924.25596795951+4010.4786679728i 5611.035916
    3.662109 -64 2536.06663493759+1549.6276030697i 2972.032921
    3.78418 -62 -169.673588676147+5536.30121748844i 5538.90064
    3.90625 -59 9964.27066228198+2820.62972423925i 10355.80233
    4.02832 -52 20824.8487983862+9696.48974311814i 22971.63992
    4.150391 -33 6400.74467693884+32197.4067574356i 32827.46616
    4.272461 -21 -220.596970853202-32191.2460204555i 32192.00185
    4.394531 -10 -17187.3463500831+1166.02326344462i 17226.8536
    4.516602 -1 -7639.67343149797-2129.70984497705i 7930.969308
    4.638672 2 1466.42840994128-878.958332955295i 1709.672493
    4.760742 1 -3850.28181590568-7384.91032329199i 8328.359415
    4.882813 -4 -1511.19621950684-2033.65536946248i 2533.666943
    5.004883 -12 -4825.03516414141+2317.86473647535i 5352.892795
    5.126953 -21 920.174634038173-809.58038655362i 1225.618929
    5.249023 -27 -7599.21116332578-2769.41229192521i 8088.118122
    5.371094 -32 -3910.04019395129-8082.89155675206i 8978.950397
    5.493164 -34 8263.74803497593+4972.3265486552i 9644.353939
    5.615234 -34 -6894.37396575423+18407.8641912939i 19656.59829
    5.737305 -34 18008.9910437607+3251.46922109927i 18300.15876
    5.859375 -35 -2607.22558704371-307.569343639487i 2625.304585
    5.981445 -38 -1314.78225934392+1298.45734226071i 1847.875499
    6.103516 -44 -5681.16564472988+21197.8432723033i 21945.93818
    6.225586 -45 1348.76287688732-362.436824089829i 1396.610808
    6.347656 -41 -168.33982671927+985.298099403542i 999.5752308
    6.469727 -34 11597.6899116714+10711.0362819031i 15787.1058
    6.591797 -23 839.171738227628+2516.67527286943i 2652.897215
    6.713867 -10 -308.022192161693+3896.83067332878i 3908.985414
    6.835938 1 -5477.11804020865+8112.4150760816i 9788.2634
    6.958008 12 -8147.03670930547+930.985494010819i 8200.057386
    7.080078 19 6539.86016522773+4754.58021342833i 8085.530532
    7.202148 24 -1107.16685875585-8243.84229657238i 8317.857553
    7.324219 27 559.152079015595-3199.5567579036i 3248.047797
    7.446289 28 -594.474012162087+517.917963306097i 788.4404656
    7.568359 26 5078.20172168708+1593.61517647288i 5322.381239
    7.69043 22 1806.94726401789+1948.5037220026i 2657.390669
    7.8125 15 -874.659729531172+1399.04012778344i 1649.9524
    7.93457 6 -858.546402938155-570.61384165573i 1030.874426
    8.056641 -19 8850.71386118693-3844.90423425739i 9649.788828
    8.178711 -31 4585.80032591416+2819.17644683753i 5383.058654
    8.300781 -41 -14594.6490472149+8319.46270103158i 16799.32262
    8.422852 -45 -6835.10641451231-16022.4318344997i 17419.44321
    8.544922 -44 -22301.9900195393+11994.7072702159i 25322.9493
    8.666992 -39 188.825422971203+2220.38864102073i 2228.403186
    8.789063 -30 1128.89298241647-80.2760211297068i 1131.743613
    8.911133 -20 -2578.41603137543-4581.82100986673i 5257.500642
    9.033203 -13 -2382.98713315738+4554.86863442898i 5140.569614
    9.155273 -8 -3193.00967330601+4826.50048863105i 5787.090611
    9.277344 -7 -3082.84224174797+1127.18853027334i 3282.448822
    9.399414 -8 -2216.62345287244-3493.60707418788i 4137.476274
    9.521484 -10 2263.15086226256+9528.68835710589i 9793.760954
    9.643555 -11 -3037.84509468465+18860.4417680526i 19103.52759
    9.765625 -10 -8487.25630006608-1241.12375145143i 8577.5234
    9.887695 -7 5620.92780893629+14128.1788032194i 15205.27099
    10.00977 -1 369.006452221661-11758.3210059515i 11764.10977
    10.13184 0 -3026.186970424-3713.8866829586i 4790.695343
    10.25391 -2 -4453.03322783916+364.13391506399i 4467.896422
    10.37598 -6 3160.06090592208+3106.86750194614i 4431.547202
    10.49805 -10 -355.464776722178+1308.78701838271i 1356.200084
    10.62012 -14 -3711.2325270108-3530.73078335065i 5122.431721
    10.74219 -17 997.36079896828-680.972767788674i 1207.664057
    10.86426 -18 5872.81330350264+9008.08308117903i 10753.39467
    10.98633 -19 310.747172170891+5978.09848655652i 5986.169503
    11.1084 -20 -9957.23430793261-12074.9903163068i 15650.93947
    11.23047 -21 3279.75391599892-1261.92789462249i 3514.14965
    11.35254 -23 -14515.0920189427-19358.674699344i 24195.99517
    11.47461 -25 -5967.90193815078+755.726609311668i 6015.561175
    11.59668 -24 874.011044702822+258.913965601253i 911.5545775
    11.71875 -21 4717.89222857021-3103.04901455166i 5646.894746
    11.84082 -14 -5487.46975928311-1524.00315949214i 5695.165493
    11.96289 5 -8506.22406866008-568.351763523324i 8525.190416
    12.08496 14 846.395867223807-4545.62164051727i 4623.749784
    12.20703 20 5132.11023827872+1609.14594775211i 5378.466899
    12.3291 20 -14444.2030040088+3611.20737322892i 14888.78165
    12.45117 14 -23750.0821871563+1758.42615860588i 23815.08905
    12.57324 4 -4791.13675464051+5836.62551318826i 7551.237566
    12.69531 -7 -18230.6298948291+2860.37071124357i 18453.66053
    12.81738 -19 12075.2548723913+310.607853559613i 12079.24904
    12.93945 -29 2408.77730316993-7565.93269376341i 7940.122519
    13.06152 -38 11.3170768221392-3078.22015520061i 3078.240959
    13.18359 -44 -2351.44646551897+8108.6421044029i 8442.711487
    13.30566 -49 -681.632248626909+2628.66793195949i 2715.606308
    13.42773 -51 -603.293431041577-4570.4671163539i 4610.111997
    13.5498 -53 2651.91862476536+1500.19965113736i 3046.846138
    13.67188 -53 -3765.8999740706+12484.428765519i 13040.05231
    13.79395 -50 -2682.83255820548+3610.94677519004i 4498.502767
    13.91602 -33 3950.584160584-12448.8798236827i 13060.69386
    14.03809 -21 -6166.27585272011-1074.17094102177i 6259.137409
    14.16016 -9 4570.98290147591-17397.7389023639i 17988.19623
    14.28223 1 -5545.24926948042-4364.08681914392i 7056.560297
    14.4043 7 1844.11495579531-399.967195005606i 1886.990654
    14.52637 8 3545.38526002874+2864.78825824759i 4558.154057
    14.64844 2 -3635.5179710427-3405.31496775236i 4981.281055
    14.77051 -7 -4995.67686658968-5309.04072520891i 7289.904031
    14.89258 -17 2887.16134378601-1984.80223324766i 3503.589663
    15.01465 -27 -1035.64160513025+3335.27805627762i 3492.367857
    15.13672 -34 -9454.16541020091-7968.80486238822i 12364.59035
    15.25879 -37 -12284.4487431461-11820.6613825841i 17048.04143
    15.38086 -37 -7287.52863469629+4509.07409628831i 8569.703776
    15.50293 -35 -6044.56832186908-13715.939286918i 14988.78903
    15.625 -33 -739.599775859514+7018.95350892884i 7057.812422
    15.74707 -33 6447.20240340576-1920.49212442638i 6727.162019
    15.86914 -38 2009.72466372105+548.416055598636i 2083.207477
    15.99121 -43 -10050.3511922003+697.394426674969i 10074.51826
    16.11328 -47 -6454.48867706416+167.413226300969i 6456.659451
    16.23535 -50 3008.4212503083-2974.4873763166i 4230.623308
    16.35742 -50 -986.804170094598+6803.57225489787i 6874.763843
    16.47949 -48 -13800.826817196+2475.68366966868i 14021.12087
    16.60156 -40 -4758.57162372462+1955.30322074959i 5144.629684
    16.72363 -28 9934.88371694111+2693.52360961623i 10293.54088
    16.8457 -7 -4613.99922767909-13824.7366625834i 14574.37246
    16.96777 22 15258.1539031384-1483.09317064485i 15330.06281
    17.08984 62 1719.80315130001-13598.9338659705i 13707.25082
    17.21191 110 383.408798889486+2373.50671222986i 2404.274614
    17.33398 163 -4700.4915323653+6918.71961505402i 8364.406838
    17.45605 216 2429.88446118614-4918.1590446243i 5485.674697
    17.57813 263 3533.39471197127-10627.7431892845i 11199.72337
    17.7002 297 3354.36544916087+2278.76517326966i 4055.18659
    17.82227 310 -7306.91969838988+1571.84559126045i 7474.073457
    17.94434 286 2389.28168862814-11591.1139204207i 11834.80413
    18.06641 244 1647.69478021255-16416.618414627i 16499.09871
    18.18848 190 -17868.8546131523-11582.61968035i 21294.43692
    18.31055 129 8802.00204181814-8564.68214691312i 12281.24669
    18.43262 65 -9063.96519098989-2736.38681602454i 9468.013403
    18.55469 5 8514.59558791478+5383.73057352358i 10073.87179
    18.67676 -48 2830.36224018543-5946.65541873506i 6585.868286
    18.79883 -92 -5881.95875394632-11307.6558898016i 12746.00018
    18.9209 -126 -4823.27456175662-5519.69668144548i 7330.14522
    19.04297 -148 3095.30951664683+3979.81424825079i 5041.811426
    19.16504 -156 -8175.44620342237+1636.61460183453i 8337.651227
    19.28711 -153 -7952.57285755027-10981.4988630257i 13558.6405
    19.40918 -140 -8161.03827259777-1792.51407145718i 8355.576125
    19.53125 -121 -6017.95181169697+8684.50253742863i 10565.8094
    19.65332 -98 16710.9000460028-10258.8665001547i 19608.63387
    19.77539 -56 1943.42010760296+4153.89400373337i 4586.03501
    19.89746 -45 13896.9710526751-3771.94954811538i 14399.7711
    20.01953 -40 -3442.09021232731+3420.05508580402i 4852.29449
    20.1416 -42 -7791.61866889957-2481.03556838098i 8177.093553
    20.26367 -47 4899.88511119852+632.082049621003i 4940.48599
    20.38574 -53 9145.35335850709-317.250945694431i 9150.854398
    20.50781 -59 -3565.52273067917+2728.1761505711i 4489.531986
    20.62988 -63 -2935.72802479986-5408.64672880124i 6154.0197
    20.75195 -64 7690.46592963375-1387.54369329848i 7814.636506
    20.87402 -65 9340.72353091303-7677.30338317467i 12090.90995
    20.99609 -65 3514.7263141235-24576.6560255617i 24826.70583
    21.11816 -65 3186.60846542855+3413.67218815843i 4669.864165
    21.24023 -64 5015.35536681251-8149.70230654968i 9569.296585
    21.3623 -61 -585.365096826057+8684.90438533353i 8704.608922
    21.48438 -55 -3507.57414864463-1257.99091904033i 3726.341042
    21.60645 -46 4553.71759512967-8507.63986663153i 9649.677717
    21.72852 -22 923.020967778911-5970.38260748328i 6041.3108
    21.85059 -11 -4097.96467952109+2296.55151519886i 4697.601875
    21.97266 -2 -4542.19042855072-7435.73449209819i 8713.30255
    22.09473 0 1462.50273074162-9020.7353523073i 9138.521792
    22.2168 -4 -5507.6950794037-7509.49875637435i 9312.748072
    22.33887 -12 -8531.08715159472-3853.27327793971i 9360.938144
    22.46094 -24 14267.0984717457+8509.76902097194i 16612.23247
    22.58301 -36 2495.08397358467-2690.87953026418i 3669.642582
    22.70508 -46 6617.0649350779+9313.58037271014i 11424.89946
    22.82715 -55 -5620.99751832683-390.559191507037i 5634.549634
    22.94922 -60 -2565.89356099192-6576.49761680743i 7059.329336
    23.07129 -62 899.564837495933+1710.33937752971i 1932.479672
    23.19336 -62 2182.99959534175+5185.09091194175i 5625.891485
    23.31543 -59 -2227.02534279749-4962.24250019543i 5439.070923
    23.4375 -55 3455.37594809429-2656.65707249914i 4358.606399
    23.55957 -48 3000.14369997867+2578.91460033274i 3956.218237
    23.68164 -27 10218.0675932234+1329.09988056577i 10304.14537
    23.80371 -15 20446.2195857369-9808.94607410802i 22677.37459
    23.92578 -3 -3149.83603706029-2813.74836355413i 4223.582237
    24.04785 5 4702.0882711203+5233.66029072718i 7035.68292
    24.16992 10 -6950.63458415753+1662.53102719462i 7146.700675
    24.29199 9 -191.774659857403-6572.83966670727i 6575.63676
    24.41406 4 9895.59550633707-11113.0141956267i 14880.25184
    24.53613 -3 6479.07240985253-1699.6709987407i 6698.30283
    24.6582 -10 -3424.84407135269-5053.24830078907i 6104.496318
    24.78027 -16 33998.2014026096+11673.1184752006i 35946.34048
    24.90234 -20 2579.64907640309-4002.46924377857i 4761.759066
    25.02441 -20 1988.17997952338-10400.1533413243i 10588.48663
    25.14648 -18 34433.4491597742-32518.2299162668i 47361.35237
    25.26855 -16 -3912.12544958129+12606.794479444i 13199.8482
    25.39063 -16 10268.1976328993+2194.5109353335i 10500.08385
    25.5127 -17 -4583.41078442889+21977.0875723693i 22449.94505
    25.63477 -23 -1760.79474792508-5887.70047677958i 6145.357195
    25.75684 -24 4589.99291139366-4163.25925762619i 6196.834883
    25.87891 -21 -11001.3572689932-2386.33061074339i 11257.19483
    26.00098 -14 -6012.87826704507+3151.10367758429i 6788.531464
    26.12305 -4 2730.91201558438-2910.47588397457i 3991.083851
    26.24512 9 6119.11746386654-3488.49528253483i 7043.663668
    26.36719 23 -410.610949979765-1336.43776176233i 1398.094147
    26.48926 35 4887.27812541252+2953.01032069022i 5710.145132
    26.61133 46 18781.1648697026+17981.5990668284i 26001.34725
    26.7334 55 9226.10409709884-5763.61807206559i 10878.43234
    26.85547 62 -2823.6900150216+9177.63812239737i 9602.201144
    26.97754 66 -8585.25513526256-4630.9406801296i 9754.599803
    27.09961 68 4171.0887628464-153.99353886889i 4173.930459
    27.22168 68 3131.42778212798+5005.30554864396i 5904.144611
    27.34375 64 4914.94857960943-2545.04742738578i 5534.797733
    27.46582 57 4889.02064832459-6889.27056906303i 8447.755434
    27.58789 34 4919.58690928443+839.211100220299i 4990.652325
    27.70996 22 7178.45338286727+69.4172070668619i 7178.789015
    27.83203 12 9952.57564166087-3534.78849791116i 10561.65194
    27.9541 7 10216.8671767217-3700.09158969039i 10866.23452
    28.07617 8 -10367.1596189787-797.054070224794i 10397.75427
    28.19824 14 -941.238530006171+13421.0303174674i 13453.99512
    28.32031 24 -3972.75810871978-2953.52785287516i 4950.367034
    28.44238 36 4468.64516993671-7576.56321178988i 8796.198029
    28.56445 48 1964.25960054496-872.719933815642i 2149.408258
    28.68652 56 -1481.52579503206-1894.33213932345i 2404.872748
    28.80859 62 1323.30581748983-2882.61619005697i 3171.847157
    28.93066 65 4027.42131644802+1735.12265279635i 4385.290535
    29.05273 68 -1581.37893588514-2009.77675111366i 2557.334927
    29.1748 70 -1035.13608933192+1355.03575350161i 1705.177005
    29.29688 73 739.852112707379+5049.01245131739i 5102.931303
    29.41895 77 -736.158562866083+12171.0168551484i 12193.25964
    29.54102 82 7657.81838843666+4082.07708472101i 8677.876226
    29.66309 81 -8000.76554376715-4013.70443323923i 8951.093373
    29.78516 77 195.971845743818-2209.58391895983i 2218.257438
    29.90723 71 4676.35600059325+3405.85748565723i 5785.168162
    30.0293 64 1694.32592483752+2037.62239583004i 2650.027428
    30.15137 58 1001.26494716765-55.4793576301801i 1002.800805
    30.27344 53 4657.44288098106+4606.79596297719i 6550.904001
    30.39551 51 179.619978282484+4123.71990553615i 4127.629973
    30.51758 49 6745.29394277859+1199.81749798425i 6851.171608
    30.63965 49 4794.67680450647-54.4559167241704i 4794.986038
    30.76172 49 4996.58810248911+2397.52917947151i 5542.024795
    30.88379 49 579.819686429223-2670.75605038402i 2732.970682
    31.00586 51 -7542.54977244857+3161.24556751224i 8178.235177
    31.12793 54 4605.01685705376-3138.37514364141i 5572.753251
    31.25 61 9.28761207207957-5043.31996524644i 5043.328517
    31.37207 71 616.428508627728+1985.02436825256i 2078.534544
    31.49414 98 1113.53145334178+423.294691320711i 1191.27272
    31.61621 111 5116.76709360167-1652.25170617101i 5376.917443
    31.73828 119 -3299.65816552413+1.83280637117264i 3299.658675
    31.86035 120 -746.971523393393-4380.12062557001i 4443.357194
    31.98242 115 -450.192905201543+2049.4385798874i 2098.302205
    32.10449 103 1931.59873040041+1310.71410425606i 2334.318984
    32.22656 88 -3124.76659693136+3070.74863887102i 4381.057348
    32.34863 73 -3382.08713794777+8297.19787411737i 8960.022655
    32.4707 57 2037.16196259677-4322.06297853055i 4778.101846
    32.59277 44 1244.99512436921+2064.67311582497i 2410.993143
    32.71484 34 217.217418117457+2368.08106924532i 2378.022573
    32.83691 27 -2454.86337957957+1214.90093141668i 2739.039701
    32.95898 21 -1239.89030543895+437.748046935466i 1314.895936
    33.08105 17 -962.23913481316+7255.11405130584i 7318.646326
    33.20313 14 880.329768434619-1543.13062771769i 1776.578913
    33.3252 13 5004.63890331626+2967.57130396369i 5818.323641
    33.44727 19 1701.0146835326+2643.78862213015i 3143.734918
    33.56934 24 2845.50920844441+7336.63085200775i 7869.12161
    33.69141 30 3325.24374655788+4804.43684258523i 5842.932427
    33.81348 34 -2693.0977866482-7856.58541760459i 8305.342263
    33.93555 34 -505.003854401609+3732.43160767937i 3766.440574
    34.05762 30 -1491.22875266924+2533.19738565963i 2939.532648
    34.17969 21 1769.91182462022+2636.95914429091i 3175.868605
    34.30176 9 2509.51797201222-2108.58426195764i 3277.774861
    34.42383 -2 3686.80035263179+426.774999161136i 3711.41937
    34.5459 -12 -2594.22082610888-3357.20251038113i 4242.733834
    34.66797 -20 5345.87684446999+732.03860386852i 5395.764983
    34.79004 -26 1838.9832831855+2105.23292860192i 2795.329176
    34.91211 -30 1438.08457770756+590.733160952163i 1554.687403
    35.03418 -32 -4721.66493534516+1738.6856251046i 5031.614797
    35.15625 -36 -9923.8733221319+332.052813734024i 9929.427012
    35.27832 -41 5190.27105873512+4930.47587069176i 7158.806184
    35.40039 -53 -473.259675041565-1665.94670744874i 1731.864069
    35.52246 -56 -615.000997637456-4285.21190216646i 4329.118533
    35.64453 -55 -3985.14054719708-697.760341603718i 4045.76503
    35.7666 -50 1527.07260066266+1874.71491525591i 2417.955074
    35.88867 -41 -1013.43431876326+2658.03669412704i 2844.680683
    36.01074 -29 2923.11148279321-1832.92285757386i 3450.244476
    36.13281 -17 -2588.9587300138+4754.12433482016i 5413.354366
    36.25488 -6 -3453.93185461619+5205.30707131477i 6246.988632
    36.37695 1 -1165.85685767968+6119.44574102028i 6229.513495
    36.49902 7 -3286.59853991086+380.75805447072i 3308.580762
    36.62109 11 4751.86332195382-7162.17036510274i 8595.166628
    36.74316 13 -5935.75495320971+346.229871442183i 5945.844094
    36.86523 13 251.690060973435+1158.87671994714i 1185.893393
    36.9873 11 5587.19193138131+5270.41053134636i 7680.75132
    37.10938 6 3325.6176103726-1966.47153938342i 3863.514308
    37.23145 -2 -1157.34139612194+3213.71946550949i 3415.762274
    37.35352 -27 -10263.2848905107+6120.31934615911i 11949.61613
    37.47559 -41 3308.68376124281+5755.85810904457i 6639.073038
    37.59766 -53 301.133423945418-164.493887203316i 343.1320124
    37.71973 -60 -3086.55215773005+12073.3757582447i 12461.66949
    37.8418 -61 -6127.26142895915-135.972419101593i 6128.769951
    37.96387 -57 126.387892985591-4833.1779906513i 4834.830234
    38.08594 -49 -3762.95755349826+1498.72994090226i 4050.437135
    38.20801 -39 2034.49318871275-4769.55673309454i 5185.348008
    38.33008 -29 1555.00662792707+130.788651123315i 1560.497127
    38.45215 -22 884.494125763053-1073.46770239613i 1390.921553
    38.57422 -18 2293.6050933806+3593.90685033276i 4263.42477
    38.69629 -16 -1399.05660227306-1700.06368148105i 2201.721122
    38.81836 -15 -498.163289350915+2212.69369819428i 2268.078496
    38.94043 -14 -5960.09351080762+2698.36504257349i 6542.468079
    39.0625 -12 -2535.00325650083+1493.76535433987i 2942.375986
    39.18457 -10 -3366.97988076542-77.3561089690642i 3367.868389
    39.30664 -7 -1805.55400815849-6143.50280616398i 6403.331321
    39.42871 -9 2606.74794969573+2986.9242928972i 3964.448462
    39.55078 -15 -997.75932298424-2331.47788103014i 2536.003308
    39.67285 -23 3829.12066506518+748.634074290236i 3901.617363
    39.79492 -31 -1950.13857864223+2119.11535496363i 2879.876797
    39.91699 -39 -536.263518536745+1992.17410770082i 2063.089004
    40.03906 -45 -3884.04397846285+3467.4055073125i 5206.601442
    40.16113 -50 713.363078231705+1152.96619218093i 1355.808955
    40.2832 -53 -837.525660052083+1548.09005552624i 1760.122738
    40.40527 -55 -2214.47392460358-2531.26735460616i 3363.214115
    40.52734 -57 -1044.77472144659-600.517525069671i 1205.062453
    40.64941 -58 -380.452662264106-1894.06703275335i 1931.899105
    40.77148 -58 6485.13981392325-808.557828919509i 6535.350348
    40.89355 -55 -2404.46278673544+227.235071732652i 2415.176406
    41.01563 -50 670.230852253394+781.459772722915i 1029.508995
    41.1377 -40 -447.374356124511+5050.15198418041i 5069.928883
    41.25977 -14 58.5880683201719-454.847835127907i 458.6056202
    41.38184 -1 -763.411267414694-716.681578383984i 1047.105175
    41.50391 6 -2685.39355483723-2953.39367250411i 3991.725533
    41.62598 9 -624.509680089188-82.2747260546564i 629.9059224
    41.74805 4 -720.196357080059-158.421580025223i 737.4145305
    41.87012 -5 1440.11822767812-2221.10285230377i 2647.118885
    41.99219 -17 -334.728586481129-984.12278595861i 1039.490685
    42.11426 -31 647.037375163048-604.672533234548i 885.5993661
    42.23633 -42 -705.303294747911+3971.01338324196i 4033.162534
    42.3584 -51 558.427731648744-507.078490578876i 754.3010852
    42.48047 -56 250.970277380595-137.324257310698i 286.0839593
    42.60254 -58 -3597.64127649128-1530.37680889195i 3909.613271
    42.72461 -57 -1025.01015272041+195.60140157953i 1043.506455
    42.84668 -55 -491.716022818346+183.960590211211i 525.0010913
    42.96875 -51 1524.84628450897-1506.41933561258i 2143.468079
    43.09082 -45 -189.583564294102-402.949045424999i 445.3199536
    43.21289 -27 -778.054500159004-256.029846907982i 819.0971174
    43.33496 -16 687.105243624112+1128.83163697201i 1321.504627
    43.45703 -6 564.438842024888+57.6600716245549i 567.3763215
    43.5791 2 -10.8673508650554+230.140947859543i 230.3973854
    43.70117 8 -1809.26798184902-246.537553814679i 1825.987786
    43.82324 8 -1597.89400716606+94.4921708429502i 1600.685487
    43.94531 4 -550.111945291363+648.129454883645i 850.1146644
    44.06738 -2 685.059496327003-999.552512418118i 1211.7804
    44.18945 -10 -163.341989804475-434.23623762473i 463.9415003
    44.31152 -16 -205.857495255981-559.503207955024i 596.1720792
    44.43359 -21 23.8403208120405+327.845603270389i 328.7112722
    44.55566 -23 -145.851628616854-3.94166905836813i 145.9048811
    44.67773 -24 22.4764578559078-335.798011399646i 336.5493955
    44.7998 -24 11.2763940079404+0.653378692624189i 11.29530724
    44.92188 -26 -108.610093016224+250.738333259205i 273.2505518
    45.04395 -30 -95.6234854264957+2.83332251176839i 95.66545187
    45.16602 -40 1.22449805015402-205.903374912507i 205.9070159
    45.28809 -43 -154.951543916627+100.632697053777i 184.7617944
    45.41016 -42 -283.688287308247-432.351348068943i 517.1138487
    45.53223 -37 -578.379138189998+23.0191926252778i 578.837033
    45.6543 -28 -279.049654797055-320.279247708474i 424.7911326
    45.77637 -15 -81.0202492928459+210.602580141542i 225.649568
    45.89844 -3 271.29598571975-404.704940089809i 487.224384
    46.02051 8 50.4170595639904-369.273701823581i 372.6995395
    46.14258 16 111.167648918074-366.861615277523i 383.3349592
    46.26465 23 37.8062259132808+479.793986730557i 481.2811864
    46.38672 27 -220.359482146431-5.29132039282969i 220.4230012
    46.50879 28 -210.585524603842-109.070881798206i 237.1554773
    46.63086 27 -475.240644152792+143.832707859926i 496.5294731
    46.75293 24 -599.985543001486-134.962796420477i 614.9777299
    46.875 18 23.0495596465563+364.888987088037i 365.6162676
    46.99707 9 121.233989794143-670.635849255876i 681.5057759
    47.11914 -16 173.745874137182-268.049287821854i 319.4339517
    47.24121 -30 -414.035901359595-283.328543233543i 501.6979081
    47.36328 -41 -79.7007533536486+114.505826247831i 139.5127031
    47.48535 -48 138.585970406168-93.207793548198i 167.0142628
    47.60742 -50 175.622061957986+183.191272688076i 253.7757889
    47.72949 -46 -436.085847892436+192.23576871263i 476.5768118
    47.85156 -39 -128.266446729341+72.5252602116649i 147.3505844
    47.97363 -29 306.618207221719-85.6521238585773i 318.3567359
    48.0957 -20 -398.863298829706-253.813191241545i 472.7716861
    48.21777 -14 341.916957473727-390.841262880209i 519.2919204
    48.33984 -11 -360.061250301286-377.910349092068i 521.9773328
    48.46191 -10 -7.82997351262856-158.963575040095i 159.1562964
    48.58398 -11 -270.858980937666+209.929287612641i 342.6877491
    48.70605 -11 -118.75236680327-181.894556747001i 217.2274255
    48.82813 -10 226.990111777232-458.331408989233i 511.4608404
    48.9502 -8 202.218253572614-36.3241549536132i 205.4547792
    49.07227 -5 -202.375570837922-127.821841240541i 239.3622668
    49.19434 -6 -65.6881522532429-164.818639228633i 177.4263712
    49.31641 -11 -58.7142782883568-351.580155814793i 356.4491162
    49.43848 -18 -506.527816414151+78.8323959543516i 512.6255704
    49.56055 -26 -234.405398460238-525.609501175965i 575.5095469
    49.68262 -34 -495.983119977138+165.356762022566i 522.8213022
    49.80469 -41 169.380686255871-3.6881251497985i 169.4208344
    49.92676 -47 607.84879737878-567.31760181794i 831.4622191
    50.04883 -51 -210.756480613972-607.840874376787i 643.3419174
    50.1709 -53 -169.868288241577-562.951541377087i 588.0218306
    50.29297 -56 262.177427241549+283.432231665385i 386.0969222
    50.41504 -57 -155.263793735264+268.812533326647i 310.4303847
    50.53711 -58 -143.422543933519-258.092547208523i 295.2656245
    50.65918 -56 -444.004242543768+34.7095795093951i 445.3588691
    50.78125 -50 -125.069633552642+463.53695071314i 480.1134428
    50.90332 -41 176.836163224882-486.45567494585i 517.600379
    51.02539 -15 190.387110023251-101.603338073873i 215.8019693
    51.14746 -2 166.226833763854-735.414339906727i 753.9665852
    51.26953 6 -76.3545182083987+103.279227000871i 128.4391341
    51.3916 8 -362.928743918772-40.5034922777884i 365.1818808
    51.51367 4 349.407107970531-127.703323306966i 372.0127227
    51.63574 -5 285.854366949389-98.4372679828795i 302.3286537
    51.75781 -18 -10.7214797993987+85.2482809140134i 85.91984362
    51.87988 -32 -218.898868031228-162.622462727781i 272.6953975
    52.00195 -44 286.78821013053-435.856595269258i 521.7455789
    52.12402 -53 143.753550605814-344.290067702394i 373.0961458
    52.24609 -59 -461.143714196593-862.389305514439i 977.9411227
    52.36816 -60 408.518292849054+163.850609212514i 440.1524937
    52.49023 -60 -236.539236690916-447.408436380418i 506.088055
    52.6123 -58 290.592924571512+290.128230287393i 410.631998
    52.73438 -54 -501.848843453875+126.608133973237i 517.5730685
    52.85645 -50 230.074864184011-546.997779669252i 593.4147067
    52.97852 -35 40.1990482302187-91.7350134348011i 100.1562588
    53.10059 -26 -201.725774937277-388.173165808175i 437.4605067
    53.22266 -17 -273.039216605866-28.5103996144543i 274.5236906
    53.34473 -10 -105.797038704362-271.039001666199i 290.9555874
    53.4668 -5 160.222416432173-33.0809431379245i 163.601869
    53.58887 -5 -470.435395413171+56.7273600101903i 473.8432807
    53.71094 -7 659.299651129412-165.835289798189i 679.8362842
    53.83301 -12 -213.621734239212+160.617344632876i 267.2679867
    53.95508 -16 646.183469047072-108.336143055351i 655.2021028
    54.07715 -18 -205.534412937453+4.69700187876499i 205.5880754
    54.19922 -17 -332.944919082828-599.428589032262i 685.6872118
    54.32129 -13 218.501265653007+125.145748342444i 251.8020282
    54.44336 -8 50.2988036062459+141.313780106412i 149.9985136
    54.56543 -4 -86.4777045438866-264.160024693527i 277.9548741
    54.6875 -1 133.258312873008-263.823653642357i 295.5684323
    54.80957 -1 218.122467814394-193.832826298833i 291.8022884
    54.93164 -8 519.712345461121-398.052740645369i 654.6349413
    55.05371 -10 460.395195275925-832.422042491657i 951.2571643
    55.17578 -9 -67.1189851432594+38.8446367012558i 77.54910681
    55.29785 -5 748.463817868091+594.363323258042i 955.7540723
    55.41992 2 -250.425874555786+227.765595558981i 338.5118686
    55.54199 13 -336.881189388811-253.655050222191i 421.6987316
    55.66406 24 -1516.10207295254-173.557776521183i 1526.003865
    55.78613 33 191.574264856066-326.093569432798i 378.203272
    55.9082 39 -105.9784985277-395.012298650424i 408.9818556
    56.03027 42 2202.85869717665-6981.79106123703i 7321.065009
    56.15234 43 85.1779295630164-133.364471413998i 158.2446268
    56.27441 43 46.6683818052103-425.336223608565i 427.8888185
    56.39648 40 -5244.48423514361-5249.23452675616i 7420.180457
    56.51855 36 255.174491520727+513.625631208171i 573.5201044
    56.64063 30 408.686826933609-62.7485492651292i 413.4758795
    56.7627 22 2079.27829317369-196.040010691352i 2088.499439
    56.88477 -1 -27.0144233313021-218.985304232565i 220.6452867
    57.00684 -15 123.513073698105-460.826356566349i 477.0916162
    57.12891 -27 -406.179664334199+1198.04394483009i 1265.026171
    57.25098 -36 -126.595569981454+201.521521825071i 237.9860544
    57.37305 -40 218.991726145753-210.214986472082i 303.5584238
    57.49512 -40 -135.703783972898-470.611218692482i 489.7861126
    57.61719 -36 44.9098788434248-107.228657994865i 116.253526
    57.73926 -29 240.022194198783+69.9012290642866i 249.993671
    57.86133 -22 1322.60225577491+68.1393344005369i 1324.356333
    57.9834 -17 254.015130279887-508.345233090731i 568.2768361
    58.10547 -14 144.524537151993+242.489645352108i 282.2916399
    58.22754 -12 -688.127462206939-46.3869045892305i 689.6891685
    58.34961 -10 185.73856739828-88.1446794647827i 205.5925581
    58.47168 -8 352.752307825333+271.200724849637i 444.953957
    58.59375 -5 270.920589431717-464.171649243932i 537.4507287
    58.71582 -2 100.95770760687-565.066505598553i 574.0144724
    58.83789 0 303.688372344543-107.638402655009i 322.1997102
    58.95996 -2 325.000681936106+110.711756299209i 343.3402631
    59.08203 -9 401.271761175996-266.743873086127i 481.8415924
    59.2041 -17 320.137859180606-453.565801057022i 555.1668081
    59.32617 -26 -649.794633893435+135.919050893867i 663.8577066
    59.44824 -34 289.3698931004+650.932760023432i 712.3541206
    59.57031 -39 -213.179933234851+18.9158399978151i 214.0175061
    59.69238 -42 212.545833482637-769.729759308961i 798.5359314
    59.81445 -44 173.035226238106-191.792054696712i 258.312566
    59.93652 -47 -292.93805903666-135.125166103382i 322.6011732
    60.05859 -52 150.56968240533+127.452545966338i 197.2698171
    60.18066 -59 409.450372889196+2.07640412652395i 409.4556378
    60.30273 -68 3.82883787444587+34.802711090943i 35.01269339
    60.4248 -76 -50.5775892555514+106.09508687885i 117.5340801
    60.54688 -79 103.408213817385+232.777637646333i 254.7129507
    60.66895 -75 478.075926183253+536.284748632677i 718.4413148
    60.79102 -37 533.0418142109-133.913573426817i 549.6056958
    60.91309 -2 -422.441542795718-25.5956226851078i 423.2162485
    61.03516 39 36.27745512994-119.874782532851i 125.2438311
    61.15723 86 374.039403954728+6.96266829528834i 374.1042027
    61.2793 132 89.0795790504098+49.219593797305i 101.7729818
    61.40137 176 194.988828043195-169.257680506856i 258.2030315
    61.52344 214 397.684931499824-34.7718932850431i 399.202191
    61.64551 242 202.473224111489+217.077367233706i 296.8467447
    61.76758 257 394.845456180352-236.903625436972i 460.4630952
    61.88965 257 122.912426516456-106.472248283641i 162.6155105
    62.01172 241 354.807448985788-91.6354401521144i 366.4496961
    62.13379 208 -38.7880091540717-26.762581976062i 47.12478592
    62.25586 160 -282.768540928269+310.166451665842i 419.7157079
    62.37793 101 168.649664025073-309.574050135932i 352.5319868
    62.5 36 -289.80613254816-318.823376490863i 430.8548942
    62.62207 -28 131.708041791663+109.565359322218i 171.3230172
    62.74414 -129 142.924393835912-212.0593608725i 255.7275012
    62.86621 -158 161.678921320696-294.381172529998i 335.8576311
    62.98828 -171 -75.2360574383766+230.339462439501i 242.3153571
    63.11035 -170 -132.565869194444-66.228177099189i 148.1886673
    63.23242 -159 139.775704141159+265.546432580829i 300.086913
    63.35449 -142 165.442166744938-41.2683304661615i 170.5115411
    63.47656 -124 36.0108871690683+323.143873052639i 325.1441937
    63.59863 -106 70.2577345102061+537.920788048037i 542.4895607
    63.7207 -89 71.6385894554145-306.244655272173i 314.5121244
    63.84277 -75 230.64137893224+93.2585251155392i 248.7822304
    63.96484 -63 88.9157316222069+78.3647093527269i 118.5201882
    64.08691 -53 -154.821477883307+37.2278290527199i 159.2344224
    64.20898 -46 103.019164714659+200.358378073716i 225.2918728
    64.33105 -42 444.485533121163+352.884622340675i 567.5340922
    64.45313 -42 94.262291004711-136.752549563626i 166.0922615
    64.5752 -44 374.80605026458-48.6407151058574i 377.9490633
    64.69727 -51 135.015277530342+36.8184636919178i 139.9454338
    64.81934 -52 430.65707050894+339.098344835823i 548.1361143
    64.94141 -50 322.186656967149+144.542349138867i 353.1242453
    65.06348 -43 -403.680294287808-302.620037352153i 504.5162703
    65.18555 -33 139.918362267453+229.047812460658i 268.4027729
    65.30762 -21 -160.628678996623+343.041095228908i 378.7859099
    65.42969 -8 145.548128954508-9.16482081020972i 145.836387
    65.55176 2 325.070804651442-265.292355190885i 419.5843917
    65.67383 10 176.883077421668-326.57160531446i 371.3982182
    65.7959 16 -241.012600300502+45.6808381528395i 245.3035109
    65.91797 21 285.378974355345+44.1616551215887i 288.7757102
    66.04004 23 283.748284543023+171.972868778725i 331.7947507
    66.16211 25 164.944739504433+59.3673225188725i 175.3032974
    66.28418 24 -249.797971809409+324.525693756266i 409.5313817
    66.40625 21 -503.359481200604+415.374179521645i 652.6151058
    66.52832 14 542.420266424568+68.0554237557704i 546.6729243
    66.65039 -7 97.9196227885617-35.0433503101663i 104.0013891
    66.77246 -21 -392.367851668681-414.630385918641i 570.8510208
    66.89453 -34 -75.675893412638+16.0781757717609i 77.36503461
    67.0166 -42 65.3354687784351+227.208387095808i 236.4156819
    67.13867 -45 437.623785150756+380.758238804254i 580.0787996
    67.26074 -43 186.226971511703-138.516970288254i 232.0935931
    67.38281 -36 -10.3366631809976+319.931657479888i 320.0985974
    67.50488 -26 -14.463650140292+484.628449436552i 484.8442339
    67.62695 -16 99.0111977832379+432.296043537042i 443.489669
    67.74902 -9 -138.008045497237+33.6817096269731i 142.0587139
    67.87109 -5 -82.2440828968326-769.45328183875i 773.8361856
    67.99316 -3 -336.63903249363+296.953842664276i 448.8957817
    68.11523 -2 43.0544713506116+93.3471546979549i 102.7977568
    68.2373 -1 864.0260809033-94.3434499728005i 869.1615241
    68.35938 0 177.209397354042-57.0996225419019i 186.1814636
    68.48145 2 255.617671111568+94.5410598689434i 272.5406498
    68.60352 7 -851.941362037833+1254.65948671256i 1516.566686
    68.72559 7 275.810726685908+233.858972987223i 361.6097015
    68.84766 4 14.3940778068501+21.9395256695738i 26.23989829
    68.96973 -1 412.438272290745+1378.07057727209i 1438.465795
    69.0918 -8 -397.594098227496+213.435862942277i 451.2603844
    69.21387 -15 -135.288792193168-177.008280924264i 222.7891129
    69.33594 -21 -289.422808802222+343.254956116123i 448.9872238
    69.45801 -26 -180.184344212766-295.336111830444i 345.962161
    69.58008 -30 160.117325757653-213.246035568383i 266.667264
    69.70215 -32 96.319316784458-109.194724266258i 145.6052835
    69.82422 -34 221.138458286138-3.5362237989193i 221.1667303
    69.94629 -36 -201.318077290745+239.758101194718i 313.0701444
    70.06836 -36 175.998171426243+247.054137506928i 303.3333203
    70.19043 -33 -101.311154480861+436.23540544696i 447.8451507
    70.3125 -26 -73.5146579436454+253.354663167735i 263.8048337
    70.43457 -15 -275.945945903595+112.075317041625i 297.8372739
    70.55664 16 -232.951613929751-335.66347081908i 408.578536
    70.67871 31 338.904311234143+36.5163511595036i 340.865921
    70.80078 44 -207.864620522113-38.9517617296201i 211.4827185
    70.92285 49 125.29315267126-203.715620180343i 239.1619284
    71.04492 48 150.827920111693+445.208378368658i 470.0633592
    71.16699 40 51.0939777114543+163.242763580889i 171.0520226
    71.28906 29 79.4105526942047+459.630335124378i 466.4397934
    71.41113 17 -15.7529094535979-69.5193278875936i 71.28177261
    71.5332 6 -77.9090846329525+102.710506739859i 128.9157619
    71.65527 -1 -292.691482704246+14.4274045884547i 293.0468462
    71.77734 -5 -168.841157051138-36.1541398861013i 172.6686368
    71.89941 -6 -110.588837589083-160.328027035667i 194.76901
    72.02148 -4 384.45983806098-260.474037200923i 464.3878671
    72.14355 -1 -63.7137419741171+252.360339023526i 260.2790457
    72.26563 3 62.1228321934876+12.5244688134667i 63.3727749
    72.3877 9 372.708818070553+292.109542022301i 473.5397001
    72.50977 27 -179.684111327665+1.80331732723733i 179.6931602
    72.63184 38 -59.4944953951521-2.63197632572388i 59.55268492
    72.75391 49 -550.977899032504+5.77226987118939i 551.0081345
    72.87598 60 -81.5558209059412+102.389647006027i 130.9006942
    72.99805 68 -14.2773846838863+95.7357387225449i 96.79450078
    73.12012 72 -35.1670077823321-197.142223796419i 200.2542755
    73.24219 71 -24.6170826062204-103.19198872473i 106.0876397
    73.36426 68 149.505235909965-23.8747747934372i 151.3995391
    73.48633 65 178.868673837062+327.219441528298i 372.9162981
    73.6084 63 -121.573119894334-138.812275185919i 184.5233623
    73.73047 63 31.8896418601615-88.8718832070885i 94.42012965
    73.85254 64 -470.446688924961+234.89657572194i 525.8293339
    73.97461 67 -3.83247031211257+113.191445170547i 113.2563071
    74.09668 69 -43.3243848336703+217.851151573606i 222.1173711
    74.21875 69 132.884861154409-301.95369838508i 329.9006249
    74.34082 66 -70.5814249356001-86.4142868376739i 111.575833
    74.46289 57 -199.605683455812+11.7078177630403i 199.9487481
    74.58496 55 60.4715985969452-105.540949411283i 121.6376021
    74.70703 56 131.100007284775-70.5979219969519i 148.90023
    74.8291 62 -37.014430422444+86.2375802514395i 93.84555561
    74.95117 73 -311.136570757014+192.745011902806i 366.0008269
    75.07324 87 -83.1697449618161+146.617256038731i 168.5640123
    75.19531 102 38.2173914021262+290.653121170985i 293.1549178
    75.31738 116 17.4566705215389-285.861952459054i 286.3944678
    75.43945 127 -156.811944181169-10.4679270408393i 157.1609472
    75.56152 135 -213.886505792572-193.60409373217i 288.4960701
    75.68359 140 -63.7174773075574+83.5628072577402i 105.08406
    75.80566 142 -92.9700513175553+68.8149844224525i 115.6673356
    75.92773 141 -47.0884246027835-118.53161434717i 127.5423982
    76.0498 136 106.481947321979-17.7809938802501i 107.9563284
    76.17188 129 32.0769753177686+128.607230710893i 132.5471695
    76.29395 119 -11.4207650065205+56.680618485928i 57.81977504
    76.41602 89 -53.5369716138297-163.166303403891i 171.7249251
    76.53809 71 -21.1161977736036+62.1282933571975i 65.61873699
    76.66016 56 -306.780376486394-133.047255600365i 334.3886535
    76.78223 44 -69.1463275147589+161.184325549229i 175.3898555
    76.9043 36 -157.46369785658-46.9748364134421i 164.3211837
    77.02637 33 89.0696501942188+99.131708815582i 133.2685195
    77.14844 34 -25.7203689687659-88.2950465320877i 91.96495323
    77.27051 36 -60.1173347582697-167.955142393888i 178.3900888
    77.39258 38 80.7316174912783-113.023197320939i 138.8950582
    77.51465 37 75.3100865210629+79.7154045598792i 109.6638266
    77.63672 34 -115.015728697007+1.20488453490185i 115.0220396
    77.75879 28 -191.013531964306+4.01234777961949i 191.0556681
    77.88086 23 3.59710742391806+110.603298013939i 110.6617762
    78.00293 18 -223.99731372449+77.2574396547745i 236.9462144
    78.125 15 170.080704623173+49.1958166770533i 177.0527449
    78.24707 13 -136.594016142414-186.797967148769i 231.4117667
    78.36914 12 59.9918561698481-89.1752717522706i 107.4767505
    78.49121 8 -199.64147330654+42.1274701326985i 204.0378436
    78.61328 1 -102.306274926785+0.402526841930047i 102.3070668
    78.73535 -8 39.6092342202559-98.564230082574i 106.2252272
    78.85742 -20 118.087956976558+71.5920426196104i 138.0948448
    78.97949 -31 -85.3321011090774+165.5061006883i 186.2091213
    79.10156 -42 -5.56454198654933+59.8836711878967i 60.14165115
    79.22363 -50 142.504982265627-94.7528228220017i 171.1308488
    79.3457 -56 -221.037133534273-28.2653701595373i 222.8370381
    79.46777 -61 29.6842194044233-309.720637886743i 311.1398824
    79.58984 -64 -310.014415822338-76.6606973149318i 319.3521575
    79.71191 -67 -0.638088671084319-78.8510227999216i 78.85360457
    79.83398 -68 27.2013426403061+151.402816078653i 153.8269344
    79.95605 -67 0.444762991403714-132.028949590786i 132.0296987
    80.07813 -62 -23.16163803129-227.646539972114i 228.8217836
    80.2002 -54 116.24788632245-30.8896068937668i 120.2819142
    80.32227 -31 -51.1829179607492+90.7443140008949i 104.1835957
    80.44434 -19 -94.4492079712382-110.366215596307i 145.2630525
    80.56641 -11 -166.207200350807-178.124782524351i 243.6252688
    80.68848 -8 -260.889895090442+191.770259161834i 323.7890821
    80.81055 -12 -148.433349917639-144.233837970981i 206.9682569
    80.93262 -21 -80.2173533457813+228.895346662344i 242.5446423
    81.05469 -33 172.913337781243-237.675791188832i 293.9197239
    81.17676 -45 80.3904301808858-205.055730538269i 220.2509339
    81.29883 -56 -267.900189349321-240.003823047705i 359.6836757
    81.4209 -63 -99.6584840779009-102.881404995095i 143.2354598
    81.54297 -68 157.432179090135-4.6862687619814i 157.5019115
    81.66504 -69 -1.74566546326461+142.60259141724i 142.6132758
    81.78711 -68 -214.277134662663-21.5433654948329i 215.3573937
    81.90918 -67 -126.1398799296+88.2900419284863i 153.9688307
    82.03125 -64 119.893696091961+235.414011448832i 264.1860238
    82.15332 -60 -22.6528095972126-328.424779233575i 329.2050811
    82.27539 -46 90.6696762455517-107.602576306699i 140.7100018
    82.39746 -37 -179.951289915063-364.805875603982i 406.7748685
    82.51953 -27 24.2161298843763+106.391956791597i 109.1131038
    82.6416 -17 -142.200734188194+8.79898220373023i 142.4727023
    82.76367 -9 165.816920052669-189.97594325018i 252.162864
    82.88574 -6 145.891127216851-117.407667935485i 187.2666054
    83.00781 -7 16.3560780884548+79.6700428259126i 81.33164829
    83.12988 -10 -141.770607769311+22.7930248875978i 143.5911808
    83.25195 -15 61.8896089637897-209.0008473497i 217.9717364
    83.37402 -19 -35.3711826772499-222.127748361834i 224.9263372
    83.49609 -22 -483.306903574828-280.142171958311i 558.6279617
    83.61816 -24 67.0206091781183-68.4027638022589i 95.76377264
    83.74023 -25 -197.528676616619-136.377015594718i 240.0338902
    83.8623 -26 343.312602038812-17.6404172604495i 343.7655117
    83.98438 -28 -37.4565568943795+55.2197292353303i 66.72489903
    84.10645 -33 -5.45353919635397-282.175510750606i 282.2282055
    84.22852 -46 26.4424492711864+37.3866563962574i 45.79263259
    84.35059 -49 -50.5665507411941+73.6424693395312i 89.33190552
    84.47266 -49 -153.054495004945-17.4425353310943i 154.0451897
    84.59473 -44 -178.881254761817-138.250299177979i 226.0788547
    84.7168 -34 -88.2720946978389-48.1743078963733i 100.5620537
    84.83887 -21 -122.547388479224+94.1629041208857i 154.5461579
    84.96094 -6 293.451500066173-341.167553202927i 450.0100913
    85.08301 6 35.3291073042465-95.7038442982705i 102.0165263
    85.20508 16 301.194111261343-276.301960451183i 408.7305543
    85.32715 23 -68.92216004963+120.04366859764i 138.4223483
    85.44922 27 -158.154804769931-131.497171948049i 205.6804524
    85.57129 29 146.709613765753+4.9889950377453i 146.7944169
    85.69336 28 -7.85843920247281+100.429487571628i 100.7364732
    85.81543 25 -148.106351626672-93.0770075879896i 174.9251861
    85.9375 20 -20.2337917134306-201.062683252771i 202.0782248
    86.05957 12 74.9573108209854-167.936156389595i 183.9052774
    86.18164 -10 159.65641059701-281.567698178417i 323.6827739
    86.30371 -23 -49.4861320527362-502.644501952254i 505.0746208
    86.42578 -35 -4.52295426248502+13.8662564044492i 14.58527278
    86.54785 -42 218.918819117284-38.9558304778469i 222.3578334
    86.66992 -44 35.6489112124539+171.083751147891i 174.7583897
    86.79199 -42 -171.788125689078-19.1034974567538i 172.847053
    86.91406 -36 -209.764596009351-295.151204294558i 362.0986318
    87.03613 -28 -11.4257862805245-120.103125945284i 120.6453872
    87.1582 -21 -194.104883554099-117.383503070996i 226.8382521
    87.28027 -16 326.356241980942-1311.11526507508i 1351.122361
    87.40234 -14 -58.7154630929037-92.0598451155775i 109.1902958
    87.52441 -14 -57.3935423471065-201.291241852928i 209.3135991
    87.64648 -15 -987.479779972007-1151.96564502994i 1517.280845
    87.76855 -15 266.372076246309+89.847677223721i 281.1168585
    87.89063 -13 168.563732432751-160.645134568245i 232.8531536
    88.0127 -9 641.800147132775+126.284322381464i 654.106382
    88.13477 -1 -27.1793973793622-88.1099366687286i 92.20672742
    88.25684 -1 -81.0288544690507-177.450163679535i 195.0749493
    88.37891 -3 -127.057264722177+437.53826772078i 455.6130861
    88.50098 -10 -4.36589733134997+45.7698395694368i 45.97759535
    88.62305 -18 78.3985701144083-222.854256335243i 236.2421541
    88.74512 -27 -0.724681355490503-318.813410707281i 318.8142343
    88.86719 -36 -0.516881918340502-145.771125667606i 145.7720421
    88.98926 -43 142.450668608831-61.1188006009358i 155.0087119
    89.11133 -48 605.533766732701-316.605187785494i 683.3081205
    89.2334 -52 26.8109554237692-286.968245191523i 288.2179749
    89.35547 -55 135.470382696568+58.044869016039i 147.3819236
    89.47754 -57 -320.510123469058+190.031176726238i 372.6105036
    89.59961 -58 97.7099211739731-100.655126182205i 140.2807297
    89.72168 -57 308.979478058+23.6817816510228i 309.8856961
    89.84375 -52 -87.7197386228962-318.11503293977i 329.9877675
    89.96582 -43 -106.751012955001-286.594805843763i 305.8306092
    90.08789 -18 136.902705745876-140.460117162775i 196.1412638
    90.20996 -5 183.868874977896-148.47120199252i 236.329137
    90.33203 3 102.90362589115-232.440597537047i 254.2002903
    90.4541 7 -23.4161405650717-277.52888425124i 278.5149857
    90.57617 4 -83.0023842932398+289.131958270124i 300.8100482
    90.69824 -4 328.193480048513+190.924601898342i 379.6882457
    90.82031 -16 19.4872556882373+45.4036783375324i 49.40897834
    90.94238 -29 -148.413830762142-306.951524076428i 340.9485347
    91.06445 -41 62.4269315944385-137.818202276887i 151.2976492
    91.18652 -50 -172.752677153425+66.427455926497i 185.0840197
    91.30859 -56 79.3349135921472-70.6871727635321i 106.2577287
    91.43066 -58 256.817282189387-132.201367305959i 288.8465301
    91.55273 -59 10.0187109405896-24.5242048926078i 26.49171936
    91.6748 -57 112.821027572661+53.4046382014654i 124.8224325
    91.79688 -55 140.697646475682+58.1738034749618i 152.2498576
    91.91895 -52 404.385153909792+34.8366752016831i 405.8829223
    92.04102 -38 223.417540877535-200.98139055026i 300.5144205
    92.16309 -29 -144.545016354023+90.1058503126039i 170.3300502
    92.28516 -18 -9.92557328585657-95.4383435649873i 95.95308451
    92.40723 -8 160.976380650901-92.7684266775918i 185.7939076
    92.5293 -1 32.5262336295288+35.4917749568585i 48.14168634
    92.65137 2 67.9539275448405-148.294869949687i 163.1229742
    92.77344 2 179.816004543504-173.723658300634i 250.0274084
    92.89551 -1 149.352464269617+27.6243489693942i 151.8856914
    93.01758 -6 134.150326962671-185.450077314701i 228.8843407
    93.13965 -10 16.1220001989568-80.5536599611908i 82.15114743
    93.26172 -13 119.686754862714-175.310627651493i 212.27043
    93.38379 -15 -18.7669923688143+25.27650317623i 31.4817664
    93.50586 -16 23.8428184920097+240.195316007986i 241.3757855
    93.62793 -17 19.4664109532286-166.294647749811i 167.4301377
    93.75 -20 -193.988314676777+3.594989999897i 194.021623
    93.87207 -25 84.0043791609213+80.3020505250984i 116.211682
    93.99414 -39 -23.0705993695418-144.381183436491i 146.2127856
    94.11621 -43 -17.0915640824774-132.729654900446i 133.8255688
    94.23828 -44 75.6807794198276+117.229193260465i 139.5358883
    94.36035 -40 -48.0957770856218+4.32063784671419i 48.28945728
    94.48242 -31 165.983205013054+73.5925816054915i 181.5662205
    94.60449 -19 100.364913804409-128.572154370827i 163.1070655
    94.72656 -5 160.776018517963+108.078919712449i 193.7265625
    94.84863 7 215.859124490888+209.063028995928i 300.5037632
    94.9707 17 -41.6736992546602-130.666747087989i 137.1513617
    95.09277 24 160.686548069422-3.20348025575203i 160.7184775
    95.21484 28 62.3892292282055+21.8123867408659i 66.09233041
    95.33691 30 -111.763676104065+2.87889891411658i 111.8007485
    95.45898 30 133.991126831966+86.8195359993253i 159.6598068
    95.58105 28 340.214869906868-40.2111566978477i 342.5829751
    95.70313 23 18.5013251847178-100.368069159248i 102.0590434
    95.8252 16 170.910232674596-145.505314627106i 224.4595826
    95.94727 -6 63.5943218949507-18.7996579243294i 66.31489211
    96.06934 -20 337.680255384871+5.47657717710726i 337.7246627
    96.19141 -33 162.507649174001-25.4560840489624i 164.4893561
    96.31348 -43 -246.529643324295-39.4624395857453i 249.668078
    96.43555 -49 123.504127628011+65.1140210406057i 139.6177112
    96.55762 -51 35.5125695829224+252.652144929783i 255.1357461
    96.67969 -48 46.2573402445768-59.7203920626521i 75.53983555
    96.80176 -43 113.024406937643-241.477542670769i 266.6194295
    96.92383 -37 -56.8380448628305-218.777553617281i 226.0402206
    97.0459 -32 -39.7317308275918+162.060734080319i 166.860097
    97.16797 -28 180.215481343155-59.578235940349i 189.8082873
    97.29004 -25 175.274760861154+5.82675044785059i 175.371585
    97.41211 -21 118.521968853243-10.0597595926605i 118.9481226
    97.53418 -15 10.7075573989081+276.482018650372i 276.6892814
    97.65625 -6 -66.2797890494218+330.599889889349i 337.1784359
    97.77832 2 297.365389479315-164.1316472631i 339.6547843
    97.90039 21 74.1770481760527-7.31082905532151i 74.53645214
    98.02246 26 -344.305053944387-87.7457548996808i 355.3101289
    98.14453 27 20.2774585289868+21.9855849544161i 29.9088828
    98.2666 22 141.176633673433+149.185323824812i 205.394992
    98.38867 14 423.094572345809+38.1230646350768i 424.8086454
    98.51074 5 53.5677709195657-102.105322179252i 115.3039587
    98.63281 -4 150.786921194587+130.405754837689i 199.3548507
    98.75488 -12 162.399794463631+270.299627468486i 315.3340797
    98.87695 -20 217.296805556369+166.608069885614i 273.8177325
    98.99902 -26 -54.9408325697357+52.4401616151756i 75.95041563
    99.12109 -31 -275.231620184866-351.796775904957i 446.6692471
    99.24316 -36 -28.7659722677443+203.829639471705i 205.8494671
    99.36523 -39 47.4412569004575+24.7032184400859i 53.48758601
    99.4873 -40 467.752946225954-378.106849738257i 601.4628904
    99.60938 -37 108.491583953572-40.2929381419641i 115.732211
    99.73145 -30 215.389697418313-82.6752875745118i 230.7117789
    99.85352 -5 9.89806716725855+1064.66862966295i 1064.714639
    99.97559 8 172.737709978993+47.001348294821i 179.0179968
    100.0977 19 -4.17917572882769+12.1625846749116i 12.86055892
    100.2197 25 791.71818431968+560.619801176753i 970.1093984
    100.3418 23 -95.8498562218861+198.103885698931i 220.0734979
    100.4639 14 -97.8883276268461+3.87889440631655i 97.96514945
    100.5859 1 -22.460832306654+289.233181107194i 290.1039849
    100.708 -14 -159.199002036253-61.0054761547785i 170.487508
    100.8301 -29 27.3952069582901-157.763184738179i 160.1240763
    100.9521 -42 54.8933002936753-56.7414780412926i 78.94852594
    101.0742 -51 92.8462415899828-46.9733621446101i 104.0524931
    101.1963 -56 21.1173630511583+287.160235153098i 287.9356589
    101.3184 -58 151.893206058699+60.4221201167384i 163.4698096
    101.4404 -57 102.825382915411+204.123620511281i 228.5596461
    101.5625 -54 45.9829404639113+139.113662512722i 146.5163537
    101.6846 -49 -89.8812143566752+137.159601843138i 163.9859417
    101.8066 -33 -178.336547972019-84.3620358097494i 197.2837485
    101.9287 -23 206.506514230029-112.237331588518i 235.0365057
    102.0508 -13 -88.4361236304305+91.4282735677849i 127.2009323
    102.1729 -6 -24.0998362222273-136.171643465174i 138.2878107
    102.2949 0 272.920220572528+181.792903018155i 327.9239338
    102.417 0 88.5105811236519+64.6699048975861i 109.6189745
    102.5391 -1 233.43987511125+229.890629126802i 327.6337538
    102.6611 -6 -57.4396466339207-74.1111367846956i 93.76445809
    102.7832 -9 -0.260386696211249+67.6107854178485i 67.61128682
    102.9053 -11 -128.38624664946+124.59736169443i 178.9064864
    103.0273 -11 -86.8980859900787+55.3069310396753i 103.0055046
    103.1494 -9 -100.035098676791-30.3941885558079i 104.5505986
    103.2715 -8 117.497062204029-222.501999784153i 251.6201493
    103.3936 -10 50.0814789921119+176.477191780022i 183.4457788
    103.5156 -16 -5.36832133062126+16.1099827264779i 16.98088388
    103.6377 -29 334.970909466434-6.45308529843405i 335.0330618
    103.7598 -64 -102.067437521244+90.8387181320842i 136.6361391
    103.8818 -78 12.5268293732584+18.769233286738i 22.5655838
    104.0039 -83 -257.904172690445+271.650554004162i 374.5778768
    104.126 -76 28.9654232238979+125.567387220824i 128.8649079
    104.248 -54 73.980745196528+54.9573283208076i 92.159962
    104.3701 -15 -103.600447523743-59.8423460546979i 119.641795
    104.4922 37 -24.9685790026918-70.7619883626341i 75.03791665
    104.6143 101 125.335040354637-22.5394470818485i 127.3455889
    104.7363 168 234.525341037808+68.2274559229484i 244.2480733
    104.8584 233 -136.245949000941-3.13710267593844i 136.2820606
    104.9805 288 0.321110043934638-115.235443041059i 115.2358904
    105.1025 328 -108.656573852877+324.462577719158i 342.1727859
    105.2246 345 76.4038176709598+47.4516216947561i 89.93997863
    105.3467 339 54.6207044934077+163.991520926201i 172.8486051
    105.4688 309 16.2553204511695-197.290166768925i 197.9586961
    105.5908 258 -91.8411202310413-39.9066835357765i 100.1365805
    105.7129 117 -133.521782127519+93.8496403014148i 163.2048446
    105.835 40 -0.545950524200833-75.0035610813428i 75.00554804
    105.957 -31 66.0804026893327-76.499770880935i 101.0882514
    106.0791 -93 -6.41165172044599+89.0177352528492i 89.24834154
    106.2012 -143 -63.8684704039264+246.404807372324i 254.5476588
    106.3232 -177 36.0094775013961+146.134012524055i 150.505256
    106.4453 -197 148.066401900772+120.085721762262i 190.6416532
    106.5674 -202 -94.5597483172745-197.569955513647i 219.0329503
    106.6895 -196 -81.3520030845096+87.7199856736267i 119.636718
    106.8115 -180 -194.073996875666-47.2832424638792i 199.7508981
    106.9336 -158 -11.5335905829742+119.137497617249i 119.6944738
    107.0557 -132 -3.97651586244219+55.764782872332i 55.90638324
    107.1777 -103 -59.8528126674513-30.0904737514331i 66.99101279
    107.2998 -76 34.4465593097798-51.6946234190628i 62.12004136
    107.4219 -52 122.97065151563+106.071416683648i 162.397434
    107.5439 -33 35.3820190802005+49.9780550857398i 61.23473903
    107.666 -14 -49.194218692905-102.545918198131i 113.7353792
    107.7881 -14 63.0386035332498+27.7780360460856i 68.88747943
    107.9102 -19 -230.562571711397+69.9825312276724i 240.9494846
    108.0322 -28 58.9175757715381+173.237280145411i 182.9820646
    108.1543 -38 -90.1954953078204+41.5469225808598i 99.30445181
    108.2764 -47 113.350134093622+10.5923176630456i 113.8439726
    108.3984 -55 -42.189610886233-53.7329994673319i 68.31689761
    108.5205 -60 -123.630524281224-77.6611638428541i 145.999188
    108.6426 -63 16.14260045868-119.138167839917i 120.2268131
    108.7646 -62 66.0778418801606+29.3020771514892i 72.28342073
    108.8867 -59 -50.5011490416918+62.6968550372148i 80.50628352
    109.0088 -55 -118.580917387047+45.1172311180214i 126.8739473
    109.1309 -50 72.1669554004474+89.5390656395835i 115.001364
    109.2529 -46 -48.2561026103698+124.217843723447i 133.261863
    109.375 -41 145.974964176367-81.4163771635436i 167.144598
    109.4971 -37 -119.689684872586-46.9191927894573i 128.557502
    109.6191 -29 7.13328604312671-128.56169819695i 128.7594424
    109.7412 -24 -66.4673207907413+100.974714115342i 120.8875412
    109.8633 -18 -50.673477754513+24.4271543907687i 56.25377516
    109.9854 -10 3.92210366547428-96.6676965477306i 96.74722969
    110.1074 -4 103.627719503683+40.7851205955537i 111.3648522
    110.2295 0 45.5967165235415+128.533194638167i 136.3812402
    110.3516 2 26.6865888477253+57.1663583613961i 63.08856119
    110.4736 0 69.7874344874071-109.541529442565i 129.8831501
    110.5957 -6 -117.100481431539+67.0785769681716i 134.952059
    110.7178 -20 -87.8563739619958-160.090072030387i 182.6131803
    110.8398 -37 -195.129942668135+121.953709042968i 230.105197
    110.9619 -56 -6.64383100981328-50.0349651283636i 50.47413423
    111.084 -73 84.4456627581338+102.172073982i 132.5526411
    111.2061 -83 -40.4487338490625-59.6953945280135i 72.10853069
    111.3281 -85 -49.5506123213678-115.958941481515i 126.1020987
    111.4502 -76 66.7156589972941-56.1745646482637i 87.21559992
    111.5723 -28 18.962105508282+83.0894549805172i 85.22569433
    111.6943 3 -104.658154489462-3.50410778071011i 104.7167994
    111.8164 33 -126.069397098052-26.7539764653577i 128.8769496
    111.9385 54 -112.540232331839+241.017268305465i 265.9974201
    112.0605 64 -99.4933585275748-61.2103839277128i 116.8145517
    112.1826 57 38.0817723953353+173.510271096848i 177.6401857
    112.3047 34 -38.4163330266838-228.018071928255i 231.2316063
    112.4268 -2 -222.598374427895+141.777071301348i 263.9143313
    112.5488 -48 -263.572643477916-20.0769436439804i 264.3361913
    112.6709 -94 -94.4432926595393-36.9483759735077i 101.4135988
    112.793 -135 96.7679644874612-42.1943461201381i 105.5670488
    112.915 -164 75.4315850691739+81.8800820384951i 111.3295642
    113.0371 -175 -122.393714499111+56.8523983311459i 134.9533866
    113.1592 -166 -16.9724714291367+109.961580075349i 111.2637132
    113.2813 -138 174.816326703119+84.0333054960763i 193.9648023
    113.4033 -96 -144.460387669196-181.927963346882i 232.3070973
    113.5254 5 11.646092683592-48.4893960381565i 49.86835673
    113.6475 51 -232.995703910784-161.240399212374i 283.3468976
    113.7695 83 88.0116047345969+41.4172081040539i 97.26987044
    113.8916 99 -84.5414531564502+95.5444127140898i 127.5773965
    114.0137 97 -0.583880518920893-172.80244513232i 172.8034316
    114.1357 81 49.6820098814338-123.393481896304i 133.0197485
    114.2578 54 57.6630108664951+41.667593259988i 71.14218966
    114.3799 24 -39.4322932780202+65.5962603979326i 76.53610345
    114.502 -4 -32.356717867832-133.380922607983i 137.2495089
    114.624 -25 -81.4879361053046-143.626296013126i 165.1326637
    114.7461 -35 -372.42516188007+42.511251717277i 374.8435777
    114.8682 -32 -1.49827823936396-67.2415340500889i 67.25822432
    114.9902 -19 -135.819414806903-25.8190860448927i 138.251722
    115.1123 0 187.29288858891-155.04058147651i 243.1382488
    115.2344 23 4.05340386138981+7.50991257286491i 8.533983227
    115.3564 45 -137.923047869469-138.8920812753i 195.7390543
    115.4785 74 63.9496136350781+7.41258043202379i 64.3777868
    115.6006 77 5.79331905058277+80.3522202684872i 80.56079597
    115.7227 75 -110.185913780608+67.9073673359982i 129.4308546
    115.8447 70 -138.435119608141-26.0247306916668i 140.8601042
    115.9668 63 -55.7540407656724+31.3465870414729i 63.961876
    116.0889 56 -2.11495347421694+116.212450443176i 116.2316939
    116.2109 53 26.4928917780062-316.444826491909i 317.5518879
    116.333 52 4.56278640957895-62.1336175651488i 62.30092657
    116.4551 54 69.8174147773068-296.728933571586i 304.8319724
    116.5771 58 38.8858976843995+82.7271910168673i 91.41061849
    116.6992 62 -145.20992894863+20.3980665671586i 146.6356184
    116.8213 65 140.038281470048-42.6373759939378i 146.3853343
    116.9434 67 69.141290371061+68.157972076284i 97.08772936
    117.0654 66 -92.3437644671452-0.313816218569428i 92.34429769
    117.1875 64 -71.6722186708508-99.4655170855437i 122.5981077
    117.3096 60 -29.9525991633621-137.02521519447i 140.2607137
    117.4316 55 -5.99345130910791E-002-195.841133136514i 195.8411423
    117.5537 57 -226.090102919659-276.848685900858i 357.4380079
    117.6758 63 40.297851870245+3.93270136936269i 40.48929495
    117.7979 73 109.179702314358-59.9708299698251i 124.5660782
    117.9199 87 126.079493673246+104.273516238675i 163.6123617
    118.042 103 -81.1416054981031+86.2849186883497i 118.4442795
    118.1641 119 -231.192337207691-124.342362988252i 262.5088951
    118.2861 132 -13.8414687996036-60.0824211460983i 61.65617235
    118.4082 142 -141.586340953871+41.491487912413i 147.5406233
    118.5303 146 -173.169321747239-935.273528480978i 951.1699044
    118.6523 145 -44.6295618912328-28.9054333520472i 53.17256691
    118.7744 140 -54.1581803776908-92.2297497078086i 106.9552955
    118.8965 131 -994.236386065997-381.035755473025i 1064.750787
    119.0186 120 234.690217518468-23.0543149043907i 235.8198457
    119.1406 109 36.9644642340586-131.164185158145i 136.2733102
    119.2627 98 505.721812028657-144.562855369217i 525.9781082
    119.3848 78 -40.2360629004343-43.3456170593831i 59.14206013
    119.5068 69 -117.049885945772-78.1572613293532i 140.7452781
    119.6289 61 118.152174400234+330.674546925153i 351.1489603
    119.751 55 -11.9600251554782+74.3384584055292i 75.29441281
    119.873 49 -17.9340335555579-137.843360588448i 139.0051136
    119.9951 44 -130.240137628216-170.992683909125i 214.9441588
    120.1172 42 -39.4048865592739-102.156573134972i 109.4929702
    120.2393 41 81.8637990064556-58.529902011263i 100.6351381
    120.3613 42 250.258902172079-413.934630765227i 483.7058989
    120.4834 44 -103.307235854439-174.335034354916i 202.6452299
    120.6055 47 81.3215581295367-9.29401960818763i 81.85092924
    120.7275 48 -84.0541729595977+252.882334271499i 266.4856075
    120.8496 48 26.8145020871931-86.6523919983805i 90.70641963
    120.9717 45 208.742843790535-115.523156139022i 238.5773972
    121.0938 40 -176.335061735394-115.448722990683i 210.7663674
    121.2158 32 -178.506619618427-120.670306926406i 215.4667868
    121.3379 12 28.4145998880895-95.22492897775i 99.37392307
    121.46 0 38.1234525135866-162.917298927519i 167.3183909
    121.582 -11 -2.16208967645404-187.943723256676i 187.9561591
    121.7041 -22 -121.226864867886-154.163343465554i 196.1180492
    121.8262 -32 91.6600368259386+169.079394051904i 192.3262952
    121.9482 -39 290.130152814745-2.99404618443045i 290.1456012
    122.0703 -45 6.34356460853267+41.4248462270494i 41.9077403
    122.1924 -50 -189.384531160709-135.623024154286i 232.937986
    122.3145 -53 -22.4274579443023-79.6030955778064i 82.7021384
    122.4365 -55 -49.9754495991939+130.938216379708i 140.1512115
    122.5586 -58 57.2945416783099-66.8778160721887i 88.06421968
    122.6807 -61 129.547872195897-208.801027424273i 245.7244804
    122.8027 -62 28.347164911922-25.4743637533349i 38.11174317
    122.9248 -61 125.184242882552-13.7671930564219i 125.9389942
    123.0469 -56 127.104177915778+9.98068719114118i 127.4954358
    123.1689 -48 252.867303042887-162.807616624244i 300.7460606
    123.291 -24 56.9774751279771-185.056569641952i 193.6294571
    123.4131 -14 -45.6160358078174+156.109667567562i 162.6377909
    123.5352 -8 5.393605207593-35.9023787802812i 36.30525829
    123.6572 -9 91.1896922412709-131.866615084299i 160.3258062
    123.7793 -15 69.4174800628417+0.295720844926016i 69.41810995
    123.9014 -26 -1.10833198275624-176.753587290666i 176.7570622
    124.0234 -38 52.7795734749651-184.250047435312i 191.660542
    124.1455 -50 130.65916554614-53.8626647187738i 141.3258794
    124.2676 -60 4.55320666750966-176.368824193877i 176.4275881
    124.3896 -67 -17.1648486523422-50.244112623889i 53.09522467
    124.5117 -71 -4.43314163545212-161.748282970409i 161.8090226
    124.6338 -72 35.2335107869169+37.1741843027207i 51.21835863
    124.7559 -71 104.873918014356+139.495241970587i 174.5206613
    124.8779 -69 -55.248646737788-107.567187578405i 120.9260634
    125 -64 -117+86i 145.2067492
    125.1221 -56 99.9405575036675-0.483833686800274i 99.94172867
    125.2441 -34 -69.7413730908859-75.6001693951081i 102.8554555
    125.3662 -22 -22.5405588159647-64.7334314896208i 68.54556108
    125.4883 -10 125.957800029003+35.2883214282431i 130.8076183
    125.6104 -1 -18.0140641903341+25.1441375245834i 30.93111961
    125.7324 1 165.38945346147-11.3928934740534i 165.7813902
    125.8545 0 34.1459574816503-85.6516948215884i 92.20715394
    125.9766 -7 172.790041032471+16.3588874055708i 173.5627019
    126.0986 -17 195.156009481038+50.0668405879571i 201.4759454
    126.2207 -26 -63.3486803286019-58.2425636564553i 86.0537711
    126.3428 -34 111.008740123117-76.3664042791729i 134.7396307
    126.4648 -39 72.1468602787521-20.0363158426679i 74.87738911
    126.5869 -41 -45.8811622623675+40.9337314085155i 61.48700202
    126.709 -41 135.874706187641-11.061318395248i 136.3242038
    126.8311 -40 222.801060436976-177.577520599657i 284.9106673
    126.9531 -39 12.855252838954-90.4742781043019i 91.3829991
    127.0752 -40 33.8831868183871-146.495624507107i 150.3630219
    127.1973 -41 36.963002644826-59.1495288473914i 69.74905252
    127.3193 -38 205.560558743551-98.7343551796822i 228.0430139
    127.4414 -33 95.9465014228894-94.8767019619325i 134.9345016
    127.5635 -24 -175.932499917293+85.9620942787772i 195.8104343
    127.6855 -13 128.926619312958-6.30078819802711i 129.0804908
    127.8076 -1 132.968668974156+166.912928245165i 213.4024193
    127.9297 9 16.0560541183701-46.4839993539667i 49.17884779
    128.0518 17 -8.12537979356242-245.327724395411i 245.4622459
    128.1738 23 -128.632315697788-79.6605492790242i 151.3012748
    128.2959 27 68.3641535366312+111.421697056131i 130.7228062
    128.418 30 123.956044362666-71.639005491654i 143.1686001
    128.54 31 151.227774363844-61.5490690823894i 163.2731688
    128.6621 30 80.7505890352361-30.7567203820777i 86.40968394
    128.7842 27 126.747452660942+151.974043362357i 197.8914516
    128.9063 22 120.798655910404+238.774480063806i 267.5921666
    129.0283 13 119.034270946223-240.634719454492i 268.4664334
    129.1504 -13 69.9218635865653-38.8890574296809i 80.00891072
    129.2725 -26 -228.646063156193+95.2290755213288i 247.6844747
    129.3945 -35 17.2030532399103+1.88695870007211i 17.30623165
    129.5166 -39 162.89974101935+40.1953223246232i 167.7855463
    129.6387 -38 345.178624730177-157.684864889668i 379.4901838
    129.7607 -30 24.7761119944394-69.7607913084692i 74.02988403
    129.8828 -19 125.404660520767+54.1098714703525i 136.5804051
    130.0049 -6 244.234433833418+135.107701015794i 279.1138648
    130.127 4 200.771276677357+3.21901098338188i 200.7970806
    130.249 12 5.34337908976031+42.4236601212686i 42.75884281
    130.3711 16 -307.465930034052-78.9302154083983i 317.4354691
    130.4932 17 73.4549520305883+126.250858410513i 146.0647433
    130.6152 16 56.9723051000058+1.59393021864886i 56.99459766
    130.7373 15 136.417205456401-443.847635865952i 464.3386456
    130.8594 13 93.064276578026-93.1366135516116i 131.663922
    130.9814 11 96.8168797798553-117.761203592058i 152.4506782
    131.1035 4 448.940003255287+668.953543528517i 805.6338932
    131.2256 -2 141.52108589225+11.0397026642067i 141.9510225
    131.3477 -10 25.8523992090545+25.7520183383816i 36.48990262
    131.4697 -19 733.038988063749+57.4824158906246i 735.2893214
    131.5918 -27 -2.35685125895492+179.492284848016i 179.5077577
    131.7139 -33 -18.8034955359456+0.902730944528905i 18.82515252
    131.8359 -37 137.793877498026+189.023547095477i 233.9167673
    131.958 -39 -79.6318451472225+60.1910399349282i 99.82079969
    132.0801 -40 -57.4740958267012-111.862280907368i 125.763435
    132.2021 -41 38.3862755204497-74.092135176368i 83.44549505
    132.3242 -42 41.1813759668049-44.4534575307656i 60.59715846
    132.4463 -44 147.363298304354+183.58408015968i 235.4125234
    132.5684 -45 162.453450893414-7.84133510240063i 162.6425843
    132.6904 -43 164.486298979428+82.5209999638315i 184.0256993
    132.8125 -38 103.598771081873+82.5090058789311i 132.4403315
    132.9346 -29 -2.5840921934207+134.377470904712i 134.4023148
    133.0566 -5 -138.157407575784+18.1393946463615i 139.3431265
    133.1787 4 53.2853709552692-143.891339564395i 153.4406998
    133.3008 10 -35.8147284616697+101.501940592198i 107.6352113
    133.4229 10 -65.786342259943-43.6155124087356i 78.93133567
    133.5449 4 266.39913320581-8.65926801229497i 266.5398302
    133.667 -6 102.261606823493+1.60528437680455i 102.2742058
    133.7891 -19 269.508686035972+0.972465602130825i 269.5104405
    133.9111 -32 -46.0729858798543+5.1640477795135i 46.36148636
    134.0332 -43 14.8512033318418+44.9103359114207i 47.30218295
    134.1553 -52 -3.88052468190924+127.162707730875i 127.2219034
    134.2773 -56 -27.6803567997604+77.100521194907i 81.91881665
    134.3994 -58 -37.2865475176689+13.9352507213502i 39.8055001
    134.5215 -58 0.689584986996977-164.02940115175i 164.0308507
    134.6436 -55 135.709835936466+82.8159484259953i 158.9831465
    134.7656 -51 23.5351400976924+10.384174038483i 25.72418881
    134.8877 -44 215.348800402939-162.576175820343i 269.8260899
    135.0098 -23 -17.6660516769452+115.820005703569i 117.1595626
    135.1318 -11 38.8451567452367+15.1730853357567i 41.70334185
    135.2539 0 -32.9421235499939+288.125809563276i 290.0028718
    135.376 7 111.969560674306+62.4016374617414i 128.1840352
    135.498 10 97.9913904973072+54.3104247822378i 112.0354178
    135.6201 7 -72.906885776298+50.8482194449823i 88.88731864
    135.7422 0 -21.7502466089842+5.95697367009146i 22.55124748
    135.8643 -8 65.8910269525172-56.8805432459614i 87.04610062
    135.9863 -18 217.855684766413-39.7677837814389i 221.4555847
    136.1084 -25 -44.4526347576083+67.747521310507i 81.02939825
    136.2305 -30 -30.5822499725742-60.7809520003172i 68.04115034
    136.3525 -32 86.4422666024816+273.775390188126i 287.0979445
    136.4746 -33 68.7457440858013+20.2035768262274i 71.65306586
    136.5967 -32 124.473588759727+93.7905693363621i 155.8536018
    136.7188 -33 -90.7911424699892-135.063924437092i 162.7430344
    136.8408 -35 -43.6237837110342+24.6255045156788i 50.09441065
    136.9629 -39 -12.8839880738373+132.209231423523i 132.83553
    137.085 -37 -21.6376585154577-44.8385724319387i 49.7864022
    137.207 -33 26.1243040209348-106.423050908051i 109.5825945
    137.3291 -25 27.3920162621042+86.4758586065264i 90.71051029
    137.4512 -15 58.9604016589381+210.793363201365i 218.8839211
    137.5732 -4 88.122365250405+43.5521340987507i 98.29720058
    137.6953 5 179.344599232905+46.4049232354109i 185.2509168
    137.8174 12 -148.823468360554-56.8012217669763i 159.2947065
    137.9395 16 23.1197302730231+146.028431829974i 147.8473024
    138.0615 18 -144.561238965151+74.1217484754805i 162.456103
    138.1836 18 77.0061869011946+82.7002393337512i 113.0012496
    138.3057 18 2.48186718643372+54.4709242199074i 54.52743576
    138.4277 16 -63.2185912545022+8.18145220692452i 63.74579547
    138.5498 13 75.1656696313496-9.11302463523885i 75.71608224
    138.6719 9 106.058522092599-2.55646626573769i 106.0893285
    138.7939 3 77.3118110443718+36.3366771134212i 85.42523181
    138.916 -13 -77.3283412424997-14.9765035587782i 78.76527165
    139.0381 -21 40.8720963019734+9.88664817623685i 42.05085098
    139.1602 -26 -98.3047280610954+158.323393560008i 186.3601795
    139.2822 -25 124.5421483855+69.0869541233723i 142.4210446
    139.4043 -19 -49.5053107988353+77.5596434013532i 92.01235831
    139.5264 -9 109.117461097757-42.4176800778997i 117.0721141
    139.6484 3 -30.4340572112159-34.6846009083691i 46.14383359
    139.7705 16 -91.0020124958935+14.7207700612884i 92.18496271
    139.8926 27 -22.9953036956191-74.800026970723i 78.25489139
    140.0146 34 79.7104854050399+23.2828930336774i 83.04128245
    140.1367 37 -11.337005720024+32.7296963867845i 34.63756233
    140.2588 37 -81.1152953953901+82.6500097122172i 115.8046426
    140.3809 35 83.6346741672129+29.3030255355477i 88.61955782
    140.5029 32 22.3996340773202+143.950737665336i 145.6830755
    140.625 28 68.3674005785773-69.3330595039363i 97.3713233
    140.7471 25 -124.244835970821+27.9528340631139i 127.3504621
    140.8691 16 -18.9273061551267-72.0240713664137i 74.46952245
    140.9912 8 -20.4729253413762+100.351050522936i 102.4181332
    141.1133 0 -10.1386922870655+46.3172694927595i 47.41394873
    141.2354 -9 -19.3606369726086-52.7879983778286i 56.22639093
    141.3574 -17 72.7683762168866-30.2207677668715i 78.79423444
    141.4795 -23 66.8206720187597+109.00322032675i 127.8542305
    141.6016 -27 31.1509329919702+6.70203537016019i 31.86373965
    141.7236 -29 -3.17355312950385-112.532599483836i 112.5773396
    141.8457 -31 -33.316075020311+98.1357135457554i 103.6367653
    141.9678 -33 -144.960800281047-109.390795196344i 181.6039088
    142.0898 -36 -81.8636945893891+196.845273274277i 213.1894137
    142.2119 -40 -30.0874821912326-4.01248219233685i 30.35385639
    142.334 -44 83.4381401396921+61.3321547307641i 103.554606
    142.4561 -45 -32.9565542117382-17.8495231404131i 37.47986048
    142.5781 -43 -67.0953642006109-52.755061339937i 85.35153422
    142.7002 -36 18.4240316209398-45.9601289435991i 49.51543591
    142.8223 -13 55.7454564556165+66.88385758516i 87.0689745
    142.9443 -1 -68.6016481459449+39.1806104587627i 79.00193899
    143.0664 6 -91.8951254644217+19.7102719624949i 93.98515258
    143.1885 8 81.914373988018+216.000554760164i 231.0112645
    143.3105 4 -86.6647555806444+56.6977824846876i 103.5635959
    143.4326 -6 143.154169163227+78.5657048814179i 163.2963139
    143.5547 -19 -109.128492828649-126.745739058346i 167.2528335
    143.6768 -33 -53.1836522580291+29.5039384827055i 60.81926712
    143.7988 -47 -181.144763012954+63.6200681959143i 191.9920265
    143.9209 -57 -44.0189080937673+10.6229385692323i 45.28256942
    144.043 -63 42.1505882524477-81.3337146214488i 91.60701515
    144.165 -65 78.5535288379388+32.7981612185846i 85.12564991
    144.2871 -64 -84.3543424949835+115.95918490846i 143.3952149
    144.4092 -59 43.0112882772331+98.6920382852841i 107.6572772
    144.5313 -52 156.112370105555-8.09219120213743i 156.3219615
    144.6533 -43 -167.901478442896-39.3249509378942i 172.4452326
    144.7754 -25 -2.15904975246776-44.619497498068i 44.67170304
    144.8975 -19 -223.626663125795+28.0682263011274i 225.3812543
    145.0195 -17 71.4463539688324+18.2827503962007i 73.74849461
    145.1416 -19 -6.91329356238555E-002+100.3885480682i 100.3885719
    145.2637 -26 -39.5097437695789-133.801537152865i 139.5129786
    145.3857 -35 0.228392364114178-88.9748319385912i 88.97512507
    145.5078 -43 69.2151228186979-3.90551464019788i 69.32522103
    145.6299 -44 43.6395882255653+81.3472737871748i 92.31355596
    145.752 -34 -77.3498937443466-23.0342284830462i 80.70676393
    145.874 -10 -131.354803004857-43.3703943907347i 138.3295897
    145.9961 27 -230.547928983344+196.165304935253i 302.7097197
    146.1182 77 -5.3218319379985-12.1323917517249i 13.2482763
    146.2402 132 -97.0183295028196+57.7739836599258i 112.9176224
    146.3623 187 81.3940067833527-182.762604953858i 200.0678738
    146.4844 234 42.696390432625+3.28849996658935i 42.82284423
    146.6064 267 -129.337713513572-40.4171806739708i 135.5056922
    146.7285 276 71.9155419993991-59.5340623357828i 93.36032219
    146.8506 253 63.385566890981+73.1939639032004i 96.82503004
    146.9727 214 -7.72481743239656+103.37337689318i 103.6616026
    147.0947 163 -104.053486422004+110.228842562999i 151.5833954
    147.2168 106 -39.7629522817019+36.3373278897864i 53.86551561
    147.3389 47 27.324721141684+76.5761485274031i 81.30526987
    147.4609 -9 -101.758076429982-225.796791677805i 247.6669079
    147.583 -59 -54.7639432394227-57.9166409228724i 79.70838585
    147.7051 -98 -67.0353497697475-196.029789405838i 207.1748451
    147.8271 -124 76.6539166915719+53.0752854863789i 93.23523408
    147.9492 -137 -69.3466659988368+80.1888630070159i 106.0151585
    148.0713 -136 73.8139134030232-108.627718241561i 131.3334496
    148.1934 -127 84.5263931564013+55.8738093209659i 101.324201
    148.3154 -111 -60.8566533486452+58.6954548447816i 84.54991825
    148.4375 -95 -95.7366592160789-36.1306388877496i 102.3275671
    148.5596 -81 -83.1623633070609-47.835210833121i 95.93844936
    148.6816 -69 -63.7134409740349-144.993777788583i 158.3748659
    148.8037 -70 -270.914521177285-96.58771076339i 287.6175649
    148.9258 -72 11.3317654760716-4.63536186688241i 12.2431813
    149.0479 -72 68.7425529780849-104.114092539312i 124.7609028
    149.1699 -69 152.408603415853-10.9715159123478i 152.8029992
    149.292 -61 17.438981477374+98.4303359738048i 99.96323882
    149.4141 -49 -202.06030873407+29.488678786222i 204.2007604
    149.5361 -36 -1.487406902331+7.74579119404934i 7.887310094
    149.6582 -23 -74.7486475642148+107.071246975142i 130.581822
    149.7803 -13 -493.448662050393-528.26260171395i 722.8782459
    149.9023 -7 -10.4067913042239+0.35625348563462i 10.41288729
    150.0244 -4 -73.4745489606915-48.2455533637776i 87.89847988
    150.1465 -2 -772.284687767755+161.085594803877i 788.9057027
    150.2686 -3 135.043384121368-129.531953865906i 187.1236026
    150.3906 -3 21.8068676408839-94.4667494969019i 96.95105073
    150.5127 -5 265.147091443442-328.926464805947i 422.4873955
    150.6348 -9 -60.0991776197162-9.57998896371816i 60.8579275
    150.7568 -13 -118.54192098457-19.8403015631785i 120.1907842
    150.8789 -20 240.837743678556+173.869551278186i 297.0411413
    151.001 -27 53.4412708884232+21.8547221079016i 57.73732166
    151.123 -33 -98.386169154535-68.6565882751081i 119.9731861
    151.2451 -38 -144.321420923343-84.7462931910381i 167.363696
    151.3672 -41 -41.3774431130101-34.5145017696562i 53.88268396
    151.4893 -43 24.4565418415677-81.59195852439i 85.17846051
    151.6113 -44 -36.1749329850157-409.780129098882i 411.3737716
    151.7334 -46 -118.677402196381-75.6950933101475i 140.7624699
    151.8555 -48 48.4072819155201-34.6943660294376i 59.55639325
    151.9775 -50 63.4243082011019+222.179508461565i 231.0549217
    152.0996 -51 14.0424850690168-72.8328184444158i 74.17419248
    152.2217 -49 119.97169646047-190.454567874323i 225.0914267
    152.3438 -43 -184.72767876707+22.3500694955234i 186.0748261
    152.4658 -33 -163.364843151824+23.2014855890133i 165.0041845
    152.5879 -6 20.4122701073113-112.44661918553i 114.2843075
    152.71 5 -3.33500140595244-102.647536549014i 102.7016991
    152.832 13 -27.4639155947141-119.502998916109i 122.6182426
    152.9541 14 -159.380784635244-18.4333743055069i 160.4432105
    153.0762 8 163.859335975282+88.9928894557026i 186.4661266
    153.1982 -2 193.13950795761-118.265176270454i 226.4718999
    153.3203 -15 41.4775369061328-3.69940943304094i 41.64218652
    153.4424 -28 -205.686824704008-4.92262546664691i 205.7457219
    153.5645 -39 -30.8454623452887-69.0655821517884i 75.64057896
    153.6865 -47 33.1792823951862+127.001399718982i 131.2639338
    153.8086 -51 -1.98351135333841-83.7113890975255i 83.7348851
    153.9307 -51 9.38013778233596-146.043552156303i 146.3444776
    154.0527 -49 -8.80326859834731-52.2645907682146i 53.00080175
    154.1748 -45 101.199902601046-72.718075866672i 124.6167679
    154.2969 -39 77.010849319664-42.0569815452984i 87.74657036
    154.4189 -31 81.3566597659884-192.288450774376i 208.7911741
    154.541 -8 -32.7787074797964-119.182761547436i 123.6081483
    154.6631 3 28.9484714763344+122.458827979229i 125.8339324
    154.7852 15 -32.9387393722375+3.24303904671289i 33.09800377
    154.9072 23 -13.7042651915454-108.641036758143i 109.5019715
    155.0293 28 87.3452136442373-24.29992345864i 90.66241022
    155.1514 28 -104.123283580854-94.4332858556159i 140.5677903
    155.2734 23 -46.2344146204197-138.199149198991i 145.7279175
    155.3955 16 54.7042238808439-81.0069947380878i 97.7480706
    155.5176 10 -60.2896285125422-117.656918705518i 132.2043487
    155.6396 6 -14.8447839307725+4.49608957800632i 15.51071989
    155.7617 5 -79.3737948947359-91.8615706634107i 121.4032433
    155.8838 6 66.0895303410514+4.89598446488361i 66.27063214
    156.0059 9 140.87193671761+18.6792074456225i 142.1049448
    156.1279 12 -78.2810805369058-35.2945965206459i 85.86987897
    156.25 14 -18.6462449961312+107.99058545351i 109.5885441
    156.3721 14 99.1248756335028-37.1633250084996i 105.8624282
    156.4941 11 -85.9348511339791-39.1482173880584i 94.43188849
    156.6162 12 -22.8102305671407-30.0072957602115i 37.69276346
    156.7383 17 91.7636423967831-22.9991912254805i 94.60194957
    156.8604 24 7.38230775480249+23.1253479811534i 24.27509397
    156.9824 35 129.045582556651-60.1173223084878i 142.3617042
    157.1045 48 -27.5919220219226-88.2985937414794i 92.50922018
    157.2266 61 121.040629595563-97.6876041921562i 155.5432481
    157.3486 72 205.13577265053-68.2919365087268i 216.2047035
    157.4707 81 -46.4415895208521+0.441582759439548i 46.44368883
    157.5928 88 77.7620833552361-77.070485241828i 109.4842514
    157.7148 94 32.4782117847262-37.2153146298055i 49.3944722
    157.8369 98 -31.1551733692689+75.9207188883624i 82.06461103
    157.959 99 94.3492488315448-71.5488161213639i 118.4103621
    158.0811 99 57.5624074967459-199.546168463419i 207.6827005
    158.2031 95 -20.470479337792-50.2086626332718i 54.2213088
    158.3252 88 -28.3144687138674-114.23070621537i 117.6875668
    158.4473 65 39.7333297576718-47.8862299786724i 62.2240188
    158.5693 53 130.435029667826-201.518362756672i 240.0478025
    158.6914 44 42.0666126663372-96.1870949089482i 104.9836041
    158.8135 40 -77.1289890505808+145.928161262447i 165.0572907
    158.9355 43 104.773472575847-47.5692580064289i 115.0665671
    159.0576 50 202.08492840852+61.0151720578818i 211.095167
    159.1797 62 10.021756089991-18.546749778984i 21.08121255
    159.3018 76 -118.15601000721-129.290070658213i 175.1478378
    159.4238 88 -113.740160265628+8.69563176763609i 114.072074
    159.5459 98 115.409736579425+71.7016266421079i 135.8695351
    159.668 104 64.4402199245775-105.939963576778i 123.9992654
    159.79 107 83.0605103367996-96.0819192271477i 127.0070218
    159.9121 108 41.8291663025672-34.3445008291055i 54.12230493
    160.0342 109 157.599496798143+73.2134210634611i 173.7751605
    160.1563 110 179.087190474532+101.489123187378i 205.8452426
    160.2783 111 -1.65902007131672-224.479304760882i 224.4854352
    160.4004 109 36.7646874841549-42.7799853368384i 56.407175
    160.5225 105 -104.120869927879+181.970242254948i 209.652867
    160.6445 97 19.3255764138789-24.828559490815i 31.4632368
    160.7666 88 139.083153806057-26.2070206875152i 141.5306737
    160.8887 78 180.208817135331-278.897757800078i 332.0529733
    161.0107 69 6.83641578313123-40.0097389866377i 40.58960205
    161.1328 61 90.3145513574982-12.7649365312858i 91.21218006
    161.2549 55 199.415228984371-24.4868250387908i 200.9130114
    161.377 49 143.757837650655-55.1072418330934i 153.9581891
    161.499 43 10.7810205196277+30.1318163482318i 32.00244928
    161.6211 36 -248.941535118315+81.099813221179i 261.8187686
    161.7432 29 142.303590427996+81.3818960524757i 163.9308539
    161.8652 22 50.8868277281552-5.44506494346902i 51.17731889
    161.9873 18 -104.786196571531-354.384745237066i 369.5520189
    162.1094 17 46.2102155033522-91.642870440201i 102.6343009
    162.2314 20 -12.3552702412427-128.341453329612i 128.9347949
    162.3535 33 666.251857432319+312.800101226764i 736.0267936
    162.4756 39 94.579957441558-60.6198302373097i 112.3393616
    162.5977 41 52.3748240699404-12.5489393420448i 53.85720077
    162.7197 37 582.161170144634-284.144526180094i 647.8037819
    162.8418 27 101.267785665987+124.017604661966i 160.1109949
    162.9639 13 21.4121926356182+53.8929626604853i 57.9908046
    163.0859 -4 182.215611752857+85.9873350840558i 201.4853617
    163.208 -21 -32.5415883998295+97.8928538003458i 103.159904
    163.3301 -36 -92.1408970253286-21.8761362720973i 94.70221879
    163.4521 -48 -9.17377650413437-61.6225006319865i 62.30161121
    163.5742 -57 12.4794992972825-63.0727842135542i 64.29552093
    163.6963 -62 246.251346287049+32.1373412219097i 248.3395543
    163.8184 -64 121.056290284246-86.6315148280465i 148.8611594
    163.9404 -65 162.371359899181-9.66791636315814i 162.6589288
    164.0625 -63 114.390876108687-30.0830065443331i 118.2804287
    164.1846 -58 76.5343950895632+87.9709585218234i 116.6036156
    164.3066 -41 -51.7003376460974+70.7180990308559i 87.60122398
    164.4287 -30 57.8350706605383-148.127373699188i 159.0176539
    164.5508 -20 64.085916884915+87.5879064444037i 108.5294711
    164.6729 -12 -42.9876874702233-6.55709334797854i 43.48490252
    164.7949 -8 188.545295558465-114.815002710472i 220.7528331
    164.917 -10 58.3774941360142-38.236241766533i 69.78496977
    165.0391 -17 212.378442420168-131.505511634064i 249.796522
    165.1611 -26 -31.6874141738707+44.2736308747988i 54.44489515
    165.2832 -35 31.0427599761957+9.8981875795192i 32.58261905
    165.4053 -43 77.394022329893+136.704128201797i 157.0918628
    165.5273 -47 17.896247233552+89.4609739972427i 91.23344526
    165.6494 -49 -15.0277806620235+56.5286045651724i 58.49202788
    165.7715 -49 -66.5532905052494-90.0292471553988i 111.9580538
    165.8936 -48 155.634492555058+8.72860792598811i 155.8790681
    166.0156 -48 -3.27331283349771+23.0866537053076i 23.31755039
    166.1377 -50 55.9763342107827-187.332563105819i 195.5168514
    166.2598 -55 58.8480405382903+90.2544450719828i 107.7448687
    166.3818 -55 39.3364127775505-2.37654192160346i 39.40813776
    166.5039 -51 107.248328511766+232.556712264505i 256.0953502
    166.626 -45 100.857217516682+27.7769532219599i 104.6123198
    166.748 -35 90.9916398929494-22.0102500970625i 93.61586212
    166.8701 -23 -47.4142646710792+85.7946065657008i 98.02462451
    166.9922 -11 -22.5310213541502-6.24825564142621i 23.38135201
    167.1143 -1 42.3012598749738-85.4837118621902i 95.37746894
    167.2363 6 118.112397348861-96.6166631807884i 152.5952752
    167.3584 12 -16.4652310007685+76.5711364892808i 78.32140688
    167.4805 15 -29.9256155878602-30.5032627109939i 42.73162183
    167.6025 17 208.65445137985+163.93786649383i 265.3531687
    167.7246 17 79.2594179445573-36.8185820118536i 87.39372583
    167.8467 15 142.731613466975+32.6013528185484i 146.4075192
    167.9688 9 -116.043630643961-47.0593557030127i 125.2226304
    168.0908 0 -60.20334766326+41.330483333259i 73.02500888
    168.2129 -25 34.6174481478677+86.7330326667671i 93.38622314
    168.335 -39 -37.1674179015199-12.9389028161536i 39.35520499
    168.457 -49 -10.1738766596245-62.428866429348i 63.25243972
    168.5791 -55 84.7641169095536+62.1036026205047i 105.0800313
    168.7012 -54 138.158015491833+117.584508693329i 181.4214814
    168.8232 -48 84.9279135057122+17.6374404578234i 86.74001267
    168.9453 -38 151.365267059805-40.7135956642266i 156.7451465
    169.0674 -25 -132.128875381692+19.1334805193328i 133.5070402
    169.1895 -14 57.8171422784367+45.2352524961422i 73.41014923
    169.3115 -6 -66.0364938576684+109.552070379817i 127.9158889
    169.4336 -1 76.9377518303319+44.8073276077978i 89.03434317
    169.5557 0 79.1339509928143+49.5046231480775i 93.34286214
    169.6777 0 -36.7759017688998+27.1674508780603i 45.72239427
    169.7998 0 13.7146368973867-51.5918065742479i 53.38357211
    169.9219 0 103.837284655069-6.51729767953381i 104.0416112
    170.0439 1 71.9994539786287-4.69122610360722i 72.15212385
    170.166 0 -65.686240605845+46.7646578151481i 80.6325953
    170.2881 -4 51.2848847351443-18.5087644189779i 54.52259864
    170.4102 -10 22.5579450301063+172.594148853505i 174.0620611
    170.5322 -19 116.625829791815+9.74469510083944i 117.0322317
    170.6543 -27 -12.7436365830759+78.190938715228i 79.2226178
    170.7764 -34 56.4862294465391-95.6078779143309i 111.0475593
    170.8984 -39 -46.0561287866336+5.21072094583256i 46.34995805
    171.0205 -43 -42.7063559340104+91.4527333637153i 100.9328256
    171.1426 -45 -25.0228534936616-22.5595549787849i 33.69089963
    171.2646 -46 70.624900193841-19.8891503829631i 73.3720303
    171.3867 -48 44.524779443454+90.3906431085822i 100.7617206
    171.5088 -49 4.36292014115531+141.442751125278i 141.5100241
    171.6309 -50 105.311376864875-21.3302853889987i 107.4498356
    171.7529 -48 92.9845437902714+80.0098135199224i 122.6690492
    171.875 -42 4.38830950291306-93.6204241390123i 93.72321525
    171.9971 -33 -55.0243924726673+110.02172129981i 123.0140761
    172.1191 -8 -42.987083275047-23.9583452694406i 49.21271824
    172.2412 3 27.2902081797405+84.2739431617906i 88.58246417
    172.3633 10 34.4001147896354+77.6271146194316i 84.90781367
    172.4854 12 -11.5195049087336+8.86818459157458i 14.53766457
    172.6074 6 71.8321474720597-50.5961239786595i 87.86253566
    172.7295 -3 91.5536745710567+39.1890809957978i 99.58845012
    172.8516 -16 39.829486221284+6.71132944603541i 40.39096329
    172.9736 -29 -15.5176588278863-63.5896751841807i 65.4556684
    173.0957 -41 33.8523032918234+136.453880171397i 140.5903263
    173.2178 -50 -95.1704377114416+29.0426787533852i 99.50321303
    173.3398 -55 43.5044791254893+161.256488261568i 167.021839
    173.4619 -57 -16.8906766586028+6.16858081547435i 17.98183381
    173.584 -57 96.8015491843248+10.1253511666608i 97.32965972
    173.7061 -54 -11.6113564327028+0.772333932035068i 11.63701413
    173.8281 -51 -93.0886990520636+16.4129799211109i 94.5245566
    173.9502 -45 -11.5437725005214-64.0182866328651i 65.05074717
    174.0723 -26 68.6110386852126+5.62082605903062i 68.8408913
    174.1943 -15 -17.9446370675699+65.8705218319648i 68.27104544
    174.3164 -5 -73.2380455364338+74.0771922383249i 104.1692936
    174.4385 3 155.368823827601+110.044677952036i 190.3924961
    174.5605 7 -59.2822035085464+76.1157325969324i 96.47789592
    174.6826 5 153.75781813641-3.46498745771603i 153.7968555
    174.8047 0 -122.713883204976-35.7326176366608i 127.8104733
    174.9268 -8 -37.7472978427937+38.7929712642795i 54.12719385
    175.0488 -17 -80.5366841764664+162.242029172192i 181.1315365
    175.1709 -24 -42.7694549269641+68.0745137989601i 80.39506019
    175.293 -29 0.208222947304691-85.4818621906571i 85.48211579
    175.415 -31 85.3481269898116-27.6949535213309i 89.72911028
    175.5371 -31 16.5130881674232+105.021324501754i 106.3116206
    175.6592 -31 79.686736660924+36.8606332092922i 87.79910182
    175.7813 -32 95.3695701037311-62.7510158811847i 114.162362
    175.9033 -35 -111.092177019033+60.936519606998i 126.7072658
    176.0254 -42 -43.0097738564407-6.45090146368232i 43.49085854
    176.1475 -42 -147.930265772821+135.706671274324i 200.7477625
    176.2695 -39 77.4211797628044-51.3244502718659i 92.88831074
    176.3916 -32 58.5297923138012+106.719911290782i 121.7163755
    176.5137 -21 -115.891486041834-68.2507271092231i 134.4953467
    176.6357 -9 -34.8378191400618-57.6469966075566i 67.35614196
    176.7578 3 33.3861581872211+5.52513285480035i 33.84025194
    176.8799 13 80.9408063302777+55.771111377383i 98.29461325
    177.002 21 -58.9925806491036+17.4618784676745i 61.52269314
    177.124 26 -105.768388351668+28.9334456930309i 109.6544402
    177.2461 29 -53.6971304337908+276.105612780604i 281.2786718
    177.3682 30 -30.8277914809514-26.7100324617173i 40.78944179
    177.4902 28 -12.3951230552794+98.3689850838848i 99.14684212
    177.6123 25 -44.2127424040988-184.309576622469i 189.5383513
    177.7344 18 18.3546101047149-4.42480345490007i 18.88042896
    177.8564 9 -130.155974812704+22.645998669571i 132.1113887
    177.9785 -16 58.5743073316995-64.1513916122315i 86.86973308
    178.1006 -29 38.7474540667895+28.1178206981545i 47.874597
    178.2227 -40 38.713550800077+69.4442571464997i 79.50625048
    178.3447 -45 -11.1712671016892+119.175362592509i 119.6978039
    178.4668 -45 3.35861245297273+63.1982943300235i 63.28747652
    178.5889 -40 66.5548632787273+40.9491806020142i 78.14336324
    178.7109 -31 -197.426581184282-71.3735899506298i 209.9319992
    178.833 -19 -57.6430316576873-9.79492140635662i 58.46930463
    178.9551 -10 -136.443827837365-112.606245589666i 176.9098208
    179.0771 -3 89.5997112074706+6.30144371679566i 89.8210245
    179.1992 0 -9.0692361174908+78.7182234853911i 79.23894088
    179.3213 2 23.2563066353221-84.2053317337095i 87.35784848
    179.4434 2 76.7957880073647-2.54719198817202i 76.83801951
    179.5654 2 -8.1060629057105+90.1603954188276i 90.52405845
    179.6875 2 -82.1871764235607+31.6432689388658i 88.06831688
    179.8096 3 -87.2267659722321+3.48664411278151i 87.29642254
    179.9316 3 -112.463780478328-48.8193309194441i 122.6027283
    180.0537 0 -198.292450594993+81.6560810702244i 214.4472232
    180.1758 -6 38.9536315033794-11.5376781821914i 40.62638829
    180.2979 -15 0.36592809276712-71.6471507374055i 71.6480852
    180.4199 -23 96.8724669154366-59.9996817840891i 113.9483947
    180.542 -31 52.5272022113383+44.5781922073199i 68.89355697
    180.6641 -37 -124.218250877078+108.521162615231i 164.9454958
    180.7861 -41 -0.445824069518181+40.713616077394i 40.71605695
    180.9082 -45 -19.4228236647864+83.2985805868106i 85.53303226
    181.0303 -48 -602.981138434837-140.817994021583i 619.2059114
    181.1523 -51 -38.0664406613173+3.75860066721192i 38.25154877
    181.2744 -55 -42.7817617017994+0.168525234852225i 42.78209363
    181.3965 -57 -516.07029704002+475.346089151526i 701.6284315
    181.5186 -57 29.6942882826325-159.298794774623i 162.0427622
    181.6406 -52 -37.1998556490439-83.188524706844i 91.12716337
    181.7627 -44 55.1676837555833-371.901595959289i 375.9711032
    181.8848 -17 -24.0679582767119+4.7724670522468i 24.53656572
    182.0068 -2 -108.610179793191+82.377558557237i 136.3166656
    182.1289 8 265.188419770518+45.0322024271875i 268.9847528
    182.251 13 43.3814411626365+20.4497546436187i 47.95979464
    182.373 13 -107.715015372545-15.1829066488509i 108.7798014
    182.4951 8 -151.511809356779+35.9650909449468i 155.7219193
    182.6172 1 -46.0214257606635+25.4110498048889i 52.5708387
    182.7393 -7 13.0316166373638-43.056953379519i 44.98582295
    182.8613 -14 -194.13995050171-250.278423500301i 316.7484959
    182.9834 -20 -90.2142631733707+25.1183371981014i 93.64584424
    183.1055 -23 11.8651081048831-63.3955676926584i 64.49634713
    183.2275 -24 158.282099020341+156.061239473121i 222.2798536
    183.3496 -24 -37.697075031283-53.7648131136788i 65.66372359
    183.4717 -23 -35.6139343793363-181.152401967371i 184.6200018
    183.5938 -22 -131.322618548916+100.467242123988i 165.3459914
    183.7158 -19 -102.371668355369+94.483163918443i 139.309105
    183.8379 -6 -48.0136023290305-55.9866044422483i 73.75503973
    183.96 2 -79.1341644782122-76.1974080736539i 109.855637
    184.082 10 -66.7822914223964-76.8110435423812i 101.7831561
    184.2041 17 -111.964208086507+56.3588274135107i 125.3487189
    184.3262 20 194.81709025443-10.0751258398327i 195.077438
    184.4482 19 119.172984998405-155.836565614088i 196.1816391
    184.5703 14 26.0005101708495-7.16592350460429i 26.96992749
    184.6924 6 -123.104039436809+104.397141386131i 161.4105562
    184.8145 -1 -63.6954318411071-33.9099421585732i 72.15949151
    184.9365 -7 95.277042665851+105.575051884485i 142.2104301
    185.0586 -11 -17.1128961453288-51.9706665927861i 54.71564128
    185.1807 -14 -87.6721766208906-126.009633966124i 153.5084311
    185.3027 -16 4.20920416941777-23.8671097733486i 24.23543539
    185.4248 -17 63.457842030223-101.560490097136i 119.7557133
    185.5469 -21 42.0986231637582-90.030030224407i 99.3866209
    185.6689 -26 -12.8940950927728-201.235871345172i 201.6485398
    185.791 -38 -94.7961282353563-62.2795528605019i 113.4241977
    185.9131 -40 71.0902962312832+69.870106326525i 99.67779079
    186.0352 -38 -21.662112231451-22.9830905565577i 31.58274146
    186.1572 -32 -38.0494326332531-80.0010322408044i 88.58851214
    186.2793 -21 56.0666102211568-26.0805263462255i 61.83573915
    186.4014 -8 -88.1637209338211-42.7557250716217i 97.98415033
    186.5234 3 -104.773716261556-59.5973543330581i 120.5378624
    186.6455 14 33.4175266135988-119.585678160824i 124.167087
    186.7676 21 -94.844170292158-52.8605318972804i 108.5801661
    186.8896 25 -16.9965396764615-1.67321662639237i 17.07870062
    187.0117 27 -72.3789440641648-30.1630645051751i 78.41251178
    187.1338 27 52.0983611035632+7.07638263316184i 52.57674791
    187.2559 25 119.179007880769-33.8020559448623i 123.8798406
    187.3779 22 -58.1767260611956+32.0657125483742i 66.42846812
    187.5 17 29.8061325481594+115.176623509137i 118.9708374
    187.6221 10 67.3177413353654-65.2363187185436i 93.74142936
    187.7441 -11 -83.5292819674286+23.5734810621317i 86.79199246
    187.8662 -24 -66.038869983288+21.7845037464725i 69.53917566
    187.9883 -36 75.615982881947-59.1073927777586i 95.97635515
    188.1104 -47 43.6197213318516+14.7975941023579i 46.06135995
    188.2324 -54 58.1014249362061-107.243364044692i 121.9709585
    188.3545 -58 -49.6521260761206-27.4457840560196i 56.73274792
    188.4766 -58 48.9930005969185-109.391727641704i 119.8618546
    188.5986 -54 88.1584815713181-132.296753636268i 158.9790832
    188.7207 -47 -8.99734877808116-0.141154247326906i 8.998455954
    188.8428 -36 10.5255483750915-101.732155618575i 102.2752104
    188.9648 -18 22.3014438617992-47.5465767479951i 52.51696258
    189.0869 7 28.2705434253384+42.9859735857212i 51.44917444
    189.209 43 33.1005179235372-134.624197836134i 138.6337583
    189.3311 87 -74.8331800650827-180.438474031669i 195.3408502
    189.4531 139 -36.7217539816425-11.6436114028306i 38.52351103
    189.5752 192 -43.2343586592688-55.4937680778345i 70.34748087
    189.6973 283 -15.6549955428352-38.2164143449859i 41.29858606
    189.8193 307 -28.9313343752188-160.290792300641i 162.880816
    189.9414 311 8.58040152050393-67.4878991013455i 68.03116797
    190.0635 294 12.4844325249587+133.544905759399i 134.1271893
    190.1855 257 41.1297708136537-83.9386127872844i 93.47378651
    190.3076 203 151.424922564082-26.5248106225363i 153.7305199
    190.4297 139 15.580644973714-23.851979635695i 28.48988295
    190.5518 68 -132.216080199351-66.6246158230784i 148.0538122
    190.6738 -2 -66.9113189283362+75.7582849196627i 101.0764183
    190.7959 -68 144.392300612999+7.89471724310326i 144.6079633
    190.918 -124 21.0763712789451-108.33247039811i 110.3636605
    191.04 -168 26.3025599955296-92.8869517529533i 96.53916546
    191.1621 -196 33.8197102587158-71.9123919996311i 79.46801196
    191.2842 -207 156.849010304348-35.5692822331607i 160.831545
    191.4063 -200 203.433161962256-34.7825331358007i 206.3852611
    191.5283 -180 -96.227559537756-138.792952123756i 168.8882079
    191.6504 -114 27.6656755787-72.3916784436596i 77.49803038
    191.7725 -80 40.5140018413297+183.67939615945i 188.0944043
    191.8945 -52 6.24111580352162-10.2016903272965i 11.95934831
    192.0166 -34 120.670299687566-98.0178131576861i 155.4632205
    192.1387 -27 -13.773792473163-294.695363073728i 295.0170747
    192.2607 -29 8.49555149949607-14.2815635373974i 16.61738404
    192.3828 -38 76.0141767292861-64.1045091746465i 99.43612603
    192.5049 -49 150.834375158735-97.9574489568979i 179.8518016
    192.627 -60 63.0334287511243-115.476272341984i 131.5598062
    192.749 -68 25.1699598017753+31.0952933159035i 40.0055514
    192.8711 -72 -132.96020438129+200.317245299394i 240.4275665
    192.9932 -73 106.864203069874-19.1019047533453i 108.558006
    193.1152 -71 66.1031910886463-1.3148960161226i 66.11626747
    193.2373 -67 -232.194792290504-216.969602793483i 317.7896004
    193.3594 -61 4.32334179507296-98.7127945869376i 98.80742431
    193.4814 -55 -60.0655582173889-71.1458714733315i 93.11072071
    193.6035 -37 682.027820363007-113.998856018905i 691.4894699
    193.7256 -26 63.1826662654138-55.0124180228796i 83.77598375
    193.8477 -16 58.8708972688621-4.16971983322853i 59.01837941
    193.9697 -8 282.188353553707-530.596069209274i 600.9679322
    194.0918 -3 144.382317154227+84.5057729344734i 167.294588
    194.2139 -2 18.0638378366158+18.4207699420175i 25.79974811
    194.3359 -5 181.025351226447-21.8386010474336i 182.3378795
    194.458 -11 33.4006494601406+78.2242514535783i 85.05666876
    194.5801 -17 -55.0187035247733+31.6376135325543i 63.46649768
    194.7021 -22 -2.10238155478351-54.5254943071749i 54.56601083
    194.8242 -25 7.54689561853397-33.0883460868632i 33.93809482
    194.9463 -26 189.963495499104-93.411265946584i 211.6879643
    195.0684 -26 52.5161097198075-113.057650598727i 124.6594326
    195.1904 -27 92.8660764429895-89.7706413925743i 129.1622089
    195.3125 -29 107.001183567078-43.1719554349741i 115.3822821
    195.4346 -32 99.7449320659701+40.4686176968468i 107.6418157
    195.5566 -40 -20.1126324191844+63.3162078259955i 66.43387808
    195.6787 -41 -49.083965734912-98.4784833133053i 110.0329377
    195.8008 -38 112.950893433124+51.6544606429452i 124.201802
    195.9229 -30 -30.2585584710071+37.1564696012325i 47.91850993
    196.0449 -19 51.9468868684718-184.323959799032i 191.5040501
    196.167 -5 33.306040102977-6.14995221479616i 33.86907468
    196.2891 7 106.736650970233-170.204798859633i 200.9039228
    196.4111 18 17.6131439525971+80.2168487327133i 82.1277399
    196.5332 27 58.5849896665775+5.55340232987306i 58.84761076
    196.6553 33 108.738049337823+83.0911833038801i 136.8506782
    196.7773 36 77.3135428802135+79.8095081469992i 111.1167922
    196.8994 38 14.1432222883584+40.3017395271665i 42.71136787
    197.0215 37 -74.1047489355113-18.1722787864103i 76.3003639
    197.1436 35 126.172945864731-54.6305928619857i 137.4922323
    197.2656 30 13.321469116524+0.709161296336291i 13.34033167
    197.3877 23 -95.2530138693021-140.71563664121i 169.9235918
    197.5098 0 119.815759213019+46.1485270413212i 128.3958827
    197.6318 -12 18.8290251223947-40.4610253885468i 44.62764572
    197.7539 -21 234.359855629447+123.8773547681i 265.0851579
    197.876 -27 73.3076948329428-43.1128999271631i 85.04551876
    197.998 -27 39.7428047380141-10.3755470804628i 41.0748403
    198.1201 -22 19.7235631154051+90.7706516718061i 92.88880528
    198.2422 -13 -15.0475830236596+10.3997247609771i 18.29163825
    198.3643 -1 -7.50059909360206-59.1716923673936i 59.64518559
    198.4863 9 45.1441554071237-121.285833742742i 129.4150232
    198.6084 18 56.1287291472422+74.1872630204117i 93.02786803
    198.7305 25 -29.4959275545895+10.2159543205621i 31.21498783
    198.8525 29 245.957828218328+5.95728520301003i 246.0299626
    198.9746 33 40.4131119370942-45.5210693364474i 60.8718931
    199.0967 36 127.251246502001-39.9823740418276i 133.3846692
    199.2188 40 -110.827594082448+36.9632582846945i 116.8290977
    199.3408 44 26.3192948610859+58.4358152983326i 64.08938907
    199.4629 50 78.376912238589+93.7152231248685i 122.1698957
    199.585 48 -38.2295334085275+22.9492665383103i 44.58885577
    199.707 43 -28.0027528745261-63.5639745785492i 69.45885856
    199.8291 36 92.527509872412-10.8921551363242i 93.16640557
    199.9512 28 186.443179235127+20.0241491154869i 187.5154011
    200.0732 20 96.1965410249787-43.4614482503168i 105.5588556
    200.1953 14 102.984682441148-74.2656055104507i 126.969386
    200.3174 10 -53.9050266558072+84.9953278017816i 100.6476907
    200.4395 8 127.073660928431+8.17930888071714i 127.3366263
    200.5615 8 12.3677637180431+166.988422525568i 167.4457967
    200.6836 8 108.393899356193+17.4189853463952i 109.7846003
    200.8057 9 94.7399385418756+42.2420569588421i 103.730648
    200.9277 11 -10.3472894984534+53.3425233280728i 54.33683092
    201.0498 15 8.28087381793466-52.3591324485256i 53.00992003
    201.1719 23 64.8746387500065-67.9251131461604i 93.92837563
    201.2939 35 45.6739023236745-9.16160378265553i 46.58369175
    201.416 65 -41.5723029751289+55.5891983307124i 69.41480639
    201.5381 80 28.1411190261241+5.5752590664038i 28.68808278
    201.6602 90 107.435570056875+120.427928812051i 161.3855252
    201.7822 94 107.219345176584-63.7018478811051i 124.7153295
    201.9043 91 76.8861385331656+80.5072094191146i 111.3233536
    202.0264 83 7.35093212861321-70.6941541727755i 71.07530962
    202.1484 73 -14.9879454762587+43.4757688819199i 45.98674798
    202.2705 61 -24.0111701822759+56.0488664101836i 60.97550098
    202.3926 51 -30.217461291287+20.2191811363929i 36.35808373
    202.5146 45 24.7619964621414-64.468772967956i 69.06069184
    202.6367 41 42.9867656969256+48.0400127737428i 64.46475667
    202.7588 41 46.860964033704+67.6657193927398i 82.30795545
    202.8809 43 38.1102252569082-71.4781093803612i 81.00314432
    203.0029 47 100.00765595795+22.744978778936i 102.5615196
    203.125 53 -23.6809119319081-67.3677972723117i 71.40872285
    203.2471 60 2.42472609821971+66.6568572761076i 66.70094391
    203.3691 80 -63.3838707607259+32.3213317766191i 71.14902361
    203.4912 92 88.4307539623348+60.568589617878i 107.1846645
    203.6133 103 66.8167126714464+31.1831439182958i 73.73507684
    203.7354 113 -2.56586483908759+10.3251721251637i 10.63921246
    203.8574 118 29.7208258580479-36.9921076344671i 47.45253963
    203.9795 119 77.0821677521848-5.71116255358996i 77.29345356
    204.1016 114 65.9065192924462-28.7578488720712i 71.90746246
    204.2236 105 -17.139431250215-19.0328613661191i 25.61269051
    204.3457 96 67.9003301304361+62.8353347701194i 92.51342674
    204.4678 87 -78.9806168006372+100.656676359997i 127.9441453
    204.5898 79 133.58422648071+115.402687786642i 176.5291078
    204.7119 73 -25.8247312287291+23.0322562590219i 34.60349074
    204.834 68 67.9690686628948-37.1563388790606i 77.46217021
    204.9561 63 -11.5921652620302+29.2117990699521i 31.42781412
    205.0781 56 -53.8173510061246+46.7624434037744i 71.29539524
    205.2002 47 -20.3854730325216+4.6655552840424i 20.91255405
    205.3223 26 76.3168945474573-45.8786017278244i 89.04557535
    205.4443 18 15.7719880863986+71.0278918843938i 72.75793451
    205.5664 14 -5.08495422537268+89.8315942742198i 89.97539713
    205.6885 14 178.567349200741+43.4992782303361i 183.7892418
    205.8105 18 13.6710233419792+85.0386251362494i 86.13050937
    205.9326 25 104.608823467678-72.9179518702939i 127.514837
    206.0547 34 -112.703594665077+35.7965917766825i 118.251834
    206.1768 40 15.5656617250445+78.7721610038455i 80.29534964
    206.2988 44 36.3304410036491+185.869701496668i 189.3870293
    206.4209 46 23.4951067054227+86.2537065998986i 89.39643137
    206.543 46 -11.9227300337604-65.3558780557799i 66.43449622
    206.665 43 50.861410254231-14.239077158494i 52.81698942
    206.7871 39 93.1775340909226+106.065655949848i 141.1806511
    206.9092 33 89.0800506297163+32.8592502956031i 94.94727879
    207.0313 24 107.381014347069-131.088906570907i 169.4549606
    207.1533 12 -70.9971312227827+109.247662854846i 130.2906155
    207.2754 -17 -31.4051375257056+27.176182792363i 41.5310435
    207.3975 -33 -38.8297222656486+212.852532594902i 216.3653114
    207.5195 -46 42.3853738966557-33.5648137303937i 54.06585467
    207.6416 -54 105.680560290469+44.4345581462433i 114.6420986
    207.7637 -56 -93.7729125555442+7.91624967494522i 94.10646172
    207.8857 -53 -64.1628504904859-25.3577764819866i 68.99194309
    208.0078 -46 55.2274630840119-5.09137933317595i 55.46165182
    208.1299 -37 90.9077654743105+26.7456034075985i 94.76048293
    208.252 -28 -44.6044188664047+70.5301977258336i 83.45096149
    208.374 -23 -30.9959373635459+80.7338832645177i 86.47952382
    208.4961 -20 104.511223655871+257.876691814428i 278.249859
    208.6182 -19 -10.9105404747048+13.7819501655092i 17.57788508
    208.7402 -19 18.3251513042196+92.404860407432i 94.20440222
    208.8623 -19 -123.308858590443-58.1726441856999i 136.341964
    208.9844 -18 24.6706619588804+7.66849569701259i 25.83500315
    209.1064 -16 -74.8786016131776+113.043454106211i 135.5936116
    209.2285 -13 -7.94685937283793-69.4293545402259i 69.882672
    209.3506 -15 88.9378310294104+7.03941629294161i 89.21598046
    209.4727 -20 78.3232239233975+42.3033731653447i 89.01742968
    209.5947 -28 68.1606150695734+100.555134244285i 121.4792347
    209.7168 -36 41.017152517398+32.0147216123651i 52.03219389
    209.8389 -44 51.237484901187+9.2524112091312i 52.06617877
    209.9609 -51 -168.323919491274+10.723476354221i 168.6651559
    210.083 -56 -20.3026320031041+32.0036118953869i 37.900238
    210.2051 -59 -150.403511502166+2.09289656025248i 150.4180723
    210.3271 -61 64.052054314777-24.1610481633289i 68.45744598
    210.4492 -62 46.0431061371615+89.0154401930631i 100.2183427
    210.5713 -63 -52.4457298433008-62.2601545302787i 81.40565964
    210.6934 -63 104.510657228179-28.00478565701i 108.1977148
    210.8154 -61 42.1326256122409+70.8953020487784i 82.47000663
    210.9375 -55 -30.235083903427+61.9965922661445i 68.97635647
    211.0596 -46 -92.9628008620612+43.5559324442661i 102.6606137
    211.1816 -20 -88.8078222879832+25.301957401914i 92.34185588
    211.3037 -7 -100.921769424634+149.738677970149i 180.573739
    211.4258 1 16.938455746352-48.773694753462i 51.6312365
    211.5479 3 -47.7280043357411-47.530506270593i 67.35808358
    211.6699 0 50.7743316132143-67.922980597099i 84.80308983
    211.792 -9 90.4793357933968+23.3364583205953i 93.44035794
    211.9141 -22 0.605676484584592+189.357583570215i 189.3585522
    212.0361 -35 23.7895755665648-1.22394937818262i 23.82104023
    212.1582 -46 65.2305190030588+124.866284462631i 140.8779955
    212.2803 -54 -540.36174405699+206.683436710988i 578.5402816
    212.4023 -59 22.2728485249172+25.1371006041821i 33.58502059
    212.5244 -61 -51.0663859437047+36.9570446508101i 63.03648882
    212.6465 -60 -147.610696835645+634.281027862242i 651.2306351
    212.7686 -58 -74.62352385207-132.048764503698i 151.6757941
    212.8906 -54 -63.5203049630295-25.3116457648075i 68.37769047
    213.0127 -49 -104.059463379619-300.119542478777i 317.6477793
    213.1348 -32 1.52321327116792+38.8550327012769i 38.8848781
    213.2568 -21 -34.62988357018+121.103375490992i 125.9573594
    213.3789 -9 239.652406572644-99.0839763726967i 259.3278048
    213.501 0 53.4641749813784+2.66209525788605i 53.53040965
    213.623 6 -73.4779436757802+50.2177651283487i 88.99905697
    213.7451 8 -111.582481297048+88.5166485965005i 142.4283933
    213.8672 5 -25.9069231310157+22.4388981444004i 34.2735002
    213.9893 0 2.902775077456-57.8243772691561i 57.89719086
    214.1113 -6 -261.849617291271-71.1191402981699i 271.3358697
    214.2334 -12 -91.83950038697+83.2920200772749i 123.9840895
    214.3555 -16 -10.6150360783586-40.4827347376565i 41.85129392
    214.4775 -18 211.679870288213+46.9851791755782i 216.8316733
    214.5996 -19 -56.2108498489952-14.3797921385587i 58.02101398
    214.7217 -20 -77.4615528195145-132.18862003192i 153.2126739
    214.8438 -22 -26.0262457727877+139.085835646582i 141.4999475
    214.9658 -26 -31.1184749812655+138.494743961769i 141.9477143
    215.0879 -36 -68.3010500471096-10.1389166334639i 69.04948275
    215.21 -38 -74.0621746515191+3.51349854611271i 74.14546774
    215.332 -36 -80.407372989383+9.53680546597093i 80.97095954
    215.4541 -30 -38.144701217363+115.048943645985i 121.2075809
    215.5762 -20 107.308979511238-91.0960719668123i 140.7611857
    215.6982 -8 -8.55757643571906-189.004412489771i 189.1980445
    215.8203 5 20.6495066189379-19.7991527079091i 28.60784109
    215.9424 16 -32.2468708231553+151.61591203962i 155.0072433
    216.0645 25 -91.9226550318239+37.2300895877397i 99.17587448
    216.1865 31 139.089268984361+26.5659663755655i 141.6035851
    216.3086 34 -24.3728105588257+3.33208275184461i 24.59952581
    216.4307 35 -108.094156665445-32.1662364558274i 112.7786038
    216.5527 33 -10.8044097979631-9.36118829674183i 14.29570276
    216.6748 29 -9.5952390925363-97.4337447488419i 97.90507254
    216.7969 23 -5.58233331939221-65.4173634327834i 65.65511316
    216.9189 15 -106.055931733341-133.552668529299i 170.5408336
    217.041 -9 -92.4199999044876+8.14419399463046i 92.77814548
    217.1631 -22 120.824729542641+28.3822124371336i 124.1135176
    217.2852 -33 -30.6801590692072+1.38695987921096i 30.71149326
    217.4072 -40 -68.9549878319958-22.5115254799495i 72.53660542
    217.5293 -42 23.5258860502072-19.7833318469529i 30.73837233
    217.6514 -39 -90.6546780206186+47.0396224675305i 102.1322512
    217.7734 -32 -85.1348014483729-9.85439126041865i 85.70322891
    217.8955 -23 -42.7721006577224-74.6944897520241i 86.0739182
    218.0176 -15 -108.693920507942+9.95931498117488i 109.1492387
    218.1396 -10 17.5341803027542+5.64069372869517i 18.41914506
    218.2617 -8 -55.8655011728795+10.5296630273096i 56.84916908
    218.3838 -7 37.3663834959616-31.2985195950481i 48.74262964
    218.5059 -7 88.0137276324705-90.8110794994387i 126.4637039
    218.6279 -7 -55.9614098114567+62.0505482957982i 83.55806324
    218.75 -6 103.346947600816+65.0756302071682i 122.1287404
    218.8721 -3 27.3650938566493-92.586881265689i 96.54625288
    218.9941 0 -28.312436274115+55.4527842279997i 62.26239095
    219.1162 0 -9.77431867492766+59.0769466297927i 59.88007121
    219.2383 -5 41.9338178829339-61.9309768441956i 74.79231896
    219.3604 -12 53.2576327823117+3.32989258442171i 53.36163073
    219.4824 -21 -16.5408980706288-95.8604401360548i 97.27705429
    219.6045 -30 -32.6788453163033-18.7910684185126i 37.69630199
    219.7266 -37 -18.8251573472107-97.3431336611798i 99.14672067
    219.8486 -43 10.8716185314667-139.229160468855i 139.6529671
    219.9707 -47 -15.1887877350226+64.2052398287176i 65.97736047
    220.0928 -50 -31.7276403518113-65.4614657741292i 72.74508
    220.2148 -52 14.5547333671508-12.5502947012865i 19.21848486
    220.3369 -53 84.8405553042521+38.5781674330392i 93.19975765
    220.459 -53 -31.1505644086242-100.174139247352i 104.9057474
    220.5811 -51 -141.444129053801-83.0248024865959i 164.0108517
    220.7031 -46 -68.1380952618205-11.6343957002819i 69.12423012
    220.8252 -36 -75.9977716747802-3.93460502320511i 76.09955595
    220.9473 -11 -11.7686614147664-2.14271123270521i 11.96213204
    221.0693 1 -80.1837469087968-129.614292280169i 152.4116073
    221.1914 9 -26.171967606303-54.4900161847766i 60.44943136
    221.3135 12 96.9087799302378+81.3174784665066i 126.5063
    221.4355 8 5.05604076370071-83.4159174313513i 83.56900639
    221.5576 -1 138.365772666341-110.36837254098i 176.992273
    221.6797 -14 -0.126864740866267-5.32340071337148i 5.324912189
    221.8018 -27 -151.90826249584+27.1049769515985i 154.3074852
    221.9238 -39 -20.6762972064489+122.930734381494i 124.6574295
    222.0459 -48 97.9295553567435-93.2640680387319i 135.2345525
    222.168 -53 -28.066495000443-59.3121747015896i 65.61754498
    222.29 -55 -16.2880383669658-80.1819013066086i 81.81954223
    222.4121 -54 2.70457649724131-27.2880246554253i 27.42172539
    222.5342 -52 142.922322382816-104.938020205294i 177.3098371
    222.6563 -49 155.042156666303-105.260016000777i 187.397282
    222.7783 -44 -145.350144310448-57.0941074915145i 156.1614599
    222.9004 -28 -31.5008551772248-66.0315305890367i 73.16055569
    223.0225 -18 128.013165708808+165.507800331246i 209.2371921
    223.1445 -8 -9.41644469092807-15.7420435464173i 18.3434284
    223.2666 0 34.6023785470779-134.062965077602i 138.4565029
    223.3887 6 -167.147420276129-218.573508437916i 275.1592969
    223.5107 8 -37.0380738851786-19.8189494200387i 42.00725739
    223.6328 5 31.2262815685837-75.2404627924722i 81.46292348
    223.7549 0 53.701756861024-148.520408713617i 157.9309675
    223.877 -6 37.3902572695486-77.2308261978897i 85.80577984
    223.999 -11 27.4055920924943+19.9444231487585i 33.89463811
    224.1211 -14 5.89070896540579+243.689299401724i 243.7604871
    224.2432 -16 123.762579029637-51.3259203387056i 133.9833052
    224.3652 -16 26.2898455735128-25.8043292275611i 36.83774406
    224.4873 -18 -321.26954387094-34.5478587826346i 323.121764
    224.6094 -21 -46.3493225629491-50.5199423627032i 68.56036959
    224.7314 -27 -88.8805429898307-11.6731622252482i 89.64381539
    224.8535 -42 502.086494864019-396.847860095855i 639.9836501
    224.9756 -48 11.769210411194-71.3183233236159i 72.28289947
    225.0977 -48 41.4192740565195-37.4085323474917i 55.81177793
    225.2197 -44 -43.3933039706026-547.006476297533i 548.7249438
    225.3418 -35 120.056739842662-3.22151971954645i 120.0999541
    225.4639 -21 40.8465726109125+11.0149723423162i 42.30569831
    225.5859 -4 149.498904968404-99.9215645645627i 179.8172451
    225.708 11 67.9116035669363+53.8818958739434i 86.69051045
    225.8301 25 -47.0468465985173+84.3967126095805i 96.62406985
    225.9521 37 -64.0986519051676-19.2726161470691i 66.93333183
    226.0742 47 -26.1625423871383+6.95273589350813i 27.07063281
    226.1963 54 124.196218860281-158.047225806768i 201.0065331
    226.3184 58 0.110730350173071-114.01340742355i 114.0134612
    226.4404 59 43.7195786127552-97.9397947007474i 107.2548598
    226.5625 56 47.6440183402278-66.426149040888i 81.74586081
    226.6846 49 133.741373453684+5.41394638334486i 133.8509088
    226.8066 25 68.5366725546296+95.9310158575354i 117.8984109
    226.9287 10 -74.6225516082477-36.6901741470072i 83.1546396
    227.0508 -2 130.09831332579-9.76930549049121i 130.4645947
    227.1729 -11 15.0302348041159+94.2875099440209i 95.47796861
    227.2949 -15 -42.7234725911676-172.656489475774i 177.8638762
    227.417 -15 11.2969364455539-54.352498032425i 55.51409565
    227.5391 -11 -14.8885776308352-180.465917413775i 181.0790355
    227.6611 -4 36.134262783703+54.9885728273225i 65.79838971
    227.7832 0 48.8436591277923+2.67060027966818i 48.91661418
    227.9053 3 113.592999234199-3.33174083444101i 113.6418496
    228.0273 4 80.9589027482589+21.8248140425203i 83.84906942
    228.1494 3 16.9769760979561+66.5741497658871i 68.70469514
    228.2715 3 -66.4518066495644+33.1591392263249i 74.26554464
    228.3936 3 75.7699298507391-100.973803967749i 126.2410051
    228.5156 6 29.5191257099602+23.0489898771832i 37.45176521
    228.6377 9 -107.94993909287-87.373159569981i 138.8785742
    228.7598 15 126.498463782516-26.3496219822327i 129.2136367
    228.8818 14 12.8694538844707-33.1979724975549i 35.60517127
    229.0039 8 237.167143999595-29.7084646025633i 239.0205997
    229.126 -1 62.5510961401969-34.0116661484499i 71.19995128
    229.248 -13 14.6804761819672-53.0068075759159i 55.00216387
    229.3701 -25 58.0421614180636+31.6584807138934i 66.1146875
    229.4922 -36 -2.13915027287021+32.8169649413107i 32.88661053
    229.6143 -43 -38.0499198120669-52.3562471725977i 64.72227604
    229.7363 -48 -0.583859481532727-116.405077277754i 116.4065415
    229.8584 -50 69.3376354646107+66.7297384876317i 96.23183304
    229.9805 -51 -2.11489829573338+43.8704324840101i 43.92138023
    230.1025 -50 200.751525225313-83.3953441427523i 217.3843562
    230.2246 -49 11.3473861406785-37.3256493444383i 39.01239894
    230.3467 -46 59.5636130222665-96.8011865998405i 113.658672
    230.4688 -40 -58.7534521032703+87.2749131798082i 105.2087383
    230.5908 -32 49.5412485801271+40.5481132284184i 64.01940954
    230.7129 -12 119.807767186731+27.6014797859743i 122.946097
    230.835 -1 4.5523266654086+35.0442559427988i 35.33869766
    230.957 5 -17.6197194102278-0.207645349788208i 17.6209429
    231.0791 8 60.1765166143932-23.9596894881806i 64.77098017
    231.2012 6 158.607076568689-59.7708658043248i 169.495608
    231.3232 0 72.8112134011026-49.298810373074i 87.93091323
    231.4453 -10 20.8377205859479-127.171152159598i 128.8670343
    231.5674 -22 5.28892966063177+119.513197856593i 119.6301686
    231.6895 -35 98.2140565396185-31.9727159469494i 103.2872474
    231.8115 -47 117.933994599884+119.431725086613i 167.8462512
    231.9336 -60 67.6985724391039-50.7183159551696i 84.58985922
    232.0557 -71 89.2826089388839-40.5906597704343i 98.07642897
    232.1777 -81 12.7186453271402+59.9135868973231i 61.24868843
    232.2998 -87 -18.8350051915043-45.0135535235295i 48.79526023
    232.4219 -87 33.6310985450823-66.6075412469627i 74.61645489
    232.5439 -77 30.4388959207554-40.8244128496872i 50.92307011
    232.666 -19 16.1943069961003+79.8585884879258i 81.48404589
    232.7881 30 47.1166507253877-32.0798118489913i 57.00081669
    232.9102 89 136.77628617135+33.1167009692344i 140.7283495
    233.0322 153 64.5327123267862-125.204301429185i 140.8566223
    233.1543 214 85.2283734964021+27.0211599171407i 89.40927654
    233.2764 265 -27.0283101818486-55.0930296397494i 61.36588194
    233.3984 299 9.17695930534637+40.016819811798i 41.05560193
    233.5205 314 18.7737383252088+56.9773384531957i 59.99058549
    233.6426 308 -52.6489851107881+24.907000214087i 58.24323388
    233.7646 282 7.00613864381762-36.4612564779211i 37.12828036
    233.8867 237 32.6549875466221+11.4659442525439i 34.60947976
    234.0088 177 110.456627217352+42.8056428995906i 118.46092
    234.1309 108 3.05638588972602-33.8601831889292i 33.99784552
    234.2529 35 114.698566964996-41.1930840388147i 121.8713725
    234.375 -36 -64.5802507361446-50.4037169558469i 81.921569
    234.4971 -102 103.72264589728+77.372159302332i 129.4018482
    234.6191 -194 -15.9609728792698+57.0765845944634i 59.26625654
    234.7412 -214 116.089880678517+18.1193597757501i 117.495411
    234.8633 -214 42.502017470194+24.7635280310174i 49.18997672
    234.9854 -198 3.53972056040607+35.5106476555441i 35.68663221
    235.1074 -169 -4.84003679994817-25.2970333296059i 25.75588965
    235.2295 -132 84.9687198581237-52.2938155602601i 99.77137115
    235.3516 -95 37.4512103604918-22.6984396677028i 43.79283413
    235.4736 -60 -54.9308196467996+29.5023389040502i 62.35208856
    235.5957 -33 99.172347742944+12.9832561083699i 100.0185958
    235.7178 -13 -0.501228221182828+115.571836228519i 115.5729231
    235.8398 0 170.799614355152+26.1748169555103i 172.7936032
    235.9619 5 -15.2782408848807+22.8604875438285i 27.4959367
    236.084 8 45.2046887301212-61.5477723959038i 76.36486214
    236.2061 7 36.8675282222696+33.2249086138511i 49.62972083
    236.3281 3 -10.1933188897128+55.1084534997617i 56.04324578
    236.4502 -2 -24.513322813207-26.6940162341065i 36.24187492
    236.5723 -20 34.0238397661554-50.1945037994628i 60.6391778
    236.6943 -31 32.0966118634976+69.9166321274024i 76.93196957
    236.8164 -41 66.0860263110299+70.6020463350887i 96.70580034
    236.9385 -47 157.904854384706-45.2341508259619i 164.2561154
    237.0605 -50 77.7439567271417+66.3919656558905i 102.2351012
    237.1826 -49 62.4965600200494-86.9498214731208i 107.0798369
    237.3047 -43 -80.14729351397+102.817063267191i 130.3646315
    237.4268 -36 56.2526017061778+78.002031826491i 96.17001699
    237.5488 -28 110.682811694869+107.388169564276i 154.217067
    237.6709 -22 34.6553268456573+53.5078120987787i 63.7501187
    237.793 -18 -48.6786128245449-28.9250564846826i 56.62390166
    237.915 -14 10.9277135635999-71.8883684319735i 72.71418321
    238.0371 -11 136.419924663057+56.0885200571138i 147.5002303
    238.1592 -8 78.0020686410244-15.8297085556825i 79.59210002
    238.2813 -4 3.00839470398023-114.392712749057i 114.4322645
    238.4033 0 -12.9288928837811+143.764799080206i 144.3449816
    238.5254 4 5.32406672951311+60.4353535202416i 60.66941274
    238.6475 3 85.4988958680906+192.696641079678i 210.8128475
    238.7695 -1 18.9475582401408-73.020525400132i 75.43876386
    238.8916 -9 118.259171547331-8.23408848564031i 118.5454844
    239.0137 -18 -57.6525613361229+62.0978841131387i 84.73467437
    239.1357 -27 -54.014855729316+8.36009788880438i 54.65799005
    239.2578 -34 23.2013964609436-43.4892002805182i 49.2911284
    239.3799 -38 92.9217041203663-30.3529337475752i 97.75348425
    239.502 -41 12.8352314202548+77.2475359488089i 78.30660876
    239.624 -42 47.7344877381724+84.7547799505896i 97.27257601
    239.7461 -43 192.670187066617+159.893071547966i 250.3749095
    239.8682 -45 10.0212611146007+22.0287760839117i 24.20108779
    239.9902 -45 108.467425685494+52.8161155241587i 120.6429629
    240.1123 -44 -161.654553837038-16.6408759980341i 162.5088106
    240.2344 -39 17.1317656073415+23.4247496721466i 29.02096294
    240.3564 -31 4.38773812852219+107.367329277771i 107.4569479
    240.4785 -5 -20.8156678880659-41.8910672521179i 46.7777035
    240.6006 8 66.2093119552878-27.5398078785583i 71.70853511
    240.7227 19 77.9369600339649+2.64593581357178i 77.98186145
    240.8447 24 89.283661131662+59.0363032469836i 107.0367098
    240.9668 22 63.652494252196+19.8975773791626i 66.68998133
    241.0889 14 82.1290955908146-26.100874117953i 86.17681807
    241.2109 3 -135.798059461079+130.658805793365i 188.4484982
    241.333 -8 4.17373647995777+24.7357041311823i 25.08535699
    241.4551 -20 -108.865985746743+111.110028645218i 155.5546249
    241.5771 -29 45.7920532793296-25.9499410803155i 52.63374949
    241.6992 -34 95.4685245395522+50.5939412234002i 108.0462219
    241.8213 -35 -61.2006149659501-33.1188623240639i 69.58717061
    241.9434 -34 55.6416164625962-37.0523774746512i 66.84959356
    242.0654 -30 68.4069310436505+47.0464223412208i 83.02333449
    242.1875 -25 31.9905233739243+87.6014068250242i 93.25985237
    242.3096 -18 -9.28105171443406+94.9910909378136i 95.44341401
    242.4316 2 -97.2397574387714+53.4488286523738i 110.9610189
    242.5537 15 0.694579175577587+207.796860354396i 207.7980212
    242.6758 27 16.9542912237158-14.7184159533717i 22.45172063
    242.7979 39 -70.6079988919709-18.128994148211i 72.89821628
    242.9199 47 10.9646954947284-103.808392238405i 104.3858556
    243.042 52 115.300242127978-29.3468434467113i 118.9763971
    243.1641 51 62.0590349568265+172.422012777692i 183.2503051
    243.2861 48 54.207622246891+25.3439469624896i 59.83963534
    243.4082 44 79.6683803463893+86.1651122426156i 117.3519382
    243.5303 40 -324.433446731839+432.52729817451i 540.6819074
    243.6523 38 16.6254452773939+3.2106602968596i 16.93262443
    243.7744 37 9.26084258707829+63.9405780462265i 64.6077451
    243.8965 38 203.452407292626+619.073264118888i 651.6475952
    244.0186 37 -91.8398419841991-57.7804146362007i 108.5040685
    244.1406 36 -62.2831933112473+32.7256163567881i 70.35738863
    244.2627 32 -284.34120527272-132.256702570706i 313.5948922
    244.3848 21 19.2203820271907+15.4301457268393i 24.64776831
    244.5068 19 80.8586653688981+117.866080828616i 142.9354287
    244.6289 21 125.316571232534-213.318491239815i 247.404571
    244.751 27 10.3717129920051+7.43006333911694i 12.75845883
    244.873 38 -39.4404797857595+99.6091812437085i 107.1332835
    244.9951 53 -50.5359523988018+151.38120045515i 159.5937039
    245.1172 69 -11.4728296970462+69.877040973976i 70.81261665
    245.2393 84 -34.1380702326339-36.8545510070326i 50.23610026
    245.3613 97 -240.274383918723+74.4707704261633i 251.5505421
    245.4834 106 -22.2368155600866+111.076618997587i 113.2805864
    245.6055 113 23.0694678556538-18.7293265606047i 29.71511435
    245.7275 117 212.086734515665-84.6051225347892i 228.3392426
    245.8496 118 -40.5415992725984+27.4301331227216i 48.94929494
    245.9717 118 -129.43293283777-71.2626962126924i 147.7540388
    246.0938 114 74.1201373015456+157.516649470062i 174.0841452
    246.2158 109 65.6708626670199+157.885383296573i 170.9984107
    246.3379 89 -52.2434979327684+51.9683949953836i 73.68919293
    246.46 78 -44.0798409258123+22.3313108588602i 49.41376145
    246.582 68 -74.2992343803139+35.7057655468482i 82.43347574
    246.7041 62 29.5105046726108+134.408352979784i 137.6098661
    246.8262 61 70.3245831147112-150.5841725797i 166.1960891
    246.9482 64 -85.6603955073402-135.85664188717i 160.60738
    247.0703 72 18.3528359097372-23.994407660929i 30.208578
    247.1924 81 43.9991393279342+141.888785260457i 148.5542044
    247.3145 89 -49.8796547003876+63.491621090192i 80.74135187
    247.4365 95 141.843782676025-31.8577601240481i 145.3773557
    247.5586 97 -52.9280880528349-1.0555984085089i 52.93861344
    247.6807 97 -115.074630641081+23.0910921580378i 117.3685186
    247.8027 96 -39.9731885334789-17.5178043146485i 43.64320416
    247.9248 95 -65.5746615411471-70.9636049003846i 96.6223031
    248.0469 95 4.0950923988903-59.5220651799254i 59.66276917
    248.1689 96 -172.899631517201-30.0982988395454i 175.4998295
    248.291 96 -89.3907115960565+56.5473868951665i 105.7747904
    248.4131 92 110.661534432241-35.631466603973i 116.2565121
    248.5352 84 -0.223265685752267+8.02313281387433i 8.026238703
    248.6572 72 -87.9259763770297+30.3705123950367i 93.02335914
    248.7793 59 8.85665212508047-15.3566238961428i 17.72755438
    248.9014 44 -58.0641418264413+114.091722640011i 128.0170525
    249.0234 30 -71.71971513523+79.5940212494859i 107.1397487
    249.1455 18 -57.1216872475851-22.8814806344428i 61.53413126
    249.2676 7 -61.9175856618854+92.1751441062586i 111.040734
    249.3896 -1 54.6569770269209+26.7803213960037i 60.86518506
    249.5117 -10 -29.5380835794738+61.0161016668041i 67.7898447
    249.6338 -17 -1.97794458653334-27.3070825299009i 27.37862343
    249.7559 -23 0.581039036250459-100.986345038832i 100.9880166
    249.8779 -27 -6.70780377182359+77.193074049085i 77.4839681
    250 -27 85 85
    250.1221 -22 -6.70780377182905-77.1930740490877i 77.4839681
    250.2441 -5 0.581039036247461+100.986345038831i 100.9880166
    250.3662 4 -1.97794458653425+27.3070825298964i 27.37862343
    250.4883 10 -29.538083579474-61.0161016668082i 67.7898447
    250.6104 11 54.6569770269232-26.7803213960071i 60.86518506
    250.7324 6 -61.9175856618856-92.1751441062596i 111.040734
    250.8545 -4 -57.1216872475835+22.8814806344413i 61.53413126
    250.9766 -18 -71.7197151352241-79.5940212494859i 107.1397487
    251.0986 -33 -58.0641418264377-114.091722640017i 128.0170525
    251.2207 -46 8.85665212508957+15.3566238961289i 17.72755438
    251.3428 -57 -87.9259763769969-30.3705123950513i 93.02335914
    251.4648 -64 -0.223265685786828-8.02313281386159i 8.026238703
    251.5869 -67 110.661534432232+35.6314666039798i 116.2565121
    251.709 -68 -89.3907115960624-56.5473868951603i 105.7747904
    251.8311 -67 -172.899631517207+30.0982988395531i 175.4998295
    251.9531 -65 4.09509239888666+59.5220651799302i 59.66276917
    252.0752 -61 -65.5746615411507+70.9636049003882i 96.6223031
    252.1973 -47 -39.9731885334832+17.5178043146512i 43.64320416
    252.3193 -37 -115.074630641085-23.0910921580326i 117.3685186
    252.4414 -27 -52.9280880528363+1.05559840851788i 52.93861344
    252.5635 -18 141.843782676019+31.8577601240504i 145.3773557
    252.6855 -10 -49.8796547004019-63.4916210901993i 80.74135187
    252.8076 -7 43.9991393279779-141.8887852605i 148.5542044
    252.9297 -8 18.3528359097199+23.994407660919i 30.208578
    253.0518 -12 -85.6603955073372+135.856641887172i 160.60738
    253.1738 -17 70.3245831147153+150.584172579702i 166.1960891
    253.2959 -21 29.510504672608-134.408352979787i 137.6098661
    253.418 -24 -74.2992343803207-35.7057655468432i 82.43347574
    253.54 -25 -44.0798409258216-22.3313108588543i 49.41376145
    253.6621 -26 -52.2434979327704-51.9683949953794i 73.68919293
    253.7842 -26 65.6708626670085-157.885383296576i 170.9984107
    253.9063 -29 74.1201373015356-157.51664947005i 174.0841452
    254.0283 -33 -129.43293283779+71.2626962127224i 147.7540388
    254.1504 -47 -40.5415992726453-27.4301331227216i 48.94929494
    254.2725 -51 212.086734515712+84.6051225347965i 228.3392426
    254.3945 -51 23.0694678556574+18.7293265605787i 29.71511435
    254.5166 -47 -22.2368155600834-111.076618997595i 113.2805864
    254.6387 -38 -240.274383918721-74.4707704261559i 251.5505421
    254.7607 -25 -34.1380702326214+36.8545510070321i 50.23610026
    254.8828 -11 -11.4728296970437-69.8770409739777i 70.81261665
    255.0049 1 -50.5359523988063-151.381200455158i 159.5937039
    255.127 12 -39.4404797857602-99.6091812437048i 107.1332835
    255.249 19 10.3717129920137-7.43006333912763i 12.75845883
    255.3711 24 125.31657123255+213.318491239809i 247.404571
    255.4932 26 80.8586653688863-117.866080828609i 142.9354287
    255.6152 26 19.2203820271652-15.4301457268575i 24.64776831
    255.7373 24 -284.341205272725+132.256702570735i 313.5948922
    255.8594 19 -62.2831933112454-32.7256163567895i 70.35738863
    255.9814 12 -91.8398419841984+57.7804146361984i 108.5040685
    256.1035 -9 203.452407292587-619.073264118902i 651.6475952
    256.2256 -22 9.2608425870759-63.9405780462245i 64.6077451
    256.3477 -33 16.6254452773933-3.21066029685926i 16.93262443
    256.4697 -40 -324.433446731866-432.527298174493i 540.6819074
    256.5918 -43 79.6683803463834-86.1651122426183i 117.3519382
    256.7139 -41 54.2076222468862-25.3439469624927i 59.83963534
    256.8359 -35 62.0590349568138-172.422012777703i 183.2503051
    256.958 -28 115.30024212798+29.3468434466981i 118.9763971
    257.0801 -21 10.9646954947202+103.808392238412i 104.3858556
    257.2021 -16 -70.6079988919581+18.1289941482155i 72.89821628
    257.3242 -15 16.9542912237196+14.7184159533715i 22.45172063
    257.4463 -14 0.694579175573324-207.796860354395i 207.7980212
    257.5684 -14 -97.2397574387774-53.4488286523642i 110.9610189
    257.6904 -13 -9.28105171443792-94.9910909378128i 95.44341401
    257.8125 -10 31.9905233739216-87.6014068250263i 93.25985237
    257.9346 -6 68.4069310436503-47.0464223412224i 83.02333449
    258.0566 4 55.6416164626007+37.0523774746648i 66.84959356
    258.1787 5 -61.200614965956+33.1188623240698i 69.58717061
    258.3008 3 95.468524539544-50.5939412234275i 108.0462219
    258.4229 -3 45.7920532793528+25.9499410803101i 52.63374949
    258.5449 -12 -108.865985746754-111.110028645252i 155.5546249
    258.667 -21 4.1737364799545-24.735704131183i 25.08535699
    258.7891 -30 -135.798059461078-130.65880579336i 188.4484982
    258.9111 -38 82.1290955908214+26.1008741179467i 86.17681807
    259.0332 -44 63.6524942521887-19.8975773791685i 66.68998133
    259.1553 -48 89.2836611316561-59.0363032469891i 107.0367098
    259.2773 -50 77.9369600339624-2.64593581357701i 77.98186145
    259.3994 -52 66.2093119552942+27.5398078785552i 71.70853511
    259.5215 -52 -20.81566788808+41.8910672521206i 46.7777035
    259.6436 -50 4.38773812849445-107.367329277782i 107.4569479
    259.7656 -45 17.1317656073452-23.4247496721579i 29.02096294
    259.8877 -36 -161.65455383706+16.6408759980422i 162.5088106
    260.0098 -11 108.467425685508-52.8161155241587i 120.6429629
    260.1318 0 10.0212611146078-22.0287760839142i 24.20108779
    260.2539 9 192.670187066613-159.893071547974i 250.3749095
    260.376 12 47.734487738164-84.7547799505851i 97.27257601
    260.498 8 12.8352314202509-77.2475359488089i 78.30660876
    260.6201 0 92.921704120374+30.3529337475686i 97.75348425
    260.7422 -13 23.2013964609451+43.4892002805184i 49.2911284
    260.8643 -26 -54.014855729331-8.3600978887971i 54.65799005
    260.9863 -38 -57.6525613361327-62.0978841131387i 84.73467437
    261.1084 -47 118.25917154735+8.23408848562394i 118.5454844
    261.2305 -52 18.9475582401428+73.0205254001359i 75.43876386
    261.3525 -54 85.4988958681197-192.696641079692i 210.8128475
    261.4746 -54 5.32406672951538-60.4353535202482i 60.66941274
    261.5967 -52 -12.9288928837852-143.764799080202i 144.3449816
    261.7188 -49 3.00839470398569+114.392712749066i 114.4322645
    261.8408 -46 78.0020686410276+15.8297085556748i 79.59210002
    261.9629 -33 136.419924663056-56.0885200571278i 147.5002303
    262.085 -24 10.9277135636074+71.8883684319758i 72.71418321
    262.207 -15 -48.6786128245462+28.9250564846902i 56.62390166
    262.3291 -6 34.6553268456546-53.5078120988019i 63.7501187
    262.4512 0 110.682811694875-107.388169564312i 154.217067
    262.5732 3 56.2526017061696-78.0020318265001i 96.17001699
    262.6953 2 -80.14729351397-102.817063267217i 130.3646315
    262.8174 -1 62.496560020043+86.9498214731337i 107.0798369
    262.9395 -5 77.7439567271501-66.3919656558855i 102.2351012
    263.0615 -9 157.904854384711+45.2341508259588i 164.2561154
    263.1836 -11 66.086026311018-70.6020463350956i 96.70580034
    263.3057 -13 32.0966118634921-69.9166321274038i 76.93196957
    263.4277 -13 34.0238397661623+50.1945037994615i 60.6391778
    263.5498 -14 -24.5133228132079+26.6940162341123i 36.24187492
    263.6719 -16 -10.1933188897322-55.1084534997708i 56.04324578
    263.7939 -21 36.8675282222639-33.2249086138547i 49.62972083
    263.916 -36 45.2046887301392+61.547772395912i 76.36486214
    264.0381 -40 -15.2782408848761-22.8604875438358i 27.4959367
    264.1602 -41 170.799614355173-26.1748169555012i 172.7936032
    264.2822 -37 -0.501228221178735-115.571836228525i 115.5729231
    264.4043 -29 99.1723477429409-12.9832561083713i 100.0185958
    264.5264 -17 -54.930819646805-29.5023389040437i 62.35208856
    264.6484 -3 37.451210360498+22.6984396676978i 43.79283413
    264.7705 9 84.9687198581355+52.293815560251i 99.77137115
    264.8926 19 -4.84003679994544+25.2970333296105i 25.75588965
    265.0146 26 3.53972056039993-35.5106476555457i 35.68663221
    265.1367 30 42.5020174702095-24.763528031031i 49.18997672
    265.2588 32 116.089880678538-18.1193597757674i 117.495411
    265.3809 32 -15.9609728792743-57.0765845944743i 59.26625654
    265.5029 29 103.722645897299-77.3721593023338i 129.4018482
    265.625 24 -64.5802507361529+50.4037169558433i 81.921569
    265.7471 17 114.698566964995+41.1930840388231i 121.8713725
    265.8691 -4 3.05638588972533+33.8601831889312i 33.99784552
    265.9912 -17 110.456627217352-42.8056428996082i 118.46092
    266.1133 -29 32.6549875466217-11.4659442525555i 34.60947976
    266.2354 -37 7.00613864382035+36.4612564779266i 37.12828036
    266.3574 -40 -52.6489851107999-24.9070002140907i 58.24323388
    266.4795 -39 18.7737383252106-56.9773384532182i 59.99058549
    266.6016 -35 9.17695930534501-40.0168198118067i 41.05560193
    266.7236 -28 -27.028310181855+55.0930296397619i 61.36588194
    266.8457 -21 85.2283734964244-27.0211599171425i 89.40927654
    266.9678 -16 64.5327123267843+125.204301429208i 140.8566223
    267.0898 -14 136.776286171364-33.1167009692272i 140.7283495
    267.2119 -14 47.1166507253834+32.079811848989i 57.00081669
    267.334 -14 16.1943069960889-79.8585884879344i 81.48404589
    267.4561 -13 30.4388959207622+40.8244128496931i 50.92307011
    267.5781 -11 33.6310985450984+66.6075412469709i 74.61645489
    267.7002 -8 -18.8350051915095+45.0135535235343i 48.79526023
    267.8223 -2 12.7186453271406-59.9135868973333i 61.24868843
    267.9443 -2 89.2826089389021+40.5906597704425i 98.07642897
    268.0664 -7 67.6985724391282+50.7183159551769i 84.58985922
    268.1885 -15 117.933994599907-119.431725086639i 167.8462512
    268.3105 -25 98.2140565396248+31.9727159469639i 103.2872474
    268.4326 -34 5.28892966063813-119.513197856605i 119.6301686
    268.5547 -41 20.8377205859388+127.171152159605i 128.8670343
    268.6768 -46 72.8112134011126+49.2988103730777i 87.93091323
    268.7988 -47 158.607076568709+59.7708658043139i 169.495608
    268.9209 -46 60.1765166144041+23.959689488172i 64.77098017
    269.043 -44 -17.6197194102351+0.207645349791846i 17.6209429
    269.165 -39 4.55232666540769-35.0442559428102i 35.33869766
    269.2871 -34 119.80776718675-27.6014797859834i 122.946097
    269.4092 -28 49.5412485801312-40.5481132284297i 64.01940954
    269.5313 -19 -58.7534521032835-87.2749131798182i 105.2087383
    269.6533 -7 59.5636130222811+96.8011865998678i 113.658672
    269.7754 20 11.3473861406719+37.3256493444405i 39.01239894
    269.8975 33 200.751525225314+83.3953441427668i 217.3843562
    270.0195 41 -2.11489829573725-43.8704324840166i 43.92138023
    270.1416 44 69.3376354646134-66.729738487646i 96.23183304
    270.2637 39 -0.583859481532727+116.405077277761i 116.4065415
    270.3857 28 -38.0499198120679+52.356247172612i 64.72227604
    270.5078 13 -2.13915027287317-32.816964941318i 32.88661053
    270.6299 -2 58.0421614180693-31.6584807138993i 66.1146875
    270.752 -16 14.6804761819667+53.006807575928i 55.00216387
    270.874 -28 62.5510961402042+34.0116661484649i 71.19995128
    270.9961 -36 237.167143999627+29.7084646025705i 239.0205997
    271.1182 -40 12.8694538844647+33.197972497557i 35.60517127
    271.2402 -40 126.498463782523+26.34962198224i 129.2136367
    271.3623 -39 -107.949939092879+87.3731595699819i 138.8785742
    271.4844 -36 29.5191257099634-23.0489898771881i 37.45176521
    271.6064 -32 75.7699298507546+100.973803967758i 126.2410051
    271.7285 -18 -66.4518066495584-33.1591392263176i 74.26554464
    271.8506 -10 16.9769760979527-66.5741497658942i 68.70469514
    271.9727 -3 80.9589027482725-21.8248140425248i 83.84906942
    272.0947 3 113.59299923421+3.33174083444464i 113.6418496
    272.2168 8 48.843659127806-2.67060027967409i 48.91661418
    272.3389 10 36.1342627837103-54.988572827335i 65.79838971
    272.4609 10 -14.8885776308525+180.465917413791i 181.0790355
    272.583 8 11.2969364455587+54.35249803243i 55.51409565
    272.7051 5 -42.7234725911781+172.656489475777i 177.8638762
    272.8271 2 15.0302348041159-94.287509944029i 95.47796861
    272.9492 -2 130.098313325802+9.76930549048711i 130.4645947
    273.0713 -6 -74.6225516082506+36.6901741470105i 83.1546396
    273.1934 -11 68.5366725546182-95.9310158575354i 117.8984109
    273.3154 -15 133.741373453691-5.41394638334941i 133.8509088
    273.4375 -19 47.6440183402319+66.4261490408926i 81.74586081
    273.5596 -23 43.7195786127536+97.9397947007492i 107.2548598
    273.6816 -28 0.110730350170343+114.013407423564i 114.0134612
    273.8037 -29 124.196218860288+158.047225806798i 201.0065331
    273.9258 -29 -26.1625423871337-6.952735893512i 27.07063281
    274.0479 -26 -64.098651905178+19.2726161470714i 66.93333183
    274.1699 -22 -47.0468465985177-84.3967126095904i 96.62406985
    274.292 -14 67.911603566945-53.881895873943i 86.69051045
    274.4141 -5 149.49890496842+99.9215645645772i 179.8172451
    274.5361 4 40.8465726109121-11.0149723423101i 42.30569831
    274.6582 14 120.056739842671+3.22151971953963i 120.0999541
    274.7803 22 -43.393303970628+547.006476297578i 548.7249438
    274.9023 26 41.4192740565245+37.4085323474965i 55.81177793
    275.0244 26 11.7692104112103+71.3183233236223i 72.28289947
    275.1465 21 502.086494864063+396.847860095913i 639.9836501
    275.2686 11 -88.8805429898455+11.673162225241i 89.64381539
    275.3906 -1 -46.3493225629554+50.5199423627173i 68.56036959
    275.5127 -11 -321.269543870973+34.5478587826292i 323.121764
    275.6348 -12 26.2898455735248+25.8043292275629i 36.83774406
    275.7568 4 123.762579029642+51.325920338712i 133.9833052
    275.8789 36 5.89070896540943-243.689299401739i 243.7604871
    276.001 81 27.4055920924943-19.9444231487714i 33.89463811
    276.123 135 37.3902572695529+77.2308261978942i 85.80577984
    276.2451 192 53.7017568610313+148.520408713624i 157.9309675
    276.3672 245 31.2262815685876+75.2404627924717i 81.46292348
    276.4893 287 -37.0380738851868+19.8189494200449i 42.00725739
    276.6113 310 -167.14742027616+218.573508437936i 275.1592969
    276.7334 310 34.6023785470852+134.062965077616i 138.4565029
    276.8555 285 -9.41644469093671+15.7420435464073i 18.3434284
    276.9775 237 128.013165708813-165.507800331261i 209.2371921
    277.0996 172 -31.5008551772189+66.0315305890456i 73.16055569
    277.2217 96 -145.350144310456+57.0941074915231i 156.1614599
    277.3438 17 155.042156666307+105.260016000785i 187.397282
    277.4658 -55 142.922322382825+104.938020205305i 177.3098371
    277.5879 -157 2.7045764972363+27.2880246554314i 27.42172539
    277.71 -179 -16.2880383669672+80.1819013066197i 81.81954223
    277.832 -184 -28.066495000442+59.3121747016064i 65.61754498
    277.9541 -175 97.9295553567445+93.264068038744i 135.2345525
    278.0762 -158 -20.6762972064444-122.930734381509i 124.6574295
    278.1982 -138 -151.90826249586-27.1049769516039i 154.3074852
    278.3203 -121 -0.126864740856718+5.32340071336785i 5.324912189
    278.4424 -110 138.365772666352+110.368372540982i 176.992273
    278.5645 -108 5.05604076370435+83.4159174313556i 83.56900639
    278.6865 -111 96.9087799302391-81.3174784665124i 126.5063
    278.8086 -117 -26.1719676062993+54.4900161847816i 60.44943136
    278.9307 -121 -80.1837469087977+129.614292280177i 152.4116073
    279.0527 -118 -11.7686614147616+2.1427112327068i 11.96213204
    279.1748 -105 -75.997771674781+3.93460502320795i 76.09955595
    279.2969 -83 -68.1380952618295+11.6343957002819i 69.12423012
    279.4189 -52 -141.444129053816+83.0248024865923i 164.0108517
    279.541 13 -31.1505644086342+100.174139247363i 104.9057474
    279.6631 36 84.8405553042626-38.5781674330492i 93.19975765
    279.7852 45 14.554733367156+12.5502947012922i 19.21848486
    279.9072 39 -31.7276403518172+65.4614657741322i 72.74508
    280.0293 16 -15.1887877350281-64.2052398287182i 65.97736047
    280.1514 -19 10.8716185314696+139.229160468856i 139.6529671
    280.2734 -62 -18.8251573472171+97.3431336611829i 99.14672067
    280.3955 -104 -32.6788453163084+18.7910684185117i 37.69630199
    280.5176 -139 -16.5408980706306+95.8604401360644i 97.27705429
    280.6396 -160 53.257632782309-3.32989258441819i 53.36163073
    280.7617 -163 41.9338178829312+61.9309768441994i 74.79231896
    280.8838 -149 -9.77431867492464-59.07694662979i 59.88007121
    281.0059 -118 -28.312436274116-55.4527842280104i 62.26239095
    281.1279 -77 27.3650938566534+92.5868812656975i 96.54625288
    281.25 -32 103.346947600821-65.0756302071663i 122.1287404
    281.3721 9 -55.9614098114624-62.0505482957952i 83.55806324
    281.4941 62 88.0137276324717+90.8110794994384i 126.4637039
    281.6162 66 37.3663834959625+31.2985195950546i 48.74262964
    281.7383 55 -55.8655011728799-10.529663027314i 56.84916908
    281.8604 33 17.5341803027603-5.6406937286929i 18.41914506
    281.9824 5 -108.693920507947-9.9593149811742i 109.1492387
    282.1045 -23 -42.7721006577229+74.6944897520261i 86.0739182
    282.2266 -46 -85.1348014483731+9.8543912604157i 85.70322891
    282.3486 -62 -90.6546780206281-47.0396224675387i 102.1322512
    282.4707 -66 23.5258860502139+19.7833318469557i 30.73837233
    282.5928 -60 -68.954987831997+22.5115254799531i 72.53660542
    282.7148 -46 -30.6801590692168-1.38695987920846i 30.71149326
    282.8369 -29 120.82472954264-28.38221243714i 124.1135176
    282.959 -11 -92.4199999044868-8.14419399462625i 92.77814548
    283.0811 4 -106.055931733347+133.552668529299i 170.5408336
    283.2031 14 -5.58233331938771+65.4173634327864i 65.65511316
    283.3252 19 -9.59523909254222+97.4337447488492i 97.90507254
    283.4473 17 -10.8044097979684+9.36118829674251i 14.29570276
    283.5693 13 -108.094156665456+32.1662364558297i 112.7786038
    283.6914 10 -24.3728105588336-3.33208275184188i 24.59952581
    283.8135 8 139.089268984372-26.5659663755664i 141.6035851
    283.9355 7 -91.922655031832-37.2300895877411i 99.17587448
    284.0576 8 -32.2468708231602-151.615912039628i 155.0072433
    284.1797 8 20.6495066189309+19.799152707908i 28.60784109
    284.3018 7 -8.55757643571019+189.004412489776i 189.1980445
    284.4238 4 107.308979511239+91.0960719668117i 140.7611857
    284.5459 0 -38.1447012173592-115.048943645986i 121.2075809
    284.668 -7 -80.4073729893885-9.53680546596041i 80.97095954
    284.79 -14 -74.0621746515239-3.51349854610885i 74.14546774
    284.9121 -20 -68.3010500471098+10.1389166334689i 69.04948275
    285.0342 -22 -31.118474981271-138.494743961775i 141.9477143
    285.1563 -20 -26.0262457727858-139.085835646593i 141.4999475
    285.2783 -13 -77.4615528195231+132.188620031932i 153.2126739
    285.4004 16 -56.2108498489978+14.3797921385635i 58.02101398
    285.5225 34 211.679870288224-46.9851791755827i 216.8316733
    285.6445 51 -10.6150360783531+40.4827347376479i 41.85129392
    285.7666 66 -91.8395003869764-83.2920200772729i 123.9840895
    285.8887 75 -261.849617291269+71.1191402981774i 271.3358697
    286.0107 77 2.9027750774585+57.8243772691623i 57.89719086
    286.1328 73 -25.9069231310211-22.4388981444063i 34.2735002
    286.2549 64 -111.582481297057-88.5166485965069i 142.4283933
    286.377 51 -73.4779436757901-50.21776512835i 88.99905697
    286.499 38 53.4641749813816-2.66209525789304i 53.53040965
    286.6211 27 239.652406572655+99.083976372699i 259.3278048
    286.7432 20 -34.6298835701837-121.103375490998i 125.9573594
    286.8652 19 1.52321327116083-38.8550327012771i 38.8848781
    286.9873 22 -104.059463379618+300.119542478787i 317.6477793
    287.1094 29 -63.5203049630277+25.3116457648167i 68.37769047
    287.2314 38 -74.6235238520684+132.048764503696i 151.6757941
    287.3535 54 -147.610696835673-634.281027862249i 651.2306351
    287.4756 59 -51.0663859437175-36.9570446508069i 63.03648882
    287.5977 62 22.2728485249157-25.1371006041804i 33.58502059
    287.7197 64 -540.361744057011-206.683436710973i 578.5402816
    287.8418 67 65.2305190030579-124.866284462643i 140.8779955
    287.9639 72 23.7895755665705+1.2239493781849i 23.82104023
    288.0859 80 0.605676484576634-189.357583570221i 189.3585522
    288.208 90 90.4793357933999-23.336458320593i 93.44035794
    288.3301 101 50.7743316132156+67.9229805970994i 84.80308983
    288.4521 113 -47.7280043357368+47.5305062705963i 67.35808358
    288.5742 124 16.9384557463468+48.7736947534615i 51.6312365
    288.6963 131 -100.92176942464-149.738677970143i 180.573739
    288.8184 134 -88.8078222879878-25.3019574019102i 92.34185588
    288.9404 131 -92.9628008620616-43.5559324442727i 102.6606137
    289.0625 123 -30.23508390343-61.9965922661476i 68.97635647
    289.1846 112 42.1326256122388-70.8953020487847i 82.47000663
    289.3066 84 104.510657228192+28.0047856570113i 108.1977148
    289.4287 71 -52.4457298433044+62.2601545302839i 81.40565964
    289.5508 63 46.0431061371614-89.015440193069i 100.2183427
    289.6729 60 64.0520543147773+24.16104816333i 68.45744598
    289.7949 63 -150.403511502168-2.09289656025453i 150.4180723
    289.917 70 -20.3026320031078-32.0036118953878i 37.900238
    290.0391 80 -168.323919491275-10.7234763542206i 168.6651559
    290.1611 93 51.2374849011859-9.25241120913415i 52.06617877
    290.2832 103 41.0171525173947-32.0147216123701i 52.03219389
    290.4053 109 68.1606150695743-100.555134244286i 121.4792347
    290.5273 111 78.3232239233947-42.3033731653499i 89.01742968
    290.6494 110 88.9378310294156-7.03941629294309i 89.21598046
    290.7715 107 -7.94685937284157+69.4293545402377i 69.882672
    290.8936 102 -74.8786016131835-113.043454106213i 135.5936116
    291.0156 96 24.6706619588794-7.66849569701373i 25.83500315
    291.1377 89 -123.308858590451+58.1726441857004i 136.341964
    291.2598 71 18.3251513042191-92.4048604074299i 94.20440222
    291.3818 59 -10.9105404747013-13.7819501655081i 17.57788508
    291.5039 45 104.511223655868-257.876691814434i 278.249859
    291.626 30 -30.9959373635517-80.7338832645145i 86.47952382
    291.748 17 -44.6044188664076-70.5301977258315i 83.45096149
    291.8701 6 90.9077654743119-26.7456034075999i 94.76048293
    291.9922 -2 55.2274630840151+5.0913793331739i 55.46165182
    292.1143 -9 -64.1628504904875+25.3577764819934i 68.99194309
    292.2363 -14 -93.7729125555501-7.91624967494477i 94.10646172
    292.3584 -20 105.680560290469-44.4345581462464i 114.6420986
    292.4805 -25 42.3853738966574+33.5648137303921i 54.06585467
    292.6025 -31 -38.8297222656497-212.852532594904i 216.3653114
    292.7246 -36 -31.4051375257044-27.1761827923627i 41.5310435
    292.8467 -38 -70.9971312227902-109.24766285484i 130.2906155
    292.9688 -36 107.381014347075+131.088906570909i 169.4549606
    293.0908 -31 89.080050629716-32.8592502956066i 94.94727879
    293.2129 -12 93.1775340909178-106.065655949855i 141.1806511
    293.335 -3 50.8614102542325+14.2390771584924i 52.81698942
    293.457 0 -11.9227300337591+65.3558780557832i 66.43449622
    293.5791 0 23.495106705421-86.2537065999034i 89.39643137
    293.7012 -7 36.3304410036469-185.869701496678i 189.3870293
    293.8232 -19 15.565661725042-78.7721610038486i 80.29534964
    293.9453 -33 -112.703594665076-35.7965917766814i 118.251834
    294.0674 -46 104.608823467679+72.9179518702917i 127.514837
    294.1895 -58 13.6710233419775-85.0386251362467i 86.13050937
    294.3115 -66 178.56734920074-43.4992782303447i 183.7892418
    294.4336 -72 -5.08495422538036-89.8315942742216i 89.97539713
    294.5557 -74 15.7719880863935-71.0278918843938i 72.75793451
    294.6777 -74 76.3168945474595+45.8786017278231i 89.04557535
    294.7998 -72 -20.3854730325215-4.66555528403823i 20.91255405
    294.9219 -68 -53.8173510061309-46.7624434037747i 71.29539524
    295.0439 -60 -11.5921652620329-29.2117990699504i 31.42781412
    295.166 -38 67.9690686628983+37.1563388790597i 77.46217021
    295.2881 -26 -25.824731228727-23.0322562590219i 34.60349074
    295.4102 -14 133.58422648071-115.402687786642i 176.5291078
    295.5322 -6 -78.9806168006402-100.656676359995i 127.9441453
    295.6543 -3 67.9003301304307-62.8353347701231i 92.51342674
    295.7764 -6 -17.1394312502153+19.0328613661226i 25.61269051
    295.8984 -14 65.9065192924478+28.7578488720688i 71.90746246
    296.0205 -24 77.0821677521876+5.71116255358419i 77.29345356
    296.1426 -34 29.720825858051+36.9921076344675i 47.45253963
    296.2646 -41 -2.56586483908965-10.3251721251627i 10.63921246
    296.3867 -46 66.8167126714493-31.1831439183007i 73.73507684
    296.5088 -48 88.4307539623365-60.5685896178842i 107.1846645
    296.6309 -47 -63.3838707607262-32.3213317766202i 71.14902361
    296.7529 -45 2.42472609822039-66.6568572761062i 66.70094391
    296.875 -45 -23.6809119319068+67.3677972723115i 71.40872285
    296.9971 -45 100.007655957946-22.7449787789365i 102.5615196
    297.1191 -46 38.1102252569101+71.4781093803591i 81.00314432
    297.2412 -44 46.8609640337004-67.6657193927467i 82.30795545
    297.3633 -39 42.9867656969212-48.0400127737489i 64.46475667
    297.4854 -31 24.7619964621439+64.468772967956i 69.06069184
    297.6074 -20 -30.2174612912901-20.2191811363931i 36.35808373
    297.7295 -8 -24.0111701822785-56.048866410188i 60.97550098
    297.8516 1 -14.9879454762598-43.4757688819215i 45.98674798
    297.9736 10 7.35093212861386+70.6941541727777i 71.07530962
    298.0957 16 76.886138533169-80.5072094191167i 111.3233536
    298.2178 20 107.219345176585+63.701847881105i 124.7153295
    298.3398 23 107.43557005687-120.427928812051i 161.3855252
    298.4619 25 28.141119026123-5.57525906640772i 28.68808278
    298.584 24 -41.5723029751341-55.5891983307104i 69.41480639
    298.7061 21 45.6739023236747+9.16160378265495i 46.58369175
    298.8281 16 64.8746387500131+67.9251131461606i 93.92837563
    298.9502 7 8.2808738179337+52.3591324485283i 53.00992003
    299.0723 -17 -10.3472894984556-53.3425233280743i 54.33683092
    299.1943 -30 94.7399385418792-42.2420569588406i 103.730648
    299.3164 -39 108.393899356199-17.4189853463968i 109.7846003
    299.4385 -42 12.3677637180426-166.988422525572i 167.4457967
    299.5605 -40 127.073660928429-8.17930888071487i 127.3366263
    299.6826 -32 -53.9050266558046-84.9953278017798i 100.6476907
    299.8047 -21 102.984682441147+74.2656055104454i 126.969386
    299.9268 -8 96.1965410249818+43.4614482503164i 105.5588556
    300.0488 2 186.443179235129-20.0241491154978i 187.5154011
    300.1709 10 92.5275098724148+10.8921551363177i 93.16640557
    300.293 15 -28.0027528745256+63.5639745785521i 69.45885856
    300.415 17 -38.2295334085314-22.94926653831i 44.58885577
    300.5371 17 78.376912238591-93.7152231248718i 122.1698957
    300.6592 16 26.3192948610836-58.4358152983363i 64.08938907
    300.7813 15 -110.827594082451-36.9632582846928i 116.8290977
    300.9033 14 127.251246502002+39.9823740418309i 133.3846692
    301.0254 8 40.4131119370947+45.5210693364456i 60.8718931
    301.1475 2 245.957828218325-5.95728520301509i 246.0299626
    301.2695 -5 -29.4959275545904-10.2159543205623i 31.21498783
    301.3916 -15 56.1287291472358-74.1872630204168i 93.02786803
    301.5137 -23 45.1441554071268+121.285833742744i 129.4150232
    301.6357 -30 -7.50059909359965+59.1716923673962i 59.64518559
    301.7578 -34 -15.0475830236597-10.3997247609785i 18.29163825
    301.8799 -37 19.7235631154035-90.7706516718094i 92.88880528
    302.002 -38 39.7428047380145+10.3755470804646i 41.0748403
    302.124 -39 73.3076948329461+43.1128999271633i 85.04551876
    302.2461 -41 234.35985562945-123.877354768104i 265.0851579
    302.3682 -42 18.8290251223915+40.4610253885455i 44.62764572
    302.4902 -43 119.815759213016-46.1485270413226i 128.3958827
    302.6123 -41 -95.2530138692983+140.715636641214i 169.9235918
    302.7344 -36 13.3214691165227-0.70916129633946i 13.34033167
    302.8564 -27 126.172945864739+54.6305928619856i 137.4922323
    302.9785 -3 -74.1047489355071+18.1722787864114i 76.3003639
    303.1006 6 14.1432222883575-40.3017395271689i 42.71136787
    303.2227 12 77.3135428802151-79.809508147001i 111.1167922
    303.3447 12 108.738049337824-83.0911833038822i 136.8506782
    303.4668 6 58.5849896665793-5.55340232987601i 58.84761076
    303.5889 -4 17.6131439525969-80.2168487327169i 82.1277399
    303.7109 -17 106.736650970232+170.204798859634i 200.9039228
    303.833 -30 33.3060401029782+6.14995221479573i 33.86907468
    303.9551 -41 51.9468868684731+184.323959799025i 191.5040501
    304.0771 -49 -30.2585584710108-37.1564696012329i 47.91850993
    304.1992 -55 112.950893433129-51.6544606429533i 124.201802
    304.3213 -57 -49.0839657349099+98.4784833133082i 110.0329377
    304.4434 -57 -20.112632419187-63.3162078259931i 66.43387808
    304.5654 -56 99.7449320659702-40.4686176968485i 107.6418157
    304.6875 -52 107.00118356708+43.1719554349743i 115.3822821
    304.8096 -45 92.8660764429916+89.7706413925743i 129.1622089
    304.9316 -24 52.5161097198095+113.057650598731i 124.6594326
    305.0537 -12 189.963495499106+93.4112659465841i 211.6879643
    305.1758 -1 7.54689561853455+33.0883460868621i 33.93809482
    305.2979 5 -2.10238155478714+54.5254943071756i 54.56601083
    305.4199 8 -55.0187035247737-31.6376135325546i 63.46649768
    305.542 6 33.4006494601401-78.2242514535836i 85.05666876
    305.6641 -1 181.025351226454+21.838601047429i 182.3378795
    305.7861 -11 18.0638378366178-18.4207699420184i 25.79974811
    305.9082 -21 144.382317154229-84.5057729344756i 167.294588
    306.0303 -29 282.188353553726+530.596069209285i 600.9679322
    306.1523 -34 58.8708972688625+4.16971983322782i 59.01837941
    306.2744 -36 63.1826662654169+55.0124180228816i 83.77598375
    306.3965 -36 682.027820363027+113.998856018899i 691.4894699
    306.5186 -35 -60.0655582173896+71.1458714733315i 93.11072071
    306.6406 -35 4.32334179507242+98.712794586939i 98.80742431
    306.7627 -36 -232.194792290506+216.969602793494i 317.7896004
    306.8848 -39 66.1031910886533+1.31489601611931i 66.11626747
    307.0068 -38 106.86420306988+19.1019047533484i 108.558006
    307.1289 -34 -132.960204381295-200.317245299397i 240.4275665
    307.251 -26 25.1699598017725-31.0952933159072i 40.0055514
    307.373 -16 63.033428751126+115.476272341986i 131.5598062
    307.4951 -4 150.834375158738+97.957448956895i 179.8518016
    307.6172 6 76.0141767292869+64.1045091746456i 99.43612603
    307.7393 15 8.49555149949273+14.2815635373973i 16.61738404
    307.8613 22 -13.7737924731636+294.695363073731i 295.0170747
    307.9834 26 120.670299687568+98.017813157685i 155.4632205
    308.1055 29 6.24111580351891+10.2016903272879i 11.95934831
    308.2275 29 40.5140018413185-183.67939615945i 188.0944043
    308.3496 28 27.6656755787121+72.3916784436587i 77.49803038
    308.4717 25 -96.2275595377518+138.792952123764i 168.8882079
    308.5938 18 203.43316196226+34.7825331358041i 206.3852611
    308.7158 9 156.849010304351+35.5692822331605i 160.831545
    308.8379 -17 33.8197102587159+71.9123919996326i 79.46801196
    308.96 -30 26.3025599955322+92.886951752957i 96.53916546
    309.082 -39 21.0763712789467+108.332470398114i 110.3636605
    309.2041 -43 144.392300612999-7.89471724310582i 144.6079633
    309.3262 -41 -66.9113189283374-75.7582849196648i 101.0764183
    309.4482 -34 -132.216080199352+66.6246158230782i 148.0538122
    309.5703 -23 15.5806449737203+23.8519796356923i 28.48988295
    309.6924 -10 151.424922564078+26.5248106225256i 153.7305199
    309.8145 0 41.1297708136577+83.9386127872836i 93.47378651
    309.9365 7 12.4844325249604-133.544905759398i 134.1271893
    310.0586 11 8.58040152050424+67.4878991013478i 68.03116797
    310.1807 12 -28.9313343752175+160.290792300645i 162.880816
    310.3027 12 -15.6549955428329+38.2164143449874i 41.29858606
    310.4248 11 -43.2343586592687+55.4937680778378i 70.34748087
    310.5469 10 -36.7217539816437+11.6436114028314i 38.52351103
    310.6689 9 -74.8331800650806+180.438474031671i 195.3408502
    310.791 4 33.1005179235393+134.624197836138i 138.6337583
    310.9131 -1 28.2705434253425-42.9859735857209i 51.44917444
    311.0352 -9 22.3014438618104+47.546576747999i 52.51696258
    311.1572 -20 10.5255483750823+101.732155618576i 102.2752104
    311.2793 -30 -8.9973487780842+0.141154247322685i 8.998455954
    311.4014 -39 88.1584815713191+132.296753636268i 158.9790832
    311.5234 -46 48.9930005969177+109.391727641703i 119.8618546
    311.6455 -50 -49.6521260761209+27.44578405602i 56.73274792
    311.7676 -53 58.1014249362077+107.243364044691i 121.9709585
    311.8896 -54 43.6197213318511-14.7975941023588i 46.06135995
    312.0117 -54 75.6159828819483+59.1073927777566i 95.97635515
    312.1338 -53 -66.038869983288-21.7845037464728i 69.53917566
    312.2559 -50 -83.5292819674278-23.5734810621321i 86.79199246
    312.3779 -43 67.317741335368+65.2363187185447i 93.74142936
    312.5 -32 29.8061325481597-115.176623509138i 118.9708374
    312.6221 -17 -58.1767260611975-32.0657125483744i 66.42846812
    312.7441 17 119.179007880771+33.8020559448604i 123.8798406
    312.8662 31 52.0983611035637-7.07638263316301i 52.57674791
    312.9883 41 -72.3789440641635+30.1630645051738i 78.41251178
    313.1104 43 -16.9965396764602+1.67321662639286i 17.07870062
    313.2324 38 -94.8441702921587+52.8605318972808i 108.5801661
    313.3545 27 33.4175266136014+119.585678160824i 124.167087
    313.4766 14 -104.773716261555+59.5973543330597i 120.5378624
    313.5986 0 -88.1637209338204+42.7557250716211i 97.98415033
    313.7207 -11 56.0666102211634+26.0805263462242i 61.83573915
    313.8428 -21 -38.0494326332428+80.0010322408127i 88.58851214
    313.9648 -28 -21.6621122314593+22.9830905565544i 31.58274146
    314.0869 -32 71.0902962312784-69.8701063265278i 99.67779079
    314.209 -35 -94.7961282353561+62.2795528605052i 113.4241977
    314.3311 -36 -12.894095092773+201.235871345173i 201.6485398
    314.4531 -35 42.0986231637589+90.0300302244081i 99.3866209
    314.5752 -31 63.4578420302215+101.560490097137i 119.7557133
    314.6973 -13 4.20920416941817+23.8671097733488i 24.23543539
    314.8193 -1 -87.6721766208906+126.009633966127i 153.5084311
    314.9414 9 -17.1128961453303+51.9706665927874i 54.71564128
    315.0635 18 95.2770426658487-105.575051884487i 142.2104301
    315.1855 23 -63.69543184111+33.9099421585704i 72.15949151
    315.3076 22 -123.104039436797-104.39714138613i 161.4105562
    315.4297 15 26.0005101708487+7.16592350459931i 26.96992749
    315.5518 4 119.17298499841+155.836565614089i 196.1816391
    315.6738 -7 194.817090254431+10.0751258398312i 195.077438
    315.7959 -17 -111.964208086506-56.3588274135106i 125.3487189
    315.918 -25 -66.7822914223976+76.8110435423834i 101.7831561
    316.04 -31 -79.1341644782147+76.1974080736545i 109.855637
    316.1621 -33 -48.0136023290295+55.9866044422508i 73.75503973
    316.2842 -34 -102.371668355373-94.4831639184436i 139.309105
    316.4063 -34 -131.322618548923-100.467242123988i 165.3459914
    316.5283 -35 -35.6139343793438+181.152401967375i 184.6200018
    316.6504 -36 -37.6970750312919+53.7648131136726i 65.66372359
    316.7725 -34 158.282099020346-156.061239473115i 222.2798536
    316.8945 -29 11.8651081048895+63.3955676926547i 64.49634713
    317.0166 -23 -90.2142631733687-25.1183371981013i 93.64584424
    317.1387 -15 -194.139950501708+250.278423500308i 316.7484959
    317.2607 -5 13.0316166373659+43.0569533795223i 44.98582295
    317.3828 2 -46.0214257606632-25.4110498048896i 52.5708387
    317.5049 10 -151.511809356781-35.9650909449487i 155.7219193
    317.627 16 -107.715015372546+15.182906648854i 108.7798014
    317.749 21 43.3814411626415-20.4497546436187i 47.95979464
    317.8711 26 265.188419770524-45.0322024271883i 268.9847528
    317.9932 30 -108.610179793196-82.3775585572364i 136.3166656
    318.1152 31 -24.0679582767144-4.77246705225354i 24.53656572
    318.2373 29 55.1676837555839+371.901595959297i 375.9711032
    318.3594 20 -37.1998556490426+83.1885247068459i 91.12716337
    318.4814 6 29.6942882826371+159.298794774624i 162.0427622
    318.6035 -38 -516.070297040035-475.346089151532i 701.6284315
    318.7256 -61 -42.781761701801-0.168525234849866i 42.78209363
    318.8477 -77 -38.0664406613168-3.75860066721032i 38.25154877
    318.9697 -82 -602.981138434849+140.81799402159i 619.2059114
    319.0918 -71 -19.4228236647888-83.2985805868145i 85.53303226
    319.2139 -43 -0.445824069519617-40.713616077395i 40.71605695
    319.3359 0 -124.218250877079-108.521162615235i 164.9454958
    319.458 55 52.5272022113394-44.5781922073228i 68.89355697
    319.5801 117 96.8724669154365+59.9996817840874i 113.9483947
    319.7021 178 0.365928092769224+71.6471507374086i 71.6480852
    319.8242 234 38.9536315033805+11.5376781821907i 40.62638829
    319.9463 279 -198.292450594999-81.6560810702233i 214.4472232
    320.0684 308 -112.463780478331+48.8193309194472i 122.6027283
    320.1904 319 -87.2267659722329-3.48664411277997i 87.29642254
    320.3125 309 -82.1871764235619-31.6432689388662i 88.06831688
    320.4346 278 -8.10606290571184-90.16039541883i 90.52405845
    320.5566 166 76.795788007365+2.54719198817483i 76.83801951
    320.6787 92 23.2563066353208+84.205331733711i 87.35784848
    320.8008 15 -9.06923611748888-78.7182234853983i 79.23894088
    320.9229 -58 89.5997112074767-6.30144371679346i 89.8210245
    321.0449 -121 -136.44382783736+112.60624558966i 176.9098208
    321.167 -168 -57.6430316576879+9.79492140635803i 58.46930463
    321.2891 -197 -197.426581184282+71.3735899506332i 209.9319992
    321.4111 -208 66.5548632787295-40.949180602017i 78.14336324
    321.5332 -204 3.35861245297257-63.1982943300264i 63.28747652
    321.6553 -190 -11.1712671016909-119.175362592509i 119.6978039
    321.7773 -169 38.7135508000773-69.4442571465022i 79.50625048
    321.8994 -145 38.7474540667922-28.1178206981538i 47.874597
    322.0215 -122 58.5743073316999+64.1513916122321i 86.86973308
    322.1436 -99 -130.155974812711-22.6459986695761i 132.1113887
    322.2656 -77 18.3546101047176+4.42480345489916i 18.88042896
    322.3877 -57 -44.2127424041022+184.30957662247i 189.5383513
    322.5098 -24 -12.3951230552788-98.3689850838827i 99.14684212
    322.6318 -12 -30.8277914809496+26.7100324617201i 40.78944179
    322.7539 -5 -53.6971304337948-276.105612780606i 281.2786718
    322.876 -4 -105.768388351673-28.933445693031i 109.6544402
    322.998 -9 -58.9925806491054-17.4618784676741i 61.52269314
    323.1201 -18 80.9408063302811-55.7711113773826i 98.29461325
    323.2422 -29 33.3861581872214-5.52513285480123i 33.84025194
    323.3643 -40 -34.8378191400663+57.6469966075575i 67.35614196
    323.4863 -50 -115.891486041836+68.2507271092244i 134.4953467
    323.6084 -57 58.5297923138066-106.719911290783i 121.7163755
    323.7305 -62 77.4211797628061+51.3244502718655i 92.88831074
    323.8525 -65 -147.930265772819-135.706671274321i 200.7477625
    323.9746 -65 -43.0097738564389+6.45090146368303i 43.49085854
    324.0967 -64 -111.092177019039-60.9365196069952i 126.7072658
    324.2188 -60 95.3695701037328+62.7510158811868i 114.162362
    324.3408 -53 79.6867366609256-36.860633209294i 87.79910182
    324.4629 -29 16.5130881674236-105.021324501757i 106.3116206
    324.585 -15 85.348126989815+27.6949535213312i 89.72911028
    324.707 -2 0.208222947303568+85.481862190659i 85.48211579
    324.8291 7 -42.7694549269615-68.0745137989663i 80.39506019
    324.9512 11 -80.5366841764622-162.2420291722i 181.1315365
    325.0732 10 -37.7472978427961-38.7929712642817i 54.12719385
    325.1953 3 -122.71388320497+35.7326176366586i 127.8104733
    325.3174 -7 153.75781813641+3.46498745771405i 153.7968555
    325.4395 -17 -59.2822035085481-76.1157325969285i 96.47789592
    325.5615 -25 155.368823827602-110.04467795204i 190.3924961
    325.6836 -30 -73.2380455364378-74.0771922383281i 104.1692936
    325.8057 -31 -17.9446370675745-65.8705218319654i 68.27104544
    325.9277 -31 68.6110386852136-5.62082605903001i 68.8408913
    326.0498 -29 -11.5437725005226+64.0182866328684i 65.05074717
    326.1719 -28 -93.0886990520682-16.4129799211149i 94.5245566
    326.2939 -29 -11.6113564327046-0.772333932034798i 11.63701413
    326.416 -33 96.8015491843258-10.1253511666574i 97.32965972
    326.5381 -32 -16.8906766586007-6.16858081547534i 17.98183381
    326.6602 -29 43.5044791254903-161.256488261562i 167.021839
    326.7822 -22 -95.1704377114423-29.0426787533844i 99.50321303
    326.9043 -12 33.8523032918213-136.453880171401i 140.5903263
    327.0264 0 -15.5176588278882+63.589675184183i 65.4556684
    327.1484 12 39.829486221286-6.71132944603625i 40.39096329
    327.2705 23 91.55367457106-39.1890809957995i 99.58845012
    327.3926 32 71.8321474720607+50.5961239786598i 87.86253566
    327.5146 38 -11.5195049087348-8.86818459157554i 14.53766457
    327.6367 43 34.4001147896415-77.627114619434i 84.90781367
    327.7588 46 27.2902081797459-84.2739431617941i 88.58246417
    327.8809 46 -42.987083275048+23.9583452694389i 49.21271824
    328.0029 43 -55.0243924726648-110.021721299805i 123.0140761
    328.125 37 4.38830950291251+93.6204241390115i 93.72321525
    328.2471 28 92.9845437902676-80.0098135199242i 122.6690492
    328.3691 2 105.311376864876+21.3302853889986i 107.4498356
    328.4912 -9 4.3629201411562-141.442751125282i 141.5100241
    328.6133 -17 44.5247794434539-90.3906431085869i 100.7617206
    328.7354 -20 70.6249001938413+19.889150382964i 73.3720303
    328.8574 -16 -25.0228534936636+22.5595549787819i 33.69089963
    328.9795 -6 -42.7063559340104-91.4527333637205i 100.9328256
    329.1016 7 -46.0561287866332-5.21072094583369i 46.34995805
    329.2236 22 56.4862294465366+95.6078779143333i 111.0475593
    329.3457 35 -12.7436365830703-78.1909387152268i 79.2226178
    329.4678 45 116.625829791813-9.74469510083634i 117.0322317
    329.5898 50 22.5579450301046-172.594148853504i 174.0620611
    329.7119 53 51.2848847351436+18.5087644189758i 54.52259864
    329.834 54 -65.6862406058485-46.7646578151503i 80.6325953
    329.9561 54 71.9994539786294+4.69122610360819i 72.15212385
    330.0781 54 103.837284655074+6.51729767953782i 104.0416112
    330.2002 55 13.7146368973837+51.5918065742498i 53.38357211
    330.3223 56 -36.7759017688998-27.1674508780625i 45.72239427
    330.4443 53 79.133950992817-49.5046231480737i 93.34286214
    330.5664 48 76.9377518303369-44.807327607796i 89.03434317
    330.6885 42 -66.0364938576638-109.552070379819i 127.9158889
    330.8105 35 57.8171422784357-45.2352524961393i 73.41014923
    330.9326 30 -132.128875381692-19.1334805193328i 133.5070402
    331.0547 26 151.365267059804+40.7135956642241i 156.7451465
    331.1768 24 84.9279135057154-17.6374404578214i 86.74001267
    331.2988 23 138.158015491838-117.584508693336i 181.4214814
    331.4209 23 84.7641169095582-62.1036026205073i 105.0800313
    331.543 23 -10.1738766596263+62.4288664293498i 63.25243972
    331.665 24 -37.1674179015203+12.9389028161519i 39.35520499
    331.7871 25 34.6174481478714-86.7330326667702i 93.38622314
    331.9092 30 -60.2033476632596-41.3304833332605i 73.02500888
    332.0313 38 -116.043630643965+47.0593557030108i 125.2226304
    332.1533 50 142.731613466975-32.6013528185416i 146.4075192
    332.2754 80 79.2594179445549+36.8185820118509i 87.39372583
    332.3975 94 208.65445137985-163.937866493831i 265.3531687
    332.5195 103 -29.9256155878608+30.5032627109917i 42.73162183
    332.6416 106 -16.4652310007683-76.5711364892867i 78.32140688
    332.7637 102 118.112397348863+96.6166631807922i 152.5952752
    332.8857 93 42.3012598749737+85.483711862194i 95.37746894
    333.0078 81 -22.5310213541507+6.24825564142429i 23.38135201
    333.1299 70 -47.4142646710798-85.7946065657029i 98.02462451
    333.252 60 90.9916398929494+22.010250097064i 93.61586212
    333.374 53 100.857217516683-27.7769532219577i 104.6123198
    333.4961 50 107.248328511773-232.556712264504i 256.0953502
    333.6182 49 39.3364127775476+2.37654192160258i 39.40813776
    333.7402 51 58.8480405382904-90.254445071983i 107.7448687
    333.8623 54 55.9763342107843+187.332563105823i 195.5168514
    333.9844 58 -3.27331283349761-23.0866537053086i 23.31755039
    334.1064 65 155.634492555063-8.7286079259849i 155.8790681
    334.2285 82 -66.553290505251+90.0292471554012i 111.9580538
    334.3506 92 -15.0277806620248-56.5286045651753i 58.49202788
    334.4727 101 17.8962472335559-89.460973997244i 91.23344526
    334.5947 107 77.3940223298931-136.704128201798i 157.0918628
    334.7168 108 31.0427599762005-9.89818757952091i 32.58261905
    334.8389 103 -31.6874141738684-44.2736308748007i 54.44489515
    334.9609 94 212.378442420167+131.505511634065i 249.796522
    335.083 81 58.3774941360168+38.2362417665348i 69.78496977
    335.2051 68 188.545295558468+114.81500271047i 220.7528331
    335.3271 55 -42.9876874702252+6.55709334797707i 43.48490252
    335.4492 45 64.0859168849205-87.587906444407i 108.5294711
    335.5713 37 57.835070660538+148.127373699192i 159.0176539
    335.6934 30 -51.7003376461033-70.7180990308586i 87.60122398
    335.8154 24 76.5343950895651-87.9709585218245i 116.6036156
    335.9375 16 114.39087610869+30.083006544334i 118.2804287
    336.0596 7 162.371359899181+9.66791636315651i 162.6589288
    336.1816 -9 121.056290284245+86.6315148280513i 148.8611594
    336.3037 -14 246.251346287049-32.137341221908i 248.3395543
    336.4258 -15 12.4794992972855+63.072784213554i 64.29552093
    336.5479 -12 -9.17377650413788+61.6225006319861i 62.30161121
    336.6699 -5 -92.1408970253293+21.8761362720971i 94.70221879
    336.792 3 -32.54158839983-97.8928538003488i 103.159904
    336.9141 12 182.215611752863-85.9873350840531i 201.4853617
    337.0361 19 21.4121926356158-53.8929626604863i 57.9908046
    337.1582 24 101.267785665992-124.01760466197i 160.1109949
    337.2803 26 582.161170144637+284.144526180103i 647.8037819
    337.4023 27 52.374824069941+12.5489393420455i 53.85720077
    337.5244 25 94.5799574415624+60.619830237314i 112.3393616
    337.6465 21 666.251857432333-312.800101226751i 736.0267936
    337.7686 16 -12.3552702412446+128.341453329609i 128.9347949
    337.8906 7 46.2102155033481+91.6428704402039i 102.6343009
    338.0127 -2 -104.786196571538+354.38474523707i 369.5520189
    338.1348 -31 50.8868277281606+5.44506494347119i 51.17731889
    338.2568 -45 142.303590428-81.3818960524742i 163.9308539
    338.3789 -56 -248.941535118319-81.0998132211841i 261.8187686
    338.501 -61 10.7810205196281-30.1318163482377i 32.00244928
    338.623 -61 143.757837650657+55.1072418330943i 153.9581891
    338.7451 -55 199.415228984378+24.4868250387916i 200.9130114
    338.8672 -45 90.3145513575019+12.7649365312871i 91.21218006
    338.9893 -33 6.83641578312736+40.0097389866382i 40.58960205
    339.1113 -23 180.208817135326+278.89775780008i 332.0529733
    339.2334 -15 139.083153806059+26.2070206875183i 141.5306737
    339.3555 -11 19.3255764138795+24.8285594908092i 31.4632368
    339.4775 -9 -104.120869927884-181.970242254957i 209.652867
    339.5996 -9 36.7646874841594+42.7799853368442i 56.407175
    339.7217 -9 -1.65902007132092+224.479304760889i 224.4854352
    339.8438 -9 179.087190474535-101.489123187378i 205.8452426
    339.9658 -9 157.599496798149-73.2134210634597i 173.7751605
    340.0879 -10 41.8291663025652+34.3445008291066i 54.12230493
    340.21 -14 83.0605103368024+96.0819192271517i 127.0070218
    340.332 -21 64.4402199245743+105.939963576783i 123.9992654
    340.4541 -30 115.409736579424-71.7016266421097i 135.8695351
    340.5762 -39 -113.74016026563-8.69563176763776i 114.072074
    340.6982 -46 -118.156010007215+129.290070658211i 175.1478378
    340.8203 -51 10.0217560899944+18.5467497789851i 21.08121255
    340.9424 -55 202.084928408525-61.0151720578849i 211.095167
    341.0645 -57 104.77347257585+47.569258006431i 115.0665671
    341.1865 -58 -77.1289890505802-145.928161262449i 165.0572907
    341.3086 -59 42.0666126663379+96.1870949089521i 104.9836041
    341.4307 -60 130.435029667827+201.518362756677i 240.0478025
    341.5527 -60 39.7333297576739+47.8862299786741i 62.2240188
    341.6748 -58 -28.3144687138686+114.230706215374i 117.6875668
    341.7969 -52 -20.4704793377956+50.2086626332711i 54.2213088
    341.9189 -43 57.5624074967476+199.546168463423i 207.6827005
    342.041 -18 94.3492488315424+71.5488161213687i 118.4103621
    342.1631 -6 -31.1551733692657-75.9207188883627i 82.06461103
    342.2852 1 32.4782117847293+37.2153146298113i 49.3944722
    342.4072 2 77.7620833552331+77.0704852418273i 109.4842514
    342.5293 -1 -46.4415895208544-0.441582759441559i 46.44368883
    342.6514 -11 205.135772650535+68.2919365087273i 216.2047035
    342.7734 -22 121.040629595565+97.6876041921559i 155.5432481
    342.8955 -35 -27.5919220219215+88.2985937414801i 92.50922018
    343.0176 -45 129.045582556654+60.11732230849i 142.3617042
    343.1396 -53 7.38230775480113-23.1253479811555i 24.27509397
    343.2617 -57 91.7636423967861+22.9991912254793i 94.60194957
    343.3838 -59 -22.8102305671421+30.0072957602131i 37.69276346
    343.5059 -58 -85.9348511339797+39.1482173880568i 94.43188849
    343.6279 -56 99.1248756335054+37.1633250085027i 105.8624282
    343.75 -52 -18.6462449961313-107.990585453512i 109.5885441
    343.8721 -47 -78.2810805369108+35.2945965206474i 85.86987897
    343.9941 -28 140.871936717616-18.6792074456239i 142.1049448
    344.1162 -16 66.0895303410517-4.89598446488534i 66.27063214
    344.2383 -5 -79.373794894736+91.8615706634127i 121.4032433
    344.3604 3 -14.8447839307722-4.49608957800507i 15.51071989
    344.4824 8 -60.2896285125463+117.656918705519i 132.2043487
    344.6045 7 54.7042238808456+81.0069947380893i 97.7480706
    344.7266 2 -46.234414620419+138.199149198994i 145.7279175
    344.8486 -5 -104.123283580856+94.4332858556131i 140.5677903
    344.9707 -13 87.3452136442424+24.2999234586396i 90.66241022
    345.0928 -20 -13.7042651915408+108.641036758149i 109.5019715
    345.2148 -24 -32.9387393722445-3.2430390467131i 33.09800377
    345.3369 -26 28.948471476334-122.458827979234i 125.8339324
    345.459 -26 -32.7787074797988+119.182761547442i 123.6081483
    345.5811 -26 81.356659765989+192.288450774381i 208.7911741
    345.7031 -26 77.0108493196662+42.0569815453025i 87.74657036
    345.8252 -29 101.199902601045+72.7180758666743i 124.6167679
    345.9473 -35 -8.80326859834791+52.2645907682127i 53.00080175
    346.0693 -35 9.38013778233287+146.043552156308i 146.3444776
    346.1914 -32 -1.98351135334036+83.7113890975259i 83.7348851
    346.3135 -26 33.179282395186-127.001399718985i 131.2639338
    346.4355 -16 -30.8454623452924+69.0655821517874i 75.64057896
    346.5576 -4 -205.686824704006+4.92262546664037i 205.7457219
    346.6797 7 41.4775369061321+3.69940943303646i 41.64218652
    346.8018 16 193.139507957616+118.265176270461i 226.4718999
    346.9238 23 163.859335975287-88.9928894557056i 186.4661266
    347.0459 28 -159.380784635246+18.4333743055081i 160.4432105
    347.168 31 -27.4639155947195+119.502998916113i 122.6182426
    347.29 32 -3.33500140595825+102.647536549016i 102.7016991
    347.4121 30 20.4122701073102+112.446619185533i 114.2843075
    347.5342 27 -163.36484315183-23.2014855890154i 165.0041845
    347.6563 20 -184.727678767074-22.3500694955245i 186.0748261
    347.7783 11 119.971696460468+190.454567874333i 225.0914267
    347.9004 -14 14.04248506901+72.8328184444158i 74.17419248
    348.0225 -28 63.4243082011103-222.179508461567i 231.0549217
    348.1445 -39 48.4072819155248+34.6943660294368i 59.55639325
    348.2666 -44 -118.677402196386+75.6950933101467i 140.7624699
    348.3887 -44 -36.1749329850164+409.780129098895i 411.3737716
    348.5107 -39 24.4565418415676+81.5919585243948i 85.17846051
    348.6328 -30 -41.3774431130098+34.5145017696544i 53.88268396
    348.7549 -18 -144.321420923347+84.7462931910355i 167.363696
    348.877 -9 -98.386169154537+68.6565882751082i 119.9731861
    348.999 -2 53.4412708884282-21.8547221079033i 57.73732166
    349.1211 1 240.837743678565-173.869551278191i 297.0411413
    349.2432 2 -118.541920984576+19.8403015631783i 120.1907842
    349.3652 1 -60.099177619719+9.57998896371375i 60.8579275
    349.4873 1 265.147091443449+328.926464805962i 422.4873955
    349.6094 1 21.8068676408802+94.4667494969075i 96.95105073
    349.7314 1 135.043384121375+129.531953865912i 187.1236026
    349.8535 0 -772.284687767788-161.085594803886i 788.9057027
    349.9756 -3 -73.4745489606966+48.2455533637736i 87.89847988
    350.0977 -10 -10.4067913042247-0.356253485635044i 10.41288729
    350.2197 -18 -493.448662050413+528.262601713973i 722.8782459
    350.3418 -26 -74.7486475642162-107.071246975149i 130.581822
    350.4639 -34 -1.48740690233132-7.74579119404719i 7.887310094
    350.5859 -40 -202.060308734078-29.4886787862261i 204.2007604
    350.708 -44 17.4389814773729-98.4303359738088i 99.96323882
    350.8301 -46 152.408603415857+10.9715159123473i 152.8029992
    350.9521 -48 68.7425529780869+104.114092539315i 124.7609028
    351.0742 -50 11.3317654760725+4.63536186687978i 12.2431813
    351.1963 -52 -270.914521177299+96.5877107633952i 287.6175649
    351.3184 -53 -63.7134409740408+144.993777788588i 158.3748659
    351.4404 -51 -83.1623633070625+47.8352108331238i 95.93844936
    351.5625 -46 -95.736659216082+36.1306388877489i 102.3275671
    351.6846 -37 -60.8566533486464-58.695454844785i 84.54991825
    351.8066 -12 84.526393156404-55.8738093209615i 101.324201
    351.9287 0 73.8139134030243+108.627718241564i 131.3334496
    352.0508 8 -69.3466659988379-80.1888630070235i 106.0151585
    352.1729 10 76.6539166915756-53.0752854863805i 93.23523408
    352.2949 5 -67.0353497697476+196.029789405841i 207.1748451
    352.417 -4 -54.763943239425+57.9166409228721i 79.70838585
    352.5391 -17 -101.758076429986+225.796791677814i 247.6669079
    352.6611 -30 27.3247211416859-76.576148527409i 81.30526987
    352.7832 -41 -39.7629522817002-36.3373278897929i 53.86551561
    352.9053 -50 -104.053486422008-110.228842563i 151.5833954
    353.0273 -56 -7.72481743239651-103.373376893186i 103.6616026
    353.1494 -58 63.3855668909844-73.1939639032009i 96.82503004
    353.2715 -58 71.9155419993975+59.5340623357835i 93.36032219
    353.3936 -57 -129.337713513581+40.4171806739671i 135.5056922
    353.5156 -53 42.6963904326274-3.28849996659127i 42.82284423
    353.6377 -47 81.394006783351+182.762604953863i 200.0678738
    353.7598 -28 -97.0183295028253-57.7739836599272i 112.9176224
    353.8818 -16 -5.32183193799716+12.1323917517292i 13.2482763
    354.0039 -5 -230.547928983356-196.165304935265i 302.7097197
    354.126 4 -131.354803004867+43.3703943907349i 138.3295897
    354.248 8 -77.3498937443514+23.0342284830486i 80.70676393
    354.3701 8 43.6395882255668-81.3472737871793i 92.31355596
    354.4922 3 69.2151228187025+3.90551464019789i 69.32522103
    354.6143 -4 0.228392364107933+88.9748319385963i 88.97512507
    354.7363 -13 -39.5097437695826+133.801537152868i 139.5129786
    354.8584 -20 -6.91329356181711E-002-100.388548068205i 100.3885719
    354.9805 -24 71.4463539688361-18.2827503962008i 73.74849461
    355.1025 -26 -223.626663125802-28.0682263011248i 225.3812543
    355.2246 -26 -2.15904975246671+44.619497498071i 44.67170304
    355.3467 -25 -167.901478442911+39.3249509378951i 172.4452326
    355.4688 -26 156.112370105562+8.09219120214121i 156.3219615
    355.5908 -28 43.0112882772355-98.6920382852876i 107.6572772
    355.7129 -36 -84.3543424949859-115.95918490847i 143.3952149
    355.835 -39 78.5535288379458-32.798161218585i 85.12564991
    355.957 -38 42.1505882524475+81.3337146214527i 91.60701515
    356.0791 -34 -44.0189080937672-10.6229385692406i 45.28256942
    356.2012 -26 -181.144763012954-63.6200681959252i 191.9920265
    356.3232 -16 -53.183652258035-29.5039384827093i 60.81926712
    356.4453 -6 -109.128492828649+126.74573905835i 167.2528335
    356.5674 3 143.15416916323-78.5657048814251i 163.2963139
    356.6895 11 -86.6647555806527-56.6977824846887i 103.5635959
    356.8115 19 81.914373988024-216.000554760179i 231.0112645
    356.9336 25 -91.8951254644306-19.7102719625036i 93.98515258
    357.0557 31 -68.6016481459536-39.1806104587694i 79.00193899
    357.1777 35 55.7454564556213-66.8838575851618i 87.0689745
    357.2998 37 18.424031620938+45.9601289436044i 49.51543591
    357.4219 37 -67.095364200619+52.7550613399308i 85.35153422
    357.5439 33 -32.9565542117423+17.8495231404123i 37.47986048
    357.666 14 83.4381401396967-61.3321547307621i 103.554606
    357.7881 2 -30.0874821912306+4.01248219233961i 30.35385639
    357.9102 -6 -81.8636945893938-196.845273274281i 213.1894137
    358.0322 -12 -144.960800281057+109.390795196348i 181.6039088
    358.1543 -13 -33.3160750203131-98.1357135457671i 103.6367653
    358.2764 -9 -3.17355312950681+112.532599483842i 112.5773396
    358.3984 -1 31.1509329919737-6.7020353701618i 31.86373965
    358.5205 7 66.8206720187702-109.003220326754i 127.8542305
    358.6426 15 72.7683762168911+30.2207677668765i 78.79423444
    358.7646 20 -19.3606369726108+52.7879983778281i 56.22639093
    358.8867 21 -10.1386922870579-46.3172694927624i 47.41394873
    359.0088 19 -20.4729253413689-100.35105052294i 102.4181332
    359.1309 17 -18.9273061551278+72.024071366413i 74.46952245
    359.2529 14 -124.244835970824-27.9528340631095i 127.3504621
    359.375 13 68.3674005785799+69.3330595039397i 97.3713233
    359.4971 13 22.3996340773156-143.950737665345i 145.6830755
    359.6191 13 83.6346741672177-29.3030255355473i 88.61955782
    359.7412 11 -81.1152953953874-82.6500097122284i 115.8046426
    359.8633 6 -11.3370057200214-32.729696386796i 34.63756233
    359.9854 0 79.7104854050409-23.2828930336758i 83.04128245
    360.1074 -9 -22.9953036956222+74.8000269707216i 78.25489139
    360.2295 -18 -91.0020124958928-14.7207700612956i 92.18496271
    360.3516 -26 -30.4340572112144+34.6846009083661i 46.14383359
    360.4736 -33 109.117461097756+42.4176800779021i 117.0721141
    360.5957 -40 -49.505310798834-77.5596434013518i 92.01235831
    360.7178 -45 124.5421483855-69.0869541233739i 142.4210446
    360.8398 -50 -98.3047280611066-158.323393560015i 186.3601795
    360.9619 -53 40.8720963019746-9.88664817624098i 42.05085098
    361.084 -55 -77.328341242509+14.9765035587737i 78.76527165
    361.2061 -52 77.3118110443728-36.3366771134177i 85.42523181
    361.3281 -45 106.058522092607+2.55646626574546i 106.0893285
    361.4502 -34 75.1656696313463+9.11302463524142i 75.71608224
    361.5723 -6 -63.2185912545023-8.18145220692864i 63.74579547
    361.6943 5 2.48186718643539-54.4709242199027i 54.52743576
    361.8164 11 77.0061869011994-82.7002393337473i 113.0012496
    361.9385 11 -144.561238965152-74.1217484754841i 162.456103
    362.0605 4 23.1197302730165-146.028431829977i 147.8473024
    362.1826 -7 -148.823468360561+56.8012217669795i 159.2947065
    362.3047 -21 179.344599232916-46.4049232354186i 185.2509168
    362.4268 -34 88.1223652504118-43.5521340987472i 98.29720058
    362.5488 -44 58.9604016589549-210.793363201375i 218.8839211
    362.6709 -51 27.3920162621154-86.4758586065305i 90.71051029
    362.793 -54 26.1243040209314+106.423050908053i 109.5825945
    362.915 -55 -21.6376585154613+44.8385724319349i 49.7864022
    363.0371 -56 -12.8839880738323-132.209231423524i 132.83553
    363.1592 -58 -43.6237837110336-24.6255045156822i 50.09441065
    363.2813 -62 -90.7911424699941+135.063924437092i 162.7430344
    363.4033 -65 124.473588759727-93.7905693363579i 155.8536018
    363.5254 -64 68.7457440858031-20.2035768262291i 71.65306586
    363.6475 -55 86.4422666024845-273.775390188141i 287.0979445
    363.7695 -39 -30.5822499725762+60.780952000316i 68.04115034
    363.8916 -15 -44.4526347576066-67.7475213105224i 81.02939825
    364.0137 16 217.855684766415+39.7677837814433i 221.4555847
    364.1357 54 65.8910269525158+56.8805432459701i 87.04610062
    364.2578 96 -21.7502466089843-5.95697367009242i 22.55124748
    364.3799 140 -72.9068857762976-50.8482194449882i 88.88731864
    364.502 184 97.9913904973076-54.3104247822372i 112.0354178
    364.624 224 111.969560674307-62.4016374617391i 128.1840352
    364.7461 255 -32.9421235499882-288.125809563276i 290.0028718
    364.8682 273 38.8451567452343-15.1730853357624i 41.70334185
    364.9902 273 -17.6660516769508-115.820005703577i 117.1595626
    365.1123 253 215.348800402939+162.576175820354i 269.8260899
    365.2344 213 23.5351400976968-10.3841740384875i 25.72418881
    365.3564 156 135.709835936475-82.8159484259832i 158.9831465
    365.4785 14 0.689584986996834+164.029401151757i 164.0308507
    365.6006 -54 -37.2865475176697-13.9352507213544i 39.8055001
    365.7227 -110 -27.6803567997565-77.1005211949081i 81.91881665
    365.8447 -151 -3.88052468190841-127.162707730873i 127.2219034
    365.9668 -173 14.8512033318466-44.9103359114206i 47.30218295
    366.0889 -177 -46.0729858798534-5.16404777951867i 46.36148636
    366.2109 -166 269.508686035971-0.972465602130541i 269.5104405
    366.333 -145 102.261606823496-1.60528437680254i 102.2742058
    366.4551 -117 266.399133205828+8.65926801229358i 266.5398302
    366.5771 -88 -65.7863422599426+43.6155124087246i 78.93133567
    366.6992 -60 -35.8147284616564-101.5019405922i 107.6352113
    366.8213 -36 53.2853709552654+143.891339564399i 153.4406998
    366.9434 -18 -138.157407575792-18.1393946463606i 139.3431265
    367.0654 -7 -2.5840921934217-134.377470904713i 134.4023148
    367.1875 -4 103.598771081875-82.5090058789308i 132.4403315
    367.3096 -7 164.486298979428-82.5209999638311i 184.0256993
    367.4316 -29 162.45345089341+7.8413351024052i 162.6425843
    367.5537 -42 147.363298304355-183.58408015968i 235.4125234
    367.6758 -52 41.1813759668048+44.453457530762i 60.59715846
    367.7979 -57 38.3862755204461+74.0921351763619i 83.44549505
    367.9199 -56 -57.4740958267039+111.862280907373i 125.763435
    368.042 -50 -79.6318451472186-60.1910399349321i 99.82079969
    368.1641 -40 137.793877498033-189.023547095474i 233.9167673
    368.2861 -29 -18.8034955359445-0.902730944529196i 18.82515252
    368.4082 -20 -2.35685125895198-179.492284848015i 179.5077577
    368.5303 -13 733.038988063749-57.4824158906124i 735.2893214
    368.6523 -11 25.8523992090539-25.7520183383813i 36.48990262
    368.7744 -10 141.521085892251-11.0397026642063i 141.9510225
    368.8965 -10 448.940003255295-668.953543528504i 805.6338932
    369.0186 -9 96.8168797798565+117.761203592057i 152.4506782
    369.1406 -7 93.0642765780262+93.1366135516107i 131.663922
    369.2627 -4 136.417205456392+443.847635865961i 464.3386456
    369.3848 2 56.9723051000174-1.59393021864026i 56.99459766
    369.5068 1 73.4549520305862-126.250858410505i 146.0647433
    369.6289 -3 -307.465930034052+78.9302154083912i 317.4354691
    369.751 -11 5.34337908976319-42.423660121275i 42.75884281
    369.873 -20 200.771276677356-3.2190109833826i 200.7970806
    369.9951 -29 244.234433833423-135.107701015796i 279.1138648
    370.1172 -37 125.404660520769-54.1098714703552i 136.5804051
    370.2393 -42 24.7761119944377+69.7607913084639i 74.02988403
    370.3613 -46 345.178624730177+157.684864889666i 379.4901838
    370.4834 -47 162.899741019357-40.1953223246277i 167.7855463
    370.6055 -48 17.2030532399208-1.8869587000799i 17.30623165
    370.7275 -48 -228.64606315621-95.2290755213513i 247.6844747
    370.8496 -47 69.9218635865758+38.889057429705i 80.00891072
    370.9717 -43 119.034270946211+240.634719454508i 268.4664334
    371.0938 -37 120.798655910399-238.7744800638i 267.5921666
    371.2158 -27 126.747452660945-151.974043362353i 197.8914516
    371.3379 -2 80.7505890352338+30.7567203820796i 86.40968394
    371.46 8 151.227774363845+61.5490690823942i 163.2731688
    371.582 16 123.956044362664+71.6390054916576i 143.1686001
    371.7041 18 68.3641535366315-111.421697056134i 130.7228062
    371.8262 14 -128.63231569779+79.6605492790206i 151.3012748
    371.9482 5 -8.12537979356375+245.32772439541i 245.4622459
    372.0703 -5 16.0560541183779+46.483999353973i 49.17884779
    372.1924 -17 132.96866897417-166.912928245191i 213.4024193
    372.3145 -27 128.926619312966+6.30078819803148i 129.0804908
    372.4365 -34 -175.932499917291-85.9620942787747i 195.8104343
    372.5586 -38 95.9465014228868+94.876701961934i 134.9345016
    372.6807 -40 205.560558743549+98.7343551796837i 228.0430139
    372.8027 -39 36.9630026448271+59.1495288473941i 69.74905252
    372.9248 -37 33.8831868183849+146.495624507114i 150.3630219
    373.0469 -33 12.8552528389508+90.474278104304i 91.3829991
    373.1689 -27 222.801060436977+177.577520599661i 284.9106673
    373.291 -6 135.874706187637+11.0613183952526i 136.3242038
    373.4131 6 -45.8811622623697-40.933731408509i 61.48700202
    373.5352 20 72.1468602787521+20.0363158426877i 74.87738911
    373.6572 32 111.008740123117+76.3664042791522i 134.7396307
    373.7793 40 -63.3486803285961+58.2425636564472i 86.0537711
    373.9014 43 195.156009481042-50.0668405879612i 201.4759454
    374.0234 41 172.790041032471-16.3588874055746i 173.5627019
    374.1455 36 34.1459574816525+85.6516948215872i 92.20715394
    374.2676 30 165.389453461475+11.3928934740537i 165.7813902
    374.3896 25 -18.0140641903329-25.1441375245864i 30.93111961
    374.5117 23 125.957800029006-35.2883214282464i 130.8076183
    374.6338 22 -22.5405588159633+64.7334314896197i 68.54556108
    374.7559 24 -69.7413730908863+75.6001693951104i 102.8554555
    374.8779 25 99.9405575036714+0.483833686802768i 99.94172867
    375 26 -117-86i 145.2067492
    375.1221 25 -55.24864673779+107.567187578402i 120.9260634
    375.2441 20 104.873918014356-139.49524197059i 174.5206613
    375.3662 20 35.233510786916-37.1741843027224i 51.21835863
    375.4883 22 -4.43314163544943+161.748282970405i 161.8090226
    375.6104 29 -17.1648486523396+50.2441126238882i 53.09522467
    375.7324 39 4.5532066675097+176.368824193876i 176.4275881
    375.8545 52 130.659165546141+53.8626647187707i 141.3258794
    375.9766 65 52.7795734749698+184.250047435313i 191.660542
    376.0986 77 -1.10833198275105+176.753587290663i 176.7570622
    376.2207 87 69.4174800628503-0.2957208449311i 69.41810995
    376.3428 95 91.1896922412895+131.866615084298i 160.3258062
    376.4648 102 5.3936052075729+35.9023787802834i 36.30525829
    376.5869 106 -45.6160358078246-156.109667567562i 162.6377909
    376.709 109 56.9774751279746+185.056569641954i 193.6294571
    376.8311 109 252.867303042882+162.807616624244i 300.7460606
    376.9531 106 127.104177915776-9.980687191138i 127.4954358
    377.0752 100 125.18424288255+13.7671930564232i 125.9389942
    377.1973 79 28.3471649119208+25.4743637533329i 38.11174317
    377.3193 66 129.547872195896+208.801027424276i 245.7244804
    377.4414 57 57.2945416783074+66.8778160721896i 88.06421968
    377.5635 51 -49.9754495991967-130.938216379709i 140.1512115
    377.6855 52 -22.4274579443075+79.6030955777993i 82.7021384
    377.8076 57 -189.38453116068+135.62302415427i 232.937986
    377.9297 67 6.34356460852616-41.4248462270576i 41.9077403
    378.0518 79 290.130152814747+2.99404618443086i 290.1456012
    378.1738 91 91.6600368259422-169.079394051904i 192.3262952
    378.2959 99 -121.226864867884+154.163343465554i 196.1180492
    378.418 104 -2.16208967645856+187.943723256679i 187.9561591
    378.54 106 38.1234525135811+162.917298927519i 167.3183909
    378.6621 107 28.4145998880881+95.2249289777526i 99.37392307
    378.7842 107 -178.506619618432+120.670306926404i 215.4667868
    378.9063 107 -176.335061735401+115.448722990688i 210.7663674
    379.0283 107 208.742843790524+115.523156139032i 238.5773972
    379.1504 106 26.8145020871704+86.6523919983727i 90.70641963
    379.2725 102 -84.0541729595813-252.882334271488i 266.4856075
    379.3945 94 81.3215581295454+9.29401960817495i 81.85092924
    379.5166 84 -103.307235854436+174.335034354912i 202.6452299
    379.6387 73 250.258902172081+413.934630765232i 483.7058989
    379.7607 61 81.8637990064617+58.5299020112655i 100.6351381
    379.8828 50 -39.4048865592711+102.156573134971i 109.4929702
    380.0049 41 -130.240137628216+170.99268390912i 214.9441588
    380.127 32 -17.9340335555568+137.843360588449i 139.0051136
    380.249 24 -11.9600251554717-74.338458405533i 75.29441281
    380.3711 16 118.152174400244-330.674546925152i 351.1489603
    380.4932 8 -117.04988594578+78.1572613293557i 140.7452781
    380.6152 1 -40.2360629004436+43.3456170593709i 59.14206013
    380.7373 -3 505.72181202865+144.562855369228i 525.9781082
    380.8594 -4 36.9644642340588+131.164185158145i 136.2733102
    380.9814 0 234.690217518471+23.0543149043899i 235.8198457
    381.1035 14 -994.236386066012+381.035755473018i 1064.750787
    381.2256 22 -54.1581803776919+92.2297497078085i 106.9552955
    381.3477 27 -44.6295618912332+28.9054333520475i 53.17256691
    381.4697 26 -173.16932174725+935.273528480981i 951.1699044
    381.5918 18 -141.586340953872-41.491487912417i 147.5406233
    381.7139 6 -13.841468799605+60.0824211460977i 61.65617235
    381.8359 -9 -231.192337207694+124.342362988245i 262.5088951
    381.958 -25 -81.1416054980987-86.2849186883543i 118.4442795
    382.0801 -39 126.079493673244-104.273516238674i 163.6123617
    382.2021 -50 109.179702314364+59.9708299698294i 124.5660782
    382.3242 -58 40.2978518702477-3.93270136936293i 40.48929495
    382.4463 -62 -226.090102919663+276.848685900859i 357.4380079
    382.5684 -64 -5.99345130961808E-002+195.841133136518i 195.8411423
    382.6904 -64 -29.952599163364+137.025215194472i 140.2607137
    382.8125 -62 -71.6722186708518+99.4655170855424i 122.5981077
    382.9346 -58 -92.3437644671451+0.313816218568732i 92.34429769
    383.0566 -43 69.1412903710607-68.1579720762747i 97.08772936
    383.1787 -33 140.038281470045+42.6373759939398i 146.3853343
    383.3008 -23 -145.209928948628-20.3980665671725i 146.6356184
    383.4229 -15 38.885897684413-82.7271910168673i 91.41061849
    383.5449 -10 69.8174147773102+296.72893357157i 304.8319724
    383.667 -9 4.56278640957795+62.1336175651459i 62.30092657
    383.7891 -13 26.4928917780073+316.444826491913i 317.5518879
    383.9111 -20 -2.11495347421282-116.212450443178i 116.2316939
    384.0332 -27 -55.7540407656725-31.3465870414777i 63.961876
    384.1553 -33 -138.435119608143+26.0247306916657i 140.8601042
    384.2773 -36 -110.185913780608-67.907367336003i 129.4308546
    384.3994 -38 5.79331905058713-80.3522202684886i 80.56079597
    384.5215 -38 63.9496136350716-7.41258043202272i 64.3777868
    384.6436 -38 -137.923047869482+138.892081275294i 195.7390543
    384.7656 -39 4.05340386139413-7.50991257286945i 8.533983227
    384.8877 -42 187.292888588898+155.040581476509i 243.1382488
    385.0098 -51 -135.819414806898+25.8190860448967i 138.251722
    385.1318 -53 -1.49827823935958+67.2415340500916i 67.25822432
    385.2539 -52 -372.425161880072-42.5112517172826i 374.8435777
    385.376 -46 -81.4879361053097+143.626296013126i 165.1326637
    385.498 -36 -32.3567178678337+133.380922607983i 137.2495089
    385.6201 -23 -39.4322932780164-65.5962603979351i 76.53610345
    385.7422 -10 57.6630108664968-41.667593259988i 71.14218966
    385.8643 1 49.6820098814255+123.393481896305i 133.0197485
    385.9863 10 -0.583880518923479+172.802445132318i 172.8034316
    386.1084 17 -84.5414531564367-95.5444127140916i 127.5773965
    386.2305 21 88.0116047345981-41.4172081040523i 97.26987044
    386.3525 23 -232.99570391077+161.240399212375i 283.3468976
    386.4746 23 11.6460926835941+48.4893960381572i 49.86835673
    386.5967 20 -144.460387669201+181.927963346883i 232.3070973
    386.7188 14 174.816326703121-84.0333054960713i 193.9648023
    386.8408 6 -16.9724714291329-109.961580075351i 111.2637132
    386.9629 -18 -122.393714499106-56.8523983311549i 134.9533866
    387.085 -31 75.4315850691792-81.8800820384932i 111.3295642
    387.207 -42 96.7679644874587+42.1943461201407i 105.5670488
    387.3291 -48 -94.4432926595345+36.9483759734941i 101.4135988
    387.4512 -49 -263.572643477906+20.0769436439636i 264.3361913
    387.5732 -45 -222.598374427901-141.777071301352i 263.9143313
    387.6953 -37 -38.4163330266777+228.018071928245i 231.2316063
    387.8174 -26 38.0817723953303-173.510271096845i 177.6401857
    387.9395 -17 -99.4933585275723+61.2103839277189i 116.8145517
    388.0615 -11 -112.540232331836-241.017268305466i 265.9974201
    388.1836 -8 -126.069397098058+26.7539764653501i 128.8769496
    388.3057 -7 -104.658154489467+3.50410778070699i 104.7167994
    388.4277 -7 18.962105508286-83.0894549805163i 85.22569433
    388.5498 -6 66.7156589972899+56.1745646482697i 87.21559992
    388.6719 -5 -49.5506123213768+115.958941481506i 126.1020987
    388.7939 -2 -40.4487338490638+59.6953945280112i 72.10853069
    388.916 2 84.4456627581384-102.172073981993i 132.5526411
    389.0381 1 -6.64383100980986+50.0349651283611i 50.47413423
    389.1602 -2 -195.12994266813-121.953709042958i 230.105197
    389.2822 -9 -87.8563739619967+160.090072030387i 182.6131803
    389.4043 -17 -117.100481431541-67.0785769681743i 134.952059
    389.5264 -25 69.7874344874048+109.541529442568i 129.8831501
    389.6484 -32 26.6865888477302-57.166358361398i 63.08856119
    389.7705 -38 45.5967165235495-128.533194638168i 136.3812402
    389.8926 -42 103.627719503684-40.7851205955489i 111.3648522
    390.0146 -45 3.92210366547152+96.6676965477292i 96.74722969
    390.1367 -48 -50.6734777545012-24.4271543907718i 56.25377516
    390.2588 -50 -66.4673207907277-100.974714115348i 120.8875412
    390.3809 -51 7.13328604312652+128.561698196944i 128.7594424
    390.5029 -49 -119.689684872577+46.9191927894605i 128.557502
    390.625 -44 145.974964176364+81.4163771635415i 167.144598
    390.7471 -34 -48.2561026103729-124.217843723445i 133.261863
    390.8691 -9 72.1669554004485-89.5390656395821i 115.001364
    390.9912 3 -118.580917387043-45.1172311180308i 126.8739473
    391.1133 12 -50.5011490416884-62.6968550372232i 80.50628352
    391.2354 14 66.0778418801604-29.3020771514865i 72.28342073
    391.3574 10 16.1426004586746+119.138167839916i 120.2268131
    391.4795 0 -123.630524281219+77.6611638428429i 145.999188
    391.6016 -12 -42.1896108862313+53.7329994673294i 68.31689761
    391.7236 -25 113.350134093615-10.5923176630427i 113.8439726
    391.8457 -37 -90.1954953078089-41.5469225808563i 99.30445181
    391.9678 -46 58.9175757715323-173.2372801454i 182.9820646
    392.0898 -52 -230.562571711394-69.9825312276687i 240.9494846
    392.2119 -54 63.0386035332494-27.7780360460888i 68.88747943
    392.334 -54 -49.1942186929108+102.545918198127i 113.7353792
    392.4561 -53 35.382019080201-49.9780550857358i 61.23473903
    392.5781 -50 122.970651515637-106.07141668364i 162.397434
    392.7002 -45 34.4465593097748+51.6946234190673i 62.12004136
    392.8223 -29 -59.8528126674501+30.0904737514283i 66.99101279
    392.9443 -19 -3.97651586243816-55.7647828723253i 55.90638324
    393.0664 -8 -11.5335905829669-119.137497617242i 119.6944738
    393.1885 0 -194.073996875651+47.283242463876i 199.7508981
    393.3105 6 -81.3520030845107-87.7199856736201i 119.636718
    393.4326 7 -94.5597483172691+197.569955513645i 219.0329503
    393.5547 4 148.06640190077-120.085721762264i 190.6416532
    393.6768 -1 36.0094775013998-146.134012524048i 150.505256
    393.7988 -8 -63.8684704039104-246.404807372325i 254.5476588
    393.9209 -14 -6.41165172043635-89.0177352528509i 89.24834154
    394.043 -18 66.0804026893279+76.4997708809356i 101.0882514
    394.165 -20 -0.545950524200677+75.0035610813372i 75.00554804
    394.2871 -21 -133.521782127509-93.8496403014147i 163.2048446
    394.4092 -22 -91.8411202310384+39.9066835357739i 100.1365805
    394.5313 -24 16.2553204511644+197.290166768917i 197.9586961
    394.6533 -28 54.6207044934068-163.991520926184i 172.8486051
    394.7754 -38 76.4038176709565-47.4516216947583i 89.93997863
    394.8975 -40 -108.656573852878-324.462577719155i 342.1727859
    395.0195 -39 0.321110043934674+115.235443041055i 115.2358904
    395.1416 -33 -136.245949000938+3.13710267593274i 136.2820606
    395.2637 -23 234.525341037806-68.2274559229429i 244.2480733
    395.3857 -11 125.335040354631+22.5394470818544i 127.3455889
    395.5078 1 -24.9685790026909+70.7619883626316i 75.03791665
    395.6299 13 -103.600447523739+59.8423460546946i 119.641795
    395.752 21 73.9807451965259-54.9573283208024i 92.159962
    395.874 27 28.9654232238972-125.567387220816i 128.8649079
    395.9961 31 -257.90417269043-271.650554004152i 374.5778768
    396.1182 32 12.5268293732534-18.7692332867403i 22.5655838
    396.2402 31 -102.067437521244-90.8387181320795i 136.6361391
    396.3623 28 334.97090946643+6.45308529843894i 335.0330618
    396.4844 22 -5.36832133061914-16.1099827264795i 16.98088388
    396.6064 14 50.0814789921164-176.477191780011i 183.4457788
    396.7285 -10 117.497062204025+222.501999784154i 251.6201493
    396.8506 -23 -100.035098676789+30.3941885558037i 104.5505986
    396.9727 -35 -86.8980859900719-55.3069310396746i 103.0055046
    397.0947 -42 -128.386246649461-124.597361694428i 178.9064864
    397.2168 -43 -0.260386696204062-67.6107854178493i 67.61128682
    397.3389 -40 -57.4396466339152+74.111136784694i 93.76445809
    397.4609 -33 233.439875111242-229.890629126798i 327.6337538
    397.583 -24 88.510581123652-64.6699048975831i 109.6189745
    397.7051 -15 272.920220572528-181.792903018157i 327.9239338
    397.8271 -9 -24.0998362222281+136.171643465171i 138.2878107
    397.9492 -6 -88.436123630422-91.4282735677864i 127.2009323
    398.0713 -4 206.506514230024+112.237331588519i 235.0365057
    398.1934 -4 -178.336547972025+84.3620358097455i 197.2837485
    398.3154 -3 -89.8812143566714-137.159601843136i 163.9859417
    398.4375 -1 45.9829404639135-139.113662512719i 146.5163537
    398.5596 0 102.825382915411-204.123620511281i 228.5596461
    398.6816 5 151.893206058696-60.4221201167312i 163.4698096
    398.8037 4 21.1173630511552-287.160235153089i 287.9356589
    398.9258 0 92.8462415899843+46.9733621446084i 104.0524931
    399.0479 -6 54.8933002936714+56.7414780412899i 78.94852594
    399.1699 -15 27.3952069582894+157.763184738176i 160.1240763
    399.292 -23 -159.199002036251+61.0054761547774i 170.487508
    399.4141 -30 -22.4608323066492-289.233181107182i 290.1039849
    399.5361 -35 -97.8883276268482-3.8788944063162i 97.96514945
    399.6582 -39 -95.849856221878-198.103885698931i 220.0734979
    399.7803 -42 791.718184319668-560.619801176735i 970.1093984
    399.9023 -44 -4.17917572882821-12.1625846749114i 12.86055892
    400.0244 -46 172.737709978996-47.0013482948116i 179.0179968
    400.1465 -47 9.89806716726235-1064.66862966292i 1064.714639
    400.2686 -45 215.389697418308+82.6752875745063i 230.7117789
    400.3906 -40 108.491583953567+40.2929381419678i 115.732211
    400.5127 -31 467.752946225939+378.106849738252i 601.4628904
    400.6348 -5 47.4412569004608-24.7032184400829i 53.48758601
    400.7568 7 -28.7659722677411-203.8296394717i 205.8494671
    400.8789 16 -275.231620184864+351.796775904946i 446.6692471
    401.001 19 -54.940832569733-52.4401616151816i 75.95041563
    401.123 15 217.29680555637-166.608069885612i 273.8177325
    401.2451 4 162.399794463634-270.299627468482i 315.3340797
    401.3672 -9 150.786921194591-130.405754837688i 199.3548507
    401.4893 -25 53.567770919562+102.105322179251i 115.3039587
    401.6113 -39 423.094572345797-38.1230646350766i 424.8086454
    401.7334 -51 141.176633673436-149.185323824806i 205.394992
    401.8555 -58 20.2774585289906-21.9855849544216i 29.9088828
    401.9775 -61 -344.305053944385+87.7457548996658i 355.3101289
    402.0996 -60 74.1770481760535+7.31082905532926i 74.53645214
    402.2217 -57 297.365389479302+164.131647263105i 339.6547843
    402.3438 -50 -66.2797890494188-330.599889889346i 337.1784359
    402.4658 -41 10.7075573989128-276.482018650367i 276.6892814
    402.5879 -15 118.52196885324+10.0597595926608i 118.9481226
    402.71 0 175.274760861152-5.82675044784538i 175.371585
    402.832 14 180.21548134315+59.5782359403551i 189.8082873
    402.9541 28 -39.7317308275935-162.060734080321i 166.860097
    403.0762 37 -56.8380448628287+218.777553617277i 226.0402206
    403.1982 41 113.024406937636+241.477542670761i 266.6194295
    403.3203 40 46.2573402445808+59.7203920626562i 75.53983555
    403.4424 35 35.5125695829296-252.652144929785i 255.1357461
    403.5645 29 123.504127628013-65.1140210406006i 139.6177112
    403.6865 23 -246.529643324291+39.4624395857402i 249.668078
    403.8086 18 162.507649173999+25.4560840489642i 164.4893561
    403.9307 15 337.680255384865-5.47657717710422i 337.7246627
    404.0527 13 63.5943218949519+18.7996579243309i 66.31489211
    404.1748 10 170.91023267459+145.505314627111i 224.4595826
    404.2969 5 18.5013251847131+100.368069159246i 102.0590434
    404.4189 -1 340.214869906866+40.2111566978485i 342.5829751
    404.541 -18 133.99112683196-86.8195359993248i 159.6598068
    404.6631 -23 -111.76367610406-2.87889891411657i 111.8007485
    404.7852 -24 62.3892292282051-21.8123867408585i 66.09233041
    404.9072 -19 160.686548069417+3.20348025574738i 160.7184775
    405.0293 -9 -41.6736992546599+130.666747087982i 137.1513617
    405.1514 3 215.859124490893-209.063028995923i 300.5037632
    405.2734 18 160.776018517963-108.07891971245i 193.7265625
    405.3955 30 100.36491380441+128.572154370827i 163.1070655
    405.5176 39 165.983205013054-73.5925816054878i 181.5662205
    405.6396 44 -48.0957770856219-4.32063784671681i 48.28945728
    405.7617 45 75.6807794198297-117.229193260466i 139.5358883
    405.8838 43 -17.0915640824792+132.729654900447i 133.8255688
    406.0059 38 -23.0705993695413+144.381183436486i 146.2127856
    406.1279 32 84.0043791609218-80.3020505250927i 116.211682
    406.25 24 -193.988314676777-3.59498999989884i 194.021623
    406.3721 14 19.4664109532239+166.294647749808i 167.4301377
    406.4941 -8 23.8428184920145-240.195316007986i 241.3757855
    406.6162 -20 -18.7669923688166-25.2765031762318i 31.4817664
    406.7383 -29 119.686754862715+175.310627651491i 212.27043
    406.8604 -34 16.122000198957+80.5536599611915i 82.15114743
    406.9824 -36 134.150326962664+185.450077314701i 228.8843407
    407.1045 -34 149.352464269617-27.6243489693939i 151.8856914
    407.2266 -29 179.816004543504+173.723658300636i 250.0274084
    407.3486 -24 67.9539275448379+148.294869949681i 163.1229742
    407.4707 -19 32.5262336295338-35.4917749568555i 48.14168634
    407.5928 -16 160.976380650906+92.7684266775969i 185.7939076
    407.7148 -14 -9.92557328586556+95.4383435649834i 95.95308451
    407.8369 -12 -144.545016354021-90.1058503126114i 170.3300502
    407.959 -9 223.417540877533+200.981390550264i 300.5144205
    408.0811 -5 404.385153909793-34.8366752016799i 405.8829223
    408.2031 0 140.697646475683-58.1738034749566i 152.2498576
    408.3252 5 112.821027572656-53.4046382014614i 124.8224325
    408.4473 7 10.0187109405882+24.5242048926045i 26.49171936
    408.5693 0 256.817282189383+132.201367305961i 288.8465301
    408.6914 -15 79.3349135921429+70.6871727635291i 106.2577287
    408.8135 -34 -172.752677153424-66.4274559265i 185.0840197
    408.9355 -52 62.4269315944347+137.818202276883i 151.2976492
    409.0576 -65 -148.413830762137+306.951524076421i 340.9485347
    409.1797 -68 19.4872556882384-45.4036783375403i 49.40897834
    409.3018 -58 328.193480048518-190.924601898332i 379.6882457
    409.4238 -34 -83.0023842932327-289.131958270127i 300.8100482
    409.5459 3 -23.4161405650722+277.528884251239i 278.5149857
    409.668 52 102.903625891141+232.44059753705i 254.2002903
    409.79 107 183.868874977887+148.471201992517i 236.329137
    409.9121 164 136.902705745874+140.460117162779i 196.1412638
    410.0342 217 -106.751012955008+286.594805843758i 305.8306092
    410.1563 262 -87.7197386228999+318.115032939769i 329.9877675
    410.2783 294 308.97947805799-23.6817816510133i 309.8856961
    410.4004 302 97.7099211739652+100.655126182198i 140.2807297
    410.5225 274 -320.510123469046-190.031176726238i 372.6105036
    410.6445 226 135.470382696577-58.0448690160391i 147.3819236
    410.7666 161 26.8109554237634+286.968245191516i 288.2179749
    410.8887 85 605.533766732695+316.605187785509i 683.3081205
    411.0107 5 142.45066860883+61.1188006009421i 155.0087119
    411.1328 -70 -0.516881918338612+145.771125667603i 145.7720421
    411.2549 -134 -0.724681355492237+318.813410707273i 318.8142343
    411.377 -181 78.398570114407+222.854256335239i 236.2421541
    411.499 -210 -4.36589733134116-45.7698395694371i 45.97759535
    411.6211 -220 -127.057264722168-437.538267720781i 455.6130861
    411.7432 -214 -81.0288544690568+177.450163679531i 195.0749493
    411.8652 -197 -27.1793973793646+88.1099366687212i 92.20672742
    411.9873 -174 641.800147132776-126.284322381447i 654.106382
    412.1094 -148 168.563732432744+160.64513456825i 232.8531536
    412.2314 -122 266.372076246316-89.8476772237148i 281.1168585
    412.3535 -75 -987.479779972036+1151.96564502992i 1517.280845
    412.4756 -56 -57.393542347113+201.291241852919i 209.3135991
    412.5977 -39 -58.7154630929056+92.0598451155759i 109.1902958
    412.7197 -26 326.35624198092+1311.11526507509i 1351.122361
    412.8418 -17 -194.104883554096+117.383503070989i 226.8382521
    412.9639 -13 -11.4257862805264+120.103125945289i 120.6453872
    413.0859 -12 -209.764596009358+295.151204294551i 362.0986318
    413.208 -15 -171.788125689076+19.1034974567535i 172.847053
    413.3301 -18 35.6489112124569-171.083751147891i 174.7583897
    413.4521 -20 218.918819117286+38.9558304778513i 222.3578334
    413.5742 -21 -4.52295426248293-13.8662564044536i 14.58527278
    413.6963 -22 -49.4861320527511+502.644501952254i 505.0746208
    413.8184 -22 159.656410597001+281.567698178419i 323.6827739
    413.9404 -22 74.9573108209863+167.936156389595i 183.9052774
    414.0625 -25 -20.2337917134326+201.062683252768i 202.0782248
    414.1846 -29 -148.10635162667+93.0770075879858i 174.9251861
    414.3066 -41 -7.85843920246992-100.42948757162i 100.7364732
    414.4287 -43 146.709613765751-4.98899503774355i 146.7944169
    414.5508 -41 -158.154804769927+131.49717194804i 205.6804524
    414.6729 -34 -68.9221600496262-120.043668597638i 138.4223483
    414.7949 -23 301.194111261347+276.301960451183i 408.7305543
    414.917 -10 35.329107304245+95.7038442982667i 102.0165263
    415.0391 3 293.451500066169+341.167553202935i 450.0100913
    415.1611 16 -122.547388479219-94.1629041208916i 154.5461579
    415.2832 25 -88.2720946978351+48.1743078963648i 100.5620537
    415.4053 31 -178.881254761817+138.250299177978i 226.0788547
    415.5273 34 -153.054495004943+17.4425353310842i 154.0451897
    415.6494 36 -50.5665507411898-73.6424693395298i 89.33190552
    415.7715 36 26.4424492711812-37.3866563962571i 45.79263259
    415.8936 34 -5.45353919636008+282.175510750601i 282.2282055
    416.0156 29 -37.4565568943769-55.2197292353315i 66.72489903
    416.1377 23 343.31260203881+17.6404172604532i 343.7655117
    416.2598 2 -197.528676616622+136.377015594718i 240.0338902
    416.3818 -11 67.0206091781202+68.4027638022644i 95.76377264
    416.5039 -22 -483.306903574837+280.142171958295i 558.6279617
    416.626 -30 -35.3711826772593+222.12774836183i 224.9263372
    416.748 -34 61.8896089637847+209.000847349701i 217.9717364
    416.8701 -32 -141.770607769308-22.7930248876002i 143.5911808
    416.9922 -26 16.3560780884601-79.670042825912i 81.33164829
    417.1143 -18 145.891127216846+117.407667935487i 187.2666054
    417.2363 -9 165.816920052667+189.97594325018i 252.162864
    417.3584 -3 -142.200734188183-8.79898220373092i 142.4727023
    417.4805 1 24.2161298843805-106.391956791597i 109.1131038
    417.6025 3 -179.951289915065+364.805875603983i 406.7748685
    417.7246 5 90.6696762455552+107.602576306703i 140.7100018
    417.8467 8 -22.6528095972289+328.424779233575i 329.2050811
    417.9688 11 119.893696091966-235.414011448826i 264.1860238
    418.0908 15 -126.139879929594-88.2900419284872i 153.9688307
    418.2129 23 -214.27713466266+21.5433654948218i 215.3573937
    418.335 23 -1.74566546325749-142.60259141724i 142.6132758
    418.457 20 157.432179090133+4.68626876198497i 157.5019115
    418.5791 14 -99.6584840778985+102.881404995086i 143.2354598
    418.7012 7 -267.900189349313+240.003823047695i 359.6836757
    418.8232 0 80.3904301808813+205.055730538264i 220.2509339
    418.9453 -5 172.913337781245+237.675791188836i 293.9197239
    419.0674 -9 -80.2173533457792-228.895346662348i 242.5446423
    419.1895 -11 -148.433349917645+144.233837970984i 206.9682569
    419.3115 -11 -260.889895090433-191.770259161846i 323.7890821
    419.4336 -10 -166.207200350814+178.124782524339i 243.6252688
    419.5557 -9 -94.4492079712463+110.366215596301i 145.2630525
    419.6777 -7 -51.1829179607445-90.7443140008959i 104.1835957
    419.7998 -3 116.247886322446+30.8896068937736i 120.2819142
    419.9219 4 -23.1616380312981+227.646539972107i 228.8217836
    420.0439 14 0.444762991402044+132.028949590787i 132.0296987
    420.166 41 27.201342640307-151.402816078649i 153.8269344
    420.2881 54 -0.638088671079903+78.8510227999229i 78.85360457
    420.4102 64 -310.014415822345+76.6606973149319i 319.3521575
    420.5322 67 29.6842194044141+309.720637886748i 311.1398824
    420.6543 65 -221.037133534276+28.2653701595278i 222.8370381
    420.7764 56 142.504982265623+94.7528228220086i 171.1308488
    420.8984 45 -5.56454198654369-59.8836711878964i 60.14165115
    421.0205 33 -85.3321011090667-165.506100688303i 186.2091213
    421.1426 23 118.087956976563-71.5920426196043i 138.0948448
    421.2646 16 39.6092342202534+98.5642300825723i 106.2252272
    421.3867 13 -102.306274926775-0.402526841933479i 102.3070668
    421.5088 13 -199.641473306533-42.1274701327012i 204.0378436
    421.6309 16 59.9918561698475+89.1752717522686i 107.4767505
    421.7529 20 -136.594016142415+186.797967148776i 231.4117667
    421.875 25 170.080704623177-49.1958166770512i 177.0527449
    421.9971 32 -223.997313724497-77.2574396547827i 236.9462144
    422.1191 50 3.59710742392451-110.603298013941i 110.6617762
    422.2412 61 -191.013531964303-4.01234777963554i 191.0556681
    422.3633 73 -115.015728697006-1.20488453491348i 115.0220396
    422.4854 84 75.3100865210631-79.7154045598798i 109.6638266
    422.6074 92 80.7316174912746+113.023197320938i 138.8950582
    422.7295 97 -60.1173347582662+167.955142393879i 178.3900888
    422.8516 98 -25.7203689687642+88.2950465320854i 91.96495323
    422.9736 95 89.0696501942153-99.1317088155791i 133.2685195
    423.0957 92 -157.463697856574+46.9748364134463i 164.3211837
    423.2178 90 -69.1463275147618-161.184325549226i 175.3898555
    423.3398 89 -306.780376486407+133.047255600363i 334.3886535
    423.4619 89 -21.1161977736024-62.128293357202i 65.61873699
    423.584 90 -53.5369716138405+163.166303403889i 171.7249251
    423.7061 90 -11.4207650065183-56.6806184859246i 57.81977504
    423.8281 88 32.0769753177789-128.607230710884i 132.5471695
    423.9502 84 106.481947321975+17.7809938802549i 107.9563284
    424.0723 70 -47.0884246027844+118.531614347165i 127.5423982
    424.1943 66 -92.9700513175525-68.814984422445i 115.6673356
    424.3164 66 -63.7174773075504-83.5628072577363i 105.08406
    424.4385 70 -213.886505792568+193.604093732168i 288.4960701
    424.5605 78 -156.811944181175+10.4679270408397i 157.1609472
    424.6826 90 17.4566705215413+285.861952459061i 286.3944678
    424.8047 102 38.217391402134-290.653121170996i 293.1549178
    424.9268 112 -83.1697449618097-146.617256038726i 168.5640123
    425.0488 119 -311.136570756997-192.745011902815i 366.0008269
    425.1709 122 -37.0144304224327-86.2375802514429i 93.84555561
    425.293 121 131.100007284772+70.597921996954i 148.90023
    425.415 117 60.4715985969397+105.540949411279i 121.6376021
    425.5371 109 -199.605683455805-11.7078177630426i 199.9487481
    425.6592 100 -70.5814249355996+86.4142868376724i 111.575833
    425.7813 87 132.884861154405+301.953698385078i 329.9006249
    425.9033 73 -43.324384833676-217.851151573597i 222.1173711
    426.0254 38 -3.83247031211107-113.191445170553i 113.2563071
    426.1475 19 -470.446688924964-234.896575721953i 525.8293339
    426.2695 3 31.8896418601607+88.8718832070887i 94.42012965
    426.3916 -7 -121.573119894335+138.812275185905i 184.5233623
    426.5137 -14 178.86867383706-327.219441528291i 372.9162981
    426.6357 -16 149.505235909965+23.8747747934474i 151.3995391
    426.7578 -14 -24.6170826062183+103.19198872473i 106.0876397
    426.8799 -9 -35.1670077823297+197.142223796411i 200.2542755
    427.002 -6 -14.277384683887-95.7357387225421i 96.79450078
    427.124 -5 -81.5558209059409-102.389647006021i 130.9006942
    427.2461 -6 -550.977899032497-5.77226987118667i 551.0081345
    427.3682 -9 -59.4944953951568+2.63197632571973i 59.55268492
    427.4902 -12 -179.684111327672-1.80331732724102i 179.6931602
    427.6123 -14 372.708818070553-292.109542022285i 473.5397001
    427.7344 -14 62.1228321934885-12.5244688134724i 63.3727749
    427.8564 -12 -63.7137419741045-252.360339023511i 260.2790457
    427.9785 -7 384.459838060981+260.474037200921i 464.3878671
    428.1006 -7 -110.588837589082+160.328027035663i 194.76901
    428.2227 -11 -168.841157051135+36.154139886101i 172.6686368
    428.3447 -18 -292.691482704249-14.4274045884557i 293.0468462
    428.4668 -28 -77.9090846329478-102.710506739858i 128.9157619
    428.5889 -38 -15.7529094535971+69.5193278875923i 71.28177261
    428.7109 -47 79.4105526942013-459.630335124378i 466.4397934
    428.833 -55 51.0939777114546-163.242763580888i 171.0520226
    428.9551 -61 150.827920111709-445.208378368667i 470.0633592
    429.0771 -65 125.293152671256+203.715620180332i 239.1619284
    429.1992 -67 -207.864620522094+38.9517617296149i 211.4827185
    429.3213 -69 338.904311234136-36.516351159503i 340.865921
    429.4434 -69 -232.951613929757+335.663470819082i 408.578536
    429.5654 -66 -275.945945903594-112.07531704162i 297.8372739
    429.6875 -60 -73.5146579436438-253.354663167733i 263.8048337
    429.8096 -50 -101.311154480862-436.235405446956i 447.8451507
    429.9316 -24 175.998171426238-247.054137506922i 303.3333203
    430.0537 -11 -201.318077290747-239.758101194713i 313.0701444
    430.1758 -2 221.138458286136+3.53622379891675i 221.1667303
    430.2979 0 96.3193167844524+109.194724266249i 145.6052835
    430.4199 -2 160.117325757656+213.246035568388i 266.667264
    430.542 -11 -180.184344212759+295.33611183044i 345.962161
    430.6641 -24 -289.422808802214-343.254956116115i 448.9872238
    430.7861 -37 -135.288792193165+177.008280924262i 222.7891129
    430.9082 -48 -397.594098227489-213.435862942271i 451.2603844
    431.0303 -57 412.438272290729-1378.07057727207i 1438.465795
    431.1523 -62 14.3940778068488-21.9395256695765i 26.23989829
    431.2744 -63 275.810726685902-233.85897298722i 361.6097015
    431.3965 -62 -851.94136203782-1254.65948671254i 1516.566686
    431.5186 -59 255.617671111565-94.5410598689461i 272.5406498
    431.6406 -56 177.209397354039+57.0996225418992i 186.1814636
    431.7627 -51 864.026080903288+94.3434499728085i 869.1615241
    431.8848 -35 43.0544713506243-93.347154697953i 102.7977568
    432.0068 -25 -336.63903249363-296.953842664261i 448.8957817
    432.1289 -14 -82.2440828968336+769.453281838735i 773.8361856
    432.251 -5 -138.008045497237-33.6817096269828i 142.0587139
    432.373 2 99.0111977832294-432.296043537041i 443.489669
    432.4951 6 -14.4636501402901-484.62844943655i 484.8442339
    432.6172 5 -10.3366631809971-319.931657479888i 320.0985974
    432.7393 1 186.226971511698+138.516970288246i 232.0935931
    432.8613 -3 437.623785150749-380.758238804259i 580.0787996
    432.9834 -7 65.3354687784354-227.208387095812i 236.4156819
    433.1055 -10 -75.6758934126257-16.0781757717748i 77.36503461
    433.2275 -11 -392.367851668698+414.630385918624i 570.8510208
    433.3496 -11 97.9196227885817+35.0433503101853i 104.0013891
    433.4717 -12 542.420266424557-68.0554237557568i 546.6729243
    433.5938 -14 -503.359481200604-415.374179521631i 652.6151058
    433.7158 -18 -249.797971809406-324.525693756256i 409.5313817
    433.8379 -31 164.94473950443-59.3673225188748i 175.3032974
    433.96 -35 283.748284543023-171.972868778721i 331.7947507
    434.082 -35 285.378974355339-44.1616551215856i 288.7757102
    434.2041 -31 -241.0126003005-45.6808381528431i 245.3035109
    434.3262 -22 176.883077421666+326.571605314458i 371.3982182
    434.4482 -10 325.07080465144+265.292355190878i 419.5843917
    434.5703 3 145.548128954519+9.16482081021573i 145.836387
    434.6924 16 -160.628678996623-343.041095228938i 378.7859099
    434.8145 26 139.91836226746-229.047812460655i 268.4027729
    434.9365 33 -403.680294287798+302.620037352155i 504.5162703
    435.0586 37 322.186656967143-144.542349138866i 353.1242453
    435.1807 39 430.657070508931-339.098344835823i 548.1361143
    435.3027 38 135.015277530343-36.8184636919157i 139.9454338
    435.4248 36 374.806050264574+48.6407151058628i 377.9490633
    435.5469 31 94.2622910047087+136.752549563626i 166.0922615
    435.6689 23 444.485533121161-352.884622340669i 567.5340922
    435.791 0 103.019164714654-200.358378073714i 225.2918728
    435.9131 -14 -154.821477883302-37.2278290527096i 159.2344224
    436.0352 -26 88.9157316222133-78.3647093527073i 118.5201882
    436.1572 -34 230.641378932223-93.2585251155643i 248.7822304
    436.2793 -38 71.6385894554179+306.244655272156i 314.5121244
    436.4014 -37 70.2577345102007-537.920788048037i 542.4895607
    436.5234 -32 36.0108871690672-323.143873052641i 325.1441937
    436.6455 -25 165.442166744939+41.2683304661591i 170.5115411
    436.7676 -18 139.775704141163-265.546432580829i 300.086913
    436.8896 -13 -132.565869194443+66.2281770991854i 148.1886673
    437.0117 -10 -75.2360574383718-230.339462439504i 242.3153571
    437.1338 -10 161.678921320696+294.381172529993i 335.8576311
    437.2559 -9 142.924393835913+212.059360872499i 255.7275012
    437.3779 -8 131.708041791662-109.565359322216i 171.3230172
    437.5 -6 -289.806132548159+318.823376490863i 430.8548942
    437.6221 -2 168.649664025073+309.574050135926i 352.5319868
    437.7441 4 -282.768540928264-310.166451665842i 419.7157079
    437.8662 5 -38.7880091540692+26.7625819760584i 47.12478592
    437.9883 2 354.807448985791+91.635440152111i 366.4496961
    438.1104 -4 122.912426516458+106.47224828364i 162.6155105
    438.2324 -13 394.84545618035+236.903625436972i 460.4630952
    438.3545 -22 202.473224111488-217.077367233708i 296.8467447
    438.4766 -31 397.684931499824+34.7718932850478i 399.202191
    438.5986 -39 194.9888280432+169.257680506854i 258.2030315
    438.7207 -44 89.0795790504209-49.2195937972998i 101.7729818
    438.8428 -48 374.039403954736-6.96266829527025i 374.1042027
    438.9648 -51 36.2774551299314+119.874782532825i 125.2438311
    439.0869 -54 -422.441542795721+25.5956226851005i 423.2162485
    439.209 -55 533.041814210893+133.913573426813i 549.6056958
    439.3311 -54 478.075926183247-536.284748632679i 718.4413148
    439.4531 -49 103.408213817382-232.777637646332i 254.7129507
    439.5752 -40 -50.5775892555565-106.095086878847i 117.5340801
    439.6973 -14 3.82883787444748-34.8027110909503i 35.01269339
    439.8193 -1 409.450372889192-2.07640412652742i 409.4556378
    439.9414 8 150.569682405322-127.452545966344i 197.2698171
    440.0635 12 -292.938059036663+135.125166103375i 322.6011732
    440.1855 11 173.035226238108+191.792054696703i 258.312566
    440.3076 4 212.545833482663+769.729759308976i 798.5359314
    440.4297 -7 -213.179933234842-18.9158399978262i 214.0175061
    440.5518 -21 289.369893100399-650.932760023427i 712.3541206
    440.6738 -35 -649.794633893428-135.919050893863i 663.8577066
    440.7959 -48 320.13785918061+453.565801057019i 555.1668081
    440.918 -58 401.271761175989+266.743873086122i 481.8415924
    441.04 -62 325.000681936094-110.71175629922i 343.3402631
    441.1621 -62 303.688372344539+107.638402655009i 322.1997102
    441.2842 -57 100.957707606867+565.066505598544i 574.0144724
    441.4063 -49 270.92058943171+464.171649243925i 537.4507287
    441.5283 -39 352.752307825315-271.200724849645i 444.953957
    441.6504 -19 185.738567398283+88.1446794647526i 205.5925581
    441.7725 -12 -688.127462206936+46.3869045892538i 689.6891685
    441.8945 -9 144.524537152011-242.489645352101i 282.2916399
    442.0166 -8 254.01513027989+508.345233090723i 568.2768361
    442.1387 -8 1322.6022557749-68.1393344005324i 1324.356333
    442.2607 -8 240.022194198779-69.9012290642777i 249.993671
    442.3828 -7 44.9098788434273+107.228657994867i 116.253526
    442.5049 -6 -135.703783972891+470.611218692476i 489.7861126
    442.627 -2 218.991726145756+210.214986472077i 303.5584238
    442.749 1 -126.595569981444-201.521521825066i 237.9860544
    442.8711 4 -406.179664334199-1198.04394483009i 1265.026171
    442.9932 5 123.513073698096+460.826356566335i 477.0916162
    443.1152 2 -27.0144233312885+218.985304232546i 220.6452867
    443.2373 -4 2079.27829317369+196.040010691344i 2088.499439
    443.3594 -14 408.686826933606+62.7485492651311i 413.4758795
    443.4814 -25 255.174491520729-513.625631208168i 573.5201044
    443.6035 -44 -5244.48423514359+5249.23452675619i 7420.180457
    443.7256 -47 46.6683818052089+425.336223608556i 427.8888185
    443.8477 -44 85.1779295630153+133.364471413993i 158.2446268
    443.9697 -36 2202.85869717669+6981.79106123701i 7321.065009
    444.0918 -23 -105.978498527696+395.01229865042i 408.9818556
    444.2139 -7 191.574264856069+326.093569432801i 378.203272
    444.3359 9 -1516.10207295253+173.557776521184i 1526.003865
    444.458 23 -336.8811893888+253.655050222197i 421.6987316
    444.5801 33 -250.425874555787-227.76559555898i 338.5118686
    444.7021 37 748.463817868084-594.36332325804i 955.7540723
    444.8242 36 -67.1189851432544-38.8446367012573i 77.54910681
    444.9463 31 460.395195275919+832.422042491652i 951.2571643
    445.0684 22 519.712345461112+398.052740645363i 654.6349413
    445.1904 11 218.122467814395+193.832826298833i 291.8022884
    445.3125 1 133.258312873009+263.823653642355i 295.5684323
    445.4346 -6 -86.47770454388+264.160024693525i 277.9548741
    445.5566 -14 50.2988036062356-141.313780106404i 149.9985136
    445.6787 -16 218.501265653006-125.145748342448i 251.8020282
    445.8008 -17 -332.944919082809+599.428589032255i 685.6872118
    445.9229 -19 -205.534412937445-4.69700187874979i 205.5880754
    446.0449 -23 646.183469047088+108.336143055348i 655.2021028
    446.167 -29 -213.621734239215-160.617344632879i 267.2679867
    446.2891 -36 659.299651129406+165.835289798194i 679.8362842
    446.4111 -41 -470.435395413164-56.7273600101868i 473.8432807
    446.5332 -44 160.222416432182+33.0809431379158i 163.601869
    446.6553 -41 -105.797038704356+271.039001666198i 290.9555874
    446.7773 -32 -273.039216605863+28.5103996144449i 274.5236906
    446.8994 -16 -201.725774937268+388.173165808184i 437.4605067
    447.0215 4 40.1990482302204+91.7350134347953i 100.1562588
    447.1436 28 230.074864184018+546.997779669237i 593.4147067
    447.2656 49 -501.84884345387-126.608133973231i 517.5730685
    447.3877 63 290.592924571503-290.128230287397i 410.631998
    447.5098 61 -236.539236690923+447.408436380428i 506.088055
    447.6318 44 408.518292849057-163.85060921251i 440.1524937
    447.7539 19 -461.143714196589+862.389305514425i 977.9411227
    447.876 -8 143.75355060581+344.290067702388i 373.0961458
    447.998 -33 286.78821013053+435.856595269255i 521.7455789
    448.1201 -50 -218.89886803122+162.622462727786i 272.6953975
    448.2422 -54 -10.721479799394-85.2482809140117i 85.91984362
    448.3643 -44 285.854366949385+98.4372679828731i 302.3286537
    448.4863 -21 349.407107970534+127.70332330696i 372.0127227
    448.6084 11 -362.928743918757+40.5034922778019i 365.1818808
    448.7305 48 -76.3545182083966-103.27922700087i 128.4391341
    448.8525 83 166.226833763857+735.414339906736i 753.9665852
    448.9746 110 190.387110023257+101.603338073879i 215.8019693
    449.0967 123 176.836163224869+486.455674945846i 517.600379
    449.2188 120 -125.069633552644-463.536950713132i 480.1134428
    449.3408 101 -444.004242543758-34.7095795093872i 445.3588691
    449.4629 28 -143.422543933507+258.092547208515i 295.2656245
    449.585 -14 -155.263793735258-268.81253332664i 310.4303847
    449.707 -50 262.17742724154-283.432231665386i 386.0969222
    449.8291 -76 -169.868288241557+562.95154137708i 588.0218306
    449.9512 -87 -210.756480613946+607.840874376787i 643.3419174
    450.0732 -82 607.848797378785+567.317601817927i 831.4622191
    450.1953 -65 169.380686255883+3.68812514980579i 169.4208344
    450.3174 -37 -495.983119977136-165.356762022569i 522.8213022
    450.4395 -7 -234.405398460241+525.609501175972i 575.5095469
    450.5615 20 -506.527816414136-78.832395954353i 512.6255704
    450.6836 39 -58.7142782883526+351.580155814772i 356.4491162
    450.8057 48 -65.6881522532474+164.818639228627i 177.4263712
    450.9277 46 -202.375570837914+127.821841240548i 239.3622668
    451.0498 35 202.218253572609+36.3241549536239i 205.4547792
    451.1719 17 226.990111777232+458.331408989214i 511.4608404
    451.2939 -1 -118.752366803267+181.894556747005i 217.2274255
    451.416 -28 -270.858980937682-209.929287612622i 342.6877491
    451.5381 -34 -7.82997351262192+158.963575040099i 159.1562964
    451.6602 -34 -360.061250301293+377.910349092079i 521.9773328
    451.7822 -31 341.916957473722+390.841262880214i 519.2919204
    451.9043 -28 -398.863298829702+253.813191241537i 472.7716861
    452.0264 -28 306.618207221712+85.6521238585803i 318.3567359
    452.1484 -34 -128.266446729332-72.525260211659i 147.3505844
    452.2705 -44 -436.085847892424-192.235768712618i 476.5768118
    452.3926 -57 175.622061957991-183.191272688067i 253.7757889
    452.5146 -69 138.585970406168+93.2077935481911i 167.0142628
    452.6367 -76 -79.7007533536323-114.505826247826i 139.5127031
    452.7588 -73 -414.035901359583+283.328543233557i 501.6979081
    452.8809 -56 173.745874137188+268.049287821848i 319.4339517
    453.0029 -23 121.233989794151+670.635849255893i 681.5057759
    453.125 24 23.0495596465568-364.888987088037i 365.6162676
    453.2471 85 -599.985543001487+134.962796420476i 614.9777299
    453.3691 220 -475.240644152787-143.832707859917i 496.5294731
    453.4912 278 -210.585524603822+109.070881798194i 237.1554773
    453.6133 320 -220.359482146422+5.29132039282388i 220.4230012
    453.7354 338 37.8062259132737-479.793986730553i 481.2811864
    453.8574 331 111.167648918071+366.861615277523i 383.3349592
    453.9795 298 50.4170595640076+369.273701823572i 372.6995395
    454.1016 242 271.29598571976+404.704940089809i 487.224384
    454.2236 171 -81.0202492928552-210.602580141545i 225.649568
    454.3457 90 -279.049654797041+320.27924770849i 424.7911326
    454.4678 8 -578.379138190002-23.0191926252624i 578.837033
    454.5898 -65 -283.688287308252+432.351348068937i 517.1138487
    454.7119 -125 -154.951543916625-100.632697053778i 184.7617944
    454.834 -168 1.22449805015374+205.903374912503i 205.9070159
    454.9561 -193 -95.6234854264945-2.83332251174975i 95.66545187
    455.0781 -200 -108.610093016225-250.738333259186i 273.2505518
    455.2002 -192 11.2763940079322-0.653378692615965i 11.29530724
    455.3223 -148 22.4764578559124+335.79801139964i 336.5493955
    455.4443 -120 -145.851628616853+3.9416690583832i 145.9048811
    455.5664 -93 23.8403208120365-327.845603270379i 328.7112722
    455.6885 -70 -205.857495255962+559.50320795503i 596.1720792
    455.8105 -51 -163.341989804479+434.236237624737i 463.9415003
    455.9326 -40 685.059496327019+999.552512418119i 1211.7804
    456.0547 -34 -550.111945291355-648.129454883644i 850.1146644
    456.1768 -32 -1597.89400716606-94.492170842919i 1600.685487
    456.2988 -35 -1809.26798184898+246.537553814705i 1825.987786
    456.4209 -40 -10.8673508650434-230.140947859537i 230.3973854
    456.543 -46 564.438842024884-57.6600716245644i 567.3763215
    456.665 -51 687.105243624093-1128.83163697202i 1321.504627
    456.7871 -54 -778.054500158989+256.029846907999i 819.0971174
    456.9092 -56 -189.583564294097+402.949045425003i 445.3199536
    457.0313 -56 1524.84628450898+1506.41933561256i 2143.468079
    457.1533 -54 -491.716022818359-183.96059021119i 525.0010913
    457.2754 -47 -1025.01015272042-195.601401579522i 1043.506455
    457.3975 -42 -3597.64127649127+1530.37680889198i 3909.613271
    457.5195 -35 250.970277380596+137.324257310693i 286.0839593
    457.6416 -28 558.427731648758+507.078490578854i 754.3010852
    457.7637 -20 -705.30329474796-3971.01338324194i 4033.162534
    457.8857 -11 647.037375163038+604.672533234554i 885.5993661
    458.0078 -1 -334.728586481116+984.122785958617i 1039.490685
    458.1299 6 1440.11822767815+2221.10285230375i 2647.118885
    458.252 13 -720.196357080064+158.421580025235i 737.4145305
    458.374 18 -624.509680089188+82.274726054681i 629.9059224
    458.4961 22 -2685.3935548372+2953.39367250414i 3991.725533
    458.6182 24 -763.411267414693+716.681578383993i 1047.105175
    458.7402 24 58.5880683201744+454.847835127905i 458.6056202
    458.8623 21 -447.374356124571-5050.15198418038i 5069.928883
    458.9844 15 670.230852253389-781.459772722922i 1029.508995
    459.1064 5 -2404.46278673544-227.235071732594i 2415.176406
    459.2285 -22 6485.13981392324+808.557828919453i 6535.350348
    459.3506 -35 -380.452662264083+1894.06703275334i 1931.899105
    459.4727 -43 -1044.77472144658+600.517525069682i 1205.062453
    459.5947 -45 -2214.47392460357+2531.2673546062i 3363.214115
    459.7168 -41 -837.525660052085-1548.09005552623i 1760.122738
    459.8389 -31 713.3630782317-1152.96619218093i 1355.808955
    459.9609 -17 -3884.04397846289-3467.40550731248i 5206.601442
    460.083 -3 -536.26351853677-1992.1741077008i 2063.089004
    460.2051 9 -1950.13857864223-2119.11535496361i 2879.876797
    460.3271 17 3829.12066506518-748.634074290283i 3901.617363
    460.4492 22 -997.759322984203+2331.47788103016i 2536.003308
    460.5713 23 2606.74794969569-2986.92429289723i 3964.448462
    460.6934 23 -1805.55400815845+6143.502806164i 6403.331321
    460.8154 22 -3366.97988076542+77.3561089691082i 3367.868389
    460.9375 21 -2535.00325650084-1493.76535433986i 2942.375986
    461.0596 19 -5960.09351080766-2698.36504257343i 6542.468079
    461.1816 14 -498.163289350946-2212.69369819427i 2268.078496
    461.3037 9 -1399.05660227306+1700.06368148108i 2201.721122
    461.4258 1 2293.60509338057-3593.90685033279i 4263.42477
    461.5479 -5 884.494125763053+1073.46770239611i 1390.921553
    461.6699 -12 1555.00662792707-130.788651123321i 1560.497127
    461.792 -17 2034.4931887128+4769.55673309454i 5185.348008
    461.9141 -20 -3762.95755349828-1498.7299409022i 4050.437135
    462.0361 -21 126.387892985635+4833.17799065132i 4834.830234
    462.1582 -21 -6127.26142895914+135.972419101663i 6128.769951
    462.2803 -20 -3086.55215773021-12073.3757582446i 12461.66949
    462.4023 -20 301.133423945414+164.493887203308i 343.1320124
    462.5244 -18 3308.68376124275-5755.8581090446i 6639.073038
    462.6465 -16 -10263.2848905107-6120.31934615895i 11949.61613
    462.7686 -10 -1157.34139612198-3213.71946550948i 3415.762274
    462.8906 -1 3325.61761037261+1966.4715393834i 3863.514308
    463.0127 10 5587.19193138125-5270.41053134644i 7680.75132
    463.1348 39 251.690060973419-1158.87671994713i 1185.893393
    463.2568 51 -5935.75495320972-346.229871442116i 5945.844094
    463.3789 58 4751.8633219539+7162.17036510269i 8595.166628
    463.501 58 -3286.59853991087-380.758054470698i 3308.580762
    463.623 52 -1165.85685767975-6119.44574102028i 6229.513495
    463.7451 41 -3453.93185461625-5205.30707131473i 6246.988632
    463.8672 29 -2588.95873001384-4754.12433482014i 5413.354366
    463.9893 18 2923.11148279323+1832.92285757382i 3450.244476
    464.1113 8 -1013.4343187633-2658.03669412705i 2844.680683
    464.2334 2 1527.07260066264-1874.71491525593i 2417.955074
    464.3555 0 -3985.14054719707+697.760341603754i 4045.76503
    464.4775 0 -615.000997637413+4285.21190216644i 4329.118533
    464.5996 3 -473.259675041578+1665.94670744878i 1731.864069
    464.7217 8 5190.27105873503-4930.47587069182i 7158.806184
    464.8438 15 -9923.87332213192-332.052813733953i 9929.427012
    464.9658 25 -4721.66493534519-1738.68562510454i 5031.614797
    465.0879 51 1438.08457770755-590.733160952182i 1554.687403
    465.21 64 1838.98328318547-2105.23292860194i 2795.329176
    465.332 77 5345.87684446998-732.038603868555i 5395.764983
    465.4541 86 -2594.22082610886+3357.20251038116i 4242.733834
    465.5762 89 3686.80035263178-426.77499916117i 3711.41937
    465.6982 87 2509.51797201223+2108.58426195761i 3277.774861
    465.8203 81 1769.9118246202-2636.95914429092i 3175.868605
    465.9424 73 -1491.22875266925-2533.19738565961i 2939.532648
    466.0645 65 -505.003854401653-3732.43160767936i 3766.440574
    466.1865 60 -2693.09778664812+7856.58541760462i 8305.342263
    466.3086 57 3325.24374655784-4804.43684258527i 5842.932427
    466.4307 58 2845.50920844432-7336.63085200778i 7869.12161
    466.5527 61 1701.01468353256-2643.78862213017i 3143.734918
    466.6748 64 5004.63890331623-2967.57130396373i 5818.323641
    466.7969 66 880.329768434624+1543.13062771768i 1776.578913
    466.9189 67 -962.239134813244-7255.11405130584i 7318.646326
    467.041 68 -1239.89030543898-437.748046935469i 1314.895936
    467.1631 71 -2454.86337957959-1214.90093141665i 2739.039701
    467.2852 76 217.217418117427-2368.08106924531i 2378.022573
    467.4072 85 1244.99512436916-2064.673115825i 2410.993143
    467.5293 96 2037.16196259681+4322.06297853052i 4778.101846
    467.6514 108 -3382.08713794787-8297.19787411735i 8960.022655
    467.7734 118 -3124.76659693139-3070.74863887101i 4381.057348
    467.8955 125 1931.59873040039-1310.71410425609i 2334.318984
    468.0176 130 -450.192905201556-2049.43857988739i 2098.302205
    468.1396 132 -746.97152339336+4380.12062557001i 4443.357194
    468.2617 132 -3299.65816552413-1.83280637114814i 3299.658675
    468.3838 129 5116.76709360167+1652.25170617095i 5376.917443
    468.5059 124 1113.53145334177-423.294691320717i 1191.27272
    468.6279 116 616.428508627695-1985.02436825256i 2078.534544
    468.75 104 9.28761207209972+5043.31996524644i 5043.328517
    468.8721 89 4605.01685705378+3138.37514364134i 5572.753251
    468.9941 51 -7542.54977244861-3161.24556751217i 8178.235177
    469.1162 33 579.819686429249+2670.75605038401i 2732.970682
    469.2383 19 4996.5881024891-2397.52917947157i 5542.024795
    469.3604 10 4794.67680450647+54.4559167241227i 4794.986038
    469.4824 8 6745.29394277855-1199.81749798431i 6851.171608
    469.6045 11 179.619978282439-4123.71990553615i 4127.629973
    469.7266 18 4657.44288098102-4606.79596297722i 6550.904001
    469.8486 26 1001.26494716767+55.4793576301625i 1002.800805
    469.9707 32 1694.3259248375-2037.62239583006i 2650.027428
    470.0928 35 4676.35600059322-3405.85748565728i 5785.168162
    470.2148 34 195.971845743823+2209.5839189598i 2218.257438
    470.3369 31 -8000.76554376711+4013.7044332393i 8951.093373
    470.459 26 7657.8183884366-4082.0770847211i 8677.876226
    470.5811 20 -736.158562866224-12171.0168551484i 12193.25964
    470.7031 15 739.85211270734-5049.01245131738i 5102.931303
    470.8252 9 -1035.13608933195-1355.03575350159i 1705.177005
    470.9473 -2 -1581.37893588514+2009.77675111364i 2557.334927
    471.0693 -11 4027.42131644798-1735.12265279641i 4385.290535
    471.1914 -21 1323.30581748985+2882.61619005695i 3171.847157
    471.3135 -32 -1481.52579503208+1894.33213932346i 2404.872748
    471.4355 -41 1964.25960054496+872.719933815603i 2149.408258
    471.5576 -48 4468.64516993679+7576.56321178981i 8796.198029
    471.6797 -53 -3972.75810871976+2953.52785287517i 4950.367034
    471.8018 -56 -941.238530006319-13421.0303174673i 13453.99512
    471.9238 -57 -10367.1596189787+797.054070224889i 10397.75427
    472.0459 -59 10216.8671767218+3700.09158969027i 10866.23452
    472.168 -60 9952.5756416609+3534.78849791107i 10561.65194
    472.29 -62 7178.45338286723-69.4172070669586i 7178.789015
    472.4121 -63 4919.58690928439-839.211100220336i 4990.652325
    472.5342 -61 4889.02064832465+6889.27056906296i 8447.755434
    472.6563 -55 4914.94857960943+2545.04742738575i 5534.797733
    472.7783 -46 3131.42778212791-5005.305548644i 5904.144611
    472.9004 -22 4171.08876284636+153.993538868813i 4173.930459
    473.0225 -11 -8585.25513526247+4630.9406801297i 9754.599803
    473.1445 -5 -2823.69001502164-9177.63812239735i 9602.201144
    473.2666 -5 9226.10409709889+5763.61807206548i 10878.43234
    473.3887 -12 18781.1648697024-17981.5990668286i 26001.34725
    473.5107 -22 4887.27812541249-2953.01032069025i 5710.145132
    473.6328 -35 -410.610949979761+1336.43776176233i 1398.094147
    473.7549 -47 6119.1174638666+3488.49528253475i 7043.663668
    473.877 -57 2730.9120155844+2910.4758839745i 3991.083851
    473.999 -65 -6012.87826704507-3151.10367758423i 6788.531464
    474.1211 -69 -11001.3572689932+2386.33061074347i 11257.19483
    474.2432 -70 4589.99291139369+4163.25925762613i 6196.834883
    474.3652 -70 -1760.79474792502+5887.70047677954i 6145.357195
    474.4873 -67 -4583.41078442914-21977.0875723692i 22449.94505
    474.6094 -63 10268.1976328993-2194.51093533358i 10500.08385
    474.7314 -55 -3912.12544958143-12606.794479444i 13199.8482
    474.8535 -33 34433.4491597746+32518.2299162664i 47361.35237
    474.9756 -20 1988.17997952345+10400.1533413242i 10588.48663
    475.0977 -9 2579.64907640312+4002.46924377853i 4761.759066
    475.2197 -1 33998.2014026094-11673.118475201i 35946.34048
    475.3418 1 -3424.84407135261+5053.2483007891i 6104.496318
    475.4639 0 6479.07240985255+1699.67099874061i 6698.30283
    475.5859 -8 9895.59550633716+11113.0141956266i 14880.25184
    475.708 -18 -191.774659857307+6572.83966670724i 6575.63676
    475.8301 -28 -6950.63458415753-1662.53102719453i 7146.700675
    475.9521 -36 4702.08827112027-5233.66029072723i 7035.68292
    476.0742 -41 -3149.83603706025+2813.74836355415i 4223.582237
    476.1963 -43 20446.2195857371+9808.94607410779i 22677.37459
    476.3184 -42 10218.0675932234-1329.09988056586i 10304.14537
    476.4404 -41 3000.14369997865-2578.91460033279i 3956.218237
    476.5625 -40 3455.3759480943+2656.65707249912i 4358.606399
    476.6846 -40 -2227.02534279741+4962.24250019543i 5439.070923
    476.8066 -41 2182.99959534172-5185.09091194176i 5625.891485
    476.9287 -38 899.564837495928-1710.33937752973i 1932.479672
    477.0508 -33 -2565.89356099184+6576.49761680742i 7059.329336
    477.1729 -24 -5620.99751832678+390.559191507095i 5634.549634
    477.2949 -14 6617.06493507785-9313.58037271024i 11424.89946
    477.417 -2 2495.08397358471+2690.87953026413i 3669.642582
    477.5391 7 14267.0984717456-8509.76902097205i 16612.23247
    477.6611 15 -8531.08715159466+3853.2732779398i 9360.938144
    477.7832 21 -5507.69507940362+7509.49875637438i 9312.748072
    477.9053 25 1462.50273074172+9020.73535230727i 9138.521792
    478.0273 27 -4542.19042855067+7435.7344920982i 8713.30255
    478.1494 29 -4097.96467952111-2296.55151519882i 4697.601875
    478.2715 28 923.02096777897+5970.38260748325i 6041.3108
    478.3936 26 4553.71759512976+8507.63986663144i 9649.677717
    478.5156 21 -3507.5741486446+1257.99091904036i 3726.341042
    478.6377 12 -585.365096826171-8684.90438533354i 8704.608922
    478.7598 -13 5015.35536681262+8149.70230654967i 9569.296585
    478.8818 -26 3186.60846542852-3413.67218815846i 4669.864165
    479.0039 -36 3514.72631412373+24576.6560255616i 24826.70583
    479.126 -40 9340.72353091312+7677.30338317454i 12090.90995
    479.248 -39 7690.46592963378+1387.54369329839i 7814.636506
    479.3701 -32 -2935.72802479979+5408.64672880127i 6154.0197
    479.4922 -21 -3565.52273067917-2728.17615057107i 4489.531986
    479.6143 -8 9145.35335850711+317.250945694324i 9150.854398
    479.7363 2 4899.88511119853-632.082049621051i 4940.48599
    479.8584 10 -7791.61866889951+2481.03556838106i 8177.093553
    479.9805 15 -3442.09021232733-3420.05508580397i 4852.29449
    480.1025 17 13896.9710526752+3771.94954811523i 14399.7711
    480.2246 17 1943.42010760295-4153.89400373339i 4586.03501
    480.3467 16 16710.9000460029+10258.8665001545i 19608.63387
    480.4688 15 -6017.95181169704-8684.50253742859i 10565.8094
    480.5908 13 -8161.03827259771+1792.51407145725i 8355.576125
    480.7129 6 -7952.57285755013+10981.4988630258i 13558.6405
    480.835 0 -8175.44620342237-1636.61460183445i 8337.651227
    480.957 -9 3095.3095166468-3979.81424825082i 5041.811426
    481.0791 -18 -4823.27456175653+5519.69668144552i 7330.14522
    481.2012 -27 -5881.95875394617+11307.6558898016i 12746.00018
    481.3232 -34 2830.36224018552+5946.65541873503i 6585.868286
    481.4453 -38 8514.59558791477-5383.73057352365i 10073.87179
    481.5674 -40 -9063.96519098983+2736.38681602462i 9468.013403
    481.6895 -41 8802.00204181823+8564.68214691306i 12281.24669
    481.8115 -42 -17868.8546131522+11582.6196803502i 21294.43692
    481.9336 -43 1647.6947802127+16416.6184146269i 16499.09871
    482.0557 -45 2389.28168862826+11591.1139204207i 11834.80413
    482.1777 -46 -7306.91969838988-1571.84559126037i 7474.073457
    482.2998 -44 3354.36544916083-2278.76517326968i 4055.18659
    482.4219 -39 3533.39471197136+10627.7431892845i 11199.72337
    482.5439 -30 2429.88446118618+4918.15904462427i 5485.674697
    482.666 -7 -4700.49153236534-6918.71961505393i 8364.406838
    482.7881 3 383.408798889458-2373.50671222985i 2404.274614
    482.9102 9 1719.80315130014+13598.9338659705i 13707.25082
    483.0322 8 15258.1539031385+1483.09317064469i 15330.06281
    483.1543 2 -4613.99922767894+13824.7366625835i 14574.37246
    483.2764 -7 9934.8837169411-2693.52360961633i 10293.54088
    483.3984 -20 -4758.57162372461-1955.30322074955i 5144.629684
    483.5205 -32 -13800.826817196-2475.68366966853i 14021.12087
    483.6426 -43 -986.804170094645-6803.57225489784i 6874.763843
    483.7646 -51 3008.42125030836+2974.48737631655i 4230.623308
    483.8867 -56 -6454.48867706414-167.413226300892i 6456.659451
    484.0088 -58 -10050.3511922002-697.394426674849i 10074.51826
    484.1309 -57 2009.72466372105-548.416055598655i 2083.207477
    484.2529 -56 6447.20240340581+1920.49212442632i 6727.162019
    484.375 -51 -739.59977585955-7018.95350892883i 7057.812422
    484.4971 -44 -6044.56832186894+13715.9392869181i 14988.78903
    484.6191 -23 -7287.52863469633-4509.07409628823i 8569.703776
    484.7412 -11 -12284.4487431459+11820.6613825843i 17048.04143
    484.8633 0 -9454.16541020082+7968.8048623883i 12364.59035
    484.9854 9 -1035.64160513027-3335.2780562776i 3492.367857
    485.1074 12 2887.16134378603+1984.80223324762i 3503.589663
    485.2295 10 -4995.6768665896+5309.04072520898i 7289.904031
    485.3516 3 -3635.51797104263+3405.31496775238i 4981.281055
    485.4736 -6 3545.3852600287-2864.78825824761i 4558.154057
    485.5957 -15 1844.11495579533+399.967195005603i 1886.990654
    485.7178 -23 -5545.24926948039+4364.08681914402i 7056.560297
    485.8398 -29 4570.98290147605+17397.7389023639i 17988.19623
    485.9619 -31 -6166.2758527201+1074.17094102184i 6259.137409
    486.084 -32 3950.58416058413+12448.8798236826i 13060.69386
    486.2061 -32 -2682.83255820553-3610.94677519i 4498.502767
    486.3281 -33 -3765.89997407069-12484.4287655189i 13040.05231
    486.4502 -35 2651.91862476533-1500.19965113738i 3046.846138
    486.5723 -38 -603.293431041502+4570.46711635393i 4610.111997
    486.6943 -37 -681.63224862694-2628.66793195946i 2715.606308
    486.8164 -32 -2351.44646551905-8108.64210440282i 8442.711487
    486.9385 -24 11.3170768222144+3078.22015520062i 3078.240959
    487.0605 -14 2408.77730317003+7565.93269376341i 7940.122519
    487.1826 -2 12075.2548723913-310.60785355972i 12079.24904
    487.3047 8 -18230.6298948291-2860.37071124343i 18453.66053
    487.4268 16 -4791.13675464056-5836.62551318821i 7551.237566
    487.5488 22 -23750.0821871562-1758.42615860563i 23815.08905
    487.6709 26 -14444.2030040088-3611.20737322877i 14888.78165
    487.793 29 5132.11023827871-1609.14594775214i 5378.466899
    487.915 30 846.395867223875+4545.62164051723i 4623.749784
    488.0371 29 -8506.22406866007+568.351763523444i 8525.190416
    488.1592 27 -5487.46975928308+1524.00315949221i 5695.165493
    488.2813 21 4717.89222857022+3103.04901455161i 5646.894746
    488.4033 12 874.011044702798-258.91396560125i 911.5545775
    488.5254 -14 -5967.90193815082-755.7266093116i 6015.561175
    488.6475 -29 -14515.0920189425+19358.6746993441i 24195.99517
    488.7695 -40 3279.75391599894+1261.92789462246i 3514.14965
    488.8916 -47 -9957.23430793247+12074.9903163069i 15650.93947
    489.0137 -48 310.747172170834-5978.09848655651i 5986.169503
    489.1357 -43 5872.81330350254-9008.08308117905i 10753.39467
    489.2578 -34 997.360798968298+680.97276778867i 1207.664057
    489.3799 -23 -3711.23252701075+3530.7307833507i 5122.431721
    489.502 -12 -355.464776722193-1308.78701838268i 1356.200084
    489.624 -2 3160.06090592201-3106.86750194613i 4431.547202
    489.7461 4 -4453.03322783915-364.133915063903i 4467.896422
    489.8682 10 -3026.18697042398+3713.88668295865i 4790.695343
    489.9902 16 369.006452221769+11758.3210059516i 11764.10977
    490.1123 21 5620.92780893614-14128.1788032195i 15205.27099
    490.2344 26 -8487.25630006606+1241.12375145149i 8577.5234
    490.3564 30 -3037.84509468486-18860.4417680526i 19103.52759
    490.4785 33 2263.15086226248-9528.68835710586i 9793.760954
    490.6006 30 -2216.62345287242+3493.6070741879i 4137.476274
    490.7227 23 -3082.84224174797-1127.18853027328i 3282.448822
    490.8447 14 -3193.00967330603-4826.50048863103i 5787.090611
    490.9668 5 -2382.98713315741-4554.86863442892i 5140.569614
    491.0889 -1 -2578.41603137535+4581.82100986674i 5257.500642
    491.2109 -8 1128.89298241645+80.276021129684i 1131.743613
    491.333 -13 188.82542297117-2220.38864102075i 2228.403186
    491.4551 -16 -22301.9900195394-11994.7072702157i 25322.9493
    491.5771 -20 -6835.10641451214+16022.4318344998i 17419.44321
    491.6992 -24 -14594.6490472149-8319.46270103144i 16799.32262
    491.8213 -28 4585.80032591414-2819.17644683758i 5383.058654
    491.9434 -31 8850.71386118695+3844.90423425733i 9649.788828
    492.0654 -31 -858.546402938151+570.613841655757i 1030.874426
    492.1875 -27 -874.65972953118-1399.04012778344i 1649.9524
    492.3096 -18 1806.94726401785-1948.50372200262i 2657.390669
    492.4316 5 5078.20172168703-1593.61517647292i 5322.381239
    492.5537 17 -594.474012162136-517.917963306042i 788.4404656
    492.6758 24 559.152079015626+3199.5567579036i 3248.047797
    492.7979 24 -1107.16685875577+8243.84229657239i 8317.857553
    492.9199 17 6539.86016522771-4754.58021342839i 8085.530532
    493.042 5 -8147.03670930547-930.985494010711i 8200.057386
    493.1641 -9 -5477.11804020871-8112.41507608154i 9788.2634
    493.2861 -25 -308.022192161738-3896.83067332877i 3908.985414
    493.4082 -40 839.171738227614-2516.67527286941i 2652.897215
    493.5303 -52 11597.6899116713-10711.0362819032i 15787.1058
    493.6523 -59 -168.339826719294-985.298099403528i 999.5752308
    493.7744 -62 1348.76287688732+362.436824089806i 1396.610808
    493.8965 -62 -5681.1656447301-21197.8432723032i 21945.93818
    494.0186 -59 -1314.78225934394-1298.45734226072i 1847.875499
    494.1406 -54 -2607.22558704372+307.569343639503i 2625.304585
    494.2627 -45 18008.9910437606-3251.4692210995i 18300.15876
    494.3848 -21 -6894.37396575441-18407.8641912939i 19656.59829
    494.5068 -10 8263.74803497588-4972.32654865528i 9644.353939
    494.6289 0 -3910.04019395122+8082.89155675213i 8978.950397
    494.751 5 -7599.21116332571+2769.41229192525i 8088.118122
    494.873 6 920.174634038183+809.580386553601i 1225.618929
    494.9951 2 -4825.0351641414-2317.86473647529i 5352.892795
    495.1172 -6 -1511.19621950684+2033.65536946247i 2533.666943
    495.2393 -15 -3850.2818159056+7384.910323292i 8328.359415
    495.3613 -23 1466.42840994124+878.958332955257i 1709.672493
    495.4834 -29 -7639.67343149797+2129.70984497714i 7930.969308
    495.6055 -32 -17187.3463500831-1166.02326344447i 17226.8536
    495.7275 -34 -220.596970852856+32191.2460204555i 32192.00185
    495.8496 -35 6400.74467693852-32197.4067574357i 32827.46616
    495.9717 -39 20824.8487983861-9696.48974311833i 22971.63992
    496.0938 -47 9964.27066228194-2820.62972423932i 10355.80233
    496.2158 -58 -169.6735886762-5536.30121748843i 5538.90064
    496.3379 -77 2536.06663493756-1549.62760306977i 2972.032921
    496.46 -77 3924.25596795946-4010.4786679728i 5611.035916
    496.582 -66 3226.8561841297-3103.18087343579i 4476.866356
    496.7041 -40 -1719.12002644841-730.12724257535i 1867.741806
    496.8262 -1 2687.86855910984+3105.33958519354i 4107.039241
    496.9482 50 307.312501380741-563.608538324287i 641.946694
    497.0703 109 -2977.59589826094-12594.0586097631i 12941.26692
    497.1924 171 -35090.2204640616+19688.1251885898i 40236.12613
    497.3145 229 -3284.44791441733-9640.36509351119i 10184.50967
    497.4365 279 4725.04595437388-2789.19559896993i 5486.863527
    497.5586 315 5978.1835094764+921.520699309485i 6048.791489
    497.6807 333 2617.25894952413-2303.57354231557i 3486.616623
    497.8027 330 1115.46493306523-1989.35033724429i 2280.740402
    497.9248 303 1700.99907961502-1131.12453627414i 2042.753188
    498.0469 256 4288.86391032226-2507.46784512026i 4968.07293
    498.1689 190 3922.18070219323-4340.38879612074i 5849.99798
    498.291 28 3967.1656089975-3860.23365778538i 5535.323555
    498.4131 -51 7814.38459559438-3590.42201281267i 8599.752138
    498.5352 -119 16846.939195841-20761.3310767956i 26736.7206
    498.6572 -169 -16527.534371266+18191.7596560368i 24578.43591
    498.7793 -197 -10047.171008139+2524.46338116382i 10359.4672
    498.9014 -205 -5758.18008507705+2811.8275792982i 6408.042777
    499.0234 -194 -1543.15370509021+3486.28676207187i 3812.54754
    499.1455 -170 -1360.98066104254+657.72680481034i 1511.579607
    499.2676 -140 -2469.81317934237-111.318963463707i 2472.32058
    499.3896 -108 -2549.77566472954+966.938092757354i 2726.962636
    499.5117 -80 -2739.20630882844+287.384821136039i 2754.240592
    499.6338 -57 -2457.67010662418-1391.95625672804i 2824.479523
    499.7559 -40 178.953043337808-2894.50766358816i 2900.034277
    499.8779 -27 -2246.27529912119-3267.41649625542i 3965.067878
  • Hey Viney,

    A quick way to verify if the PGA caps have anything to do with it would be to measure the internally generated test signal. This signal is amplified by the PGA and converted with the modulator, so it is essentially the same as an externally generated signal.

    Regards,
    Brian
  • Brian,
    But what can be a source for the 25 Hz signal?
    How to measure the internally generated signal?
    Regards
    Viney Chaddha
  • Hey Viney,

    I'm not sure. My guess is that it is something in the room you are in like a machine, fan, transformer, etc. The test signal will tell us definitively. You can configure the test signal source and frequency in the CONFIG2 register and multiplex it to any channel using the input mux settings found in the CHnSET register.

    Regards,
    Brian Pisani
  • Brian,

    I set the config2 register as 0xA2 and CH1Set and CH2Set as 0x05, I could not see any particular frequency especially 25 Hz region. Attached are the screen shots of the noise. I am foxed as to why do I see the noise (kind of sine wave superimposed on the signal) in the ECG capture, the king I showed earlier. The FFT is also attached and do not see any peak in that region.

    Regards and look-forward to your suggestions.

    Viney Chaddha

        

  • Hey Viney,

    That looks like the DC test signal so I'd have to think the noisy signal is being generated in your environment somewhere and is not from the chip.

    Regards,