ADS1294 -- spikes in acquired data

Other Parts Discussed in Thread: ADS1294R

Hello, all.  We are using an ADS1294R connected to an STMicrocontroller SOC to acquire ECG signals.  We are finding occasional abnormal values in either or both of the 2 ECG channels (channels 1 and 2).  The ADS is configured in SDATAC mode with the START line tied low and the START command issued.  The ADS is set to sample in low-power mode at 1 Ksps from an ECG simulator.  The SOC reads data using the RDATA command at 1 ms intervals.  The vast majority of samples are correct but these abnormal values are causing spikes in our waveforms.  They occur anywhere between 30 seconds and 10 minutes apart.  I have been able to trigger the logic analyzer when a spike is detected and the SPI transactions are attached.   Only a single RDATA response has an abnormal value, in this case channel 2--the previous 1 ms and subsequent 1ms readings are fine.  According to the data sheet, reading data using the RDATA command at any time should not cause a data corruption.   I've attached the logic analyzer traces.  The channel 2 data value 0x4FDED5 represents about .25 V but our signals are 1mV or less.  Anyone have any ideas?

Register setup:

00: 0xD0
01: 0x04
02: 0x00
03: 0xC6
04: 0x00
05: 0x10
06: 0x00
07: 0x00
08: 0x80
09: 0x00
0A: 0x00
0B: 0x00
0C: 0x00
0D: 0x02
0E: 0x06
0F: 0x00
10: 0x00
11: 0x00
12: 0x00
13: 0x00
14: 0x00
15: 0x00
16: 0xE2
17: 0x00
18: 0x07
19: 0x3F

Here's the bad channel 1 sample:

These are the prior and following samples:

  • Hey Ron,

    When I've seen this before, it was been because someone was using RDATAC mode and waited too long to retrieve a sample and a new conversion overwrote their value mid-acquisition. You shouldn't have trouble with that since you are in SDATAC mode which means data won't get loaded until the device decodes the RDATA command. So I have nothing I can point to immediately and say that's what I think it is.

    I can, however, help you debug. I'd like to determine whether this is an interface issue or if maybe your simulator is actually producing the spikes at the input for whatever reason. The way we can tell is that this device will filter data with a 3rd order sinc filter which will actually take 3 sample periods to settle. That means that if the input data actually had a spike in it that it would take another 2 conversions for the output to settle to the levels it was at before. Is that what you see?

    If not then it is probably an interface issue and I we can investigate that further.

    Regards,
    Brian Pisani
  • Thanks for the quick response, Brian.  I don't see any settling in channel 2 value of the subsequent sample.  Here are the three samples for channels 2 and 3 from the logic analyzer traces:


    0xFFE3BA, 0x002B62

    0x4FDED5, 0x002F95   <- bad channel 2

    0xFFDED5, 0x002F95

    The third sample is almost identical except for 3 bits of channel 2.  That suggests to me that the ADS was loading the same acquired sample as the second sample so no intervening data update.  Is it possible the second latch in the sequence latched incomplete data?  The documentation says its not possible to load corrupted data though with RDATA so I didn't think I needed to synchronize with _DRDY.

    Ron

  • Hey Ron,

    Out of curiosity, if you are synchronized to DRDY, does the problem persist? I will check with a colleague who will know more than me.

    Even if unrelated to this issue, I could see problems with not being synchronized with DRDY exactly. If the timing is off be even a little bit you can imagine a scenario where you would not actually collect the device's output correctly. For instance if your clock is slightly faster you collect two of the same sample or if you clock a little too slow you might miss a sample every once and a while.

    Brian
  • I don't know if the problem goes away when synchronized to DRDY. We are in a very time-sensitive loop at 1 ms intervals, trying to reduce power consumption and waiting for DRDY will consume that time. Our updates are triggered by an external clock not synchronized to DRDY. Reading the same sample again or missing a sample is not really an issue for us but getting a bad value is. We can also sample at 2 Ksps in the ADS if we need less latency in the acquired values. The ADS documentation indicates we should not get corrupted data using RDATA. Is that true?
  • Hi Brian,

    I am another engineer working on this project and would like to add that we have used different ECG simulators and have seen the spikes consistently on all of them. In my mind, this would eliminate the simulator as the culprit. 

    Thanks,

    -Konstantin

  • Hi, Brian. Have you heard back from your colleague? We're still looking for any suggestions you might have.

    Ron
  • Hey Ron,

    Sorry for the delay! I haven't forgotten about you. Just give me a little more time to consult with him. He was very busy yesterday.

    Brian
  • Brian, I tried a few alternatives today.  I tried reading all channels instead of 3 but that didn't make a difference.  I now read the data twice in the same SPI bus transaction.  In the logic analyzer capture you can see that channel 3 has a large value, 0xBAD034 in the first read but 0x001E74 in the second.  Of interest is that DRDY is not asserted anywhere in my capture spanning 3 ms.  Fortunately we can impose thresholds for the detection of these anomalies and use the second reading if necessary.  It seems that there is some small window where the acquired data is not yet valid when the RDATA issued.

    Ron

  • Hey Ron,

    After speaking with the chip designer, he says you really should synchronize collections to DRDY since it is likely that if the timing is just right that a new sample corrupts the current conversion result. There is no logic in place to prevent this. The idea in mind was that SDATAC mode could be used to read registers without corruption but that reading data without synchronization is generally unsupported.

    I understand that your application is time sensitive so it may be tough to synchronize conversions to the DRDY signal at this point, but this device provides the ability to provide the master clock which gives you control over timing. Let me know how I can help.

    Regards,
    Brian Pisani
  • Thanks for clarifying that, Brian. You may want to revise section 9.5.2.8 of the data sheet. This is a bit misleading: "When reading data with the RDATA command, the read operation can overlap the occurrence of the next DRDY without data corruption." It implies to me that the data results are double-buffered but apparently that is not the case.

    Ron